When displaying a sine-wave where several cycles/division are present, when the READOUT is switched on, there is noticeable "noise" (for lack of a better definition) in the trace. It appears as tiny blank spots in the trace that move rapidly along the trace. The noise is not in sync with the triggering and appears at many points along the trace. Switching the readout off stops the effect.
The power supply has fresh filter caps as well as the horizontal board and the HV board so I don't think this is PS ripple and more likely digital noise from the READOUT circuitry but that's just a guess.
I've looked for places in the READOUT board and/or interface board where there might be bypass filtering components that might have been designed to eliminate this but I don't see anything. I do note that switching on the READOUT connects it to the +15V rail and possibly that's the source of the noise.
I've mostly observed this with sine-waves but it might be present with other waveforms. As a side note, with mostly square-waves, with the GRATICULE ILLUMINATION turned up, I can see a much smaller presentation of similar noise in the trace. This is much less noticeable but could be related.
Any ideas what I can look for as a cause for this?
Thanks, Barry - N4BUQ
|
Correction: The READOUT switch connects to the -15V rail through R5006, not the +15V rail.
Thanks, Barry - N4BUQ
toggle quoted message
Show quoted text
----- Original Message ----- From: "n4buq" <n4buq@...> To: "tekscopes" <[email protected]> Sent: Thursday, December 30, 2021 9:14:41 AM Subject: [TekScopes] 7704A - Noise on sine-wave traces When displaying a sine-wave where several cycles/division are present, when the READOUT is switched on, there is noticeable "noise" (for lack of a better definition) in the trace. It appears as tiny blank spots in the trace that move rapidly along the trace. The noise is not in sync with the triggering and appears at many points along the trace. Switching the readout off stops the effect.
The power supply has fresh filter caps as well as the horizontal board and the HV board so I don't think this is PS ripple and more likely digital noise from the READOUT circuitry but that's just a guess.
I've looked for places in the READOUT board and/or interface board where there might be bypass filtering components that might have been designed to eliminate this but I don't see anything. I do note that switching on the READOUT connects it to the +15V rail and possibly that's the source of the noise.
I've mostly observed this with sine-waves but it might be present with other waveforms. As a side note, with mostly square-waves, with the GRATICULE ILLUMINATION turned up, I can see a much smaller presentation of similar noise in the trace. This is much less noticeable but could be related.
Any ideas what I can look for as a cause for this?
Thanks, Barry - N4BUQ
|
Do not do anything.
There is one beam.? It has to be "borrowed" to write the readouts.
There are two ways:
1) Borrow the beam when it's needed.? This chops tiny holes in the trace.? it's going to be obvious at some sweep speeds, but will work on slow sweeps.
2) refresh the readouts on the sweep retrace.? This eliminates the "noise" but for slow sweep speeds, you will get a flickering readout.
There's generally a switch on the readout board to control this behavior.
Harvey
toggle quoted message
Show quoted text
On 12/30/2021 10:14 AM, n4buq wrote: When displaying a sine-wave where several cycles/division are present, when the READOUT is switched on, there is noticeable "noise" (for lack of a better definition) in the trace. It appears as tiny blank spots in the trace that move rapidly along the trace. The noise is not in sync with the triggering and appears at many points along the trace. Switching the readout off stops the effect.
The power supply has fresh filter caps as well as the horizontal board and the HV board so I don't think this is PS ripple and more likely digital noise from the READOUT circuitry but that's just a guess.
I've looked for places in the READOUT board and/or interface board where there might be bypass filtering components that might have been designed to eliminate this but I don't see anything. I do note that switching on the READOUT connects it to the +15V rail and possibly that's the source of the noise.
I've mostly observed this with sine-waves but it might be present with other waveforms. As a side note, with mostly square-waves, with the GRATICULE ILLUMINATION turned up, I can see a much smaller presentation of similar noise in the trace. This is much less noticeable but could be related.
Any ideas what I can look for as a cause for this?
Thanks, Barry - N4BUQ
|
Hi Harvey,
I'll see if that resolves the issue. The way that the blanks move along the trace, it seemed a more random effect than I'd expect when writing to the same locations each time but maybe not.
Thanks! Barry - N4BUQ
toggle quoted message
Show quoted text
----- Original Message ----- From: "Harvey White" <madyn@...> To: "tekscopes" <[email protected]> Sent: Thursday, December 30, 2021 9:42:50 AM Subject: Re: [TekScopes] 7704A - Noise on sine-wave traces Do not do anything.
There is one beam.? It has to be "borrowed" to write the readouts.
