HP 83592B hangs 8350B sweeper


 

My 8350B with 83592B plug-in turns on?with flickering displays that go blank?and hangs the 8350B with no error message,?and no response to any switches on either the plug-in or mainframe. While flickering for a few seconds the 8350B displays show -310 MHz start and 20 GHz stop.


It does the same to a second known working 8350B. All data and address lines are active when hung so I don't understand why the 8350B does not respond.


If I press CW while the displays are flickering it responds. I may have to power down and up again to then enter a CW frequency, it comes up showing CW. Once in this state I can run tests and all front panel entries work. I ran all the A3 tests which show the ROMs and address decoders are OK.


The frequency in CW is high by 13 to 47 MHz going from?50 MHz to 18 GHz (it covers 0.01 to 20 GHz).


The power is way off. The display shows -2 to +23 dBm. Is this normal? (no opt 002 attenuator).

?-2 is actually -17 dbM?and power does not increase until about +2.5 dBm. At +10 dbM?it is actually +9.5 dBm.


Tomorrow I will start a full calibration. Suggestions and tips would be appreciated.

Is there a battery backed memory in the 8350B? It seems to remember the setup for quite?a while after being turned off. Is the save-n non-volatile?


Does anyone have a spare cover for the PCBs for sale? The plate that goes on top of the 83592B and has many screws. it also holds the big heat sink in place. It is obvious that someone else's fingers have been inside this plug-in.


 

I'd immediately suspect a shorted tantalum capacitor on one of the supply
lines. I had one with nearly-identical symptoms a few years ago, and that
was the cause.

-Dave

--
Dave McGuire, AK4HZ
New Kensington, PA



On October 9, 2015 9:36:34 PM "bunge@... [hp_agilent_equipment]"
<hp_agilent_equipment@...> wrote:

My 8350B with 83592B plug-in turns on with flickering displays that go
blank and hangs the 8350B with no error message, and no response to any
switches on either the plug-in or mainframe. While flickering for a few
seconds the 8350B displays show -310 MHz start and 20 GHz stop.


It does the same to a second known working 8350B. All data and address
lines are active when hung so I don't understand why the 8350B does not
respond.


If I press CW while the displays are flickering it responds. I may have to
power down and up again to then enter a CW frequency, it comes up showing
CW. Once in this state I can run tests and all front panel entries work. I
ran all the A3 tests which show the ROMs and address decoders are OK.


The frequency in CW is high by 13 to 47 MHz going from 50 MHz to 18 GHz (it
covers 0.01 to 20 GHz).


The power is way off. The display shows -2 to +23 dBm. Is this normal? (no
opt 002 attenuator).
-2 is actually -17 dbM and power does not increase until about +2.5 dBm. At
+10 dbM it is actually +9.5 dBm.


Tomorrow I will start a full calibration. Suggestions and tips would be
appreciated.
Is there a battery backed memory in the 8350B? It seems to remember the
setup for quite a while after being turned off. Is the save-n non-volatile?


Does anyone have a spare cover for the PCBs for sale? The plate that goes
on top of the 83592B and has many screws. it also holds the big heat sink
in place. It is obvious that someone else's fingers have been inside this
plug-in.


 

Suspect A3 in your 83592B
Look at manual 83592-90074
8-57 Troubleshooting the A3 Digital Interface Assembly

Dennis


--------------------------------------------

On Fri, 10/9/15, bunge@... [hp_agilent_equipment] <hp_agilent_equipment@...> wrote:

Subject: [hp_agilent_equipment] HP 83592B hangs 8350B sweeper
To: hp_agilent_equipment@...
Date: Friday, October 9, 2015, 9:36 PM






























