this is understood to be a add on thread to the recent discussion about reverse power protection / DC blocks ,? instrument protection often is non existent in hobby labs for a variety of reasons which sometimes leads to a costly learning experience , here is a brief summary of items that can protect your equipment from damage which should be present in every lab :
1.) DC Block : most Spectrum analyzers have a allergy to DC on the input connector ,? depending on which range is used DC can damage the input attenuator , YIG filter and mixer , a DC block is? ? ? ?cheap "life insurance " 2.) 2 or 3dB Power attenuators ,? easy hack if you test power amplifiers , add a low value pad in front of your expensive 30dB 18GHz Pad , if you have a oops moment you blow the ( often cheaper) low value pad instead of the more expensive one? 3.) m-f adapters : just add a precision type (3.5mm, 2.92mm, 2.4mm) m-f adapter in front of your noise source or power sensor , to protect the connectors of your equipment 4.) torque wrench :? one of the most underappreciated items in hobby labs , one 8lbs for precision and 5lbs for SMA protects connectors from over torque and also ensures good and repeatable? ? ? ?measuremnts by avoiding under torquing? |
It's worth adding that a DC block only provides some measure of safety. The ramp up voltage of a DC signal is an AC signal and can pass right through your series cap. If you look at DC blocking filter designs like the HP 3048A option K23, there's a limiter after the series cap. It seems like the "right" devices for this sort of thing are RF power limiters. Unfortunately, the cost of many of these devices new from reputable manufacturers probably matches or exceeds what many people paid for their second-hand spectrum analyzer. Maybe there are good second-hand market options, but I haven't looked into to it. If you want something to reliably protect a low frequency SA, that shouldn't be too hard to design and build. But, if you want to get the full 22 GHz range out of your 8566, I expect that wouldn't be so easy.
|
Lothar baier
开云体育Its not really a problem in real life since most DC blocks only have capacitors that are 100n or lower, the short current pulse doesn’t hurt anything!On Mar 20, 2022, at 06:52, Matt Huszagh via groups.io <huszaghmatt@...> wrote:
|
On Sun, Mar 20, 2022 at 08:59 AM, Lothar baier wrote:
Its not really a problem in real life since most DC blocks only have capacitors that are 100n or lower, the short current pulse doesn’t hurt anything!It seems to me that saying "100n is small enough to protect the circuit" is not the answer, without stating how big the DC level is. Isn't it the ENERGY absorbed by the delicate components that is key? A different variation on Protection: The AC line voltage here is routinely 125V. Some older equipment might be "stressed" by that. I use a buck transformer to bring it down to about 115V. The transformer, coincidentally, came from an HP spec-an that I disassembled for parts. Pete |
开云体育An older Uninterruptable Power Supply makes a very convenient way to reduce line voltage. Find one of the old "loaf of bread" sized units that uses the large 60 cycle transformer, rather than the newer ones that use a switching inverter. Remove everything except the transformer, fuse holder, outlets and switch. Use one half of the? primary winding along with the 120 volt secondary to make an autotranformer. It will drop 120 volts to about 109, which is much easier on older gear designed for 110 or 115 volts.? ???? Bruce Gentry, KA2IVY
On 3/20/22 13:45, saipan59 (Pete)
wrote:
On Sun, Mar 20, 2022 at 08:59 AM, Lothar baier wrote: |
开云体育greenboxmaven via groups.io wrote on 3/20/2022 2:02 PM: Hello-- Dual-primary transformers can serve as isolation transformers. Use one 120 V primary as the AC input winding, and? the second 120 V winding to power the device you're testing. Note that tested devices using? half-wave rectification? may require significant transformer derating. 73-- Brad? AA1IP |
Mini-Circuits sells limiters as well as DC blocks, but they only go up to 8 GHz.? Not sure about the price, but they are most likely competitive.? Assuming you actually need one. Jim Ford
------ Original Message ------
From: "Matt Huszagh" <huszaghmatt@...>
Sent: 3/20/2022 6:52:14 AM