There are two ways:
1) Borrow the beam when it's needed.? This chops tiny holes in the trace.? it's going to be obvious at some sweep speeds, but will work on slow sweeps.
2) refresh the readouts on the sweep retrace.? This eliminates the "noise" but for slow sweep speeds, you will get a flickering readout.
There's generally a switch on the readout board to control this behavior.
Harvey
On 12/30/2021 10:14 AM, n4buq wrote:
When displaying a sine-wave where several cycles/division are present, when the READOUT is switched on, there is noticeable "noise" (for lack of a better definition) in the trace. It appears as tiny blank spots in the trace that move rapidly along the trace. The noise is not in sync with the triggering and appears at many points along the trace. Switching the readout off stops the effect.
The power supply has fresh filter caps as well as the horizontal board and the HV board so I don't think this is PS ripple and more likely digital noise from the READOUT circuitry but that's just a guess.
I've looked for places in the READOUT board and/or interface board where there might be bypass filtering components that might have been designed to eliminate this but I don't see anything. I do note that switching on the READOUT connects it to the +15V rail and possibly that's the source of the noise.
I've mostly observed this with sine-waves but it might be present with other waveforms. As a side note, with mostly square-waves, with the GRATICULE ILLUMINATION turned up, I can see a much smaller presentation of similar noise in the trace. This is much less noticeable but could be related.
Any ideas what I can look for as a cause for this?
Thanks, Barry - N4BUQ
|
There's likely no relationship between the chopping frequency and the sweep speed.? I'd go for random.
You're thinking that this behaves like digital, perhaps?
Harvey
toggle quoted message
Show quoted text
On 12/30/2021 10:46 AM, n4buq wrote: Hi Harvey,
I'll see if that resolves the issue. The way that the blanks move along the trace, it seemed a more random effect than I'd expect when writing to the same locations each time but maybe not.
Thanks! Barry - N4BUQ
----- Original Message -----
From: "Harvey White" <madyn@...> To: "tekscopes" <[email protected]> Sent: Thursday, December 30, 2021 9:42:50 AM Subject: Re: [TekScopes] 7704A - Noise on sine-wave traces Do not do anything.
There is one beam.? It has to be "borrowed" to write the readouts.
There are two ways:
1) Borrow the beam when it's needed.? This chops tiny holes in the trace.? it's going to be obvious at some sweep speeds, but will work on slow sweeps.
2) refresh the readouts on the sweep retrace.? This eliminates the "noise" but for slow sweep speeds, you will get a flickering readout.
There's generally a switch on the readout board to control this behavior.
Harvey
On 12/30/2021 10:14 AM, n4buq wrote:
When displaying a sine-wave where several cycles/division are present, when the READOUT is switched on, there is noticeable "noise" (for lack of a better definition) in the trace. It appears as tiny blank spots in the trace that move rapidly along the trace. The noise is not in sync with the triggering and appears at many points along the trace. Switching the readout off stops the effect.
The power supply has fresh filter caps as well as the horizontal board and the HV board so I don't think this is PS ripple and more likely digital noise from the READOUT circuitry but that's just a guess.
I've looked for places in the READOUT board and/or interface board where there might be bypass filtering components that might have been designed to eliminate this but I don't see anything. I do note that switching on the READOUT connects it to the +15V rail and possibly that's the source of the noise.
I've mostly observed this with sine-waves but it might be present with other waveforms. As a side note, with mostly square-waves, with the GRATICULE ILLUMINATION turned up, I can see a much smaller presentation of similar noise in the trace. This is much less noticeable but could be related.
Any ideas what I can look for as a cause for this?
Thanks, Barry - N4BUQ
|
Yes - I was thinking it was digital noise coming from the readout character generation circuits.
I tried the switch in the GATED position and that does indeed eliminate [almost] all of the gaps in the trace. Oddly, there are still two or three gaps that stay in the same position along the trace. Furthermore, adjusting the HOLDOFF on the time-base moves the gaps along the trace in "steps". I'm not sure if that's normal but perhaps it is.
I like the readout in this mode. Yes, it gets odd with slow sweeps but when in FREE RUN, the readout characters bounce/wave under certain waveforms and input frequencies. I think someone (maybe that was you?) that told me there are past posts regarding how to fix that but I don't think I ever found that (or wasn't able to get it adjusted properly). In any case, in GATED, that's eliminated (albeit it becomes a bit of an issue having the time-base selected by the GATE SELECTOR switch).
While looking through the manual, I found that removing Q3416 causes the system to display full rows in the upper and lower rows allowing proper centering, etc., of the readout characters. Fun stuff.
