Hi all,
one of my other projects I'm working on is a 5342A which does not count on the high frequency input.
I checked the sampler according to the service manual. I used a 34401A for the resistance meausrement and got 22k in forward bias and >3M in reverse bias. This is quite different from the values stated in the service manual; however, when I look at the schematic, I think it should be possible to test the sampler with the diode test function. I did that, and I found out that in forward bias, I have approx. 0.4 volts, and reverse bias is open. So it looks like a pretty nice diode. However I cannot compare to other samplers, and the service manual also tells that the resistance measured can vary greatly if not exact 1mA measuring current is used.
Does that sampler look good to you guys, or is it probably broken? because if it is broken I think it makes no sense to invest additional time to fix things.

Tobias, HB9FSX


 

Hi Tobias
Check out?http://www.azurelectronics.com/Repair%20HP%205342A%20Microwave%20Frequency.htm as it may help.
Regards
Steve

On ?Monday?, ?October? ?16?, ?2017? ?08?:?56?:?32? ?AM? ?BST, Tobias Pluess tobias.pluess@... [hp_agilent_equipment] wrote:


?

Hi all,
one of my other projects I'm working on is a 5342A which does not count on the high frequency input.
I checked the sampler according to the service manual. I used a 34401A for the resistance meausrement and got 22k in forward bias and >3M in reverse bias. This is quite different from the values stated in the service manual; however, when I look at the schematic, I think it should be possible to test the sampler with the diode test function. I did that, and I found out that in forward bias, I have approx. 0.4 volts, and reverse bias is open. So it looks like a pretty nice diode. However I cannot compare to other samplers, and the service manual also tells that the resistance measured can vary greatly if not exact 1mA measuring current is used.
Does that sampler look good to you guys, or is it probably broken? because if it is broken I think it makes no sense to invest additional time to fix things.

Tobias, HB9FSX


 

开云体育

Hi Tobias,

Do check out Azure, it’s a good site! I have a couple of 5342A parts units and should have a good sampler and other bits.

73,

Steve
WB0DBS



On Oct 17, 2017, at 2:31 AM, Steve A starmux64@... [hp_agilent_equipment] <hp_agilent_equipment@...> wrote:

?

Hi Tobias
Check out? as it may help.
Regards
Steve

On ?Monday?, ?October? ?16?, ?2017? ?08?:?56?:?32? ?AM? ?BST, Tobias Pluess tobias.pluess@... [hp_agilent_equipment] <hp_agilent_equipment@...> wrote:


?

Hi all,
one of my other projects I'm working on is a 5342A which does not count on the high frequency input.
I checked the sampler according to the service manual. I used a 34401A for the resistance meausrement and got 22k in forward bias and >3M in reverse bias. This is quite different from the values stated in the service manual; however, when I look at the schematic, I think it should be possible to test the sampler with the diode test function. I did that, and I found out that in forward bias, I have approx. 0.4 volts, and reverse bias is open. So it looks like a pretty nice diode. However I cannot compare to other samplers, and the service manual also tells that the resistance measured can vary greatly if not exact 1mA measuring current is used.
Does that sampler look good to you guys, or is it probably broken? because if it is broken I think it makes no sense to invest additional time to fix things.

Tobias, HB9FSX


 

Hi,

Your measurements with the diode mode suggest the sampler is OK, digital meters don't use the same amount of current to test resistance as older analogue meters and that's the reason you didn't get their readings.


I agree with Steve, you may need to replace a number of the 1826-0371, either on the A8 main VCO, A4 offset VCO or both.


I'd also check the A5 output to see if you're getting drive to the sampler driver and then check the sampler driver A26 itself.


I'd check in that order, check the VCOs are outputting signal, check the A5 to ensure it's outputting a signal to?A26 and it should alternate between the two VCOs, then check A26.