Subject: Re: [HP-Agilent-Keysight-equipment] Protecting Equipment It's worth adding that a DC block only provides some measure of safety. The ramp up voltage of a DC signal is an AC signal and can pass right through your series cap. If you look at DC blocking filter designs like the HP 3048A option K23, there's a limiter after the series cap. It seems like the "right" devices for this sort of thing are RF power limiters. Unfortunately, the cost of many of these devices new from reputable manufacturers probably matches or exceeds what many people paid for their second-hand spectrum analyzer. Maybe there are good second-hand market options, but I haven't looked into to it. If you want something to reliably protect a low frequency SA, that shouldn't be too hard to design and build. But, if you want to get the full 22 GHz range out of your 8566, I expect that wouldn't be so easy. |
Lothar baier
开云体育A few words about limiters , Generally limiters are used in systems to protect receivers usually in RADAR or EW applications ,? in test systems the use of limiters s usually frowned upon ( with very few exceptions)? and for good reason ! First of there is a cost penalty that is not insignificant , the cost for a connectorized limiter usually starts at about $600 and can go up into the $2K range . Limiters also degrade VSWR and add additional system loss . But the by far biggest problem is that limiters are non linear devices and as such excellent harmonics generators , harmonics will start to occur before the limiter hits the hard threshold and worsens as the power increases so if you put one in front a spectrum analyzer you have to be cautious not to end up with tst results that are worse than they actually are ? From: [email protected] <[email protected]>
On Behalf Of Jim Ford via groups.io
Sent: Sunday, March 20, 2022 2:53 PM To: [email protected] Subject: Re: [HP-Agilent-Keysight-equipment] Protecting Equipment ? Mini-Circuits sells limiters as well as DC blocks, but they only go up to 8 GHz.? Not sure about the price, but they are most likely competitive.? Assuming you actually need one. ? Jim Ford ? ------ Original Message ------ From: "Matt Huszagh" <huszaghmatt@...> Sent: 3/20/2022 6:52:14 AM Subject: Re: [HP-Agilent-Keysight-equipment] Protecting Equipment ?
|
开云体育Limiters are very commonly recommended for Conducted Immunity testing in EMC labs where the Spec An input comes from the Line Impedance Stabilization Network (LISN) because the turn on/off transients can be quite high. The levels expected to be measured typically should be much less than the distortion level of the limiter. The recommended limiter in this case actually has an intentional 10 dB insertion loss. ? Tom, N8ZM ? From: [email protected] <[email protected]> On Behalf Of Lothar baier
Sent: Sunday, March 20, 2022 9:03 PM To: [email protected] Subject: Re: [HP-Agilent-Keysight-equipment] Protecting Equipment ? A few words about limiters , Generally limiters are used in systems to protect receivers usually in RADAR or EW applications ,? in test systems the use of limiters s usually frowned upon ( with very few exceptions)? and for good reason ! First of there is a cost penalty that is not insignificant , the cost for a connectorized limiter usually starts at about $600 and can go up into the $2K range . Limiters also degrade VSWR and add additional system loss . But the by far biggest problem is that limiters are non linear devices and as such excellent harmonics generators , harmonics will start to occur before the limiter hits the hard threshold and worsens as the power increases so if you put one in front a spectrum analyzer you have to be cautious not to end up with tst results that are worse than they actually are ? From: [email protected] <[email protected]> On Behalf Of Jim Ford via groups.io ? Mini-Circuits sells limiters as well as DC blocks, but they only go up to 8 GHz.? Not sure about the price, but they are most likely competitive.? Assuming you actually need one. ? Jim Ford ? ------ Original Message ------ From: "Matt Huszagh" <huszaghmatt@...> Sent: 3/20/2022 6:52:14 AM Subject: Re: [HP-Agilent-Keysight-equipment] Protecting Equipment ?
|
开云体育I mis-spoke. I meant Conducted Emissions. Sorry. ? Tom, N8ZM ? From: [email protected] <[email protected]> On Behalf Of n8zmTWH@...