Thanks again, Barry - N4BUQ
toggle quoted message
Show quoted text
----- Original Message ----- From: "Harvey White" <madyn@...> To: "tekscopes" <[email protected]> Sent: Thursday, December 30, 2021 9:59:43 AM Subject: Re: [TekScopes] 7704A - Noise on sine-wave traces There's likely no relationship between the chopping frequency and the sweep speed.? I'd go for random.
You're thinking that this behaves like digital, perhaps?
Harvey
On 12/30/2021 10:46 AM, n4buq wrote:
Hi Harvey,
I'll see if that resolves the issue. The way that the blanks move along the trace, it seemed a more random effect than I'd expect when writing to the same locations each time but maybe not.
Thanks! Barry - N4BUQ
----- Original Message -----
From: "Harvey White" <madyn@...> To: "tekscopes" <[email protected]> Sent: Thursday, December 30, 2021 9:42:50 AM Subject: Re: [TekScopes] 7704A - Noise on sine-wave traces Do not do anything.
There is one beam.? It has to be "borrowed" to write the readouts.
There are two ways:
1) Borrow the beam when it's needed.? This chops tiny holes in the trace.? it's going to be obvious at some sweep speeds, but will work on slow sweeps.
2) refresh the readouts on the sweep retrace.? This eliminates the "noise" but for slow sweep speeds, you will get a flickering readout.
There's generally a switch on the readout board to control this behavior.
Harvey
On 12/30/2021 10:14 AM, n4buq wrote:
When displaying a sine-wave where several cycles/division are present, when the READOUT is switched on, there is noticeable "noise" (for lack of a better definition) in the trace. It appears as tiny blank spots in the trace that move rapidly along the trace. The noise is not in sync with the triggering and appears at many points along the trace. Switching the readout off stops the effect.
The power supply has fresh filter caps as well as the horizontal board and the HV board so I don't think this is PS ripple and more likely digital noise from the READOUT circuitry but that's just a guess.
I've looked for places in the READOUT board and/or interface board where there might be bypass filtering components that might have been designed to eliminate this but I don't see anything. I do note that switching on the READOUT connects it to the +15V rail and possibly that's the source of the noise.
I've mostly observed this with sine-waves but it might be present with other waveforms. As a side note, with mostly square-waves, with the GRATICULE ILLUMINATION turned up, I can see a much smaller presentation of similar noise in the trace. This is much less noticeable but could be related.
Any ideas what I can look for as a cause for this?
Thanks, Barry - N4BUQ
|
Hi Barry, please reread message 187015. Albert
|
I remember now that this involves adjustment of the LF Compensation. The calibration steps for that calls for a standardizer fixture and I don't have one of those. Is there a way to make that adjustment without a standardizer fixture?
Thanks, Barry - N4BUQ
toggle quoted message
Show quoted text
----- Original Message ----- From: "Albert Otten" <aodiversen@...> To: "tekscopes" <[email protected]> Sent: Thursday, December 30, 2021 12:16:56 PM Subject: Re: [TekScopes] 7704A - Noise on sine-wave traces Hi Barry, please reread message 187015. Albert
|
1) AKA Chop mode, 2) AKA Alt mode.? ?Both useful but at different sweep speeds.? ? ? Jim FordSent from my T-Mobile 4G LTE Device
toggle quoted message
Show quoted text
-------- Original message --------From: Harvey White <madyn@...> Date: 12/30/21 7:46 AM (GMT-08:00) To: [email protected] Subject: Re: [TekScopes] 7704A - Noise on sine-wave traces Do not do anything.There is one beam.? It has to be "borrowed" to write the readouts.There are two ways:1) Borrow the beam when it's needed.? This chops tiny holes in the trace.? it's going to be obvious at some sweep speeds, but will work on slow sweeps.2) refresh the readouts on the sweep retrace.? This eliminates the "noise" but for slow sweep speeds, you will get a flickering readout.There's generally a switch on the readout board to control this behavior.HarveyOn 12/30/2021 10:14 AM, n4buq wrote:> When displaying a sine-wave where several cycles/division are present, when the READOUT is switched on, there is noticeable "noise" (for lack of a better definition) in the trace.? It appears as tiny blank spots in the trace that move rapidly along the trace.? The noise is not in sync with the triggering and appears at many points along the trace.? Switching the readout off stops the effect.>> The power supply has fresh filter caps as well as the horizontal board and the HV board so I don't think this is PS ripple and more likely digital noise from the READOUT circuitry but that's just a guess.>> I've looked for places in the READOUT board and/or interface board where there might be bypass filtering components that might have been designed to eliminate this but I don't see anything.? I do note that switching on the READOUT connects it to the +15V rail and possibly that's the source of the noise.>> I've mostly observed this with sine-waves but it might be present with other waveforms.? As a side note, with mostly square-waves, with the GRATICULE ILLUMINATION turned up, I can see a much smaller presentation of similar noise in the trace.? This is much less noticeable but could be related.>> Any ideas what I can look for as a cause for this?>> Thanks,> Barry - N4BUQ>>> >>>
|
It is indeed interesting to slow the sweep speed down to about 0.5s or more and watch the readout written after the sweep completes. I had not seen that before today.