To check the sampler driver A26?you really need a?spectrum analyser. As most spectrum analysers have a capacitor (DC block) in the front end you need to terminate the A26 module with 50 Ohm and then measure what you see across that with the spec an, if you see lots of harmonics then all is good. You will not see anything?(apart from the 325MHz drive) without a dc return for the step recovery diode.


If the A26 is not working that could either be the SRD or it could be U1 on the A26 module - I have seen faulty U1 ICs in the past, most likely due to heat - they run very hot.


For the 1826-0371, you can buy fakes/replicas/copies off ebay or do what I do, I make a small smd pcb with 2 x 2sc3356 transistors, 2 x 100 ohm resistors and a 120 ohm resistor in the emitter lead?(all smd components)?with pins to fit in the DIP footprint. You'll see the schematic for the 1826-0371 on a number of the 5342A schematics. I've not had any issue with my made up boards yet.


If you find it is U1 on the A5 module I can also give you an alternative circuit for that but it is not quite optimised yet ?and?requires short leads to minimise parasitics.


Good luck

Greg


 

开云体育

Hi all,

problem solved: as I said, the counter was counting fine on the low frequency input, but on the high frequency input, it displayed only zeros. The reason was that
a) the -5.2V operating voltage was out of the spec
b) the solder joint between the semirigid coax to the front panel's N connector cracked, thus the semirigid was no longer connected to the connector. I re-soldered and now it counts fine (only tested with 2 GHz so far).

I wonder whether it is possible to replace the ROM. It would be nice to add option 004, the DAC. I have couple of DAC80 here, however, opt. 004 needs a different ROM, unfortunately.

I also read that there was an option 005, which extends the frequency range to 24 GHz. Does someone know what option 005 actually was? was it a different sampler, or something that can easily be changed?

Tobias


From: hp_agilent_equipment@... [hp_agilent_equipment@...]
Sent: Wednesday, October 18, 2017 04:37
To: hp_agilent_equipment@...
Subject: [hp_agilent_equipment] Re: 5342A: blown sampler?

?

Hi,

Your measurements with the diode mode suggest the sampler is OK, digital meters don't use the same amount of current to test resistance as older analogue meters and that's the reason you didn't get their readings.


I agree with Steve, you may need to replace a number of the 1826-0371, either on the A8 main VCO, A4 offset VCO or both.


I'd also check the A5 output to see if you're getting drive to the sampler driver and then check the sampler driver A26 itself.


I'd check in that order, check the VCOs are outputting signal, check the A5 to ensure it's outputting a signal to?A26 and it should alternate between the two VCOs, then check A26.


To check the sampler driver A26?you really need a?spectrum analyser. As most spectrum analysers have a capacitor (DC block) in the front end you need to terminate the A26 module with 50 Ohm and then measure what you see ac ross that with the spec an, if you see lots of harmonics then all is good. You will not see anything?(apart from the 325MHz drive) without a dc return for the step recovery diode.


If the A26 is not working that could either be the SRD or it could be U1 on the A26 module - I have seen faulty U1 ICs in the past, most likely due to heat - they run very hot.


For the 1826-0371, you can buy fakes/replicas/copies off ebay or do what I do, I make a small smd pcb with 2 x 2sc3356 transistors, 2 x 100 ohm resistors and a 120 ohm resistor in the emitter lead?(all smd components)?with pins to fit in the DIP footprint. You'll see the schematic for the 1826-0371 on a number of the 5342A schematics. I've not had any issue with my made up boards yet.


If you find it is U1 on the A5 module I can also give you an alternative circuit for that but it is not quite optimised yet ?and?requires short leads to minimise parasi tics.


Good luck

Greg


Dr. David Kirkby - Kirkby Microwave Ltd
 

The connnector on the higher frequency model was 3.5 mm, not N. I don't know if there are other changes

On 10/20/17 01:05 PM, Tobias Pluess tobias.pluess@... [hp_agilent_equipment] wrote:
Hi all,

problem solved: as I said, the counter was counting fine on the low frequency input, but on the high frequency input, it displayed only zeros. The reason was that
a) the -5.2V operating voltage was out of the spec
b) the solder joint between the semirigid coax to the front panel's N connector cracked, thus the semirigid was no longer connected to the connector. I re-soldered and now it counts fine (only tested with 2 GHz so far).