Sent: Sunday, March 20, 2022 9:22 PM To: [email protected] Subject: Re: [HP-Agilent-Keysight-equipment] Protecting Equipment ? Limiters are very commonly recommended for Conducted Immunity testing in EMC labs where the Spec An input comes from the Line Impedance Stabilization Network (LISN) because the turn on/off transients can be quite high. The levels expected to be measured typically should be much less than the distortion level of the limiter. The recommended limiter in this case actually has an intentional 10 dB insertion loss. ? Tom, N8ZM ? From: [email protected] <[email protected]> On Behalf Of Lothar baier ? A few words about limiters , Generally limiters are used in systems to protect receivers usually in RADAR or EW applications ,? in test systems the use of limiters s usually frowned upon ( with very few exceptions)? and for good reason ! First of there is a cost penalty that is not insignificant , the cost for a connectorized limiter usually starts at about $600 and can go up into the $2K range . Limiters also degrade VSWR and add additional system loss . But the by far biggest problem is that limiters are non linear devices and as such excellent harmonics generators , harmonics will start to occur before the limiter hits the hard threshold and worsens as the power increases so if you put one in front a spectrum analyzer you have to be cautious not to end up with tst results that are worse than they actually are ? From: [email protected] <[email protected]> On Behalf Of Jim Ford via groups.io ? Mini-Circuits sells limiters as well as DC blocks, but they only go up to 8 GHz.? Not sure about the price, but they are most likely competitive.? Assuming you actually need one. ? Jim Ford ? ------ Original Message ------ From: "Matt Huszagh" <huszaghmatt@...> Sent: 3/20/2022 6:52:14 AM Subject: Re: [HP-Agilent-Keysight-equipment] Protecting Equipment ?
|
Lothar baier
开云体育Those are transient limiters that are especially designed for this purpose , usually they only cover to 30 or 100MHz t match the frequency range of the LISN ( I only have seen the ones made by ETS that went to 100MHz not sure if they make higher frequency ones ) As far as the limiter itself concerns usually one attenuator of 3Db is used on the input followed by a diode limiter which is followed by another attenuator and then another ?limiter . ? ? From: [email protected] <[email protected]>
On Behalf Of n8zmTWH via groups.io
Sent: Sunday, March 20, 2022 8:22 PM To: [email protected] Subject: Re: [HP-Agilent-Keysight-equipment] Protecting Equipment ? Limiters are very commonly recommended for Conducted Immunity testing in EMC labs where the Spec An input comes from the Line Impedance Stabilization Network (LISN) because the turn on/off transients can be quite high. The levels expected to be measured typically should be much less than the distortion level of the limiter. The recommended limiter in this case actually has an intentional 10 dB insertion loss. ? Tom, N8ZM ? From:
[email protected] <[email protected]>
On Behalf Of Lothar baier ? A few words about limiters , Generally limiters are used in systems to protect receivers usually in RADAR or EW applications ,? in test systems the use of limiters s usually frowned upon ( with very few exceptions)? and for good reason ! First of there is a cost penalty that is not insignificant , the cost for a connectorized limiter usually starts at about $600 and can go up into the $2K range . Limiters also degrade VSWR and add additional system loss . But the by far biggest problem is that limiters are non linear devices and as such excellent harmonics generators , harmonics will start to occur before the limiter hits the hard threshold and worsens as the power increases so if you put one in front a spectrum analyzer you have to be cautious not to end up with tst results that are worse than they actually are ? From:
[email protected] <[email protected]>
On Behalf Of Jim Ford via groups.io ? Mini-Circuits sells limiters as well as DC blocks, but they only go up to 8 GHz.? Not sure about the price, but they are most likely competitive.? Assuming you actually need one. ? Jim Ford ? ------ Original Message ------ From: "Matt Huszagh" <huszaghmatt@...> Sent: 3/20/2022 6:52:14 AM Subject: Re: [HP-Agilent-Keysight-equipment] Protecting Equipment ?