Thanks, Barry - N4BUQ
toggle quoted message
Show quoted text
----- Original Message ----- From: "Jim Ford" <james.ford@...> To: "tekscopes" <[email protected]> Sent: Thursday, December 30, 2021 6:52:39 PM Subject: Re: [TekScopes] 7704A - Noise on sine-wave traces 1) AKA Chop mode, 2) AKA Alt mode.? ?Both useful but at different sweep speeds. ? ? ? Jim FordSent from my T-Mobile 4G LTE Device -------- Original message --------From: Harvey White <madyn@...> Date: 12/30/21 7:46 AM (GMT-08:00) To: [email protected] Subject: Re: [TekScopes] 7704A - Noise on sine-wave traces Do not do anything.There is one beam.? It has to be "borrowed" to write the readouts.There are two ways:1) Borrow the beam when it's needed.? This chops tiny holes in the trace.? it's going to be obvious at some sweep speeds, but will work on slow sweeps.2) refresh the readouts on the sweep retrace.? This eliminates the "noise" but for slow sweep speeds, you will get a flickering readout.There's generally a switch on the readout board to control this behavior.HarveyOn 12/30/2021 10:14 AM, n4buq wrote:> When displaying a sine-wave where several cycles/division are present, when the READOUT is switched on, there is noticeable "noise" (for lack of a better definition) in the trace.? It appears as tiny blank spots in the trace that move rapidly along the trace.? The noise is not in sync with the triggering and appears at many points along the trace.? Switching the readout off stops the effect.>> The power supply has fresh filter caps as well as the horizontal board and the HV board so I don't think this is PS ripple and more likely digital noise from the READOUT circuitry but that's just a guess.>> I've looked for places in the READOUT board and/or interface board where there might be bypass filtering components that might have been designed to eliminate this but I don't see anything.? I do note that switching on the READOUT connects it to the +15V rail and possibly that's the source of the noise.>> I've mostly observed this with sine-waves but it might be present with other waveforms.? As a side note, with mostly square-waves, with the GRATICULE ILLUMINATION turned up, I can see a much smaller presentation of similar noise in the trace.? This is much less noticeable but could be related.>> Any ideas what I can look for as a cause for this?>> Thanks,> Barry - N4BUQ>>> >>>
|
I see where Table 4-1, Item 3 (Signal Standardizer Calibration Fixture) states that "Calibrated 7000-series plug-in units with suitable signal sources may be substituted if lower performance is acceptable." I just don't quite see where the instructions outline how to set those up in place of the standardizer - particularly for low-frequency linearity and/or vertical high-frequency compensation.
Thanks, Barry - N4BUQ
toggle quoted message
Show quoted text
----- Original Message ----- From: "n4buq" <n4buq@...> To: "tekscopes" <[email protected]> Sent: Thursday, December 30, 2021 5:11:41 PM Subject: Re: [TekScopes] 7704A - Noise on sine-wave traces I remember now that this involves adjustment of the LF Compensation. The calibration steps for that calls for a standardizer fixture and I don't have one of those. Is there a way to make that adjustment without a standardizer fixture?
Thanks, Barry - N4BUQ
----- Original Message -----
From: "Albert Otten" <aodiversen@...> To: "tekscopes" <[email protected]> Sent: Thursday, December 30, 2021 12:16:56 PM Subject: Re: [TekScopes] 7704A - Noise on sine-wave traces Hi Barry, please reread message 187015. Albert
|
I think that's where the "lower performance" comes in.
I think what you have to make the assumption of, is that the subject "standard" plugin is somehow correct and in calibration.
Using this as a reference makes all things equally wrong.
Harvey
toggle quoted message
Show quoted text
On 12/30/2021 8:13 PM, n4buq wrote: I see where Table 4-1, Item 3 (Signal Standardizer Calibration Fixture) states that "Calibrated 7000-series plug-in units with suitable signal sources may be substituted if lower performance is acceptable." I just don't quite see where the instructions outline how to set those up in place of the standardizer - particularly for low-frequency linearity and/or vertical high-frequency compensation.
