Re: 465B Horizontal Trace Issues - Help requested
Tom,
I have the correct manual and measured DC voltages with 115.0 VAC line input through my Variac. Here are the DC voltages I measured: TP 4338 +55V, tolerance +54.62 to +55.39 Reading: +56.94 TP 4339 +15V, tolerance +14.75 to -15.26 Reading: +15.49 TP 4337 +5V, tolerance +4.92 to +5.09 Reading: +5.172 TP 4439 -8V, tolerance -7.86 to -8.14 Reading: -8.27 TP 4340 +110V tolerance +106.7 to +113.3 Reading: +113.5 It looks like all the readings are just outside of the acceptable range. I took the readings several times over a few minute span, and they were the same. I'm not sure how to measure ripple without another scope. Is there a way with a DMM (Fluke 117)? I'm assuming I took the measurements in the right place. I was expecting the TPs to be marked with the TP# (e.g. TP4338), but they were only marked with the voltage. These were all together on the main board on the bottom of the unit. I also shot some Deoxit into the position switch to see if that would help, but it did not seem to make a difference. What next? |
Re: 7623A Storage problem
I remember that is mounted in the HV metallic shield/box
toggle quoted message
Show quoted text
-----Mensaje original----- De: [email protected] [mailto:[email protected]] En nombre de Colin Herbert via groups.io Enviado el: viernes, 25 de septiembre de 2020 18:47 Para: [email protected] Asunto: Re: [TekScopes] 7623A Storage problem I have found that F814 (2A) on the LV power supply was blown and continues to do so, possibly because of a failure of Q1195 or C1198 (47uF). My question is how can I get to either of these components? I can see from the Manual that C1198 is on the HV board and that I might have to get in there to look at it. But Q1195 isn't on the HV board because it is heatsinked, but where is it located? I have a transistor checker and can thus check it out, if I can locate it. Colin. -----Original Message-----
From: [email protected] [mailto:[email protected]] On Behalf Of Roger Evans via groups.io Sent: 23 September 2020 14:43 To: [email protected] Subject: Re: [TekScopes] 7623A Storage problem Colin, I had the afterthought about low voltages since the regulation of the power transistor Q1195 relies on +5V, +15V and +130V being OK. The components I mentioned are all on the low voltage side of the transformer T1199 and can be checked with the power off. Q1195 is attached by a plug and socket so it can effectively be tested out of circuit, Q1190 can be checked as a pair of diodes for open circuit / short circuit. The two capacitors are likely to be short circuit if they have failed and L1198 will be open circuit if the fuse didn't do its job in time. At least you don't have to wait 30 minutes for the high voltage to leak away! Regards, Roger |
Re: Tektronix 475A B Delayed only triggering with A Trigger Slope
Thanks Bob. I did not see any problems in that circuit. Now I admit that I am still learning about how these circuits work and often I am just checking voltages, looking at signals and testing components to find problems. Im not a engineer. I also may not have described the operation of the Tek 475A incorrectly.
What I found was the second tunnel diode on the B trigger, was also bad. Once replaced I was able to trigger with the B Slope knob. However it does not light the triggered bulb, only A slope seems to do this, this may be normal operation. Hopefully I can get replacement tunnel diodes in the future. I dont think I have many used spare ones left. |
TDS5/6/700 application software packages?
