Linear fit yields 22.52 GHz/mA.
Note that with this sensitivity, which is well within the 23 +/- 1 mA per GHz spec, that at the low end the output is BELOW the 3.8 GHz cavity oscillator.? This plot was taken with the following settings:
1)? PLL off 2)? Source in CW mode 3)? Default pretune constants (test 45) loaded 4)? HP 34401A inserted at 40 ohm sense resistor 5)? HP 5342A counter connected directly to YIG oscillator output
Linearity of YIG oscillator looks great.
How does the firmware deal with the YTO being below the cavity oscillator?
|
Is it not possible to insert even small graphs into these messages?
Let's try attaching...
toggle quoted message
Show quoted text
On 3/23/2019 6:00 PM, Peter Gottlieb via Groups.Io wrote: Linear fit yields 22.52 GHz/mA.
Note that with this sensitivity, which is well within the 23 +/- 1 mA per GHz spec, that at the low end the output is BELOW the 3.8 GHz cavity oscillator.? This plot was taken with the following settings:
1)? PLL off 2)? Source in CW mode 3)? Default pretune constants (test 45) loaded 4)? HP 34401A inserted at 40 ohm sense resistor 5)? HP 5342A counter connected directly to YIG oscillator output
Linearity of YIG oscillator looks great.
How does the firmware deal with the YTO being below the cavity oscillator?
|
On 3/23/19 5:03 PM, Peter Gottlieb wrote: Is it not possible to insert even small graphs into these messages? Let's try attaching... Worked fine. Did not show inline, but as attachment.
|
Hi Peter,
Your plot shows? y = 0.0444x + 0.0656. So, there is a 65.6 MHz offset at 0 mA main coil current. This could in part be caused by a FM coil current which does not show up across the 40 Ohm resistor. Basically, I think you have confirmed that your YTO works and produces an output well below 3.8 GHz. The only exception to this statement would be a case where the YTO stops working in a narrow range of frequencies, but that almost always happens near the very low end of the oscillator's range (well below 3.8 GHz).
The question now is, how to allocate the 65.6 MHz offset. How much of it is due to some FM coil current and how much is due to some other problem. Keep in mind what I said about the FM coil current, it should not exceed 150 mA, which corresponds to approximately 35 MHz tuning range in either direction. I guess, I would somehow check to make sure the coil isn't being fried with too much current.
I have no idea how the firmware deals with pretune constants in CW mode and/or PLL off.
Vladan
|
I found that the 34401A had excessive internal resistance while in the loop so redid the YTO measurements using instead a 0.102 ohm shunt resistor.? I had redone the chart but flipped it so the trend line would show mA/GHz directly (attached).? This doesn't show the MHz offset directly but the linearity of the YTO is remarkable.? I have to run out and do some errands but later today I will check the output to the FM coil.? However, a NOS replacement phase lock board behaves identically so I don't expect to see this as a board fault.
In the analysis I posted earlier, I calculated the worst case tolerance stack up both positive and negative.? The firmware would have to be able to handle this range.
Peter
toggle quoted message
Show quoted text
On 3/24/2019 1:34 PM, pianovt via Groups.Io wrote: Hi Peter,
Your plot shows? y = 0.0444x + 0.0656. So, there is a 65.6 MHz offset at 0 mA main coil current. This could in part be caused by a FM coil current which does not show up across the 40 Ohm resistor. Basically, I think you have confirmed that your YTO works and produces an output well below 3.8 GHz. The only exception to this statement would be a case where the YTO stops working in a narrow range of frequencies, but that almost always happens near the very low end of the oscillator's range (well below 3.8 GHz).
The question now is, how to allocate the 65.6 MHz offset. How much of it is due to some FM coil current and how much is due to some other problem. Keep in mind what I said about the FM coil current, it should not exceed 150 mA, which corresponds to approximately 35 MHz tuning range in either direction. I guess, I would somehow check to make sure the coil isn't being fried with too much current.
I have no idea how the firmware deals with pretune constants in CW mode and/or PLL off.
Vladan
|
I don't know if this will help, but here goes.
First, don't worry about the apparent offset in coil tuning current. YTOs are only specified to operate over a certain range, like an octave or two, but can run a ways beyond at either end. It doesn't help (other than mathematically to get the tuning chart numbers) to extrapolate down to zero coil current - the YTO will stall out somewhere a little below its lowest spec frequency and coil current. Any slope variations or offsets extrapolated at zero just represent the character of the magnetic structure in the YTO.