My 8350B with 83592B plug-in turns on?with
flickering displays that go blank?and hangs the 8350B
with no error message,?and no response to any switches
on either the plug-in or mainframe. While flickering for a
few seconds the 8350B displays show -310 MHz start and 20
GHz stop.
It does the same to a second known
working 8350B. All data and address lines are active when
hung so I don't understand why the 8350B does not
respond.
If I press CW while the displays are
flickering it responds. I may have to power down and up
again to then enter a CW frequency, it comes up showing CW.
Once in this state I can run tests and all front panel
entries work. I ran all the A3 tests which show the ROMs and
address decoders are OK.
The frequency in CW is high by 13 to 47
MHz going from?50 MHz to 18 GHz (it covers 0.01 to 20
GHz).
The power is way off. The display shows
-2 to +23 dBm. Is this normal? (no opt 002
attenuator).?-2 is actually -17
dbM?and power does not increase until about +2.5 dBm.
At +10 dbM?it is actually +9.5
dBm.
Tomorrow I will start a full
calibration. Suggestions and tips would be appreciated.
Is there a battery backed memory in the
8350B? It seems to remember the setup for quite?a while
after being turned off. Is the save-n
non-volatile?
Does anyone have a spare cover for the
PCBs for sale? The plate that goes on top of the 83592B and
has many screws. it also holds the big heat sink in place.
It is obvious that someone else's fingers have been
inside this plug-in.


 

As an addition, you have removed your nicad battery pack, right?

The nicad pack in the 8350B main frame, when overcharged (which is always)
vents caustic fumes and they fume all over the plugin, and the main frame
board next to the battery.

They cause a nightmare of problems.

The battery is next to worthless as it is, it should be removed on sight,
and all of the mess cleaned up.

-Chuck Harris

Dave McGuire Mcguire@... [hp_agilent_equipment] wrote:


I'd immediately suspect a shorted tantalum capacitor on one of the supply
lines. I had one with nearly-identical symptoms a few years ago, and that
was the cause.

-Dave


Cliff Barber
 

?
Hi,
Try a burst of aerosol Contact Cleaner on the contacts on the back of the plugin also on the mating connector inside the mainframe.
In my experience that seems to suppress the error messages that you have described.
It might need several iterations of the process. The act of pushing the plugin into the mainframe will help by wiping the mating contacts, so that should help.
Good luck
Cliff? G4BGP
?

----- Original Message -----
Sent: Saturday, October 10, 2015 2:36 AM
Subject: [hp_agilent_equipment] HP 83592B hangs 8350B sweeper

?

My 8350B with 83592B plug-in turns on?with flickering displays that go blank?and hangs the 8350B with no error message,?and no response to any switches on either the plug-in or mainframe. While flickering for a few seconds the 8350B displays show -310 MHz start and 20 GHz stop.


It does the same to a second known working 8350B. All data and address lines are active when hung so I don't understand why the 8350B does not respond.


If I press CW while the displays are flickering it responds. I may have to power down and up again to then enter a CW frequency, it comes up showing CW. Once in this state I can run tests and all front panel entries work. I ran all the A3 tests which show the ROMs and address decoders are OK.


The frequency in CW is high by 13 to 47 MHz going from?50 MHz to 18 GHz (it covers 0.01 to 20 GHz).


The power is way off. The display shows -2 to +23 dBm. Is this normal? (no opt 002 attenuator).

?-2 is actually -17 dbM?and power does not increase until about +2.5 dBm. At +10 dbM?it is actually +9.5 dBm.


Tomorrow I will start a full calibration. Suggestions and tips would be appreciated.

Is there a battery backed memory in the 8350B? It seems to remember the setup for quite?a while after being turned off. Is the save-n non-volatile?


Does anyone have a spare cover for the PCBs for sale? The plate that goes on top of the 83592B and has many screws. it also holds the big heat sink in place. It is obvious that someone else's fingers have been inside this plug-in.

No virus found in this message.
Checked by AVG -
Version: 2015.0.6140 / Virus Database: 4435/10790 - Release Date: 10/10/15


 

¿ªÔÆÌåÓý

Thanks for the replies.

Cliff: I tried the contact cleaner. No improvement but it was worth a try. There are no error messages.

?

Chuck: I found the battery. Slight corrosion mainly on the battery, a little on the metal clip. Cleaned it up and borrowed a battery from my HP35. I found a supplier of bad HP35 batteries that only lasted a month, also a supplier of good ones that have lasted several years and hold a charge. I will try to find them again and post a link. These batteries are used in the HP8672A and HP8566A, probably others. Anyway this is not the problem because both my 8350B work fine with other plugins. There is no sign of damage to the 8350B or the 83592B, they are clean and shiny like new. That is not to say there is no problem caused by chemical fumes. A scrub in soapy water and rinse would be a last resort but I have done it before.

?

Dennis: Yes that was the first place I looked. It passed all the tests. See below.

?

Dave: I will re-check the capacitors. My resistance measurements showed a drop when A7 was removed but it was masked by the 22 ohms of the RPG lamp. I will try my ESR meter.

