Get a service manual. You need to get into the diagnostic mode.
Press and hold these three buttons, delta V, delta T and Trigger slope and release them together.?
Press trigger mode button to select test.
To see hrs, select Exer 05.?
Press and release Upper trigger coupling to initiate Exer 05.?

Note that these two numbers can be changed easily.?
cslim


Re: Cleaning dirty pots on 5xx series plug-ins

Albert
 

Archives is just short for "old messages".
Albert

OK. Sounds interesting. I see links to "Files", "Photos", "Database", and others but no "archives." Would someone please point me to the forum archives.

Thanks,
Tom


First post - 2465A on the way

 

Hi everyone. I'm new here but have some experience (as a user) with Tek scopes and other measurement gear. I am expecting my own 2465A to arrive early next week. Bought it from seller tharitis - seems to have a good record on eBay. Serial number is B014481, which seems like it makes this an early unit. Based on the serial number and what I've read here I think this is not in the range where the U800 problem has been seen. But the serial number is so low that I suspect it has probably seen a lot of use.

I've also spent some time finding many of the messages dealing with replacement of caps etc, which I can do if / when needed, but I wanted to find out from the experienced folks here what I can do on arrival of the scope to wring it out for any obvious problems. I have P6131 and P6133 probes but no gear to generate calibrated frequencies or voltages. I do have power supplies but they are not cal'd, and my ability to generate frequencies is limited to the audio band.

I was also wondering if there was a reliable way to check actual hours on the CRT.

Thanks to all in advance. Hoping I get something I can work with.


Re: Could please any one give a clue about this nice pulsed line generator??

Craig Sawyers
 

BNC have poor performance above 1.8 to 2GHz so they may slow the
risetime of the pulse.
Well yes - SMA is preferable. But Amphenol say:

"Amphenol 50 ¦¸ BNC connectors are miniature, lightweight units designed to
operate up to 11 GHz and typically yield low reflection through 4 GHz."

They state that the VSWR is a maximum of 1.3 up to 4GHz. GR874 is excellent
(at least when new!), with 1.03 maximum to 5GHz with solid air line. N-type
(standard in EMC test labs) is 1.3 maximum to 11GHz.

SMA (for hardline) is 1.05 + 0.005f (f in GHz) and goes to 18GHz. So 1.14
at 18GHz. Impressive; but terminating hard line to an SMA is not easy. It
is nothing like as good with flexible lines (1.62 at 18GHz with RG178)

Craig


Re: Could please any one give a clue about this nice pulsed line generator??

Craig Sawyers
 

Hi Albert..probably I put some meters of RG147 as fixed(solder) internal
line
in order to obtain a 100nS pulse width.
What is the secret about this GR connectors ?, I have the temptation to
replace by a BNC, any wrong with this change?.
Gabriel.
My version of this generator has a fixed solid line coiled up inside. There
is also a front panel SMA to attach more line if needed.

Nothing special about GR connectors. Essential if you have a lot of GR
gear; they were really GR's solution to a 5GHz connector, introduced in
1943. Then later they introduced the GR900 series that went higher in
frequency. Both "hermaphrodite" in that there was not a male and female
connector - everything was identical. There were lots of in-house
connectors designed - anyone remember the Rhode&Schwarz Dezifix? All
superceded by the N-type for this frequency range. But really, a high
quality, branded BNC would be fine (all BNC's are not created equal!)

Craig


Re: Cleaning dirty pots on 5xx series plug-ins

 

--- In TekScopes@..., tubesnthings@... wrote:

Dan;
There's a bunch of material on this subject in the forum's archives.
More was added, recently. Check it out.
Bernd
OK. Sounds interesting. I see links to "Files", "Photos", "Database", and others but no "archives." Would someone please point me to the forum archives.
Thanks,
Tom


Re: WTB: Readout IC (155-0021-01), or entire readout board

Rob
 

¿ªÔÆÌåÓý

Yep Good Luck. It wasn¡¯t too long ago that a list of vendors was delineated and discussed here. Wish I could remember. You may want to run a limited search for say the last 9-12 months. Anyway, Talon is worth remembering as well as a Greek company and of course classic Tek page¡­. and ¡­..and¡­

?