The proper range, slope, and offset needed for tuning a particular YTO are typically handled by the YTO driver circuit, which can make it so the net result is a nice convenient GHz/V type system, at its input. There are usually sufficient adjustments (pots and/or DAC settings) in the driver circuitry to accommodate the tolerances of the YTO, the 3.8 GHz oscillator in this case, and the rest of the parts, to make everything come out right, during calibration.
There may be circuitry included to improve the tuning linearity (due to magnetic saturation), and compensate for the dynamic response (magnetic delay and hysteresis). Ideally, the YTO frequency is perfectly and instantly proportional to the magnetic field, but in reality the field depends on an iron core, coil inductance, AC core losses, and so forth.
The FM coil may be used in a PLL, for fine tuning. Its effect is very small (but much faster), compared to the main coil. It too has to be properly driven, of course.
In a down-converting generator made for reaching low frequency, it would normally be set up to avoid spectral inversion - that's why there's a minimum output frequency that can be expected. If the 3.8 GHz LO is free-running, its drift may be enough to limit how far down you can go at the low end, regardless of the YTO's tuning precision. For sources that can get way down toward zero (DC even), both the LO and YTO need to be precisely controlled (synthesized).
The coarse and fine tuning, compensation, and calibration all contribute to successful operation, so there may be quite a few possible suspects.
If you have schematics, you should be able to find all of these elements, and track down the proper conditions for diagnostics and repair.
Good luck.
Ed
|
I checked the FM coil drive and it is right about at zero so that seems like it's not the issue.
Nobody seems to know how the firmware works so we are left to guess and try things.? There is some range which the firmware can correct for and I heard from someone who needed to nudge his phase lock board to get his system into that range.? I tried the same thing and can get some measure of success as well, although apparently I have another problem causing me grief.? At this point I have this temporary fix in which permits the unit to run on default pretune values but won't pass test 48 which sets more precise pretune values.
Below about 50 MHz I have a PLL instability which gets worse as the frequency decreases.? The PLL begins to lose lock below about 25 MHz.? I took two photos of the unit output when set to 25 MHz in CW mode showing the instability, then I put it into source tune mode which opens the phase lock loop.? See attached.? Notice how the output is nice and stable as soon as the loop is opened.
Not sure what to make of this.? The YTO main coil drive doesn't look any noisier at 25 MHz than it does at higher frequencies.
toggle quoted message
Show quoted text
On 3/24/2019 3:28 PM, Ed Breya via Groups.Io wrote: I don't know if this will help, but here goes.
First, don't worry about the apparent offset in coil tuning current. YTOs are only specified to operate over a certain range, like an octave or two, but can run a ways beyond at either end. It doesn't help (other than mathematically to get the tuning chart numbers) to extrapolate down to zero coil current - the YTO will stall out somewhere a little below its lowest spec frequency and coil current. Any slope variations or offsets extrapolated at zero just represent the character of the magnetic structure in the YTO.
The proper range, slope, and offset needed for tuning a particular YTO are typically handled by the YTO driver circuit, which can make it so the net result is a nice convenient GHz/V type system, at its input. There are usually sufficient adjustments (pots and/or DAC settings) in the driver circuitry to accommodate the tolerances of the YTO, the 3.8 GHz oscillator in this case, and the rest of the parts, to make everything come out right, during calibration.
There may be circuitry included to improve the tuning linearity (due to magnetic saturation), and compensate for the dynamic response (magnetic delay and hysteresis). Ideally, the YTO frequency is perfectly and instantly proportional to the magnetic field, but in reality the field depends on an iron core, coil inductance, AC core losses, and so forth.
The FM coil may be used in a PLL, for fine tuning. Its effect is very small (but much faster), compared to the main coil. It too has to be properly driven, of course.
In a down-converting generator made for reaching low frequency, it would normally be set up to avoid spectral inversion - that's why there's a minimum output frequency that can be expected. If the 3.8 GHz LO is free-running, its drift may be enough to limit how far down you can go at the low end, regardless of the YTO's tuning precision. For sources that can get way down toward zero (DC even), both the LO and YTO need to be precisely controlled (synthesized).
The coarse and fine tuning, compensation, and calibration all contribute to successful operation, so there may be quite a few possible suspects.
If you have schematics, you should be able to find all of these elements, and track down the proper conditions for diagnostics and repair.
Good luck.