?

The hanging occurs when starting in start/stop mode. Once I get it into CW it remains on for hours. I think it best to carry on with the full calibration. Also INSTR PRESET freezes the 8350B with blank start/stop displays. Cycling the ?power and catching CW is the only way to recover. Once the CW setup is recovered it comes on reliably in that mode.

?

Is -2 to +23 dBm normal for this plug-in? (no opt 002 attenuator)

Does anyone have a spare cover for the PCBs for sale? The plate that goes on top of the 83592B and has many screws. it also holds the big heat sink in place. It is obvious that someone else's fingers have been inside this plug-in.

Peter.

?

----- Original Message -----

Sent: Saturday, October 10, 2015 2:36 AM

Subject: [hp_agilent_equipment] HP 83592B hangs 8350B sweeper

My 8350B with 83592B plug-in turns on?with flickering displays that go blank?and hangs the 8350B with no error message,?and no response to any switches on either the plug-in or mainframe. While flickering for a few seconds the 8350B displays show -310 MHz start and 20 GHz stop.

It does the same to a second known working 8350B. All data and address lines are active when hung so I don't understand why the 8350B does not respond.

If I press CW while the displays are flickering it responds. I may have to power down and up again to then enter a CW frequency, it comes up showing CW. Once in this state I can run tests and all front panel entries work. I ran all the A3 tests which show the ROMs and address decoders are OK.

The frequency in CW is high by 13 to 47 MHz going from?50 MHz to 18 GHz (it covers 0.01 to 20 GHz).

The power is way off. The display shows -2 to +23 dBm. Is this normal? (no opt 002 attenuator). ?-2 is actually -17 dbM?and power does not increase until about +2.5 dBm. At +10 dbM?it is actually +9.5 dBm.

Does anyone have a spare cover for the PCBs for sale? The plate that goes on top of the 83592B and has many screws. it also holds the big heat sink in place. It is obvious that someone else's fingers have been inside this plug-in.

?

Reply to sender

?

Reply to group

?

?

(5)

¡¤???????? 5

¡¤???????? 4



.



 

¿ªÔÆÌåÓý

New clues:

I started through the Section 5 adjustments and got to 5.2 step 7 with minor adjustments. Entering 13.5 GHz freezes the displays. They start to flicker at 13.370 GHz. I can no longer select any frequency above this.

The second observation is that the A9 reference resistor heat sink is too hot to touch. It shows Q1 and Q3 power TO3 devices? on the schematic but has 3 on the heat sink and shows 3 on the component diagram, fig 8-70 in my manual (83592-90079). Q2 & Q3 are identical but not directly connected together. Q3 is supposed to be a voltage regulator. Q1 is supposed to be the coil current driver. It is difficult to tell which device is making the heat. The parts list does not show any of the ¡°transistors/regulators¡±. Q1 is marked 1820-0430, Q2 & Q3 are marked ¡° RCA 4-0080? HYP8432¡±.

Anyone familiar with this A9 assembly or have a schematic that shows the three devices?


 

On 10/10/2015 12:27 PM, 'Peter' bunge@... [hp_agilent_equipment] wrote:
New clues:

I started through the Section 5 adjustments and got to 5.2 step 7 with
minor adjustments. Entering 13.5 GHz freezes the displays. They start to
flicker at 13.370 GHz. I can no longer select any frequency above this.

The second observation is that the A9 reference resistor heat sink is
too hot to touch. It shows Q1 and Q3 power TO3 devices on the schematic
but has 3 on the heat sink and shows 3 on the component diagram, fig
8-70 in my manual (83592-90079). Q2 & Q3 are identical but not directly
connected together. Q3 is supposed to be a voltage regulator. Q1 is
supposed to be the coil current driver. It is difficult to tell which
device is making the heat. The parts list does not show any of the
¡°transistors/regulators¡±. Q1 is marked 1820-0430, Q2 & Q3 are marked ¡°
RCA 4-0080 HYP8432¡±.
To me this further suggests a power supply problem. As you increase
the YIG coil current by calling for higher frequencies, one or more of
Q1/Q2/Q3 is heating up too much, and probably dragging one of the power
supplies down. How do the rails look when you go up into that frequency
range?

-Dave

--
Dave McGuire, AK4HZ
New Kensington, PA


 

¿ªÔÆÌåÓý

Good thought Dave, this is where I¡¯m headed next but it is difficult without a proper schematic.