Hey y'all,
I've always been curious to see the applications that were available for my TDS784D back in the day. Have any of you even seen these applications, or know where I might find them? In the datasheet on Tek's website, the applications are described thusly: "Application Software Packages These Java based application packages reduce the cost, time and complexity common to many application-specific test procedures. These application-specific capabilities are easily installed via floppy disks and require a hard disk drive (Option HD or Option 2M). TDSPWR1 gives designers interested in power consumption applications the ability to automatically calculate True Power, Apparent Power, Power Factor, Instantaneous Power and Energy to eliminate manual calculations. TDSJIT1V2 performs a suite of jitter measurements with unprecedented accuracy on single-shot waveforms. Jitter information can be displayed as statistics, histograms, time trends or jitter spectrums. This software makes timing measurements by providing statistics on specific, fully time-correlated data and clock edges. It also performs automated AC timing measurements on all parameters in single-shot acquisitions. TDSCEM1 performs a suite of automated eye diagram measurements on SONET/SDH and Fibre Channel standards. These measurements include eye height, eye width and quality factor. TDSCPM1 performs automated parametric measurements of pulse amplitude, spectral power and pulse symmetry to standards requirements. Standards include ANSI T1.102 and ITU-T G.703." Siggi |
Re: Does anyone sell TDS460A nvram chip with options activated
Hey Roger,
out of curiosity, what did you find? All I found was mentions of options software and this eBay listing (). I have a TDS874D for which the procedure is well understood, and I amused myself at one point by adding the "Communication Mask Testing" option to mine. It's not of any use to me, but hey, why not :). Siggi On Fri, Sep 25, 2020 at 11:47 AM Roger Evans via groups.io <very_fuzzy_logic@...> wrote: I had a very quick look around and it seems that upgrades on the TDS460A |
Re: 7623A Storage problem
I have found that F814 (2A) on the LV power supply was blown and continues to do so, possibly because of a failure of Q1195 or C1198 (47uF). My question is how can I get to either of these components? I can see from the Manual that C1198 is on the HV board and that I might have to get in there to look at it. But Q1195 isn't on the HV board because it is heatsinked, but where is it located? I have a transistor checker and can thus check it out, if I can locate it.
toggle quoted message
Show quoted text
Colin. -----Original Message-----
From: [email protected] [mailto:[email protected]] On Behalf Of Roger Evans via groups.io Sent: 23 September 2020 14:43 To: [email protected] Subject: Re: [TekScopes] 7623A Storage problem Colin, I had the afterthought about low voltages since the regulation of the power transistor Q1195 relies on +5V, +15V and +130V being OK. The components I mentioned are all on the low voltage side of the transformer T1199 and can be checked with the power off. Q1195 is attached by a plug and socket so it can effectively be tested out of circuit, Q1190 can be checked as a pair of diodes for open circuit / short circuit. The two capacitors are likely to be short circuit if they have failed and L1198 will be open circuit if the fuse didn't do its job in time. At least you don't have to wait 30 minutes for the high voltage to leak away! Regards, Roger |
Re: Does anyone sell TDS460A nvram chip with options activated
Which GPIB-USB adapter do you have and what operating system?
toggle quoted message
Show quoted text
Vince - K8ZW. On 09/25/2020 08:34 AM, Paul via groups.io wrote:
Hi Roger, --
K8ZW |
TDS784D with faint display and blurry image
Dear all,
I have bought a TDS784D at Ebay some time ago and had no time to test it. When I finally had time to turn it on, I realized that the display is very faint and the image is out of focus. The intensity adjustment available on the menu is already at 100%. I found a very brief adjustment instruction in the service manual and I fumbled on the pots ( after taking a long time to find them) but they seem to be arranged in a different order than the one described on the manual. There is a list of the appropriate voltages to the display module on the display troubleshooting section and all of them are ok. I found one person on Ebay selling a new display module and he mentions that there is a time consuming adjustment procedure for this module, but I could not find it anywhere. I have uploaded one photo of the menu screen to show how out of focus the display is (the file is TDS784D dim display). My questions are: 1 - Where can I find a detailed adjustment procedure for the display module? 2 - Is there a way to check if the problem is the CRT being "exausted"? 3 - There is a company offering the NewScope-T1 Color LCD replacement Kit as a replacement. is this a good alternate in case I need to replace the CRT? |
Re: Front cover for 2225 ?
On Fri, Sep 25, 2020 at 05:31 PM, <tekscopegroup@...> wrote:
P/N 200-3232-00 Raymond |
Re: Front cover for 2225 ?