Well you get the point. I hope this finds you well.

Rob

?

From: TekScopes@... [mailto:TekScopes@...] On Behalf Of chrismh_somenumber
Sent: Thursday, February 28, 2013 10:55 PM
To: TekScopes@...
Subject: [TekScopes] Re: WTB: Readout IC (155-0021-01), or entire readout board

?

?

Thanks, I didn't know about that site. I checked Sphere and Tucker.

I ordered one from SMC. Hopefully he has it.

Chris

--- In TekScopes@..., "Rob" wrote:
>
> Looks like the chip can be obtained here:
>
>
>
>
>
>
>
> I did not search further at the other "normal parts vendors" (Sphere, Talon,
> etc.)... No particular association with SMC in a monetary way. I am however
> a satisfied customer and I do chat with the owner from time to time.
>
>
>
> Anyway, not 100% sure he has 155-0021-01 left nor how his price compares to
> others out there. Personally I have done enough business with SMC that if he
> has it I stop my search and purchase just based on my time being valuable
> and results of enough price shopping I am satisfied.
>
>
>
> Anyway, hopefully helpful. Sorry for rambling on so.
>
> Rob
>
>
>
> From: TekScopes@... [mailto:TekScopes@...] On Behalf
> Of chrismh_somenumber
> Sent: Thursday, February 28, 2013 8:35 PM
> To: TekScopes@...
> Subject: [TekScopes] Re: WTB: Readout IC (155-0021-01), or entire readout
> board
>
>
>
>
>
> The readout board part # is 670-1900-XX. Looks like it was used in lots of
> scopes, not just those I listed.
>
> Thanks
>
> --- In TekScopes@... ,
> "chrismh_somenumber" wrote:
> >
> > By swapping in the above IC from another scope, I've determined that the
> custom IC on my 7844 has died.
> >
> > Would anyone happen to have one they would part with, or have an entire
> working readout board from a 7844/7904A?
> >
> > Thanks,
> > Chris
> >
>


Re: WTB: Readout IC (155-0021-01), or entire readout board

chrismh_somenumber
 

Thanks, I didn't know about that site. I checked Sphere and Tucker.

I ordered one from SMC. Hopefully he has it.

Chris

--- In TekScopes@..., "Rob" <rgwood@...> wrote:

Looks like the chip can be obtained here:







I did not search further at the other "normal parts vendors" (Sphere, Talon,
etc.)... No particular association with SMC in a monetary way. I am however
a satisfied customer and I do chat with the owner from time to time.



Anyway, not 100% sure he has 155-0021-01 left nor how his price compares to
others out there. Personally I have done enough business with SMC that if he
has it I stop my search and purchase just based on my time being valuable
and results of enough price shopping I am satisfied.



Anyway, hopefully helpful. Sorry for rambling on so.

Rob



From: TekScopes@... [mailto:TekScopes@...] On Behalf
Of chrismh_somenumber
Sent: Thursday, February 28, 2013 8:35 PM
To: TekScopes@...
Subject: [TekScopes] Re: WTB: Readout IC (155-0021-01), or entire readout
board





The readout board part # is 670-1900-XX. Looks like it was used in lots of
scopes, not just those I listed.

Thanks

--- In TekScopes@... <mailto:TekScopes%40yahoogroups.com> ,
"chrismh_somenumber" wrote:

By swapping in the above IC from another scope, I've determined that the
custom IC on my 7844 has died.

Would anyone happen to have one they would part with, or have an entire
working readout board from a 7844/7904A?

Thanks,
Chris


Re: WTB: Readout IC (155-0021-01), or entire readout board

Rob
 

¿ªÔÆÌåÓý

Looks like the chip can be obtained here:

?

?

I did not search further at the other ¡°normal parts vendors¡± (Sphere, Talon, etc.)... ?No particular association with SMC in a monetary way. I am however a satisfied customer and I do chat with the owner from time to time. ?

?