|
开云体育Yes, they are because the inductors in the LISN can generate some nasty spikes when device power is cycled. ? Most conducted EMC specs cut off at 30 MHz, or at least did up to the last time I worked in that area about 10 years ago. ? Tom, N8ZM ? From: [email protected] <[email protected]> On Behalf Of Lothar baier
Sent: Sunday, March 20, 2022 10:14 PM To: [email protected] Subject: Re: [HP-Agilent-Keysight-equipment] Protecting Equipment ? Those are transient limiters that are especially designed for this purpose , usually they only cover to 30 or 100MHz t match the frequency range of the LISN ( I only have seen the ones made by ETS that went to 100MHz not sure if they make higher frequency ones ) As far as the limiter itself concerns usually one attenuator of 3Db is used on the input followed by a diode limiter which is followed by another attenuator and then another ?limiter . ? ? From: [email protected] <[email protected]> On Behalf Of n8zmTWH via groups.io ? Limiters are very commonly recommended for Conducted Immunity testing in EMC labs where the Spec An input comes from the Line Impedance Stabilization Network (LISN) because the turn on/off transients can be quite high. The levels expected to be measured typically should be much less than the distortion level of the limiter. The recommended limiter in this case actually has an intentional 10 dB insertion loss. ? Tom, N8ZM ? From: [email protected] <[email protected]> On Behalf Of Lothar baier ? A few words about limiters , Generally limiters are used in systems to protect receivers usually in RADAR or EW applications ,? in test systems the use of limiters s usually frowned upon ( with very few exceptions)? and for good reason ! First of there is a cost penalty that is not insignificant , the cost for a connectorized limiter usually starts at about $600 and can go up into the $2K range . Limiters also degrade VSWR and add additional system loss . But the by far biggest problem is that limiters are non linear devices and as such excellent harmonics generators , harmonics will start to occur before the limiter hits the hard threshold and worsens as the power increases so if you put one in front a spectrum analyzer you have to be cautious not to end up with tst results that are worse than they actually are ? From: [email protected] <[email protected]> On Behalf Of Jim Ford via groups.io ? Mini-Circuits sells limiters as well as DC blocks, but they only go up to 8 GHz.? Not sure about the price, but they are most likely competitive.? Assuming you actually need one. ? Jim Ford ? ------ Original Message ------ From: "Matt Huszagh" <huszaghmatt@...> Sent: 3/20/2022 6:52:14 AM Subject: Re: [HP-Agilent-Keysight-equipment] Protecting Equipment ?
|
On Sunday 20 March 2022 02:02:21 pm greenboxmaven via groups.io wrote:
An older Uninterruptable Power Supply makes a very convenient way toInteresting! I happen to have a pile of those... -- Member of the toughest, meanest, deadliest, most unrelenting -- and ablest -- form of life in this section of space, ?a critter that can be killed but can't be tamed. ?--Robert A. Heinlein, "The Puppet Masters" - Information is more dangerous than cannon to a society ruled by lies. --James M Dakin |
开云体育I made two of these to use with older
communications receivers. Power transformers in the day often
used paper that becomes acidic and crackely now that it's
seventy years old, National recivers are especially vulnerable.
I wonder if vacuum impregnating them with silicon oil or winding
varnish would help keep them from? developing the intra-winding
shorts they are so prone to. I can't recall ever seeing one of
them have an inter-winding short, an open winding, or a short to
ground. ?? Bruce Gentry, KA2IVY
On 3/21/22 10:59, Roy J.
Tellason, Sr. wrote:
On Sunday 20 March 2022 02:02:21 pm greenboxmaven via groups.io wrote:An older Uninterruptable Power Supply makes a very convenient way to reduce line voltage. Find one of the old "loaf of bread" sized units that uses the large 60 cycle transformer, rather than the newer ones that use a switching inverter. Remove everything except the transformer, fuse holder, outlets and switch. Use one half of the? primary winding along with the 120 volt secondary to make an autotranformer. It will drop 120 volts to about 109, which is much easier on older gear designed for 110 or 115 volts. ???? Bruce Gentry, KA2IVYInteresting! I happen to have a pile of those... |
开云体育Tom, ? That is also why MIL-STD-461 CE102 requires a 20dB Pad on the signal path from the LISN to the Receiver. They do not recommend a Limiter due to non-linearities that Lothar points out. ? Limiters can really hurt the Noise Figure of a receiver input at well. ? Ross ? From: [email protected] <[email protected]> On Behalf Of n8zmTWH@...
