Keyboard Shortcuts
ctrl + shift + ? :
Show all keyboard shortcuts
ctrl + g :
Navigate to a group
ctrl + shift + f :
Find
ctrl + / :
Quick actions
esc to dismiss
Likes
Search
8596E Spectrum Analyzer Fault
Hi guys,
?
In the course of investigating some IFTs today, I fired up my 8596E, which started-up normally and displayed the usual min thru ~12Ghz spectrum. However, when I tried to enter the frequency range I wanted to study, it froze up and just would not respond. I was able to enter the start frequency okay, but the very second I entered the stop one, it just locked-up and became super-sluggish and unresponsive to any further input. At the instant, I got a "FREQ UNCAL" message appear on the screen. A *possible* clue is the 'OVEN' message takes a lot longer to go away than it has previously when it was all working fine. However, even when fully warmed up, I am still experiencing the thing freezing-up as described above.
Any ideas what the fault might be?
?
Cheers,
Jinxie |
toggle quoted message
Show quoted text
On Feb 19, 2025, at 10:12?AM, Jinxie via groups.io <paul666@...> wrote:
|
开云体育Yes, it can. ?I’ve had several similar type HP front panels do that. Disassembly, lots of q-tips with IPA cleaning panel holes and keys then putting everything back together solves it, if that’s the problem. ?No real cost except an hour of your time.?Peter On Feb 19, 2025, at 2:45?PM, Jinxie via groups.io <paul666@...> wrote:
|
This doesn't sound like a dirty keyboard issue to me. ?Before digging into the keyboard for cleaning, I would get it in this state and then unplug the keyboard to see if the issue continues.
?
I think the processor has gotten swamped dealing with something unexpected. ?Maybe it's busy responding to a peripheral issuing high speed erroneous interrupts, or stuck in a high-priority loop waiting for a response from a peripheral.
?
Missed interrupts because the processor is busy with higher priority tasks would explain why the oven is taking longer than usual. ?There's no sensor checking the oven temperature; it's just a timer and that's likely interrupt driven.
?
The keyboard appears to be polled (and not interrupt driven) which is why I think it's not a primary suspect.? I definitely agree the power supply is the first thing to check.
?
-mark
?
?
On Wed, Feb 19, 2025 at 01:11 PM, Jinxie wrote:
|
Thank you, Mark. I would have thought if it was the power supply then the issue would arise spontaneously at any time and not just precisely when I try to enter a frequency limit. But I may be wrong. For sure the keys are very dirty judging by the amount of pressure required to get them to actuate, so I shall clean them first in any event, since they certainly need it. I'll follow up with the other tests once that's out of the way and if the problem persists.
?
J. |
Okay, well, I've been able to bring on the fault condition by simply rotating the datum dial whilst in the 'stop frequency' mode, so dirty keys are now looking increasingly unlikely (though they will require attention).
Since the consensus (such as it is!) is to investigate the PSU, I'll do that next. I'll scope the power rails of the individual low voltage supplies out of the main PSU and check their voltage levels are within spec. That should trap 99% of PSU problems. I'll also check the ESRs of the larger electrolytics and any suspicious bulges whilst I'm at it. Anything else I should look at? |
Response to commands is usually postponed until the end of the current sweep. If the synth is having trouble locking, or the sweep is slow due to narrowband, or perhaps crossing band edges, command response will be affected. Try adjusting start / stop / span to see if there is a particular range that is slowing things down. And set autocouple to all... |
Instead of START / STOP, try entering as CENTER and SPAN. ?Just to see if it freezes when it has enough info to start a sweep, or if it is related specifically to the sweep defined by entering the STOP frequency.
?
Does it sweep when first powered on, or after preset..?
?
Trying to get at whether this is a problem with the sweep synthesis set up, or something related to keypress handling... |
to navigate to use esc to dismiss