Anyway, not 100% sure he has 155-0021-01 left nor how his price compares to others out there. Personally I have done enough business with SMC that if he has it I stop my search and purchase just based on my time being valuable and results of enough price shopping I am satisfied.

?

Anyway, hopefully helpful. Sorry for rambling on so.

Rob ?

?

From: TekScopes@... [mailto:TekScopes@...] On Behalf Of chrismh_somenumber
Sent: Thursday, February 28, 2013 8:35 PM
To: TekScopes@...
Subject: [TekScopes] Re: WTB: Readout IC (155-0021-01), or entire readout board

?

?

The readout board part # is 670-1900-XX. Looks like it was used in lots of scopes, not just those I listed.

Thanks

--- In TekScopes@..., "chrismh_somenumber" wrote:
>
> By swapping in the above IC from another scope, I've determined that the custom IC on my 7844 has died.
>
> Would anyone happen to have one they would part with, or have an entire working readout board from a 7844/7904A?
>
> Thanks,
> Chris
>


Re: WTB: Readout IC (155-0021-01), or entire readout board

chrismh_somenumber
 

The readout board part # is 670-1900-XX. Looks like it was used in lots of scopes, not just those I listed.

Thanks

--- In TekScopes@..., "chrismh_somenumber" <hogancm@...> wrote:

By swapping in the above IC from another scope, I've determined that the custom IC on my 7844 has died.

Would anyone happen to have one they would part with, or have an entire working readout board from a 7844/7904A?

Thanks,
Chris


WTB: Readout IC (155-0021-01), or entire readout board

chrismh_somenumber
 

By swapping in the above IC from another scope, I've determined that the custom IC on my 7844 has died.

Would anyone happen to have one they would part with, or have an entire working readout board from a 7844/7904A?

Thanks,
Chris


Re: Could please any one give a clue about this nice pulsed line generator??

 

BNC have poor performance above 1.8 to 2GHz so they may slow the risetime of
the pulse. SMA is a much better choice, and this is what Tek used in the
later versions of the HAPG. But at the time it was originally built SMA was
probably not available. GR874 is capable of >8GHz. Soldering the RG-147
directly to the PC board is OK, but long lengths of coax make poor delay
lines (charge lines). They suffer from "dribble up" effects. The HAPGs use
special copper hard line for their internal delay lines. If the RG-147 is
short (a few feet) it "dribble up" won't matter.

Read the Theory of Operation in the manual to learn how a length of charge
line works. The pulse width is going to be twice as long as the length of
the line since the pulse has to travel out to the end and return to the
beginning before the avalanche, which is using the charge stored in the
line, can turn off. If you know the velocity of propagation of the
electrical signal in the particular 50 Ohm coax you are using (it must be 50
ohm) and you multiply the length times the velocity of propagation and then
double the result that should give you an idea of how wide the pulse will
be.

Dennis

-----Original Message-----
From: iglesia_cristiana_arpas_eternas, Sent: Thursday, February 28, 2013
3:22 PM

Hi Albert..probably I put some meters of RG147 as fixed(solder) internal
line in order to obtain a 100nS pulse width.
What is the secret about this GR connectors ?, I have the temptation to
replace by a BNC, any wrong with this change?.
Gabriel.
--- In TekScopes@..., "Albert" <aodiversen@...> wrote:

Correction: in my early XF unit the charge line is soldered to the board.
Probably a peltola jack has been removed there. In my later XH unit the
board and the line have SMA connectors.
Albert

--- The cable has a peltola connector at one end. I have another unit in
which the cable is soldered to the board.
---
Albert


Re: Could please any one give a clue about this nice pulsed line generator??

 

Hi Albert..probably I put some meters of RG147 as fixed(solder) internal line in order to obtain a 100nS pulse width.
What is the secret about this GR connectors ?, I have the temptation to replace by a BNC, any wrong with this change?.
Gabriel.

--- In TekScopes@..., "Albert" <aodiversen@...> wrote:

Correction: in my early XF unit the charge line is soldered to the board. Probably a peltola jack has been removed there. In my later XH unit the board and the line have SMA connectors.
Albert


--- The cable has a peltola connector at one end. I have another unit in which the cable is soldered to the board.
---
Albert


Re: Could please any one give a clue about this nice pulsed line generator??