I wonder whether it is possible to replace the ROM. It would be nice to add option 004, the DAC. I have couple of DAC80 here, however, opt. 004 needs a different ROM, unfortunately.

I also read that there was an option 005, which extends the frequency range to 24 GHz. Does someone know what option 005 actually was? was it a different sampler, or something that can easily be changed?

Tobias

________________________________
From: hp_agilent_equipment@... [hp_agilent_equipment@...]
Sent: Wednesday, October 18, 2017 04:37
To: hp_agilent_equipment@...
Subject: [hp_agilent_equipment] Re: 5342A: blown sampler?



Hi,

Your measurements with the diode mode suggest the sampler is OK, digital meters don't use the same amount of current to test resistance as older analogue meters and that's the reason you didn't get their readings.


I agree with Steve, you may need to replace a number of the 1826-0371, either on the A8 main VCO, A4 offset VCO or both.


I'd also check the A5 output to see if you're getting drive to the sampler driver and then check the sampler driver A26 itself.


I'd check in that order, check the VCOs are outputting signal, check the A5 to ensure it's outputting a signal to A26 and it should alternate between the two VCOs, then check A26.


To check the sampler driver A26 you really need a spectrum analyser. As most spectrum analysers have a capacitor (DC block) in the front end you need to terminate the A26 module with 50 Ohm and then measure what you see ac ross that with the spec an, if you see lots of harmonics then all is good. You will not see anything (apart from the 325MHz drive) without a dc return for the step recovery diode.


If the A26 is not working that could either be the SRD or it could be U1 on the A26 module - I have seen faulty U1 ICs in the past, most likely due to heat - they run very hot.


For the 1826-0371, you can buy fakes/replicas/copies off ebay or do what I do, I make a small smd pcb with 2 x 2sc3356 transistors, 2 x 100 ohm resistors and a 120 ohm resistor in the emitter lead (all smd components) with pins to fit in the DIP footprint. You'll see the schematic for the 1826-0371 on a number of the 5342A schematics. I've not had any issue with my made up boards yet.


If you find it is U1 on the A5 module I can also give you an alternative circuit for that but it is not quite optimised yet and requires short leads to minimise parasi tics.


Good luck

Greg


--
Dr. David Kirkby Ph.D CEng MIET
Kirkby Microwave Ltd
Registered office: Stokes Hall Lodge, Burnham Rd, Althorne, Essex, CM3 6DT, UK.
Registered in England and Wales, company number 08914892.

Tel: 07910 441670 / +44 7910 441670 (0900 to 2100 GMT only please)


 

Hello Tobias,

Congrats on the repair! Happy for you the sampler was fine! I can only comment on the frequency range. I have a number of HP equipment (including the HP5342A) that are spec'd to 20-22 GHz and with an (expensive) option that range can be extended to 26 GHz. The only hw change is the input connector that goes from N to APC3.5. No other change! Try the counter at 24 GHz and you will see it will work!
Good luck!

Harke


On Friday, October 20, 2017, 2:05:31 PM GMT+2, Tobias Pluess tobias.pluess@... [hp_agilent_equipment] wrote:


?

Hi all,

problem solved: as I said, the counter was counting fine on the low frequency input, but on the high frequency input, it displayed only zeros. The reason was that
a) the -5.2V operating voltage was out of the spec
b) the solder joint between the semirigid coax to the front panel's N connector cracked, thus the semirigid was no longer connected to the connector. I re-soldered and now it counts fine (only tested with 2 GHz so far).

I wonder whether it is possible to replace the ROM. It would be nice to add option 004, the DAC. I have couple of DAC80 here, however, opt. 004 needs a different ROM, unfortunately.

