#qmx SSB Firmware beta 1_01_006 release - calibration
#qmx
Hans, I think calibration in _006 is better, but there are still some issues.
I am attaching some screenshots from my QMX-lowband calibration that may help in diagnosing the phase-error cals that don't finish, and also a USB IMD optimization that was strange, and fixed itself when I repeated it.
?
I did the initial calibration from a factory-reset condition, using the 'R' option to do all of the optimizations, which I then reviewed.
The 80m Phase Error plot didn't complete - it stopped at about 700 on the x-axis.? And retrying it with the 'O' repeated the same results.
Then I retried a few more times and captured the attached 5 shots of the progression.? Note that it for some reason started with a high DAC value, not the one corresponding to the x-axis, which is why it stopped early.? Completely repeatable.
?
The 20m USB IMD optimization looked crazy.? The other IMDs looked OK.? So I retried 20m using the 'O', and it cleaned itself up.? See the initial and retry results in the last two attachments.
?
I hope this is helpful- I will check the cal process/results on my other QMX*, and let you know if anything else interesting shows up.
Stan KC7XE
?
?
Cal-80m-PE-1.PNG
Cal-80m-PE-2.PNG
Cal-80m-PE-3.PNG
Cal-80m-PE-4.PNG
Cal-80m-PE-finished-incomplete.PNG
1st-try-USB-IMD-20m.PNG
2nd-try-USB-IMD-20m.PNG
|
Notes on calibration of my other QMX* devices.?? Some (hopefully) useful info:
?
Calibration of my QMX high-band:
- all completed normally and looked good, except 30m phase error had an irregular shape, large downs and ups near the beginning similar to the irregular shapes that occurred with earlier versions.
?
Calibration of my QMX+ rev2:?
- The 160m phase error plot stopped very early, at about DAC=750.? The symptoms were exactly the same as the 80m on my QMX low-band, where the calibration DAC values started at 2xxx rather than 0.? Retries do the same; it never completes.
- The 20m phase error plot stopped early, about DAC=1500.? But one retry allowed it to complete.? I didn't see the DAC values on the first run to see if they started high.
?
Calibration of my QMX+ rev3 (done overnight, so cooler temperatures):
- 40m, 30m, and 6m PE calibration stopped early; 40m at DAC~1000,? 30m at DAC~1200, and 6m at DAC~700.
- a single 'O' retry fixed the 40m and 6m PE sweeps.
- 'O' retries did not fix the 30m; it stopped at the same place, repeatable - See series of screenshots below.?
-? Note: 1) the new startup measurement messages are not present;
-? Note: 2) it stops after a "change greater than 30 degrees" message.? The change >30 degrees was in fact a measurement of 0.1 following a measurement of 359.8, so it was a phase rollover, not a large change.? So the change from near 0 to 359.7 at DAC=250 worked, but the opposite change from 359.8 to 0.1 at DAC=1250 did not work.
- I power-cycled the QMX+ (first power-cycle since start of calibration) to see if behavior would change on 30m; no change.?
- 160m USB IMD optimization flatlined with all values at 0 (I never saw that before).? A single 'O' retry fixed this (I did this before any of he above retries).
- All other PE and IMD completed and looked to have reasonable values.
?
I hope this info is useful in some way -
Stan KC7XE
?
Cal_PE_30m_retry_1.PNG
Cal_PE_30m_retry_2.PNG
Cal_PE_30m_retry_3.PNG
Cal_PE_30m_retry_4.PNG
Cal_PE_30m_retry_5.PNG
|
Just had another instance of phase error calibration ending early, at DAC=750.
This time it was on the 6m band, as part of a full 'R' calibration of all bands on my QMX+ r3.
All other bands completed this time.
And it has that same characteristic of the DAC values being used starting at a high value (in this case near 2000), rather than near 0 as labeled on the x-axis of the plot.? See attached screenshots:
?
Stan KC7XE |
"?The change >30 degrees was in fact a measurement of 0.1 following a measurement of 359.8, so it was a phase rollover, not a large change."
I wonder if this could be the root cause of calibration issues. -Steve K1RF ------ Original Message ------