Thanks, Barry - N4BUQ
----- Original Message -----
From: "n4buq" <n4buq@...> To: "tekscopes" <[email protected]> Sent: Thursday, December 30, 2021 5:11:41 PM Subject: Re: [TekScopes] 7704A - Noise on sine-wave traces I remember now that this involves adjustment of the LF Compensation. The calibration steps for that calls for a standardizer fixture and I don't have one of those. Is there a way to make that adjustment without a standardizer fixture?
Thanks, Barry - N4BUQ
----- Original Message -----
From: "Albert Otten" <aodiversen@...> To: "tekscopes" <[email protected]> Sent: Thursday, December 30, 2021 12:16:56 PM Subject: Re: [TekScopes] 7704A - Noise on sine-wave traces Hi Barry, please reread message 187015. Albert
|
Hi Barry,
For minimizing that wavy readout forget about the signal standardizer. Display your signal (calibrator or whatever) at full height (8 div) and the time base for the most nervous readout. Then rotate R4465 (Vert. Amp.) for the best setting (most quiet readout). Note that this setting is also best for your ordinary signal response. If the result is still unsatisfactory then perhaps a slight readjustment of R4453 might help (but remember the original setting, just in case).
Albert
toggle quoted message
Show quoted text
On Fri, Dec 31, 2021 at 02:13 AM, n4buq wrote: I see where Table 4-1, Item 3 (Signal Standardizer Calibration Fixture) states that "Calibrated 7000-series plug-in units with suitable signal sources may be substituted if lower performance is acceptable." I just don't quite see where the instructions outline how to set those up in place of the standardizer - particularly for low-frequency linearity and/or vertical high-frequency compensation.
Thanks, Barry - N4BUQ
----- Original Message -----
From: "n4buq" <n4buq@...> To: "tekscopes" <[email protected]> Sent: Thursday, December 30, 2021 5:11:41 PM Subject: Re: [TekScopes] 7704A - Noise on sine-wave traces I remember now that this involves adjustment of the LF Compensation. The calibration steps for that calls for a standardizer fixture and I don't have one of those. Is there a way to make that adjustment without a standardizer fixture?
|
I can confirm Albert's procedure, I have adjusted several scopes with the wavy readout issue as suggested,I think 7904 , 7104 and 7603 for this same issue.
No CAL or standardizer needed.
Jon
|
Hi Albert,
That works! I was just hesitant to start changing the settings without knowing more about what they affect. Very nice to have the wavy readouts stable.
Thanks, Barry - N4BUQ
toggle quoted message
Show quoted text
----- Original Message ----- From: "Albert Otten" <aodiversen@...> To: "tekscopes" <[email protected]> Sent: Friday, December 31, 2021 6:53:41 AM Subject: Re: [TekScopes] 7704A - Noise on sine-wave traces Hi Barry,
For minimizing that wavy readout forget about the signal standardizer. Display your signal (calibrator or whatever) at full height (8 div) and the time base for the most nervous readout. Then rotate R4465 (Vert. Amp.) for the best setting (most quiet readout). Note that this setting is also best for your ordinary signal response. If the result is still unsatisfactory then perhaps a slight readjustment of R4453 might help (but remember the original setting, just in case).
Albert
On Fri, Dec 31, 2021 at 02:13 AM, n4buq wrote:
I see where Table 4-1, Item 3 (Signal Standardizer Calibration Fixture) states that "Calibrated 7000-series plug-in units with suitable signal sources may be substituted if lower performance is acceptable." I just don't quite see where the instructions outline how to set those up in place of the standardizer - particularly for low-frequency linearity and/or vertical high-frequency compensation.
Thanks, Barry - N4BUQ
----- Original Message -----
From: "n4buq" <n4buq@...> To: "tekscopes" <[email protected]> Sent: Thursday, December 30, 2021 5:11:41 PM Subject: Re: [TekScopes] 7704A - Noise on sine-wave traces I remember now that this involves adjustment of the LF Compensation. The calibration steps for that calls for a standardizer fixture and I don't have one of those. Is there a way to make that adjustment without a standardizer fixture?