I also read that there was an option 005, which extends the frequency range to 24 GHz. Does someone know what option 005 actually was? was it a different sampler, or something that can easily be changed?

Tobias


From: hp_agilent_equipment@... [hp_agilent_equipment@...]
Sent: Wednesday, October 18, 2017 04:37
To: hp_agilent_equipment@...
Subject: [hp_agilent_equipment] Re: 5342A: blown sampler?

?

Hi,

Your measurements with the diode mode suggest the sampler is OK, digital meters don't use the same amount of current to test resistance as older analogue meters and that's the reason you didn't get their readings.


I agree with Steve, you may need to replace a number of the 1826-0371, either on the A8 main VCO, A4 offset VCO or both.


I'd also check the A5 output to see if you're getting drive to the sampler driver and then check the sampler driver A26 itself.


I'd check in that order, check the VCOs are outputting signal, check the A5 to ensure it's outputting a signal to?A26 and it should alternate between the two VCOs, then check A26.


To check the sampler driver A26?you really need a?spectrum analyser. As most spectrum analysers have a capacitor (DC block) in the front end you need to terminate the A26 module with 50 Ohm and then measure what you see ac ross that with the spec an, if you see lots of harmonics then all is good. You will not see anything?(apart from the 325MHz drive) without a dc return for the step recovery diode.


If the A26 is not working that could either be the SRD or it could be U1 on the A26 module - I have seen faulty U1 ICs in the past, most likely due to heat - they run very hot.


For the 1826-0371, you can buy fakes/replicas/copies off ebay or do what I do, I make a small smd pcb with 2 x 2sc3356 transistors, 2 x 100 ohm resistors and a 120 ohm resistor in the emitter lead?(all smd components)?with pins to fit in the DIP footprint. You'll see the schematic for the 1826-0371 on a number of the 5342A schematics. I've not had any issue with my made up boards yet.


If you find it is U1 on the A5 module I can also give you an alternative circuit for that but it is not quite optimised yet ?and?requires short leads to minimise parasi tics.


Good luck

Greg


 

开云体育

Hi Harke,

oh, interesting. So the upgrade to 26 GHz, which would actually be a 5343A, changes only the connector? :-)
This is funny, because I use a HP 8563E spectrum analyzer at work, which operates fine at 26.5 GHz using a N connector!

Best
Tobias



From: hp_agilent_equipment@... [hp_agilent_equipment@...]
Sent: Friday, October 20, 2017 14:13
To: hp_agilent_equipment@...
Subject: Re: RE: [hp_agilent_equipment] Re: 5342A: blown sampler?

?

Hello Tobias,

Congrats on the repair! Happy for you the sampler was fine! I can only comment on the frequency range. I have a number of HP equipment (including the HP5342A) that are spec'd to 20-22 GHz and with an (expensive) option that range can be extended to 26 GHz. The only hw change is the input connector that goes from N to APC3.5. No other change! Try the counter at 24 GHz and you will see it will work!
Good luck!

Harke


On Friday, October 20, 2017, 2:05:31 PM GMT+2, Tobias Pluess tobias.pluess@... [hp_agilent_equipment] wrote:


?

Hi all,

problem solved: as I said, the counter was counting fine on the low frequency input, but on the high frequency input, it displayed only zeros. The reason was that
a) the -5.2V operating voltage was out of the spec
b) the solder joint between the semirigid coax to the front panel's N connector cracked, thus the semirigid was no longer connected to the connector. I re-soldered and now it counts fine (only tested with 2 GHz so far).

I wonder whether it is possible to replace the ROM. It would be nice to add option 004, the DAC. I have couple of DAC80 here, however, opt. 004 needs a different ROM, unfortunately.

I also read that there was an option 005, which extends the frequency range to 24 GHz. Does someone know what option 005 actually was? was it a different sampler, or something that can easily be changed?

