Re: Need for Termination Feedthrus
On Sun, Jan 20, 2013 at 11:08 AM, Philip <ndpmcintosh@...> wrote:
Sure but the question remains--why? If I know that the signal amplitude is within range for the task and my cables are short then why would I want to stick an additional 50 ohm feedthru in the signal path?To avoid reflections from the unterminated input of the scope. Try connecting a fast rise generator (10MHz square wave would be fine) to the scope through an unterminated cable, then add a terminator. You'll see a huge difference. |
Re: Anybody know the difference between a 475M and 475?
On Sun, Jan 20, 2013 at 11:00 AM, Steve <ditter2@...> wrote:
TekScopes@..., David DiGiacomo wrote:Well, that's confusing. There are several sources that list theDavid, military OS-261B(V)1/U as a 475M, and then several others that list it as 474/04. (The OS-261C(V)1/U is the 475/04/07.) Do you think there's any electrical difference between the 475M and the 475/04? |
Re: designspark
I have run KiCad under MacPort wine very well on an older Mac with no issues.
toggle quoted message
Show quoted text
On 2013-01-20, at 7:08 AM, willprice94 <will.price94@...> wrote:
I don't know how powerful your mac is, but you could try running a light linux distribution in a virtual machine. |
Re: 2465 - impending U800 failure?
I sold an apparently working U800 to a list member from an old 2445. Though the scope seemed to function just fine on initial testing, he later emailed this back:
toggle quoted message
Show quoted text
"I left FB after doing just a quick test at 1ms/Div. After more testing I discovered that it is very noisy at 10X when time base is faster than 5ns/Div after a quick 2 minute warmup. This Horz Amp is not usable." So I refunded his money. What I take this to mean is that the U800 can be failing or not meeting specs even without total failure. Just something to keep in mind. Chip --- In TekScopes@..., "victor_j_silva" wrote:
|
Ground leads on 6108 Passive Probes
I have a a number of P6108 passive probes without ground leads. Looking in the little plastic sleeve where they go I see the metal shows signs of wear indicating ground leads have been installed in the past.
I have a tektronix ground lead with alligator clip, but, I see no way that the little metal contact can be installed in the plastic sleeve. Are there special clips that only go with certain probes? |
Re: Need for Termination Feedthrus
Sure but the question remains--why? If I know that the signal amplitude is within range for the task and my cables are short then why would I want to stick an additional 50 ohm feedthru in the signal path?
toggle quoted message
Show quoted text
And thanks for the book recommendation. Ant other will be welcome also. --- In TekScopes@..., David wrote:
|
Re: Anybody know the difference between a 475M and 475?
TekScopes@..., David DiGiacomo wrote:
David, I respectrully dissagree. This model is the contract version of the standard 475, with option 4 or option 4&7. It shows the scope operating with the 1106 power supply in an opt 7 configuration, and the front cover is the standard plastic one shipped with all 475s. The military recently retired this scope, and there are many of them showing up on e-pay now. They can be identified by the aluminum plate on the rear which identifies the contract. The 475M is actually marked "475M" on the face plate. The cover is as I mentioned hard deep drawn aluminum, with rubber gasket and retaining hasps to keep it in place. I don't think it was ever offered in option 7, (mine is not, and the two others I have seen are standard.) - Steve |
Re: Possible P6042 part...
--- In TekScopes@..., "Albert" wrote:
The same seller was offering the core and lid as seperate listings. The core and lid are a matched set, and the offest calibration is based on the pairing. You can not replace one without the other, unless you plan to pick your own offset bias resistor. - Steve |
Re: 2465 - impending U800 failure?
Sounds very much like the U800 issue.
About six months ago a posted about baking some U800s that had failed this way and was able to rejuvenate them. I put one of the U800s in a test scope and it's still working to this day. I'm amazed. --Victor --- In TekScopes@..., "cmjones012003" wrote: [SNIP]
|
Earlier posts tektronix 2247a probs
Hi All
????? Being new to the group. I went back and looked at some earlier posts for problems with 2247A's. I too have had this 2247a for several years now. And, have also had a problem with channel A's sweep not displayed. B works perfect. If there are any takers on this I would love to hear if anyone has had a similar problem and found a fix?? I also have several other scopes that need fixin. One at a time.
?