|
Hi Barry , I have been watching these postings with interest . One of my 7844's has this or a remarkably similar problem . There is the wavy / jumping about readout but in my case it is made worse by using the position control -- did this happen with yours . I would be pleased to know how similar my case is and maybe if its not the same there might be suggestions of where the problem might be . many thanks in advance for help from anyone reading this Brian? (UK)
toggle quoted message
Show quoted text
On Friday, 31 December 2021, 17:16:11 GMT, n4buq <n4buq@...> wrote: Hi Albert, That works!? I was just hesitant to start changing the settings without knowing more about what they affect.? Very nice to have the wavy readouts stable. Thanks, Barry - N4BUQ ----- Original Message ----- From: "Albert Otten" <aodiversen@...> To: "tekscopes" <[email protected]> Sent: Friday, December 31, 2021 6:53:41 AM Subject: Re: [TekScopes] 7704A - Noise on sine-wave traces Hi Barry,
For minimizing that wavy readout forget about the signal standardizer.? Display your signal (calibrator or whatever) at full height (8 div) and the time base for the most nervous readout. Then rotate R4465 (Vert. Amp.) for the best setting (most quiet readout). Note that this setting is also best for your ordinary signal response. If the result is still unsatisfactory then perhaps a slight readjustment of R4453 might help (but remember the original setting, just in case).
Albert
On Fri, Dec 31, 2021 at 02:13 AM, n4buq wrote:
I see where Table 4-1, Item 3 (Signal Standardizer Calibration Fixture) states that "Calibrated 7000-series plug-in units with suitable signal sources may be substituted if lower performance is acceptable."? I just don't quite see where the instructions outline how to set those up in place of the standardizer - particularly for low-frequency linearity and/or vertical high-frequency compensation.
Thanks, Barry - N4BUQ
----- Original Message -----
From: "n4buq" <n4buq@...> To: "tekscopes" <[email protected]> Sent: Thursday, December 30, 2021 5:11:41 PM Subject: Re: [TekScopes] 7704A - Noise on sine-wave traces I remember now that this involves adjustment of the LF Compensation.? The calibration steps for that calls for a standardizer fixture and I don't have one of those.? Is there a way to make that adjustment without a standardizer fixture?
|
Hi Brian,
I presume you're referring to the vertical positioning control and not the horizontal positioning control but I presume the vertical one. I did not notice that made much, if any, difference. I don't fully understand this but I think the waviness is caused by overshoot/undershoot of the beam as it jumps from the trace to the readout. If that's the case, then I would think that different vertical positions of the trace would affect this.
Overall, I think I prefer the GATED setting (draws the readout after the traces); however, at least for the 7704A, there's a switch that selects which B slot drives the readout so one must choose one or the other and I'm pretty sure I'm going to forget that setting when I have just one time-base in the B position which will disable the readout. Maybe not but for now, I'm going to leave it in the GATED position.
Thanks, Barry - N4BUQ
toggle quoted message
Show quoted text
----- Original Message ----- From: "Brian via groups.io" <brianas1948@...> To: "tekscopes" <[email protected]> Sent: Friday, December 31, 2021 11:26:08 AM Subject: Re: [TekScopes] 7704A - Noise on sine-wave traces Hi Barry , I have been watching these postings with interest . One of my 7844's has this or a remarkably similar problem . There is the wavy / jumping about readout but in my case it is made worse by using the position control -- did this happen with yours . I would be pleased to know how similar my case is and maybe if its not the same there might be suggestions of where the problem might be . many thanks in advance for help from anyone reading this Brian? (UK)
On Friday, 31 December 2021, 17:16:11 GMT, n4buq <n4buq@...> wrote:
Hi Albert,
That works!? I was just hesitant to start changing the settings without knowing more about what they affect.? Very nice to have the wavy readouts stable.
Thanks, Barry - N4BUQ
----- Original Message -----
From: "Albert Otten" <aodiversen@...> To: "tekscopes" <[email protected]> Sent: Friday, December 31, 2021 6:53:41 AM Subject: Re: [TekScopes] 7704A - Noise on sine-wave traces Hi Barry,
For minimizing that wavy readout forget about the signal standardizer.? Display your signal (calibrator or whatever) at full height (8 div) and the time base for the most nervous readout. Then rotate R4465 (Vert. Amp.) for the best setting (most quiet readout). Note that this setting is also best for your ordinary signal response. If the result is still unsatisfactory then perhaps a slight readjustment of R4453 might help (but remember the original setting, just in case).
Albert
On Fri, Dec 31, 2021 at 02:13 AM, n4buq wrote:
I see where Table 4-1, Item 3 (Signal Standardizer Calibration Fixture) states that "Calibrated 7000-series plug-in units with suitable signal sources may be substituted if lower performance is acceptable."? I just don't quite see where the instructions outline how to set those up in place of the standardizer - particularly for low-frequency linearity and/or vertical high-frequency compensation.