Tobias


From: hp_agilent_equipment@... [hp_agilent_equipment@...]
Sent: Wednesday, October 18, 2017 04:37
To: hp_agilent_equipment@...
Subject: [hp_agilent_equipment] Re: 5342A: blown sampler?

?

Hi,

Your measurements with the diode mode suggest the sampler is OK, digital meters don't use the same amount of current to test resistance as older analogue meters and that's the reason you didn't get their readings.


I agree with Steve, you may need to replace a number of the 1826-0371, either on the A8 main VCO, A4 offset VCO or both.


I'd also check the A5 output to see if you're getting drive to the sampler driver and then check the sampler driver A26 itself.


I'd check in that order, check the VCOs are outputting signal, check the A5 to ensure it's outputting a signal to?A26 and it should alternate between the two VCOs, then check A26.


To check the sampler driver A26?you really need a?spectrum analyser. As most spectrum analysers have a capacitor (DC block) in the front end you need to terminate the A26 module with 50 Ohm and then measure what you see ac ross that with the spec an, if you see lots of harmonics then all is good. You will not see anything?(apart from the 325MHz drive) without a dc return for the step recovery diode.


If the A26 is not working that could either be the SRD or it could be U1 on the A26 module - I have seen faulty U1 ICs in the past, most likely due to heat - they run very hot.


For the 1826-0371, you can buy fakes/replicas/copies off ebay or do what I do, I make a small smd pcb with 2 x 2sc3356 transistors, 2 x 100 ohm resistors and a 120 ohm resistor in the emitter lead?(all smd components)?with pins to fit in the DIP footprint. You'll see the schematic for the 1826-0371 on a number of the 5342A schematics. I've not had any issue with my made up boards yet.


If you find it is U1 on the A5 module I can also give you an alternative circuit for that but it is not quite optimised yet ?and?requires short leads to minimise parasi tics.


Good luck

Greg


 

开云体育

I’ve done the connector change and, using a good quality SMA, I have two 5342As that count reliably to 26GHz and one that just won’t count reliably above about 21GHz. I don’t have a 5343A to compare to but I wonder if maybe they were just hand-selected 5342As?

73,

Steve
WB0DBS



On Oct 20, 2017, at 7:38 AM, Tobias Pluess tobias.pluess@... [hp_agilent_equipment] <hp_agilent_equipment@...> wrote:

?

Hi Harke,

oh, interesting. So the upgrade to 26 GHz, which would actually be a 5343A, changes only the connector? :-)
This is funny, because I use a HP 8563E spectrum analyzer at work, which operates fine at 26.5 GHz using a N connector!

Best
Tobias



From: hp_agilent_equipment@... [hp_agilent_equipment@...]
Sent: Friday, October 20, 2017 14:13
To: hp_agilent_equipment@...
Subject: Re: RE: [hp_agilent_equipment] Re: 5342A: blown sampler?

?

Hello Tobias,

Congrats on the repair! Happy for you the sampler was fine! I can only comment on the frequency range. I have a number of HP equipment (including the HP5342A) that are spec'd to 20-22 GHz and with an (expensive) option that range can be extended to 26 GHz. The only hw change is the input connector that goes from N to APC3.5. No other change! Try the counter at 24 GHz and you will see it will work!
Good luck!

Harke


On Friday, October 20, 2017, 2:05:31 PM GMT+2, Tobias Pluess tobias.pluess@... [hp_agilent_equipment] <hp_agilent_equipment@...> wrote:


?

Hi all,

problem solved: as I said, the counter was counting fine on the low frequency input, but on the high frequency input, it displayed only zeros. The reason was that
a) the -5.2V operating voltage was out of the spec
b) the solder joint between the semirigid coax to the front panel's N connector cracked, thus the semirigid was no longer connected to the connector. I re-soldered and now it counts fine (only tested with 2 GHz so far).

I wonder whether it is possible to replace the ROM. It would be nice to add option 004, the DAC. I have couple of DAC80 here, however, opt. 004 needs a different ROM, unfortunately.