Sent: Sunday, March 20, 2022 7:22 PM To: [email protected] Subject: Re: [HP-Agilent-Keysight-equipment] Protecting Equipment ? Limiters are very commonly recommended for Conducted Immunity testing in EMC labs where the Spec An input comes from the Line Impedance Stabilization Network (LISN) because the turn on/off transients can be quite high. The levels expected to be measured typically should be much less than the distortion level of the limiter. The recommended limiter in this case actually has an intentional 10 dB insertion loss. ? Tom, N8ZM ? From: [email protected] <[email protected]> On Behalf Of Lothar baier ? A few words about limiters , Generally limiters are used in systems to protect receivers usually in RADAR or EW applications ,? in test systems the use of limiters s usually frowned upon ( with very few exceptions)? and for good reason ! First of there is a cost penalty that is not insignificant , the cost for a connectorized limiter usually starts at about $600 and can go up into the $2K range . Limiters also degrade VSWR and add additional system loss . But the by far biggest problem is that limiters are non linear devices and as such excellent harmonics generators , harmonics will start to occur before the limiter hits the hard threshold and worsens as the power increases so if you put one in front a spectrum analyzer you have to be cautious not to end up with tst results that are worse than they actually are ? From: [email protected] <[email protected]> On Behalf Of Jim Ford via groups.io ? Mini-Circuits sells limiters as well as DC blocks, but they only go up to 8 GHz.? Not sure about the price, but they are most likely competitive.? Assuming you actually need one. ? Jim Ford ? ------ Original Message ------ From: "Matt Huszagh" <huszaghmatt@...> Sent: 3/20/2022 6:52:14 AM Subject: Re: [HP-Agilent-Keysight-equipment] Protecting Equipment ?
|
开云体育In my experience, non-military EMC, if the gadget you are testing can trigger the distortion of a limiter on a CE test, you are WAY above the limits you need to meet. It does make it harder to find the real offending frequencies, however. Then that 20 dB pad becomes beneficial as more than just a protective device.. ? Everything in front of the input screws up the noise figure; that’s just physics. Unless it’s a preamp, and then you have to watch for overload problems, much like the limiter effect. ? Tom, N8ZM ? From: [email protected] <[email protected]> On Behalf Of si_emi_01
Sent: Monday, March 21, 2022 3:50 PM To: [email protected] Subject: Re: [HP-Agilent-Keysight-equipment] Protecting Equipment ? Tom, ? That is also why MIL-STD-461 CE102 requires a 20dB Pad on the signal path from the LISN to the Receiver. They do not recommend a Limiter due to non-linearities that Lothar points out. ? Limiters can really hurt the Noise Figure of a receiver input at well. ? Ross ? From: [email protected] <[email protected]> On Behalf Of n8zmTWH@... ? Limiters are very commonly recommended for Conducted Immunity testing in EMC labs where the Spec An input comes from the Line Impedance Stabilization Network (LISN) because the turn on/off transients can be quite high. The levels expected to be measured typically should be much less than the distortion level of the limiter. The recommended limiter in this case actually has an intentional 10 dB insertion loss. ? Tom, N8ZM ? From: [email protected] <[email protected]> On Behalf Of Lothar baier ? A few words about limiters , Generally limiters are used in systems to protect receivers usually in RADAR or EW applications ,? in test systems the use of limiters s usually frowned upon ( with very few exceptions)? and for good reason ! First of there is a cost penalty that is not insignificant , the cost for a connectorized limiter usually starts at about $600 and can go up into the $2K range . Limiters also degrade VSWR and add additional system loss . But the by far biggest problem is that limiters are non linear devices and as such excellent harmonics generators , harmonics will start to occur before the limiter hits the hard threshold and worsens as the power increases so if you put one in front a spectrum analyzer you have to be cautious not to end up with tst results that are worse than they actually are ? From: [email protected] <[email protected]> On Behalf Of Jim Ford via groups.io ? Mini-Circuits sells limiters as well as DC blocks, but they only go up to 8 GHz.? Not sure about the price, but they are most likely competitive.? Assuming you actually need one. ? Jim Ford ? ------ Original Message ------ From: "Matt Huszagh" <huszaghmatt@...> Sent: 3/20/2022 6:52:14 AM Subject: Re: [HP-Agilent-Keysight-equipment] Protecting Equipment ?