Thanks, Barry - N4BUQ
----- Original Message -----
From: "n4buq" <n4buq@...> To: "tekscopes" <[email protected]> Sent: Thursday, December 30, 2021 5:11:41 PM Subject: Re: [TekScopes] 7704A - Noise on sine-wave traces I remember now that this involves adjustment of the LF Compensation.? The calibration steps for that calls for a standardizer fixture and I don't have one of those.? Is there a way to make that adjustment without a standardizer fixture?
|
I think what I wrote is confusing. This is probably clearer:
Overall, I think I prefer the GATED setting (draws the readout after the traces); however, at least for the 7704A, there's a switch that selects which HORIZONTAL slot drives the readout so one must choose the A or the B slot and I'm pretty sure I'm going to forget that setting when I have just one time-base in the B position which will disable the readout. Maybe not but for now, I'm going to leave it in the GATED position.
Thanks, Barry - N4BUQ
toggle quoted message
Show quoted text
----- Original Message ----- From: "n4buq" <n4buq@...> To: "tekscopes" <[email protected]> Sent: Friday, December 31, 2021 12:07:42 PM Subject: Re: [TekScopes] 7704A - Noise on sine-wave traces Hi Brian,
I presume you're referring to the vertical positioning control and not the horizontal positioning control but I presume the vertical one. I did not notice that made much, if any, difference. I don't fully understand this but I think the waviness is caused by overshoot/undershoot of the beam as it jumps from the trace to the readout. If that's the case, then I would think that different vertical positions of the trace would affect this.
Overall, I think I prefer the GATED setting (draws the readout after the traces); however, at least for the 7704A, there's a switch that selects which B slot drives the readout so one must choose one or the other and I'm pretty sure I'm going to forget that setting when I have just one time-base in the B position which will disable the readout. Maybe not but for now, I'm going to leave it in the GATED position.
Thanks, Barry - N4BUQ
----- Original Message -----
From: "Brian via groups.io" <brianas1948@...> To: "tekscopes" <[email protected]> Sent: Friday, December 31, 2021 11:26:08 AM Subject: Re: [TekScopes] 7704A - Noise on sine-wave traces Hi Barry , I have been watching these postings with interest . One of my 7844's has this or a remarkably similar problem . There is the wavy / jumping about readout but in my case it is made worse by using the position control -- did this happen with yours . I would be pleased to know how similar my case is and maybe if its not the same there might be suggestions of where the problem might be . many thanks in advance for help from anyone reading this Brian? (UK)
On Friday, 31 December 2021, 17:16:11 GMT, n4buq <n4buq@...> wrote:
Hi Albert,
That works!? I was just hesitant to start changing the settings without knowing more about what they affect.? Very nice to have the wavy readouts stable.
Thanks, Barry - N4BUQ
----- Original Message -----
From: "Albert Otten" <aodiversen@...> To: "tekscopes" <[email protected]> Sent: Friday, December 31, 2021 6:53:41 AM Subject: Re: [TekScopes] 7704A - Noise on sine-wave traces Hi Barry,
For minimizing that wavy readout forget about the signal standardizer.? Display your signal (calibrator or whatever) at full height (8 div) and the time base for the most nervous readout. Then rotate R4465 (Vert. Amp.) for the best setting (most quiet readout). Note that this setting is also best for your ordinary signal response. If the result is still unsatisfactory then perhaps a slight readjustment of R4453 might help (but remember the original setting, just in case).
Albert
On Fri, Dec 31, 2021 at 02:13 AM, n4buq wrote:
I see where Table 4-1, Item 3 (Signal Standardizer Calibration Fixture) states that "Calibrated 7000-series plug-in units with suitable signal sources may be substituted if lower performance is acceptable."? I just don't quite see where the instructions outline how to set those up in place of the standardizer - particularly for low-frequency linearity and/or vertical high-frequency compensation.
Thanks, Barry - N4BUQ
----- Original Message -----
From: "n4buq" <n4buq@...> To: "tekscopes" <[email protected]> Sent: Thursday, December 30, 2021 5:11:41 PM Subject: Re: [TekScopes] 7704A - Noise on sine-wave traces I remember now that this involves adjustment of the LF Compensation.? The calibration steps for that calls for a standardizer fixture and I don't have one of those.? Is there a way to make that adjustment without a standardizer fixture?