On Thu, Sep 24, 2020 at 08:51 AM, Raymond Domp Frank wrote:
I also have a 2247A and I can can confirm that front covers are completely different from other 22xx series ones, as would be the one for the 2252 which uses the same as the 2247A. No stamped PN anywhere on that one that I could find. |
Re: Does anyone sell TDS460A nvram chip with options activated
Hi Roger,
actually I have an USB based GPIB adapter, but without controlling Software to read/write I won't be able to do much. The only I GPIB SW I use with this is Pete Millet's package for the HP 8903 distortion analyzer. Embarking on a big DIY effort to figure out the commands etc.. not really in my skill set. That's why I was looking for "spare parts" approach of swapping chips. Thanks Paba |
Re: Tek 7854 intermittent issues
OK here's some thoughts:
toggle quoted message
Show quoted text
1) The -5V supply measured at [2] at the input to L68 should have about 100mV peak to peak saw-tooth ripple at about 50kHz which is before the output filtering network (see the voltage and waveform conditions page just prior to the schematic). So the ripple at TP92 should be LESS than 100mV. Were you measuring its output at TP92 or on one of the logic boards (if the latter you may be picking up digital hash). 2) I'd take a good look at the electrolytic caps on the auxiliary regulator board, especially C68, C84, C86 and C92. There are a couple more at the input but I don't think they will be the cause of problems. Also look for any suspect caps on the -5V rail on the logic boards HtH David -----Original Message-----
From: [email protected] [mailto:[email protected]] On Behalf Of Nick Corvid Sent: 24 September 2020 22:47 To: [email protected] Subject: Re: [TekScopes] Tek 7854 intermittent issues Any ideas what could be causing all that ripple on the -5 line? |
Re: Tek 571 curve tracer
An update:
My initial indication of a problem was glitches in the display, such as "_" characters showing up on parts of the screen where they didn't belong. These glitches were stored in video RAM and were thus visible until the screen image in RAM was updated. My guess for the most likely culprit was power supply noise, and indeed there is a LOT of noise on the supplies, but it is not line noise. In fact, it is nearly all switching noise from the digital circuitry. If I hold the microcontrollers in reset, the supplies get very quiet. When the microcontrollers are running, there is significant noise on all the supplies: -15V rail: 1.2Vpp +5V rail: 1.1Vpp +12V rail: 0.5Vpp +15V rail: 0.7Vpp The noise on each of the supplies is all the same noise (triggering on one channel produces a solid lock on the others). It is asynchronous to the 14.xxx MHz pixel clock of the NS405, so I believe it is all from the 8032 (12 MHz XTAL). The lowest observable frequency in the noise is around 2 MHz (only looking in the time domain at the moment). I've isolated this down to just the main board; the CRT, front panel, and collector supply are out of the picture. The main board contains the 8032, NS405, base/gate drive circuit (DAC), and emitter test circuit (ADC). Supplies are all linear, using 78xx and 79xx regulators. At this high of frequency, electrolytics aren't contributing much to the bypass, right? Each rail only has 220uF after the regulator. I've checked half of them, and while they've aged they aren't trash yet. Anybody seen a ceramic bypass cap go bad? Dave Casey On Thu, Sep 24, 2020 at 1:11 AM Dave Casey via groups.io <polara413= [email protected]> wrote: I'm not so sure they are. DC leakage current is well below 100uA at 20V. |
Does anyone sell TDS460A nvram chip with options activated
Hello
I figure i can pop out my nvram chip and replace with a another that has the FFT option enabled. Anyone sell these. If you have such chip in yours would you consider copying it and selling it to me. The reason I ask it no Dos machines or nvram programmers or isa-gpib around here. Thanks Paba |
Re: 453 potentiometer/switch
On Fri, Sep 25, 2020 at 02:08 AM, Richard P wrote:
Remove the aluminum cover sideplate, mounted with two nuts accepting Allen screws (Astig and Trace Rotation) and the variable time/div knob. Underneath, you'll find the nut holding the pot. It's mounted on a recessed part in the extrusion. Raymond |