|
Lothar baier
开云体育What many people seem to forget is the fact that the non linear behavior of a limiter kicks in way before the limiter is actually “triggered” ?although the distortions are not as bad ? From: [email protected] <[email protected]>
On Behalf Of si_emi_01 via groups.io
Sent: Monday, March 21, 2022 2:50 PM To: [email protected] Subject: Re: [HP-Agilent-Keysight-equipment] Protecting Equipment ? Tom, ? That is also why MIL-STD-461 CE102 requires a 20dB Pad on the signal path from the LISN to the Receiver. They do not recommend a Limiter due to non-linearities that Lothar points out. ? Limiters can really hurt the Noise Figure of a receiver input at well. ? Ross ? From:
[email protected] <[email protected]>
On Behalf Of n8zmTWH@... ? Limiters are very commonly recommended for Conducted Immunity testing in EMC labs where the Spec An input comes from the Line Impedance Stabilization Network (LISN) because the turn on/off transients can be quite high. The levels expected to be measured typically should be much less than the distortion level of the limiter. The recommended limiter in this case actually has an intentional 10 dB insertion loss. ? Tom, N8ZM ? From:
[email protected] <[email protected]>
On Behalf Of Lothar baier ? A few words about limiters , Generally limiters are used in systems to protect receivers usually in RADAR or EW applications ,? in test systems the use of limiters s usually frowned upon ( with very few exceptions)? and for good reason ! First of there is a cost penalty that is not insignificant , the cost for a connectorized limiter usually starts at about $600 and can go up into the $2K range . Limiters also degrade VSWR and add additional system loss . But the by far biggest problem is that limiters are non linear devices and as such excellent harmonics generators , harmonics will start to occur before the limiter hits the hard threshold and worsens as the power increases so if you put one in front a spectrum analyzer you have to be cautious not to end up with tst results that are worse than they actually are ? From:
[email protected] <[email protected]>
On Behalf Of Jim Ford via groups.io ? Mini-Circuits sells limiters as well as DC blocks, but they only go up to 8 GHz.? Not sure about the price, but they are most likely competitive.? Assuming you actually need one. ? Jim Ford ? ------ Original Message ------ From: "Matt Huszagh" <huszaghmatt@...> Sent: 3/20/2022 6:52:14 AM Subject: Re: [HP-Agilent-Keysight-equipment] Protecting Equipment ?
|
开云体育Tom, ? That’s true except the most stringent limits for DC powered MIL-STD-461 CE-102 Conducted Emissions are much greater (94dBuV down to 60dBuV from 10kHz to 10MHz), than those of RE-102 Radiated Emissions (60dBuV to 24dBuV to 69dBuV from 10kHz to 18GHz). They also modify the CE-102 Limits for AC powered equipment by increasing the limit line by 6dB, 9dB, 10dB and 12dB based on Line Voltage too. ? The LISN Measurement Output Capacitor (such as the 0.25uF for the Standard MIL-STD-461 50uH LISN), is a High Pass Filter. MIL-STD-461 CE-102 starts at 10kHz, the LISN is designed for 10kHz to 10MHz. The Calibration Factors are all entered in the Program (like TILE!), you are using and added/subtracted from the measured value. ? Unless you are testing a poorly designed product or the Power Supply and the emissions are there because: 1. The Power Supply is operating in Discontinuous Conduction Mode (often seen as spaced emissions lower - 10s of kHz, than the switching frequency of the Power Supply). Happens in some types of Regulators and also when the Output Load is light. 2. The emissions are harmonics of the AC/DC Rectifiers Commutation from Bulk Capacitance Charging, ? After you check/fix the above issues, you should not see the need to add a limiter. ? ? If you have those issues, the power supply is suspect (grounding/PFC Circuitry/inadequate Bulk Capacitance/Type of Diodes, etc.), or a gross test setup issue (LISN Grounding/poor Cable Shield Termination/etc.), may be at fault. ? It could be that your Receiver is not powered through an Isolation Transformer to prevent an Alternate Return Path (Ground Loop), as well. This is especially true for measurements below 10kHz (like MIL-STD-461 CE-101 and others). ? Ross ? From: [email protected] <[email protected]> On Behalf Of n8zmTWH@...