Ed
|
I have not had to deal with freq. tune issues in an 8753, but did have to deal with them in 8566 and 8340 repairs. There are coarse adjust for both the high and low end freq. tune points measured in voltage (typically 1v/GHz, in the 8566 and 8340). In fixed tune mode you could set the end point frequencies within 1 MHz and typically within 50 KHz in PLL unlocked mode - when PLL is locked it was exact, within much less than 1 KHz. The pretune 1024 bit DAC¡¯s were normally set to 511 bit for coarse adjust then corrected for the precise tune and would be typically in the center 40% of the DAC range more often +/- 10% of center. The 8340¡¯s also used midpoint DAC corrections. This correction would be applied to the YIG FM coil fine tune. Now that the YIG tune is set statically to be correct, you now have to include the swept tune points, the span tune and, the sweep time that are added to the tune ramp. Any one of these can add offset to the original tune ramp, because in the case of the above SA and SWPR they all had pretune DAC circuits. On the SA and Sweepers I encountered many issues with DAC¡¯s with offsets, typically the LSB¡¯s (2,4,8). You can troubleshoot the span issues with start sweep/single sweep with a very slow sweep and measure the swept freq. start point with your SA. Do this at several spans looking for differences in start points, always set for the same start point. The sweep times can be checked by setting the same span and start point and change the swept time through its range looking for a change in the start frequency. Assuming 1V/GHz tune then you should see about 65mv offset, and if I recall correctly you said it was 65MHz too low then your offset should be -65mv. If the actual V/GHz is different, then correct as needed. Best regards, Don Bitters
|
Hi Peter
I don't know this instrument, but here a couple of things I would check..
I cant remember if you already checked this, but can you connect a coupler to the 3.8GHz oscillator, to tap of a small bit and check that this is always stable?? I guess it will be, but will be good to get it verified.
The YTO unlock instability, you should see on the FM coil drive.
Next I would follow the signal all the way to the PLL circuit. Check that it don't get attenuated to much at lower frequencies so the PLL can't lock to it.? ??
Good luck with this.?
Br, Askild
?
toggle quoted message
Show quoted text
On Sun, Mar 24, 2019 at 10:29 PM Peter Gottlieb < hpnpilot@...> wrote: I checked the FM coil drive and it is right about at zero so that seems like
it's not the issue.
Nobody seems to know how the firmware works so we are left to guess and try
things.? There is some range which the firmware can correct for and I heard from
someone who needed to nudge his phase lock board to get his system into that
range.? I tried the same thing and can get some measure of success as well,
although apparently I have another problem causing me grief.? At this point I
have this temporary fix in which permits the unit to run on default pretune
values but won't pass test 48 which sets more precise pretune values.
Below about 50 MHz I have a PLL instability which gets worse as the frequency
decreases.? The PLL begins to lose lock below about 25 MHz.? I took two photos
of the unit output when set to 25 MHz in CW mode showing the instability, then I
put it into source tune mode which opens the phase lock loop.? See attached.?
Notice how the output is nice and stable as soon as the loop is opened.
Not sure what to make of this.? The YTO main coil drive doesn't look any noisier
at 25 MHz than it does at higher frequencies.
On 3/24/2019 3:28 PM, Ed Breya via Groups.Io wrote:
> I don't know if this will help, but here goes.
>
> First, don't worry about the apparent offset in coil tuning current. YTOs are
> only specified to operate over a certain range, like an octave or two, but can
> run a ways beyond at either end. It doesn't help (other than mathematically to
> get the tuning chart numbers) to extrapolate down to zero coil current - the
> YTO will stall out somewhere a little below its lowest spec frequency and coil
> current. Any slope variations or offsets extrapolated at zero just represent
> the character of the magnetic structure in the YTO.
>
> The proper range, slope, and offset needed for tuning a particular YTO are
> typically handled by the YTO driver circuit, which can make it so the net
> result is a nice convenient GHz/V type system, at its input. There are usually
> sufficient adjustments (pots and/or DAC settings) in the driver circuitry to
> accommodate the tolerances of the YTO, the 3.8 GHz oscillator in this case,
> and the rest of the parts, to make everything come out right, during calibration.
>
> There may be circuitry included to improve the tuning linearity (due to
> magnetic saturation), and compensate for the dynamic response (magnetic delay
> and hysteresis). Ideally, the YTO frequency is perfectly and instantly
> proportional to the magnetic field, but in reality the field depends on an
> iron core, coil inductance, AC core losses, and so forth.
>
> The FM coil may be used in a PLL, for fine tuning. Its effect is very small
> (but much faster), compared to the main coil. It too has to be properly
> driven, of course.
