Keyboard Shortcuts
Likes
Search
3457A carbon composite resistors in signal path?
Having a closer look at the boards, I've noticed that the resistors are all on the various inputs, there are two more carbon composition resistors between the front binding posts and hybrid (for other functions and ranges) and also on the trigger input from the respective BNC on the rear panel. There is also one across the control data lines to the optional relay card too, so I think it is likely that they were selected for surge rating to handle any transients that may come in.
The True RMS converter board also has a few, but two are 100Meg resistors, which I believe would have been chosen due to production limitations of the time making such high value resistors unavailable in 1/4 watt size metal and carbon film resistors. As for calibration, I assumed this would be able to get done as the other models are still serviced by Keysight. For the 3457A, they just say to contact them, so I will have to do so and see what they say.... Jared |
The drift in the resistor values doesn't look too bad as it is about what you expect for carbon composition resistors of that age so I wouldn't replace them.
The small drift in values is going to have negligible impact of the accuracy of the meter as you will only get the input leakage current, probably a few pA, of the gates of the input FETs and their associated protection diodes flowing through them. The error contribution from the drift in these resistors is therefore only a few nanovolts and will barely nudge the LSD of the readings on the most sensitive range. You are likely to get far greater errors from thermal voltages in your test leads. I specifically wouldn't replace the carbon composition resistors with your suggested metal film types as these are likely to have significantly poorer surge overload performance than the original carbon composition types. If you really must replace these resistors you should use the original carbon composition types, if you can find them, or the more modern ceramic composition types, if you can afford them. |
开云体育I know from personal experience that the ceramics and film under surge current will typically be destroyed ( almost like RIFA caps just not as messy) ..hence? the Carbon comps...搁别苍é别 On 5/9/24 6:49 AM, Jared Cabot via
groups.io wrote:
|
Jared, The drift you're seeing is not excessive from that type (and age) of resistor. You could still replace them with brand new ones (the "for" for that would be long term?service and peace of mind - they'll only continue drifting from this point on, though regularly running a long-stored unit will help those resistors, as well as all parts, do better), you'd just have to be mindful of their role when choosing a type or resistor (per above points). I'd do the replacement, obviously, before sending the units in for calibration.? I checked on the US portal for Keysight services (which is where I generate quotations, etc.) and, for some reason, I do see the 3456A, I certainly?see the 3458A and also the 34401A (the latter two I know for a fact they are supported), but I don't see the 3457A. I'd be curious if they support it for cal where you are.? Radu.? On Thu, May 9, 2024 at 6:49?AM Jared Cabot via <jaredcabot=[email protected]> wrote:
|
Well, I lifted one leg of each resistor, let them cool to ambient and measured the resistance. |
The resistors are in a high impedance part of the circuit so the absolute value will not affect the accuracy of the readings. |
The carbon composition resistors were probably used with the
thought that they are rugged, and will only introduce a variation that is less than system noise. However, carbon composition resistors that are old, and subject to humidity do strange things. Those over 100K seem to always go down in value, and those below 100K go up in value. I don't know why, but they do. Old carbon composition resistors typically move much more than their tolerance would suggest... very much more..., which may push their effect from negligible into very noticeable. In the case when the resistor has been left in an unused state over a long time, in a device, when it is first powered on, and for many hours(days) thereafter, it will drift.... But, only in one direction. As an example, I have a 432B power meter that took 2 months of being powered on 24/7 before its drift slowed to an almost stop. -Chuck Harris On Wed, 08 May 2024 16:10:50 -0700 "Jared Cabot via groups.io" <jaredcabot@...> wrote: Thanks for the replies. |
Jared, A couple of points:
Also, I couldn't find this data, but 1ppm/C doesn't strike me as excessive from this instrument (particularly?given that the 3456A seems to have a TC of about 2ppm/C).? Now, if not too difficult to work on, you could try replacing them with some good types - maybe something like CMF60s or even Mills WWs - and see if you observe an improvement of performance.? Radu.? On Wed, May 8, 2024 at 4:10?PM Jared Cabot via <jaredcabot=[email protected]> wrote:
|
Thanks for the replies. ? I am inclined to leave the resistors alone, but I just keep wondering how much their temperature coefficient and value drift with age would be affecting the accuracy of the instrument? |
There's no need to change those 51 k resistors. HP had it all figured out, for proper and safe input overload handling and ESD. You'll probably notice they are fairly large (maybe 1 or 2W) in power rating. That comes into play in the event of sustained over-voltage on the input connection, where the voltage clamps in the front-end limit the circuit voltage, and the resistors limit the current and take the rest of the total voltage, and have to dissipate the power. To appreciate the size needed, just say, take the specified maximum input overload voltage squared over the total R, for total power. Normally, these parts seldom need to do their thing, but they're always ready. If they appear overheated or burnt, then the unit may have been subject to severe abuse.
Ed |
These resistors look like they are part of an overload protection circuit in conjunction with the clamping diodes in U101.? Carbon composition resistors are used here for their much better surge handling capacity than film type resistors. I wouldn't replace them with film resistors if you want to retain the best input protection for the (probably unobtainum) hybrid amplifier module.
The drift of a carbon composition resistors would have negligible effect on the accuracy of the circuit as they are in series with some very high impedance amplifier inputs. |
开云体育a lot of times carbon comps were used as they could handle surge I better and not go open搁别苍é别 On 5/7/24 6:08 PM, Richard Knoppow via
groups.io wrote:
?? composition resistors actually have more spurious reactance than film resistors. I have no idea why these were used but suspect carbon or metal film resistors would work at least as well. |
composition resistors actually have more spurious reactance than film resistors. I have no idea why these were used but suspect carbon or metal film resistors would work at least as well.
toggle quoted message
Show quoted text
On 5/7/2024 5:36 PM, Jared Cabot via groups.io wrote:
I recently performed a few repairs on a friends 3457A multimeter (new RAM battery, replacement current range switching relays and added thermal insulation on the LM339 vreg) and have also just acquired one of my own that I'll do the RAM battery replacement, LM339 thermal insulation and maybe power supply capacitor replacement on. --
Richard Knoppow Los Angeles WB6KBL SKCC 19998 |
I recently performed a few repairs on a friends 3457A multimeter (new RAM battery, replacement current range switching relays and added thermal insulation on the LM339 vreg) and have also just acquired one of my own that I'll do the RAM battery replacement, LM339 thermal insulation and maybe power supply capacitor replacement on.
I noticed there are a number of 51K carbon composition resistors in the signal path, which I found a bit strange... You can see them in the input switching section highlighted in yellow of the linked image below, R101, R102, R201, R202, R203, R204 As carbon composition resistors are prone to drift and noise with age, would it be of any benefit to replace these with metal film resistors? Or is there some reason carbon composite were used here (high pulse energy handling and low inductance etc)? Thanks! Jared |