On Mon, 11 Mar 2013 06:35:34 -0000, "Michael" <af7u@...>
wrote:
>This post is a follow up from post #7360, and my 7L13
with center frequency LED readout all zeros. I am a bit
further along with my investigation. I now understand the
operation 'somewhat' more. I have looked over the diagrams
and taken some voltage readings, I also understand the
basic DVM logic and readout operation which sends data to
the LCD and the CRT. I have an article from a Jan, 1975
'73' magazine which talks about the Fairchild 3814 digital
voltmeter. I must admit I am not too clear on the 'Analog'
section of the meter. On board 'A2100', diagram 12, the
input to pin 2 of U2110 comes from plug P2540 pin 4.
Looking over the A2500 board shows that pin 4 goes to
chassis ground. Why would pin 2 from U2110 go all the way
back to A2500 to reach chassis ground.
U2110 (LM308 style low bias current bipolar operational
amplifier) is
configured as a differential amplifier. The signal ground
that the
tuning voltage uses may not be at quite the same potential
as the
signal ground that the readout uses so the differential
amplifier
subtracts the difference. The differential amplifier will
also
subtract low frequency noise between the grounds.
>Anyway, the A2000 board seems to be working, albeit
the frequency of the clock oscillator is not 200kHz as the
diagram shows, it is running at 153.7kHz, though amplitude
is good at 5 volts, I would think that would affect
calibration but not operation. Back on A2100, the tuning
control voltage at R2104 is good, it swings from + - 9.6
volts. The output of U2110 swings + - 10.7 volts. The DCPL
Supply rails 15, -15, and 5 volts are all good. The
reference voltage at the cathode of VR2115 (Zener) is 3.1,
not the 3.3 volts is shows. Voltage at TP 2100, (output of
U2130) is 11.65, a bit higher than the listed 11.3 volts.
The voltages around the N-channel FET (Q2120) is Drain =
3.37 volts, Source = -3.1 volts, and Gate = -3.5 volts,
are those in the ball park?
Those voltages all sound good to me and they are mostly
the ones I
would have asked you to measure. In the worst case they
add up to a
calibration problem and not a complete failure.
>Needless to say there is NO signals present, waveform,
1, 2, 3, and 4 are all missing. Lines 'E' and 'F' are both
quiet, the voltages on 'E' are 7.08 volts, on 'F' the
voltage is 5.12 volts which swings a bit higher when I
adjust center frequency, up to 5.44 volts. Of course 'E'
has no signal as would be expected without drive from
U2145 there is nothing to trigger the multivibrator U2030
on board A2000 sending a signal back up line 'F'. I keep
thinking of this in terms of a frequency counter, I keep
thinking there should be a driving input signal. I try to
keep in mind it's a voltmeter, and keep in mind the
circuit description given in the manual. It describes a
ramp signal coming from the output of U2110, I don't
understand that statement. The 'Tune' voltage is just a DC
voltage, plus to minus. The minus input of U2120 goes to
chassis ground??? What 'RAMP' are they speaking about?
E is 7 volts? That is quite a trick since the pull-up
resistor
(R2149) is connected to +5V and the output of U2145 (LM311
style
comparator) is wired open collector (pin 1 is the emitter
of the
output transistor and wired to ground) so it can only pull
down and
not pull up. On the receiver side, U2010D is part of a
74L04 and also
has a 5 volt supply voltage so even if it failed, it can
not pull up
to 7 volts
U2145 (LM311) is probably bad but you might want to look
for a bad
socket or open trace. Nothing should be able to make its
output go to
7 volts except an open on pin 4 or failed part.
The operational amplifiers and comparator are all TO-99 8
pin metal
cans which are difficult to find these days. If you can
not find
metal cans, I would try replacing them with 8 pin DIP
parts which
conveniently have the same pinout. I have done this before
by cutting
the DIP leads short and soldering little wires to them.
Cutoff 1/4
watt resistor leads work well.
Texas Instruments still sells TO-99 LM311 comparators but
they are $10
each.
U2010 may be damaged as well. I think it can be replaced
without
problems with a 74LS04 or even a 74HCT04 which are both
more common.
If you have to order the LM311, pick up an LM308 and 74L04
replacement
as well just in case. The LM308 operational amplifiers can
be
directly replaced with the Linear Technology LT1008. An
LT1012 or
lots of other devices can be used if you pull the
compensation
capacitor. JFET or MOSFET input devices should work as
well.
>Lastly the manual's circuit description on page 5-10
talks about a positive going ramp from U2110, aren't these
just DC levels, what 'Ramp' are they talking about?
I think the ramp they talk about is the output of U2140
(another
LM308) which is configured as an integrator. The frequency
(and
slope) of the ramp is proportional to the measured
voltage.
>Okay, I see there are a lot of questions here. I had
better post this and hope for the best. It's a great 7L13,
it works well, but without a 'Center' frequency display,
it's pretty much just a toy. Also, I do appreciate the
feedback~!~ I am not accustom to working on test
equipment, my experience has been mostly all in Hi-Fi, but
this has been very interesting.
I had to add a few pages of the 7L13 manual to my book of
stuff I do
not have so it took a bit longer to respond to your post.
The 7000
series spectrum analyzers look like fun. I will have to
keep an eye
out for one.