54110D scope power supply problems


nn2027299
 

Dear All,

I have a 54110D scope which exhibits power supply problems. I've
followed the troubleshooting guide in the 54111D service manual
which leads me to believe that the analog board is faulty.

Has anyone got any schematics for these scopes, or a spare board
kicking around they'd like to sell?

Thanks in advance,

Ed.


guyroyet
 

I own the manual of servicing the 54111D model .There is only a
block diagram for the 3 supplies:
One is general and the 2 others are for "digital" and for "analogic"
I will receive a second 54111D scope .
I suspect HP to have done economy on electrochimical capacitors .
Mine has a large ripple superposed and the trace oscillate .
Try to change these large capacitors .I can send the diagram ,but
there is no schematic .I am in France and if you are in the States
it is difficult to meet !But if you go to France I will glad to meet
you.I live in the French Riviera .
I know that HP editated a second service manual for this device ,
perhaps more complete .Mine is numbered 90902.
Best regards.Guy Royet,88 bd de Bazeilles,83480 Puget sur Argens ,
France .Email : guy.royet@...

--- In hp_agilent_equipment@..., "nn2027299"
<silvermoon.group@n...> wrote:
Dear All,

I have a 54110D scope which exhibits power supply problems. I've
followed the troubleshooting guide in the 54111D service manual
which leads me to believe that the analog board is faulty.

Has anyone got any schematics for these scopes, or a spare board
kicking around they'd like to sell?

Thanks in advance,

Ed.