Sent: Monday, March 21, 2022 2:16 PM To: [email protected] Subject: Re: [HP-Agilent-Keysight-equipment] Protecting Equipment ? In my experience, non-military EMC, if the gadget you are testing can trigger the distortion of a limiter on a CE test, you are WAY above the limits you need to meet. It does make it harder to find the real offending frequencies, however. Then that 20 dB pad becomes beneficial as more than just a protective device.. ? Everything in front of the input screws up the noise figure; that’s just physics. Unless it’s a preamp, and then you have to watch for overload problems, much like the limiter effect. ? Tom, N8ZM ? From: [email protected] <[email protected]> On Behalf Of si_emi_01 ? Tom, ? That is also why MIL-STD-461 CE102 requires a 20dB Pad on the signal path from the LISN to the Receiver. They do not recommend a Limiter due to non-linearities that Lothar points out. ? Limiters can really hurt the Noise Figure of a receiver input at well. ? Ross ? From: [email protected] <[email protected]> On Behalf Of n8zmTWH@... ? Limiters are very commonly recommended for Conducted Immunity testing in EMC labs where the Spec An input comes from the Line Impedance Stabilization Network (LISN) because the turn on/off transients can be quite high. The levels expected to be measured typically should be much less than the distortion level of the limiter. The recommended limiter in this case actually has an intentional 10 dB insertion loss. ? Tom, N8ZM ? From: [email protected] <[email protected]> On Behalf Of Lothar baier ? A few words about limiters , Generally limiters are used in systems to protect receivers usually in RADAR or EW applications ,? in test systems the use of limiters s usually frowned upon ( with very few exceptions)? and for good reason ! First of there is a cost penalty that is not insignificant , the cost for a connectorized limiter usually starts at about $600 and can go up into the $2K range . Limiters also degrade VSWR and add additional system loss . But the by far biggest problem is that limiters are non linear devices and as such excellent harmonics generators , harmonics will start to occur before the limiter hits the hard threshold and worsens as the power increases so if you put one in front a spectrum analyzer you have to be cautious not to end up with tst results that are worse than they actually are ? From: [email protected] <[email protected]> On Behalf Of Jim Ford via groups.io ? Mini-Circuits sells limiters as well as DC blocks, but they only go up to 8 GHz.? Not sure about the price, but they are most likely competitive.? Assuming you actually need one. ? Jim Ford ? ------ Original Message ------ From: "Matt Huszagh" <huszaghmatt@...> Sent: 3/20/2022 6:52:14 AM Subject: Re: [HP-Agilent-Keysight-equipment] Protecting Equipment ?
|
开云体育Lothar, ? Correct. The same thing happens with Zeners below 6.2V and TVS Diodes (way too much Capacitance for RF Applications). ? Look at the Breakover point of the RF Limiter (or TVS), it is a curve. Its Impedance changes below the breakover voltage as current flows through it and approaches the knee. If it is not specified, it can be obtained via a Curve Tracer or manually generated I-V Curves that you can produce with a simple Power Supply UUT and Resistance Standard or Resistance Decade. ? Many Signal Integrity issues develop when they are not used properly and the curve is well understood. ? Ross ? From: [email protected] <[email protected]> On Behalf Of Lothar baier
Sent: Monday, March 21, 2022 2:28 PM To: [email protected] Subject: Re: [HP-Agilent-Keysight-equipment] Protecting Equipment ? What many people seem to forget is the fact that the non linear behavior of a limiter kicks in way before the limiter is actually “triggered” ?although the distortions are not as bad ? From: [email protected] <[email protected]> On Behalf Of si_emi_01 via groups.io ? Tom, ? That is also why MIL-STD-461 CE102 requires a 20dB Pad on the signal path from the LISN to the Receiver. They do not recommend a Limiter due to non-linearities that Lothar points out. ? Limiters can really hurt the Noise Figure of a receiver input at well. ? Ross ? From: [email protected] <[email protected]> On Behalf Of n8zmTWH@... ? Limiters are very commonly recommended for Conducted Immunity testing in EMC labs where the Spec An input comes from the Line Impedance Stabilization Network (LISN) because the turn on/off transients can be quite high. The levels expected to be measured typically should be much less than the distortion level of the limiter. The recommended limiter in this case actually has an intentional 10 dB insertion loss. ? Tom, N8ZM ? From: [email protected] <[email protected]> On Behalf Of Lothar baier ? A few words about limiters , Generally limiters are used in systems to protect receivers usually in RADAR or EW applications ,? in test systems the use of limiters s usually frowned upon ( with very few exceptions)? and for good reason ! First of there is a cost penalty that is not insignificant , the cost for a connectorized limiter usually starts at about $600 and can go up into the $2K range . Limiters also degrade VSWR and add additional system loss . But the by far biggest problem is that limiters are non linear devices and as such excellent harmonics generators , harmonics will start to occur before the limiter hits the hard threshold and worsens as the power increases so if you put one in front a spectrum analyzer you have to be cautious not to end up with tst results that are worse than they actually are ? From: [email protected] <[email protected]> On Behalf Of Jim Ford via groups.io ? Mini-Circuits sells limiters as well as DC blocks, but they only go up to 8 GHz.? Not sure about the price, but they are most likely competitive.? Assuming you actually need one. ? Jim Ford ? ------ Original Message ------ From: "Matt Huszagh" <huszaghmatt@...> Sent: 3/20/2022 6:52:14 AM Subject: Re: [HP-Agilent-Keysight-equipment] Protecting Equipment ?
