¿ªÔÆÌåÓý

hp8595A faulty


 

Hi to all , I have a spectrum analyzer hp8595A and I suspect the RF assembly has a problem.
Can I replace this assembly , whith one from an hp8595E ?
Many thanks in advance for your help
Best regards

Maurizio IZ1MDJ


 

Maurizio
Sorry.... NO ..... dramatically different beasts the 95A and the 95E
Dave
NR1DX
manuals@...

On 2/22/2019 2:50 AM, Maurizio IZ1MDJ wrote:
Hi to all , I have a spectrum analyzer hp8595A and I suspect the RF assembly has a problem.
Can I replace this assembly , whith one from an hp8595E ?
Many thanks in advance for your help
Best regards

Maurizio IZ1MDJ
--
Dave
Manuals@...
www.ArtekManuals.com


Maurizio IZ1MDJ
 

Hi Dave many thanks , I will try to isolate the fault subassy.
Maurizio
IZ1MDJ


 

What are the faults? Are there any error messages? Is it in the low band or high band only??
Is there a frequency or span problem or an amplitude problem?


On Sat, Feb 23, 2019 at 10:54 AM, Maurizio IZ1MDJ wrote:
Hi Dave many thanks , I will try to isolate the fault subassy.
Maurizio
IZ1MDJ


 

It's a long story . Initially the power supply was? missing one voltage ( I don't remember my be - 12V).
Solved the problem? , a cold junction on pcb board of the supply ,now? the spectrum doesn't display? the calibrator signal.
The calibrator output is into specifications Freq and Amplitude.
Next week I will able to reserve more time to check again the spectrum ,
following the troubleshoting procedure as explained in the manual.
Many thanks