My manual (download) has a whole lot of extra stuff at the end of the manual that I did not print.

There may be something in there.

Peter.

?

From: hp_agilent_equipment@... [mailto:hp_agilent_equipment@...]
Sent: October-10-15 12:44 PM
To: hp_agilent_equipment@...
Subject: Re: [hp_agilent_equipment] HP 83592B hangs 8350B sweeper

?

?

On 10/10/2015 12:27 PM, 'Peter' bunge@... [hp_agilent_equipment] wrote:
> New clues:
>
> I started through the Section 5 adjustments and got to 5.2 step 7 with
> minor adjustments. Entering 13.5 GHz freezes the displays. They start to
> flicker at 13.370 GHz. I can no longer select any frequency above this.
>
> The second observation is that the A9 reference resistor heat sink is
> too hot to touch. It shows Q1 and Q3 power TO3 devices on the schematic
> but has 3 on the heat sink and shows 3 on the component diagram, fig
> 8-70 in my manual (83592-90079). Q2 & Q3 are identical but not directly
> connected together. Q3 is supposed to be a voltage regulator. Q1 is
> supposed to be the coil current driver. It is difficult to tell which
> device is making the heat. The parts list does not show any of the
> ¡°transistors/regulators¡±. Q1 is marked 1820-0430, Q2 & Q3 are marked ¡°
> RCA 4-0080 HYP8432¡±.

To me this further suggests a power supply problem. As you increase
the YIG coil current by calling for higher frequencies, one or more of
Q1/Q2/Q3 is heating up too much, and probably dragging one of the power
supplies down. How do the rails look when you go up into that frequency
range?

-Dave

--
Dave McGuire, AK4HZ
New Kensington, PA


 

I agree - YIG coil loads on the supply may be dragging it down. But first, does it produce the right frequency at lower settings where the problem doesn't show? This would confirm that the coil drive is at least OK.

Also, as I recall, there are two YIG device coils to drive - the oscillator, which should go around 2-6 or 2-7 GHz, and the YIG-tuned multiplier (YTM) filter coil. Both go up with frequency, but the oscillator will be periodic in tuning, repeating its range while the multiplication factor "N" changes with bands. The YTM coil only goes up and up, probably taking the bulk of the current.

There's also a power amplifier to drive the YTM - it may be juiced up another notch as N is increased. 13.5 GHz may be where N changes from 2 to 3.

Ed


 

I just noticed the original posting - it looks like the frequency is reasonably close, but amplitude is way off. Depending on the band (N), it could just be that alignment is needed, once the basic problem is fixed.

To get enough output power available for leveling, the YTO and YTM tuning have to be properly set up to track each other through the entire range of coverage. It could be that someone before had grossly misadjusted it while trying to fix it. Once you figure out the setting/overload problem, the rest should be OK after adjustment.

If the fundamental band around 2-7 GHz (N=1) is OK, but the higher ones (N=2 or 3) are not, it could be a problem with the YTM or power amp.

Ed


 

¿ªÔÆÌåÓý

Thanks Ed, I have sorted out a few things. The drawings in the manual of A9 are back view but can be mistaken for front views because the transistors are drawn that way.

This makes the schematic correct with Q1 the YIG tuning coil driver (see A8 description), Q2 the SYTM (Switched YIG Tuned Multiplier) driver (see A7 SYTM Driver), and Q3 the 78H05 regulator.

I checked the collector voltages of ?Q1(pin2) and Q2(pin10 on motherboard)

?

Select Freq???????? Actual freq???????? pin 2????????????????????? pin 10

50 MHz???????????????? 50.4??????????????????????? +5.63v?????????????????? +15.98v

100 MHz?????????????? 101???????????????????????? +5.44???????????????????? +15.98

500 MHz?????????????? 503???????????????????????? +3.92???????????????????? +15.98

1 GHz??????????????????? 1.006???????????????????? +2.029?????????????????? +15.97

2 GHz??????????????????? 2.012???????????????????? -1.77????????????????????? +15.98

3 GHz??????????????????? 3.002???????????????????? +8.84???????????????????? +15.99

4 GHz??????????????????? 4.007???????????????????? +5.05???????????????????? +15.16

5 GHz??????????????????? 5.013???????????????????? +1.254?????????????????? +14.34

12 GHz????????????????? 12.026?????????????????? -2.53????????????????????? +8.52

13 GHz????????????????? 13.031?????????????????? -4.42????????????????????? +7.70

14 GHz????????????????? no O/P????????????????? +2.56???????????????????? +6.91

?