|
Lothar baier
开云体育Years ago I had this discussion with a PhD student ,? his argument was that the limiter doesn’t matter unless it is triggered so I set up a experiment doing a power sweep of the limiter and capturing the output spectrum for each point ,? he was quite surprised ? From: [email protected] <[email protected]>
On Behalf Of si_emi_01 via groups.io
Sent: Monday, March 21, 2022 4:16 PM To: [email protected] Subject: Re: [HP-Agilent-Keysight-equipment] Protecting Equipment ? Lothar, ? Correct. The same thing happens with Zeners below 6.2V and TVS Diodes (way too much Capacitance for RF Applications). ? Look at the Breakover point of the RF Limiter (or TVS), it is a curve. Its Impedance changes below the breakover voltage as current flows through it and approaches the knee. If it is not specified, it can be obtained via a Curve Tracer or manually generated I-V Curves that you can produce with a simple Power Supply UUT and Resistance Standard or Resistance Decade. ? Many Signal Integrity issues develop when they are not used properly and the curve is well understood. ? Ross ? From:
[email protected] <[email protected]>
On Behalf Of Lothar baier ? What many people seem to forget is the fact that the non linear behavior of a limiter kicks in way before the limiter is actually “triggered” ?although the distortions are not as bad ? From:
[email protected] <[email protected]>
On Behalf Of si_emi_01 via groups.io ? Tom, ? That is also why MIL-STD-461 CE102 requires a 20dB Pad on the signal path from the LISN to the Receiver. They do not recommend a Limiter due to non-linearities that Lothar points out. ? Limiters can really hurt the Noise Figure of a receiver input at well. ? Ross ? From:
[email protected] <[email protected]>
On Behalf Of n8zmTWH@... ? Limiters are very commonly recommended for Conducted Immunity testing in EMC labs where the Spec An input comes from the Line Impedance Stabilization Network (LISN) because the turn on/off transients can be quite high. The levels expected to be measured typically should be much less than the distortion level of the limiter. The recommended limiter in this case actually has an intentional 10 dB insertion loss. ? Tom, N8ZM ? From:
[email protected] <[email protected]>
On Behalf Of Lothar baier ? A few words about limiters , Generally limiters are used in systems to protect receivers usually in RADAR or EW applications ,? in test systems the use of limiters s usually frowned upon ( with very few exceptions)? and for good reason ! First of there is a cost penalty that is not insignificant , the cost for a connectorized limiter usually starts at about $600 and can go up into the $2K range . Limiters also degrade VSWR and add additional system loss . But the by far biggest problem is that limiters are non linear devices and as such excellent harmonics generators , harmonics will start to occur before the limiter hits the hard threshold and worsens as the power increases so if you put one in front a spectrum analyzer you have to be cautious not to end up with tst results that are worse than they actually are ? From:
[email protected] <[email protected]>
On Behalf Of Jim Ford via groups.io ? Mini-Circuits sells limiters as well as DC blocks, but they only go up to 8 GHz.? Not sure about the price, but they are most likely competitive.? Assuming you actually need one. ? Jim Ford ? ------ Original Message ------ From: "Matt Huszagh" <huszaghmatt@...> Sent: 3/20/2022 6:52:14 AM Subject: Re: [HP-Agilent-Keysight-equipment] Protecting Equipment ?
|