Hi Tony, well, it is a symptom. This loop lacks conditions required for locking. You need to trace the 50 MHz signal from the reference oscillator to the A8 power splitter. Next trace the 60 to 110 MHz oscillator signal to the A8 mixer, see block diagrams BD1 - 3. Are they present in the first place? And then move on, what do you find? G?ran |
It is the SYMPTOM - probably a problem in the phase detector or anything between the pD and the oscillator drive.
toggle quoted message
Show quoted text
You need to verify the PD inputs and output. Cheers! Bruce Quoting Tony <tonycox01@...>: Hi Goran, |
Seems like a reasonable plan. The observation you gave is an indicator that the PD is not controlling the VCO. Disd the VCO frequency change when the drive voltage changed?
toggle quoted message
Show quoted text
Cheers! Bruce Quoting Tony <tonycox01@...>: Thanks Bruce. These are complex beasts - it¡¯s a bit hard to know where to start the trouble shooting. I¡¯m going to check the 50MHz through the freq multipliers and if there is nothing amiss there I¡¯ll move to the PD section. |
I looked at this the other way around and stepped through the frequencies observing?the tune voltage (is that what you mean by drive voltage?). It was constant through the tuning range starting at 500MHz at about -12V and suddenly swung to?+13 at about 750MHz and stayed there for the rest of the range up to 990MHz. Tony On Mon, 20 Feb 2023 at 21:58, Bruce <bruce@...> wrote: Seems like a reasonable plan.? The observation you gave is an? |
QuotiTony -
toggle quoted message
Show quoted text
Yes - drive = tune. My GUESS (that and $4 will get you a Starbucks) is that the problem is the PD. Check that the drive is adequate for both signals and try stepping your way through the PD circuitry. Hopefully it is discrete - if not it is a bit more of a problem to analyze. What I would expect to find is that one of the two outputs is working and the other has a problem. Cheers! Bruce ng Tony <tonycox01@...>: I looked at this the other way around and stepped through the frequencies |
Can you send a picture of the PD schematic ??
toggle quoted message
Show quoted text
Cheers! Bruce Quoting Tony <tonycox01@...>: I looked at this the other way around and stepped through the frequencies |
Hi Tony, On schematic 13 you find the phase detector U19 (ECL EXOR circuit). On pins 4 and 5 you shall have 10 MHz signals which are locked and 90 degrees out of phase. It is assumed that you have a 10 MHz reference signal connected to the instrument. On schematic 10 you find the phase detector U17A and U17B. On pins 3 and 11 you shall have 100 (or 250) Hz signals which are locked and in phase. G?ran |