They seem reasonable, but maybe not the SYTM (Q2 pin 10). The bands change at 2.4, 7.0, and 13.5. It seems strange the frequencies are close up to 13 GHz.

A rough calculation has each transistor dissipating 1 to 2 watts but the regulator could be higher with 7 volts across it, so 7 to 20 watts depending on the current. I cannot measure the current of the 5v supply without cutting traces and there is no wire to hang a DC current meter onto. The heatsink gets very hot at any CW frequency. I don¡¯t think the unreg 5v should be as high as the 12v I measure. The top cover screws onto the top of the heatsink and maybe is needed to dissipate the heat, another 24 sq in of metal close to the 8350B top lid! I¡¯m going to drop this investigation for now.

I don¡¯t think the lock up is due to a power supply, I have checked them all.

I can get it running in CW but as soon as I press START the 8350B locks up with no display, no error, and no response to buttons. I need to switch my attention to the sweep circuits before I can calibrate anything. I know this has been meddled with because the top cover is missing and an SMA connector was loose.

Peter.

?

?

?

From: hp_agilent_equipment@... [mailto:hp_agilent_equipment@...]
Sent: October-10-15 5:31 PM
To: hp_agilent_equipment@...
Subject: RE: [hp_agilent_equipment] HP 83592B hangs 8350B sweeper

?

?

I just noticed the original posting - it looks like the frequency is reasonably close, but amplitude is way off. Depending on the band (N), it could just be that alignment is needed, once the basic problem is fixed.

To get enough output power available for leveling, the YTO and YTM tuning have to be properly set up to track each other through the entire range of coverage. It could be that someone before had grossly misadjusted it while trying to fix it. Once you figure out the setting/overload problem, the rest should be OK after adjustment.

If the fundamental band around 2-7 GHz (N=1) is OK, but the higher ones (N=2 or 3) are not, it could be a problem with the YTM or power amp.

Ed


 

Hi - I have a cover. Send me your address and I'll mail it to you no charge.

Dan in Chandler, AZ


 

¿ªÔÆÌåÓý

Ed, I can sweep reliably within any band and see the frequency move up on my 8566 spectrum analyzer (1 to 2, 3 to 5, 7.5 to 10, and 15 to 18 GHz). I used a 10 sec sweep so it was visible.

If I overlap a band the 8350 freezes. I suspected the bandswitch DAC which starts at 0v and has one step to +3.5v which is wrong. However there is a test (shift 56, page 8-127 step 2a) and the DAC looks OK. The TV buffer is OK as well. I am not sure where the bandswitch DAC gets its address and data. I think it is getting the wrong data. The Address is almost certainly from the 8350 but data could have been read from the switches on A7 or A8. Do you know?

I am struggling a bit with downloaded manuals but making progress. Tucker has an original manual on Ebay for $100. Global has A and C manuals for about $35 but no B version.

Does anyone have an original HP 83592B manual for sale at $40?

Peter.

?

?

From: hp_agilent_equipment@... [mailto:hp_agilent_equipment@...]
Sent: October-10-15 5:31 PM
To: hp_agilent_equipment@...
Subject: RE: [hp_agilent_equipment] HP 83592B hangs 8350B sweeper

?

?

I just noticed the original posting - it looks like the frequency is reasonably close, but amplitude is way off. Depending on the band (N), it could just be that alignment is needed, once the basic problem is fixed.

To get enough output power available for leveling, the YTO and YTM tuning have to be properly set up to track each other through the entire range of coverage. It could be that someone before had grossly misadjusted it while trying to fix it. Once you figure out the setting/overload problem, the rest should be OK after adjustment.

If the fundamental band around 2-7 GHz (N=1) is OK, but the higher ones (N=2 or 3) are not, it could be a problem with the YTM or power amp.

Ed


 

¿ªÔÆÌåÓý

Has anyone noticed that boards in HP 8350 plug-ins have part numbers different from the plug-in model?

My 83592 has two boards, A6 and A7, that have part numbers starting with 83595-x. The others start 83592-x.

?