 

I will be surprised if there is a manual since it does not even have a part
number.
Sometimes you have to settle for the next best thing.
Dennis

-----Original Message-----
From: TekScopes@... [mailto:TekScopes@...] On Behalf
Of iglesia_cristiana_arpas_eternas
Sent: Thursday, February 28, 2013 3:18 PM
To: TekScopes@...
Subject: [TekScopes] Re: Could please any one give a clue about this nice
pulsed line generator??


Very useful info indeed..thanks a lot Dennis.
I'm not feel "safe men" until obtain the right info about..I just like to
have "all" the service manual of my gear.
I was send a mail to Artekmanual , but not response yet.
Regards Gabriel.
--- In TekScopes@..., "Dennis Tillman" <dennis@...> wrote:

I think you are being overly concerned about your purchase not working.
These are rugged instruments and not likely to fail. The PV509 manual
is close enough to tell you everything you will need to know about
your purchase. But here is some additional information on the HAPG
that I kept from years ago.



From: "Craig Sawyers" <c.sawyers@>

Date: Tue May 4, 2004 2:25 pm

Subject: Wierd Pulse generator craig_sawyers



Hi Group



Well I took delivery of the strange "High Amplitude Pulse Generator".
This

is a TM500 style plug-in, and I bought it kind of out of curiosity,
since I

had seen nothing like it. It also has no identifying type number at
all -

so it was either a development unit, or an internal use only test
fixture -

hopefully one of the ex-Tek guys on the list will recognize it - I've
put

some pics on the yahoo gallery.



This is the only thing I've seen that uses the strange attenuator that
the

7A29 uses. This has a huge bandwidth - way above 1GHZ, and probably
*way*

above so that it doesn't cramp the 7A29's style. I suspect that the
7A29

has a 1.5GHz bandwidth, so the attenuator is possible approaching 5GHz.



It also has a rigid charge line inside, estimated at around 3m (10') long.

This line isn't secured at all, and is prevented from shorting to the
back

of the PCB with a layer of self-adhesive plastic film - very
prototype-ish.

Date on the PCB is copyright 1978, and component date codes indicate
late

1982 build.



The unit is basically a pre-trigger pulse generator that uses an
avalanche

transistor to give a fast rising pulse. The pre-trigger can be
adjusted with a front

panel control from 50-125ns early, so that a sampling time base gets
time to

crank itself into action.



Using a 7104/7A29 gives a measured 410ps rise, implying a rise time of
the

pulse gen itself is 240ps (because the 7104/7A29 has a measured rise
time of

330ps).



Preshoot is 2%, and overshoot is un-measurably small, the very
slightest

ripple on the pulse top. One of the cleanest rises I have ever seen.
Front

panel says rise time <=300ps. Pulse length as expected (from the
internal

charge line) is about 30ns and the rep rate is 110kHz. Amplitude can
be

adjusted from 250mV to 25V in switched steps, with a variable control too.



Pretrigger gives a 2.5V pulse 250ns long with a 1.4ns rise.



I'll try it out on a sampler (S4) later and get a better grip on the
pulse

rise and what the pre-trigger does on a 7T11. I always find triggering
a

7T11 from external sources a bit fiddly, not surprising when you read
the

manual in detail and see that each triggering mode has very different

frequency and amplitude characteristics.



This is a comparison between the "Unknown Unit" and a Type 111
pre-trigger

pulse generator:



111

===

Fixed output level of at least 10V, not precise

<500ps rise

Switchable polarity

2ns pulse length without external charge line

Rep rate variable from 10Hz to 100kHz

Pre-trigger 30-250ns early, 300ns duration, <5ns rise, +10V



The unknown one

===============

Switched and variable output level from 250mV to 25V

<300ps rise

Fixed polarity

30ns pulse length without external charge line

Fixed rep rate of 110kHz (probably 100kHz nominal)