I also read that there was an option 005, which extends the frequency range to 24 GHz. Does someone know what option 005 actually was? was it a different sampler, or something that can easily be changed?

Tobias


From: hp_agilent_equipment@... [hp_agilent_equipment@...]
Sent: Wednesday, October 18, 2017 04:37
To: hp_agilent_equipment@...
Subject: [hp_agilent_equipment] Re: 5342A: blown sampler?

?

Hi,

Your measurements with the diode mode suggest the sampler is OK, digital meters don't use the same amount of current to test resistance as older analogue meters and that's the reason you didn't get their readings.


I agree with Steve, you may need to replace a number of the 1826-0371, either on the A8 main VCO, A4 offset VCO or both.


I'd also check the A5 output to see if you're getting drive to the sampler driver and then check the sampler driver A26 itself.


I'd check in that order, check the VCOs are outputting signal, check the A5 to ensure it's outputting a signal to?A26 and it should alternate between the two VCOs, then check A26.


To check the sampler driver A26?you really need a?spectrum analyser. As most spectrum analysers have a capacitor (DC block) in the front end you need to terminate the A26 module with 50 Ohm and then measure what you see ac ross that with the spec an, if you see lots of harmonics then all is good. You will not see anything?(apart from the 325MHz drive) without a dc return for the step recovery diode.


If the A26 is not working that could either be the SRD or it could be U1 on the A26 module - I have seen faulty U1 ICs in the past, most likely due to heat - they run very hot.


For the 1826-0371, you can buy fakes/replicas/copies off ebay or do what I do, I make a small smd pcb with 2 x 2sc3356 transistors, 2 x 100 ohm resistors and a 120 ohm resistor in the emitter lead?(all smd components)?with pins to fit in the DIP footprint. You'll see the schematic for the 1826-0371 on a number of the 5342A schematics. I've not had any issue with my made up boards yet.


If you find it is U1 on the A5 module I can also give you an alternative circuit for that but it is not quite optimised yet ?and?requires short leads to minimise parasi tics.


Good luck

Greg


 

I have a 5342A with option 005 which extends operation from 18GHz to 24GHz. It has the original N-type connector though. So I expect there must be other hardware changes, or as Steve K suggests selected parts.? The option cost $500 extra.

Steve


On ?Friday?, ?October? ?20?, ?2017? ?02?:?00?:?22? ?PM? ?BST, Steve - Home Steve-Krull@... [hp_agilent_equipment] wrote:


?

I’ve done the connector change and, using a good quality SMA, I have two 5342As that count reliably to 26GHz and one that just won’t count reliably above about 21GHz. I don’t have a 5343A to compare to but I wonder if maybe they were just hand-selected 5342As?


73,

Steve
WB0DBS



On Oct 20, 2017, at 7:38 AM, Tobias Pluess tobias.pluess@... [hp_agilent_equipment] <hp_agilent_equipment@...> wrote:

?

Hi Harke,

oh, interesting. So the upgrade to 26 GHz, which would actually be a 5343A, changes only the connector? :-)
This is funny, because I use a HP 8563E spectrum analyzer at work, which operates fine at 26.5 GHz using a N connector!

Best
Tobias



From: hp_agilent_equipment@... [hp_agilent_equipment@...]
Sent: Friday, October 20, 2017 14:13
To: hp_agilent_equipment@...
Subject: Re: RE: [hp_agilent_equipment] Re: 5342A: blown sampler?

?

Hello Tobias,

Congrats on the repair! Happy for you the sampler was fine! I can only comment on the frequency range. I have a number of HP equipment (including the HP5342A) that are spec'd to 20-22 GHz and with an (expensive) option that range can be extended to 26 GHz. The only hw change is the input connector that goes from N to APC3.5. No other change! Try the counter at 24 GHz and you will see it will work!
Good luck!

