HP 8935 has false alarm for overvoltage duplex out


 

opened the I/O module on the underside next to the I/O panel.
What are the silk screen number (s) for the detection? diodes?
I see some of tht SOT23 packages labeled CR1 to 4 and cr200 and cr201.
I sure wish schematics were available for this thing!


 

On Mon, May 6, 2024 at 11:07 AM, rab@... wrote:
BAS70-05?
Thansk you very much. I'll get some before I open the beast.


 

¿ªÔÆÌåÓý

The Diodes (2) to detect duplex port power were HSMS-2800-TR1 but are no longer available. The replacement is: BAS70-05?? the Mouser elect part is: 771-BAS70-05-T/R Good luck...Rick

_._,_._,_


 

What kind of diodes are they?
I tried putting a load on the duplex port, thinking that might make it happy, but unfortunately it didn't help.


 

Jim, duplex alarm is generally the 2 diodes right at the duplex output (in the input module).? Are leaky/open.

Rick

On May 3, 2024 2:14 PM, "J Mcvey wrote:

I get the duplex overvoltage message with the alarm sound which drives me nuts.


 

I get the duplex overvoltage message with the alarm sound which drives me nuts.
I am not even using the duplex out!? I mostly use RF I/O for my application. I tried measurement reset, but it continues with the alarm and message with no relief .
Is there a way to shut off the alarm?
I suppose this problem is telling me the power supply needs attention or something.

The test set seems to work otherwise.

Any help or insights would be greatly appreciated.