开云体育

ctrl + shift + ? for shortcuts
© 2025 开云体育

Retrofitting 3.5mm connector from HP 8485A sensor onto an existing HP 8481A sensor - possible?


 

On Thu, Mar 20, 2025 at 11:50 PM, Sven Schnelle wrote:
But exactly as you said, the
levels where low by at least 3dB above 19GHz.

?

?
I can see that... as N-connectors...say, precision ones... they might go to 25 or 26 GHz...but N connectors are usually to 18 GHz, where... AFAIK...TE11 can start happening.
?

The 85025A... with an N connector... is spec t to about 18 GHz... if I recall correctly.
?

For frequencies beyond 18 GHz, 3.5 mm, 2.92 mm (K), or SMA connectors are preferred...AFAIK.
I've never tried an 805025A past 20 GHz; but I'm surprised the N connector maintains TEM... and if it goes to TE11... then 'works to 40 GHz" doesn't make sense to me.


 

After doing some testing yesterday, I discovered that the 8481A that was to be the recipient of the dead 8485A's connector etc. assembly was itself dead (8481A won't zero and responds about 20 dB down incident signal, so dynamic range is something like 20 dB, not 50 dB.) I don't immediately recall the 8481A's history, however, I thought I had stored it away in working condition (and I'm the only person that would have access to it.) Given that unexpected result, and the current-era prices for used 8481A, my project idea won't be able proceed. So all that said, I'm going to keep eye out for 8485A (basically as a reference against which to calibrate 8673D synthesized signal generator.)

Thanks again for the contributions to the question, it is very much helpful.


 

I think it is likely to work - BUT, the assembly and disassembly will be (at lest) challenging.
?
The thermistors may be similar, and possibly the thermistor board in the cartridge (you need to check this carefully before you start.? If the cartridge are the same, it would be straight forward,? if only the thermistor board is the same it will be a bit more difficult.? It might be reasonable o pick up a dead 8481 to practice disassembly on.
?
HP did sell the cartridges some time ago but they found that many attempts to replace the cartridges resulted in damage to the matching components in the head.
?
I would look carefully at the part list for the 8485 to get an idea as to how difficult the disassembly / reassembly will be.? It is relatively easy to get the cartridge out of the 8481,??
?
keep us posted as to how it works out - pictures would be very desirable.


 

"F1EKU via groups.io" <rfconsulting.fr@...> writes:

From experience, the 8481A series does not perform up well beyond 18
GHz. The response drops very quickly towards 19/20
GHz, and at 26.5 GHz, the response difference is more than 3 to 4
dB. The N connector is not responsible for this poor
performance, as 85025A detectors easily perform up to 40 GHz with only
a few glitches due to VSWR beyond 18.6 GHz.
Conversely, an 8485A often performs up to 30 or even 33 GHz.
I can confirm that. Last week i tried to do flatness calibration on my
83752A sweeper. As i only have a 8485D that accepts a max input level
of -20dBm, i tried the 8481A instead. But exactly as you said, the
levels where low by at least 3dB above 19GHz.


 

Thank you for detail understanding!


 

From experience, the 8481A series does not perform up well beyond 18 GHz. The response drops very quickly towards 19/20 GHz, and at 26.5 GHz, the response difference is more than 3 to 4 dB. The N connector is not responsible for this poor performance, as 85025A detectors easily perform up to 40 GHz with only a few glitches due to VSWR beyond 18.6 GHz. Conversely, an 8485A often performs up to 30 or even 33 GHz.


 

Have a HP 8485A sensor I bought used, turns out it had been damaged due to overload, seller refunded, now have a physically nice but electrically dead HP 8485A sensor.

Have a HP 8481A sensor surplus to my needs.

I'm interested in the possibility of converting my 8481A to an 8485A (3.5mm male connector), albeit not calibrated, in lieu of the cost of buying a working 8485A.

Found a Keysight publication that seems to be quite comprehensive coverage of this era sensor:



This is titled "Keysight 8480 series coaxial power sensors operating and service manual." I wasn't aware the manual existed; it is current publication from Keysight.

The manual has pretty extensive "how to disassemble and reassemble" and other maintenance activities.

The underlying thermocouple sensor element appears to be the same part number for 8481A and 8485A, so this brings me to my question - could I create a thermocouple sensor with a 3.5mm HP male connector that could subsequently be recalibrated and used as an 8485A?

I've searched and haven't found any reports of the practicality or result of actually trying this procedure.

Thanks for any comments, observations, or requests for clarification.