Cheers, Frank Walden
|
Re: RM503 Troubleshooting tips needed for beginner
Hey Ken, ? I probably missed your earlier posts. ?I am on vacation on ?a little itty bitty laptop so can only be so helpful this week... First, a bit of background, because I don't think we've addressed this. ?From my experience with this scope, and other 5xx series Tek scopes I've had, and other
discussions posted in this group, there would seem to be three common failure modes in the RM503/504 scope. ? ?The first is probably the most common. ?The filter capacitors (C611, C612) get dried up and start to fail. ?This is why the first priority should be to get the 500V rail in operation. ?That voltage should be relatively independent of the state of anything else in the circuit. ?If you pull V620 out and you do not have close to 500VDC on the 500V line, we have a problem in the first power supply and its likely those filter caps. ?You only have 260V. ?Is that with the input
switched to European 220V? ?If so, great since at 110V you should then have the full 500V. ? I would concentrate on getting this rail up to par if it is not, though it seems there may be other problems. ? The second most common problem is a breakdown in the coils of the second transformer, T620. ?This one oscillates at several tens of KHz driven by V620 and generates all the other voltages the circuits need. ?The problem is the insulation of the transformer windings break down with time (and possibly long-term absorption of atmospheric water vapor) and the coils short out or open up or short to
ground. ?The circuit relies on the inductances of the coils and their capacitances to oscillate properly, so if any one coil (e.g. the HV coil) shorts out, the rest of the voltages go out too. ? The third failure mode is the HV shorting out through a breakdown in the insulation of ?T601. ?This is because the CRT filament is driven by T601 *raised* to -3KV by the HV circuit of T620. ?We've seen evidence of this failure mode in some scopes sold on e-bay. ?One solution that has been found is to get a cheap 6.3 V filament transformer, hopefully with >3KV isolation, and replace the T601 filament
coils with that. ? ? ? What I would concentrate on is getting the +500V up and running. ?But from what you indicate below, T620 may be no good. ?That;'s harder to deal with. Question 1: ...a tough one. ?Here's where the wisdom of this group will come in very handy. ?Could we replace V692 with a series of silicon diodes? ?Would this work? My first thought would be to run a chain of 1N4007 diodes in series. ?You will need at least 4 of them since they have a peak reverse voltage of 1KV. ?Wired in series, they should be able to rectify the output of T620. ?Two questions for the group: ?will they be low enough capacitance to function in place of V692, and do we need some high-value resistors in parallel
with the diodes to ensure a uniform voltage across each? ? Ken, if I were you, I would leave V692 soldered in place until you can determine if you have 3KV ac on the ouput of the T620 HV coil. ?Those HV rectifier tubes are very delicate and the leads can break off very easily. ? Question 2: Probably not. ? Get the RCA handbook mentioned below. ?Gas regulators are basically fancy neon lamps. ?They maintain a constant voltage within a certain current range, but they have a startup and a dropout voltage. ?The tube and its series dropping resistor must have at least the startup voltage to fire, and at least the dropout voltage to operate. ?When operating, there is a current through the series resistor ?(R659, 100K). ?In this case and the voltage supplied to that resistor must be at least 85v plus I*100K where I is the operating current of the tube (see handbook). ?These are generally pretty robust tubes - if it is glowing it is probably working. ?It should have a purplish-orange glow in operation. ?Check R659 to
make sure it's still 100K. ? If it gets warm, check it warm to make sure it is not opening up or becoming much higher resistance as it warms up. ?If it's glowing and it's at 97V it's not good. ?But I still suspect that the 85V rail will fall in to line when the +250 works. ?You could try temporarily jumpin another 100K resistor in parallel with R659. ?That will pull V659 up a little harder, and give it some more voltage. ?It should then glow and run at 85V. ? Question 4: Oscillator: ?V620. ?It oscillates at ?~30KHz or so (could be different - I'm going from vague memory of the measurement on my working scope) to generate the AC which drives the high frequency transformer. ?The tube is a beam power pentode, so it dissipates a lot of power and should get hot. ?What you describe sounds normal and is the filament glow. ?What I was seeing in a defective circuit was the outside envelope of the guts of the tube, normally grey, get glowing hot. ?I am not sure if this element is the plate (anode) of the tube or an external shield, and it might glow a little in normal operation, but not much. ?The glowing spot starts out as an oval and grows to encompass the plate and the envelope gets very hot as well. ?Eventaully it will weaken the metal structures, possibly warp them, and drive out gasses that ruin the vacuum and make the problem
worse. ?Just be aware of the potential for this problem and keep an eye on that tube while you are troubleshooting. Question 5: By the way, get a copy of the RCA Vacuum tube handbook from the mid 1950's. ?It's fun reading and very helpful as well. ?I did a brief check and it looks like you might even be able to read it on google books. ?Also I would not be surprised if Archive.org has a copy. ?But it's better in my opinion to have a paper
copy in hand. The encouraging news, at least to me, is that the secondary voltages of that transformer are all in the correct proportion if low. ?Note the +250 is at +160, which is 2.5x the +100, which is at +60. ?I think the bad news is you have an open in the HV circuit. ?4.7 MOhm to ground sounds like too much to me for the HV coil of T620. ?? I hope this measurement is wrong because T620 is "Unobtanium" without a donor scope, or re-winding the thing (an ordeal). ?After verifying the +500V, I
would try to verify voltage on the HV side of this transformer, maybe by substituting some diodes for V692. Ok, being dragged to the beach now. ?Good luck and keep me posted. ?Also: ?danube@... if I don't respond here. ?Sometimes I lose messages in the hundreds of posts. ? Dan From: whippletreefarms To: TekScopes@... Sent: Thursday, January 17, 2013 1:56 AM Subject: [TekScopes] Re: RM503 Troubleshooting tips needed for beginner
?