|
Hi Barry , thank you for the reply . I had heard / seen it mentioned that it was LF compensation being way off causing this but like you I hesitate to just blindly adjust a pot and see if it makes a difference . I had looked at the circuit and seen that some of the pots one might be adjusting are in series with the dreaded tant capacitors , I wonder if with that effect when using the vert position control it means the tant cap is bad . I'll have to tweak the pot and see then maybe change the tant if none or not enough effect . thank you?Brian (UK)
toggle quoted message
Show quoted text
On Friday, 31 December 2021, 18:07:51 GMT, n4buq <n4buq@...> wrote: Hi Brian, I presume you're referring to the vertical positioning control and not the horizontal positioning control but I presume the vertical one.? I did not notice that made much, if any, difference.? I don't fully understand this but I think the waviness is caused by overshoot/undershoot of the beam as it jumps from the trace to the readout.? If that's the case, then I would think that different vertical positions of the trace would affect this. Overall, I think I prefer the GATED setting (draws the readout after the traces); however, at least for the 7704A, there's a switch that selects which B slot drives the readout so one must choose one or the other and I'm pretty sure I'm going to forget that setting when I have just one time-base in the B position which will disable the readout.? Maybe not but for now, I'm going to leave it in the GATED position. Thanks, Barry - N4BUQ ----- Original Message ----- From: "Brian via groups.io" <brianas1948@...> To: "tekscopes" <[email protected]> Sent: Friday, December 31, 2021 11:26:08 AM Subject: Re: [TekScopes] 7704A - Noise on sine-wave traces Hi Barry , I have been watching these postings with interest . One of my 7844's has this or a remarkably similar problem . There is the wavy / jumping about readout but in my case it is made worse by using the position control -- did this happen with yours . I would be pleased to know how similar my case is and maybe if its not the same there might be suggestions of where the problem might be . many thanks in advance for help from anyone reading this Brian? (UK)
? ? On Friday, 31 December 2021, 17:16:11 GMT, n4buq <n4buq@...> wrote:
Hi Albert,
That works!? I was just hesitant to start changing the settings without knowing more about what they affect.? Very nice to have the wavy readouts stable.
Thanks, Barry - N4BUQ
----- Original Message -----
From: "Albert Otten" <aodiversen@...> To: "tekscopes" <[email protected]> Sent: Friday, December 31, 2021 6:53:41 AM Subject: Re: [TekScopes] 7704A - Noise on sine-wave traces Hi Barry,
For minimizing that wavy readout forget about the signal standardizer.? Display your signal (calibrator or whatever) at full height (8 div) and the time base for the most nervous readout. Then rotate R4465 (Vert. Amp.) for the best setting (most quiet readout). Note that this setting is also best for your ordinary signal response. If the result is still unsatisfactory then perhaps a slight readjustment of R4453 might help (but remember the original setting, just in case).
Albert
On Fri, Dec 31, 2021 at 02:13 AM, n4buq wrote:
I see where Table 4-1, Item 3 (Signal Standardizer Calibration Fixture) states that "Calibrated 7000-series plug-in units with suitable signal sources may be substituted if lower performance is acceptable."? I just don't quite see where the instructions outline how to set those up in place of the standardizer - particularly for low-frequency linearity and/or vertical high-frequency compensation.
Thanks, Barry - N4BUQ
----- Original Message -----
From: "n4buq" <n4buq@...> To: "tekscopes" <[email protected]> Sent: Thursday, December 30, 2021 5:11:41 PM Subject: Re: [TekScopes] 7704A - Noise on sine-wave traces I remember now that this involves adjustment of the LF Compensation.? The calibration steps for that calls for a standardizer fixture and I don't have one of those.? Is there a way to make that adjustment without a standardizer fixture?
?
|
Hi Brian,
The spot has to jump from the current trace vertical position to the readout vertical position. The undershoot or overshoot will be a fixed percentage of the jump size. If you move the trace (waveform) further away from the readout position then in absolute sense the jumps (say between top trace to readout and bottom trace to readout) will increase - and so will their differences - and the wavy behavior will increase. It will also increase when you increase the displayed signal amplitude (with same vertical centering). In your 7844 the trimmers to adjust are R1749 and R1764 (for Beam 1), see calibration step G-6e, page 5-33, "LF COMP 1" and "LF COMP 2". I think you'd best use the same setting for both, though I didn't see such mentioned at first reading.
Albert
toggle quoted message
Show quoted text
On Fri, Dec 31, 2021 at 07:15 PM, Brian wrote: Hi Barry , thank you for the reply . I had heard / seen it mentioned that it was LF compensation being way off causing this but like you I hesitate to just blindly adjust a pot and see if it makes a difference . I had looked at the circuit and seen that some of the pots one might be adjusting are in series with the dreaded tant capacitors , I wonder if with that effect when using the vert position control it means the tant cap is bad . I'll have to tweak the pot and see then maybe change the tant if none or not enough effect . thank you?Brian (UK)
|