Pre-trigger 50-125ns early (measured 30-130ns), 250ns duration, 1.4ns
rise,

+2.5V

From: TekScopes@... [mailto:TekScopes@...] On
Behalf Of iglesia_cristiana_arpas_eternas
Sent: Tuesday, February 26, 2013 3:26 PM



Here the link to this almost new pulsed generator, at first glance is
a transmission line pulser with 300ps ..but no model or number at all.
I just brought it with the hope work right...but a good service manual
will be very appreciated and gratefully.
Link to Pulse Gen.
<
232112
187?autorefresh=true&pt=LH_DefaultDomain_0&hash=item3a7e99ce3b&nma=tru
e&si=w
4%252BkkAkTYIq9sTkPYLC8H98HEpU%253D&orig_cvip=true&rt=nc&_trksid=p2047
675.l2
557>

Regards Gabriel.

_._,___



------------------------------------

Yahoo! Groups Links


Re: Could please any one give a clue about this nice pulsed line generator??

 

Very useful info indeed..thanks a lot Dennis.
I'm not feel "safe men" until obtain the right info about..I just like to have "all" the service manual of my gear.
I was send a mail to Artekmanual , but not response yet.
Regards Gabriel.

--- In TekScopes@..., "Dennis Tillman" <dennis@...> wrote:

I think you are being overly concerned about your purchase not working.
These are rugged instruments and not likely to fail. The PV509 manual is
close enough to tell you everything you will need to know about your
purchase. But here is some additional information on the HAPG that I kept
from years ago.



From: "Craig Sawyers" <c.sawyers@>

Date: Tue May 4, 2004 2:25 pm

Subject: Wierd Pulse generator craig_sawyers



Hi Group



Well I took delivery of the strange "High Amplitude Pulse Generator". This

is a TM500 style plug-in, and I bought it kind of out of curiosity, since I

had seen nothing like it. It also has no identifying type number at all -

so it was either a development unit, or an internal use only test fixture -

hopefully one of the ex-Tek guys on the list will recognize it - I've put

some pics on the yahoo gallery.



This is the only thing I've seen that uses the strange attenuator that the

7A29 uses. This has a huge bandwidth - way above 1GHZ, and probably *way*

above so that it doesn't cramp the 7A29's style. I suspect that the 7A29

has a 1.5GHz bandwidth, so the attenuator is possible approaching 5GHz.



It also has a rigid charge line inside, estimated at around 3m (10') long.

This line isn't secured at all, and is prevented from shorting to the back

of the PCB with a layer of self-adhesive plastic film - very prototype-ish.

Date on the PCB is copyright 1978, and component date codes indicate late

1982 build.



The unit is basically a pre-trigger pulse generator that uses an avalanche

transistor to give a fast rising pulse. The pre-trigger can be adjusted with
a front

panel control from 50-125ns early, so that a sampling time base gets time to

crank itself into action.



Using a 7104/7A29 gives a measured 410ps rise, implying a rise time of the

pulse gen itself is 240ps (because the 7104/7A29 has a measured rise time of

330ps).



Preshoot is 2%, and overshoot is un-measurably small, the very slightest

ripple on the pulse top. One of the cleanest rises I have ever seen. Front

panel says rise time <=300ps. Pulse length as expected (from the internal

charge line) is about 30ns and the rep rate is 110kHz. Amplitude can be

adjusted from 250mV to 25V in switched steps, with a variable control too.



Pretrigger gives a 2.5V pulse 250ns long with a 1.4ns rise.



I'll try it out on a sampler (S4) later and get a better grip on the pulse

rise and what the pre-trigger does on a 7T11. I always find triggering a

7T11 from external sources a bit fiddly, not surprising when you read the

manual in detail and see that each triggering mode has very different

frequency and amplitude characteristics.