I know that replacing the A6 board (83595-60106) with a good one (83592-60106) gets the plug-in working, even with the 83595-x A7 board.

?

On the bad A6 board I have found that a pot (R63, SRD bias) was missing and the Band 0 jumper was in the B1 position. The board also has two extra pots which are not shown on the schematic or component location. There are seven single turn pots on the top left, viewed from component side. They are shown in the 83595 manual. Did someone try to get it running with the board from a 83595 (0.01 to 26.5 GHz) plug-in?

I need to look carefully at every component. The 60106 in the number is the same (83592-60106) and the boards look identical but I¡¯m sure there are component differences as I have found. I do not believe that this is the original board for the plug-in.

?

I have seen instruments put together with odd parts before. I have an HP8656A chassis that is part 8656A and part 8657A inside. I gave up troubleshooting when I discovered this. This is a very dishonest practice to do this and sell the items as-is, for parts, etc.

?

?


 

¿ªÔÆÌåÓý

Peter,

I have an 83594A and an 83595A wth boards that have mismatched part numbers so it may not have been anything nefarious but if so its contagious! There are a couple of boards that don't match any of the manuals I've been able to find, and there are switches that open parts of loops, and pots for adjustments, and none are mentioned in any manual. The "joys" of working on older gear!

Steve
WB0DBS



On Oct 16, 2015, at 10:47 AM, 'Peter' bunge@... [hp_agilent_equipment] <hp_agilent_equipment@...> wrote:

?

Has anyone noticed that boards in HP 8350 plug-ins have part numbers different from the plug-in model?

My 83592 has two boards, A6 and A7, that have part numbers starting with 83595-x. The others start 83592-x.

?

I know that replacing the A6 board (83595-60106) with a good one (83592-60106) gets the plug-in working, even with the 83595-x A7 board.

?

On the bad A6 board I have found that a pot (R63, SRD bias) was missing and the Band 0 jumper was in the B1 position. The board also has two extra pots which are not shown on the schematic or component location. There are seven single turn pots on the top left, viewed from component side. They are shown in the 83595 manual. Did someone try to get it running with the board from a 83595 (0.01 to 26.5 GHz) plug-in?

I need to look carefully at every component. The 60106 in the number is the same (83592-60106) and the boards look identical but I¡¯m sure there are component differences as I have found. I do not believe that this is the original board for the plug-in.

?

I have seen instruments put together with odd parts before. I have an HP8656A chassis that is part 8656A and part 8657A inside. I gave up troubleshooting when I discovered this. This is a very dishonest practice to do this and sell the items as-is, for parts, etc.

?

?


 

¿ªÔÆÌåÓý

GREAT PROGRESS!

Thanks Steve, you are quite correct. My A7 board is marked 83595 and works in this 83592 plug-in.

However I suspected that the A6 board that was causing all the trouble was actually from an 83595 not an 83592. I removed the 4L and 4H pots, installed the R63 pot, moved the Band jumper, ?and concentrated on the Sweep Control and Bandswitch circuits.

The A3 SW1 forces the signals to go through the variable gain amplifier for troubleshooting. Using this setting I could sweep band 0 but was finding that the TP7 signal was going positive by 2.4 volts. It should start at 0v and go negative to -10v max. There was an offset in the variable gain amplifier! I had also noticed that the gain seemed high. The schematics showed the gain resistors to be the same as the ones on the bad A6 board but they were different from the ones on the good A6 board. The parts list agreed with the resistors on the good A6 board so I assumed them to be correct. This was not the first error on the schematics I had found.

Troubleshooting: Set sweep from 1 MHz to 1.1 MHz so TP5 is essentially at 0v. Check TP1 (BandSw DAC O/P) is 1.195v on good A6 and 1.194 on bad A6. So far so good. Check U5-6 is -10.05v on good A6 but is? -12.73v on bad board. TP7 is 0v on the good A6 but is +2.4v on the bad board. Conclusion: the gain is too high and is tripping U14 in the Bandswitch Comparator.

Change all 4 gain resistors:

R29???????? B0?????????? was 109.73k??????? change to 82.541k (TP7 is now 0v with the above test)

R27???????? B1?????????? was 57.014k??????? change to 42.884k

R25???????? B2?????????? was 40.423k??????? change to 30.423k

R23???????? B3????????? was 40.423k??????? change to 30.423k

Now the plug-in sweeps from 0.01 to 20 GHz and the INSTR PRESET no longer hangs the 8350B.