Harke


On Friday, October 20, 2017, 2:05:31 PM GMT+2, Tobias Pluess tobias.pluess@... [hp_agilent_equipment] <hp_agilent_equipment@...> wrote:


?

Hi all,

problem solved: as I said, the counter was counting fine on the low frequency input, but on the high frequency input, it displayed only zeros. The reason was that
a) the -5.2V operating voltage was out of the spec
b) the solder joint between the semirigid coax to the front panel's N connector cracked, thus the semirigid was no longer connected to the connector. I re-soldered and now it counts fine (only tested with 2 GHz so far).

I wonder whether it is possible to replace the ROM. It would be nice to add option 004, the DAC. I have couple of DAC80 here, however, opt. 004 needs a different ROM, unfortunately.

I also read that there was an option 005, which extends the frequency range to 24 GHz. Does someone know what option 005 actually was? was it a different sampler, or something that can easily be changed?

Tobias


From: hp_agilent_equipment@... [hp_agilent_equipment@...]
Sent: Wednesday, October 18, 2017 04:37
To: hp_agilent_equipment@...
Subject: [hp_agilent_equipment] Re: 5342A: blown sampler?

?

Hi,

Your measurements with the diode mode suggest the sampler is OK, digital meters don't use the same amount of current to test resistance as older analogue meters and that's the reason you didn't get their readings.


I agree with Steve, you may need to replace a number of the 1826-0371, either on the A8 main VCO, A4 offset VCO or both.


I'd also check the A5 output to see if you're getting drive to the sampler driver and then check the sampler driver A26 itself.


I'd check in that order, check the VCOs are outputting signal, check the A5 to ensure it's outputting a signal to?A26 and it should alternate between the two VCOs, then check A26.


To check the sampler driver A26?you really need a?spectrum analyser. As most spectrum analysers have a capacitor (DC block) in the front end you need to terminate the A26 module with 50 Ohm and then measure what you see ac ross that with the spec an, if you see lots of harmonics then all is good. You will not see anything?(apart from the 325MHz drive) without a dc return for the step recovery diode.


If the A26 is not working that could either be the SRD or it could be U1 on the A26 module - I have seen faulty U1 ICs in the past, most likely due to heat - they run very hot.


For the 1826-0371, you can buy fakes/replicas/copies off ebay or do what I do, I make a small smd pcb with 2 x 2sc3356 transistors, 2 x 100 ohm resistors and a 120 ohm resistor in the emitter lead?(all smd components)?with pins to fit in the DIP footprint. You'll see the schematic for the 1826-0371 on a number of the 5342A schematics. I've not had any issue with my made up boards yet.


If you find it is U1 on the A5 module I can also give you an alternative circuit for that but it is not quite optimised yet ?and?requires short leads to minimise parasi tics.


Good luck

Greg


 

开云体育

My counter counts reliably up to 20 GHz. I could not yet test higher frequencies.
I wonder whether somebody has a ROM dump of it?
It would be nice if I could add the DAC option, but this requires a newer ROM which my unit does not have.

Tobias


-------- Original message --------
From: "Steve A starmux64@... [hp_agilent_equipment]"
Date: 10/21/17 10:48 (GMT+01:00)
To: "Steve - Home Steve-Krull@... [hp_agilent_equipment]"
Subject: Re: [hp_agilent_equipment] Re: 5342A: blown sampler?

?

I have a 5342A with option 005 which extends operation from 18GHz to 24GHz. It has the original N-type connector though. So I expect there must be other hardware changes, or as Steve K suggests selected parts.? The option cost $500 extra.

Steve


On ?Friday?, ?October? ?20?, ?2017? ?02?:?00?:?22? ?PM? ?BST, Steve - Home Steve-Krull@... [hp_agilent_equipment] wrote:


?

I’ve done the connector change and, using a good quality SMA, I have two 5342As that count reliably to 26GHz and one that just won’t count reliably above about 21GHz. I don’t have a 5343A to compare to but I wonder if maybe they were just hand-selected 5342As?