This is a comparison between the "Unknown Unit" and a Type 111 pre-trigger

pulse generator:



111

===

Fixed output level of at least 10V, not precise

<500ps rise

Switchable polarity

2ns pulse length without external charge line

Rep rate variable from 10Hz to 100kHz

Pre-trigger 30-250ns early, 300ns duration, <5ns rise, +10V



The unknown one

===============

Switched and variable output level from 250mV to 25V

<300ps rise

Fixed polarity

30ns pulse length without external charge line

Fixed rep rate of 110kHz (probably 100kHz nominal)

Pre-trigger 50-125ns early (measured 30-130ns), 250ns duration, 1.4ns rise,

+2.5V

From: TekScopes@... [mailto:TekScopes@...] On Behalf
Of iglesia_cristiana_arpas_eternas
Sent: Tuesday, February 26, 2013 3:26 PM



Here the link to this almost new pulsed generator, at first glance is a
transmission line pulser with 300ps ..but no model or number at all.
I just brought it with the hope work right...but a good service manual will
be very appreciated and gratefully.
Link to Pulse Gen.
<
187?autorefresh=true&pt=LH_DefaultDomain_0&hash=item3a7e99ce3b&nma=true&si=w
4%252BkkAkTYIq9sTkPYLC8H98HEpU%253D&orig_cvip=true&rt=nc&_trksid=p2047675.l2
557>

Regards Gabriel.

_._,___


Re: Cleaning dirty pots on 5xx series plug-ins

 

¿ªÔÆÌåÓý

Dan;
There's a bunch of material on this subject in the forum's archives.
More was added, recently. Check it out.
Bernd
?
In a message dated 2/27/2013 8:25:32 A.M. Pacific Standard Time, kaboomdk@... writes:

?

Howdy folks,

? I have another cleaning question. ?I've got a bunch of type L and type CA (and other) 5-series plug ins and the variable attenuator pots have become "dirty". ?They are intermittent over much of their range, and usually not useful in the "cal" position.

? ?Is there any Tek-net wisdom particular to these pots I should follow in cleaning them? ?What is their failure mechanism?

? I have cleaned these with tuner cleaner or other freon based sprays in the past, and the effect is definitely not permanent as I find I have to clean them aga in (albeit a few years later). ?Is there any benefit to taking them out of the plug-ins, sonicating them in a cleaning bath and re-installing them? ?Anything I should avoid?

? Thanks again for the help.

? Dan


Re: Could please any one give a clue about this nice pulsed line generator??

 

¿ªÔÆÌåÓý

I think you are being overly concerned about your purchase not working. These are rugged instruments and not likely to fail. The PV509 manual is close enough to tell you everything you will need to know about your purchase. But here is some additional information on the HAPG that I kept from years ago.

?

From: "Craig Sawyers"

Date: Tue May 4, 2004? 2:25 pm

Subject: Wierd Pulse generator? craig_sawyers

?

Hi Group

?

Well I took delivery of the strange "High Amplitude Pulse Generator". This

is a TM500 style plug-in, and I bought it kind of out of curiosity, since I

had seen nothing like it. It also has no identifying type number at all -

so it was either a development unit, or an internal use only test fixture -

hopefully one of the ex-Tek guys on the list will recognize it - I've put

some pics on the yahoo gallery.

?

This is the only thing I've seen that uses the strange attenuator that the

7A29 uses. This has a huge bandwidth - way above 1GHZ, and probably *way*

above so that it doesn't cramp the 7A29's style. I suspect that the 7A29

has a 1.5GHz bandwidth, so the attenuator is possible approaching 5GHz.

?

It also has a rigid charge line inside, estimated at around 3m (10') long.

This line isn't secured at all, and is prevented from shorting to the back

of the PCB with a layer of self-adhesive plastic film - very prototype-ish.

Date on the PCB is copyright 1978, and component date codes indicate late

1982 build.

?

The unit is basically a pre-trigger pulse generator that uses an avalanche

transistor to give a fast rising pulse. The pre-trigger can be adjusted with a front

panel control from 50-125ns early, so that a sampling time base gets time to

crank itself into action.

?

Using a 7104/7A29 gives a measured 410ps rise, implying a rise time of the

pulse gen itself is 240ps (because the 7104/7A29 has a measured rise time of

330ps).

?

Preshoot is 2%, and overshoot is un-measurably small, the very slightest

ripple on the pulse top. One of the cleanest rises I have ever seen. Front

panel says rise time <=300ps. Pulse length as expected (from the internal

charge line) is about 30ns and the rep rate is 110kHz. Amplitude can be

adjusted from 250mV to 25V in switched steps, with a variable control too.