>
> In a down-converting generator made for reaching low frequency, it would
> normally be set up to avoid spectral inversion - that's why there's a minimum
> output frequency that can be expected. If the 3.8 GHz LO is free-running, its
> drift may be enough to limit how far down you can go at the low end,
> regardless of the YTO's tuning precision. For sources that can get way down
> toward zero (DC even), both the LO and YTO need to be precisely controlled
> (synthesized).
>
> The coarse and fine tuning, compensation, and calibration all contribute to
> successful operation, so there may be quite a few possible suspects.
>
> If you have schematics, you should be able to find all of these elements, and
> track down the proper conditions for diagnostics and repair.
>
> Good luck.
>
> Ed
>
|
I'm writing up a full repair report but in the meanwhile, yes, it was the FM coil where I saw the instability.? This was a complex fault with an easy end resolution (once I knew what to do).
Peter
toggle quoted message
Show quoted text
On 3/25/2019 3:03 AM, Askild wrote: Hi Peter
I don't know this instrument, but here a couple of things I would check..
I cant remember if you already checked this, but can you connect a coupler to the 3.8GHz oscillator, to tap of a small bit and check that this is always stable? I guess it will be, but will be good to get it verified.
The YTO unlock instability, you should see on the FM coil drive.
Next I would follow the signal all the way to the PLL circuit. Check that it don't get attenuated to much at lower frequencies so the PLL can't lock to it.
Good luck with this.
Br, Askild
On Sun, Mar 24, 2019 at 10:29 PM Peter Gottlieb <hpnpilot@... <mailto:hpnpilot@...>> wrote:
I checked the FM coil drive and it is right about at zero so that seems like it's not the issue.
Nobody seems to know how the firmware works so we are left to guess and try things.? There is some range which the firmware can correct for and I heard from someone who needed to nudge his phase lock board to get his system into that range.? I tried the same thing and can get some measure of success as well, although apparently I have another problem causing me grief. At this point I have this temporary fix in which permits the unit to run on default pretune values but won't pass test 48 which sets more precise pretune values.
Below about 50 MHz I have a PLL instability which gets worse as the frequency decreases.? The PLL begins to lose lock below about 25 MHz.? I took two photos of the unit output when set to 25 MHz in CW mode showing the instability, then I put it into source tune mode which opens the phase lock loop. See attached. Notice how the output is nice and stable as soon as the loop is opened.
Not sure what to make of this.? The YTO main coil drive doesn't look any noisier at 25 MHz than it does at higher frequencies.
On 3/24/2019 3:28 PM, Ed Breya via Groups.Io wrote: > I don't know if this will help, but here goes. > > First, don't worry about the apparent offset in coil tuning current. YTOs are > only specified to operate over a certain range, like an octave or two, but can > run a ways beyond at either end. It doesn't help (other than mathematically to > get the tuning chart numbers) to extrapolate down to zero coil current - the > YTO will stall out somewhere a little below its lowest spec frequency and coil > current. Any slope variations or offsets extrapolated at zero just represent > the character of the magnetic structure in the YTO. > > The proper range, slope, and offset needed for tuning a particular YTO are > typically handled by the YTO driver circuit, which can make it so the net > result is a nice convenient GHz/V type system, at its input. There are usually > sufficient adjustments (pots and/or DAC settings) in the driver circuitry to > accommodate the tolerances of the YTO, the 3.8 GHz oscillator in this case, > and the rest of the parts, to make everything come out right, during calibration. > > There may be circuitry included to improve the tuning linearity (due to > magnetic saturation), and compensate for the dynamic response (magnetic delay > and hysteresis). Ideally, the YTO frequency is perfectly and instantly > proportional to the magnetic field, but in reality the field depends on an > iron core, coil inductance, AC core losses, and so forth. > > The FM coil may be used in a PLL, for fine tuning. Its effect is very small > (but much faster), compared to the main coil. It too has to be properly > driven, of course. > > In a down-converting generator made for reaching low frequency, it would > normally be set up to avoid spectral inversion - that's why there's a minimum > output frequency that can be expected. If the 3.8 GHz LO is free-running, its > drift may be enough to limit how far down you can go at the low end, > regardless of the YTO's tuning precision. For sources that can get way down > toward zero (DC even), both the LO and YTO need to be precisely controlled > (synthesized). > > The coarse and fine tuning, compensation, and calibration all contribute to > successful operation, so there may be quite a few possible suspects. > > If you have schematics, you should be able to find all of these elements, and > track down the proper conditions for diagnostics and repair. > > Good luck. > > Ed >
|