So the V620 and V634 tested OK. My buddy wasn't able to test the V659, but it shouldn't be the main problem at this point anyway.
My plan is to next replace the V692 diode tube. I still would like to know about my questions 1, 2, 4, and 5 below before I search for the diode tube though. Thanks, Ken --- In TekScopes@..., "whippletreefarms" wrote: > > > > Thanks for all the input everyone. I got another DMM and an HV probe. My old meter did indeed start giving erroneous results. So here is where we stand now: > > Normal situation (nothing removed and no wires lifted, measured with new meter) > -3kV rail = 0V > +500V rail = +260V to +270V > +250V rail = +150.4V > +85V rail = +97.8V > +100V rail = +60.1V > +12.6V rail = +4.97V > -100V rail = -57.2V > > I know the HV probe is working, because I measured the +500V rail with it both before and after measuring the -3kV, and it corresponded to the direct measurement. > > I'm guessing that my old scope somehow shorted the +500V to ground and that was the spark I initially saw. It hasn't repeated since, so this must have been a red herring. > > I removed the -3KV rail from the end of the diode tube, and the voltage at the end of the tube still measured 0V. > > Per Albert's comments: > * When removing both the V620 and V634, the +500V rail increases to +476V. > * With R628/C628 shorted to ground, the +500V rail reads +315V. > * The voltage decays to about +120V about 20 seconds after power off. > * The V659 is a 87V voltage regulator, so I would be surprised that it hasn't ignited by +97.8V. > > Per David's comment: > * C611/C612 probably isn't an issue now that I am getting consistent (although incorrect) +500V readings. > * On a related note, however, they held voltage a LONG time after shutoff, so long in fact, that I had to actually remove the white/red T620 wire to measure the resistance! > > Per Dan's comments: > * The V620 tube has an orange glow on a vertical element. I'm not sure if this is the same as the plate glow you mentioned. It does get hot - too hot to touch. > * See comments above on removing V620, and also measuring decay time > * -3kV to ground measures 4.7 MOhms to ground > * I removed the CRT filament (white/yellow wire, pin 14) from everything, and measured that it was open to ground. > * I measured the other CRT lines to ground, and they all seemed reasonable based on the circuit diagram. > > I suppose it is possible I added to my woes by bunging up the V692 lead when lifting wires and thus causing 0V on the -3kV rail. Or maybe it has always been 0V, but since I couldn't measure before, I don't know. > > So my new round of questions: > 1. Is there a way to measure the input to V692? The power-off T620 measurements all seemed correct. > 2. Does the +97V on the +85V rail indicate V659 is bad? > 3. I plan to take V620, V634 and V659 to the hardware tech at work. He says he has two tube testers in his attic and can measure most tubes. Thank goodness for his "things were so much better in the old days" curmudgeonly attitude! > 4. (tube circuit newbie question) A couple of you mentioned the "oscillator". Which tube is that? The V620 or the V634? And what frequency does it oscillate at? > 5. What to try next? I will say I am learning an awful lot, and am not ready to give up yet! > > Thanks again, > Ken > > |
Re: Anybody know the difference between a 475M and 475?
The 475M was developed for specific contracts, and never appeared in the Tek catalog. So I suspect the only way to get the specs is from the manual. There was a dedicated manual printed for this scope. They seam to be very rare. Some of the vendors of scanned manuals on CDROM offer it.The 475M service manual is Army TM 11-6625-2735-14-1. You can download it here: |
Re: 2465 - impending U800 failure?
Tom Jobe
¿ªÔÆÌåÓýHi David,
The 2224 conversion to a 2232 has come up before on this
group.
Denis Cobley provided lots of information about this issue
years ago, and Jack Rouse discovered that it came down to
the?presence of jumper W3400 on the storage board to turn?a 2224 into
a 2232.