?

Pretrigger gives a 2.5V pulse 250ns long with a 1.4ns rise.

?

I'll try it out on a sampler (S4) later and get a better grip on the pulse

rise and what the pre-trigger does on a 7T11. I always find triggering a

7T11 from external sources a bit fiddly, not surprising when you read the

manual in detail and see that each triggering mode has very different

frequency and amplitude characteristics.

?

This is a comparison between the "Unknown Unit" and a Type 111 pre-trigger

pulse generator:

?

111

===

Fixed output level of at least 10V, not precise

<500ps rise

Switchable polarity

2ns pulse length without external charge line

Rep rate variable from 10Hz to 100kHz

Pre-trigger 30-250ns early, 300ns duration, <5ns rise, +10V

?

The unknown one

===============

Switched and variable output level from 250mV to 25V

<300ps rise

Fixed polarity

30ns pulse length without external charge line

Fixed rep rate of 110kHz (probably 100kHz nominal)

Pre-trigger 50-125ns early (measured 30-130ns), 250ns duration, 1.4ns rise,

+2.5V

From: TekScopes@... [mailto:TekScopes@...] On Behalf Of iglesia_cristiana_arpas_eternas
Sent: Tuesday, February 26, 2013 3:26 PM

Here the link to this almost? new pulsed generator, at first glance is a transmission line pulser with 300ps ..but no model or number at all.
I just brought it with the hope work right...but a good service manual will be very appreciated and gratefully.
?

Regards Gabriel.

_._,___


Re: 7k Readout Board Swap

chrismh_somenumber
 

Thanks for the info. I'm going to give it a shot this evening.

Chris

--- In TekScopes@..., David <davidwhess@...> wrote:

The list I have says that the 7844 and 7104 use the same readout board
which is Tektronix part number 670-1900-XX so as far as I know, you
can just swap the boards.

There is a PDF called Tektronix7xxx_Assemblies.pdf in the files
section which lists which assemblies are used in which oscilloscopes.

On Thu, 28 Feb 2013 03:45:16 -0000, "chrismh_somenumber"
<hogancm@...> wrote:

I have a 7844 that works fine except for the complete lack of readout.

In looking at voltages and signals around the U2126 timer chip, all is OK except for pins 7 and 10. I have a junker 7104 that uses the same IC (and its readout works). I was going to swap the 7844's IC into the 7104, but the 7104's is soldered in.

I'm wondering if I can just swap the entire 7844 board into the 7104.

I have looked at the circuit diagrams for both and they are largely the same. The signals in and out are the same, they even use the same jumper designations.

Has anyone tried this before?

Thanks,
Chris


Re: Could please any one give a clue about this nice pulsed line generator??

Albert
 

Correction: in my early XF unit the charge line is soldered to the board. Probably a peltola jack has been removed there. In my later XH unit the board and the line have SMA connectors.
Albert

--- The cable has a peltola connector at one end. I have another unit in which the cable is soldered to the board.
---
Albert


Re: Could please any one give a clue about this nice pulsed line generator??

Albert
 

Hi Gabriel,

Your HAPG version is a rather early one. I have one with a sticker (not the original label) saying SN160. The boards are exactly the same version XF, with the same hand written signature GUB. The only difference in board components seems to be that yours has a variable R75. Further mine has an internal open ended charge line. You might make the same modification, with the advantage that you can use the GR connector to export the pretrigger signal to the front panel. Simply reconnect the GR connector to the empty socket (about 2 cm away from Q60, to the rear side). Pretrigger lead time can be adjusted with R75. Best avalanching can be set by R100. R10 sets -18V IIRC. C102 influences rise time and/or aberrations.
The internal charge line is a rigid 20 ns cable, probably very high quality. It might be difficult to find something like that. The cable has a peltola connector at one end. I have another unit in which the cable is soldered to the board. I suppose the internal charge line was a later modification, copied from the PG509.
The most vulnerable component in your unit might be the attenuator. Same problems as in the 7A29.

Albert