RF Alignment Failure in Agilent CXA signal analyzer N9000A


 

I have Agilent CXA Signal Analyzer N9000A, I'm getting these errors, please help.
?
I referred manual and it suggested to check 50Mhz with -24dB, But when I check in my equipment its showing 50Mhz with -60dB. While I trace it out it was across H270(HMC270MS8G) where I was getting 50Mhz with -60dB but near FPGA it is exact 50Mhz with -24dB, so does it represent there is problem with H270 or any other if yes please help me, And any blue print of this will help a-lot.
?


 

开云体育

Hi

Are the Cal signal at w9 ok-about -25 ?—if so I guess you have the same fault we have seen in the MXA series-bad switch at Rf Front end..

Regards

Hardy

?

Fra: [email protected] <[email protected]> P? vegne af ipother6 via groups.io
Sendt: 2. april 2025 16:46
Til: [email protected]
Emne: [HP-Agilent-Keysight-equipment] RF Alignment Failure in Agilent CXA signal analyzer N9000A

?

I have Agilent CXA Signal Analyzer N9000A, I'm getting these errors, please help.

?

I referred manual and it suggested to check 50Mhz with -24dB, But when I check in my equipment its showing 50Mhz with -60dB. While I trace it out it was across H270(HMC270MS8G) where I was getting 50Mhz with -60dB but near FPGA it is exact 50Mhz with -24dB, so does it represent there is problem with H270 or any other if yes please help me, And any blue print of this will help a-lot.

?


 

Hi, Hardyhansendk.
Thank you for your response, As you mentioned about W9 cal signals, When I tested there I'm getting 50Mhz with -60dB not -25dB, it would be helpful if you share info regarding that. Once again thank you for your response.
?
Regards,
San.