The 2224 has W3400 removed, the 2232 has W3400 installed (the
jumper is just a piece of wire).
It was believed that the firmware was the same for the 2221A,
2224 and 2232, and that the firmware chip determined which scope?it was
installed in by the presence/absence of?some?jumpers. Someone studied
the ROM image and said that this all seemed to be true.
Also, you needed to cut the thin plastic face overlay to
expose the additional switches?across the top, just to the right of the
CRT.
A pointed hobby knife and some extra square buttons from
another 222x was all it took.
I did one or two of these 2224 to 2232 conversions using this
information and?everything seemed to work perfectly.
Jumper W3401 is on the A10 board and is shown at location 3A
on the 2232 schematic number <15> titled "Digital Acquisition and
Memory".
Physically W3401 is located at 2F on the A10 board
layout.
You might?consider checking that your 2224/2232 has W3401
installed before you do your bandwidth testing.
tom jobe...
PS Every time I go through the Tekscopes message archive, I
marvel at all of the fine posts put up by Denis Cobley. There must be a thousand
of them and they are all informative. I sure wish Denis was back on
Tekscopes.
?
?
?
?
|
Re: designspark
I don't know how powerful your mac is, but you could try running a light linux distribution in a virtual machine.
toggle quoted message
Show quoted text
For example: are both very lightweight, I'm sure you'd have no trouble running them in virtualbox (). I personally use Arch Linux and have been using the gEDA suite. I initially tried eagle (too restrictive) and KiCAD (had a few irritating UI bugs, too annoying for me to use). Both CrunchBang and lubuntu are based on Ubuntu/Debian and so have a rich software repository, of which I'm certain KiCAD is present, if not I'm sure there is a ppa repository that can be added to provide packages. --- In TekScopes@..., "scott_dixon" wrote:
|
Re: Possible P6042 part...
Albert
Why would someone remove the coil/core section from "a working probe" ?
toggle quoted message
Show quoted text
Albert --- In TekScopes@..., "Alex" wrote:
|
495P For Sale
I bought mine used from a local reconditioner several years ago and it has had several years of light use since then.
It has three flaws (other than these, it appears to work just fine and I have enjoyed using it): 1. Just recently, the green shift key failed to respond. While I believe it should be simple to fix and someone else can probably do the work quite easily and cost effectively, it is beyond my skill level to repair the problem. 2. There is a small crack in the RF Atten knob. 3. The surface of the screen has some discoloration. This is slightly visible in pictures, but it is not apparent when using it unless you look carefully for it. Otherwise the display is bright and clear, and the control knobs are in great shape. It comes with the operating manual on CD. I also have the companion TR503 Tracking Generator for sale. I have pictures I can share. I would prefer to sell via the group. Please contact me off line. Gordon, KA2NLM |
Re: 2465 - impending U800 failure?
I have a working 2232 that I suspect may have originally been a 2224
which is the uncommon 60 MHz version. Besides some cosmetic discrepancies, the fastest analog sweep speeds have an awful lot of horizontal displacement although that may be a calibration issue. Once my other 2232 is up and running, I will do a careful comparison between them. I do not have any data on the 2224 to use but it may be possible to tell by checking the vertical response temperature compensation networks which are different between the 100 MHz and 60 MHz oscilloscopes which use the same hardware. On Sun, 20 Jan 2013 06:21:59 -0000, "dehav7" <yachadm@...> wrote: 2465b helpful? My a*s! |
Re: Anybody know the difference between a 475M and 475?
--- In TekScopes@..., Matt Jodz wrote:
I can only partially answer this question and I even own one. I have never had the need to open it up as it is the only 475 scope I have ever obtained that is fully functional. I suspect it is very close to a standard 475. Unlike the 465M, which is a totally different scope from the base 465, my 475M looks almost identical to a stock 475. It still is fan cooled with the foam filter in the rear. It does have a metal front cover, similar to a 453 with a gasket and area to hold two probes. The cover clamps to the sides of the scope housing with aluminum hasps. The 475M was developed for specific contracts, and never appeared in the Tek catalog. So I suspect the only way to get the specs is from the manual. There was a dedicated manual printed for this scope. They seam to be very rare. Some of the vendors of scanned manuals on CDROM offer it. As it is a military model, it probably has some enhanced environmental specifications. Howver, it may not be a large step up from the stock 475. The 400 series were considered "portable" scopes, and were designed ot and tested to meet MIL-T2800C Class 3 environmental and dynamics specifications. These are considerably tighter than the class 5 that other Tek instruments were designed to. (Modern instruments don't even meet class 5!) If it works, it is probably a great scope. Troubleshooting it may be difficult due to the rareness of the manual. - Steve |