73,

Steve
WB0DBS



On Oct 20, 2017, at 7:38 AM, Tobias Pluess tobias.pluess@... [hp_agilent_equipment] <hp_agilent_equipment@...> wrote:

?

Hi Harke,

oh, interesting. So the upgrade to 26 GHz, which would actually be a 5343A, changes only the connector? :-)
This is funny, because I use a HP 8563E spectrum analyzer at work, which operates fine at 26.5 GHz using a N connector!

Best
Tobias



From: hp_agilent_equipment@... [hp_agilent_equipment@...]
Sent: Friday, October 20, 2017 14:13
To: hp_agilent_equipment@...
Subject: Re: RE: [hp_agilent_equipment] Re: 5342A: blown sampler?

?

Hello Tobias,

Congrats on the repair! Happy for you the sampler was fine! I can only comment on the frequency range. I have a number of HP equipment (including the HP5342A) that are spec'd to 20-22 GHz and with an (expensive) option that range can be extended to 26 GHz. The only hw change is the input connector that goes from N to APC3.5. No other change! Try the counter at 24 GHz and you will see it will work!
Good luck!

Harke


On Friday, October 20, 2017, 2:05:31 PM GMT+2, Tobias Pluess tobias.pluess@... [hp_agilent_equipment] <hp_agilent_equipment@...> wrote:


?

Hi all,

problem solved: as I said, the counter was counting fine on the low frequency input, but on the high frequency input, it displayed only zeros. The reason was that
a) the -5.2V operating voltage was out of the spec
b) the solder joint between the semirigid coax to the front panel's N connector cracked, thus the semirigid was no longer connected to the connector. I re-soldered and now it counts fine (only tested with 2 GHz so far).

I wonder whether it is possible to replace the ROM. It would be nice to add option 004, the DAC. I have couple of DAC80 here, however, opt. 004 needs a different ROM, unfortunately.

I also read that there was an option 005, which extends the frequency range to 24 GHz. Does someone know what option 005 actually was? was it a different sampler, or something that can easily be changed?

Tobias


From: hp_agilent_equipment@... [hp_agilent_equipment@...]
Sent: Wednesday, October 18, 2017 04:37
To: hp_agilent_equipment@...
Subject: [hp_agilent_equipment] Re: 5342A: blown sampler?

?

Hi,

Your measurements with the diode mode suggest the sampler is OK, digital meters don't use the same amount of current to test resistance as older analogue meters and that's the reason you didn't get their readings.


I agree with Steve, you may need to replace a number of the 1826-0371, either on the A8 main VCO, A4 offset VCO or both.


I'd also check the A5 output to see if you're getting drive to the sampler driver and then check the sampler driver A26 itself.


I'd check in that order, check the VCOs are outputting signal, check the A5 to ensure it's outputting a signal to?A26 and it should alternate between the two VCOs, then check A26.


To check the sampler driver A26?you really need a?spectrum analyser. As most spectrum analysers have a capacitor (DC block) in the front end you need to terminate the A26 module with 50 Ohm and then measure what you see ac ross that with the spec an, if you see lots of harmonics then all is good. You will not see anything?(apart from the 325MHz drive) without a dc return for the step recovery diode.


If the A26 is not working that could either be the SRD or it could be U1 on the A26 module - I have seen faulty U1 ICs in the past, most likely due to heat - they run very hot.


For the 1826-0371, you can buy fakes/replicas/copies off ebay or do what I do, I make a small smd pcb with 2 x 2sc3356 transistors, 2 x 100 ohm resistors and a 120 ohm resistor in the emitter lead?(all smd components)?with pins to fit in the DIP footprint. You'll see the schematic for the 1826-0371 on a number of the 5342A schematics. I've not had any issue with my made up boards yet.


If you find it is U1 on the A5 module I can also give you an alternative circuit for that but it is not quite optimised yet ?and?requires short leads to minimise parasi tics.


Good luck

Greg