Conclusion: The A6 board was from an HP 83595. I can now proceed with the calibration bu expect some problems because the resistors I used were 1%, not 0.1%, and are not the exact values. I may not have enough range of adjustment. I may also run into problems in the SRD and PIN diode bias circuits because of component values.

Some of the analog switches are normally closed (PMI SW06) but are shown open on the schematic. U2A pins 1 & 2 are switched, U2A-2 goes to U13-11.

Peter.

?

?

From: hp_agilent_equipment@... [mailto:hp_agilent_equipment@...]
Sent: October-16-15 12:10 PM
To: hp_agilent_equipment@...
Subject: Re: [hp_agilent_equipment] HP 83592B hangs 8350B sweeper

?

?

Peter,

?

I have an 83594A and an 83595A wth boards that have mismatched part numbers so it may not have been anything nefarious but if so its contagious! There are a couple of boards that don't match any of the manuals I've been able to find, and there are switches that open parts of loops, and pots for adjustments, and none are mentioned in any manual. The "joys" of working on older gear!

Steve

WB0DBS

?

?


On Oct 16, 2015, at 10:47 AM, 'Peter' bunge@... [hp_agilent_equipment] <hp_agilent_equipment@...> wrote:

?

Has anyone noticed that boards in HP 8350 plug-ins have part numbers different from the plug-in model?

My 83592 has two boards, A6 and A7, that have part numbers starting with 83595-x. The others start 83592-x.

?

I know that replacing the A6 board (83595-60106) with a good one (83592-60106) gets the plug-in working, even with the 83595-x A7 board.

?

On the bad A6 board I have found that a pot (R63, SRD bias) was missing and the Band 0 jumper was in the B1 position. The board also has two extra pots which are not shown on the schematic or component location. There are seven single turn pots on the top left, viewed from component side. They are shown in the 83595 manual. Did someone try to get it running with the board from a 83595 (0.01 to 26.5 GHz) plug-in?

I need to look carefully at every component. The 60106 in the number is the same (83592-60106) and the boards look identical but I¡¯m sure there are component differences as I have found. I do not believe that this is the original board for the plug-in.

?

I have seen instruments put together with odd parts before. I have an HP8656A chassis that is part 8656A and part 8657A inside. I gave up troubleshooting when I discovered this. This is a very dishonest practice to do this and sell the items as-is, for parts, etc.

?

?


 

¿ªÔÆÌåÓý

Has anyone done a calibration on one of these plug-ins?

Step 20 in adjustments: What does A vs B mode on an oscilloscope mean? I think HP means trigger on B and look at A ??(A + B or A ¨C B will not work. The waveforms are completely different time/div and shapes)

A6-TP8 does not have positive sweeps as shown. They go negative from 0 to -10v.

Step 22: A6R37 will not affect TP8 an far as I can see.

Am I missing something?

?

From: hp_agilent_equipment@... [mailto:hp_agilent_equipment@...]
Sent: October-17-15 5:54 PM
To: hp_agilent_equipment@...
Subject: RE: [hp_agilent_equipment] HP 83592B hangs 8350B sweeper

?

?

GREAT PROGRESS!

Thanks Steve, you are quite correct. My A7 board is marked 83595 and works in this 83592 plug-in.

However I suspected that the A6 board that was causing all the trouble was actually from an 83595 not an 83592. I removed the 4L and 4H pots, installed the R63 pot, moved the Band jumper, ?and concentrated on the Sweep Control and Bandswitch circuits.

The A3 SW1 forces the signals to go through the variable gain amplifier for troubleshooting. Using this setting I could sweep band 0 but was finding that the TP7 signal was going positive by 2.4 volts. It should start at 0v and go negative to -10v max. There was an offset in the variable gain amplifier! I had also noticed that the gain seemed high. The schematics showed the gain resistors to be the same as the ones on the bad A6 board but they were different from the ones on the good A6 board. The parts list agreed with the resistors on the good A6 board so I assumed them to be correct. This was not the first error on the schematics I had found.

Troubleshooting: Set sweep from 1 MHz to 1.1 MHz so TP5 is essentially at 0v. Check TP1 (BandSw DAC O/P) is 1.195v on good A6 and 1.194 on bad A6. So far so good. Check U5-6 is -10.05v on good A6 but is? -12.73v on bad board. TP7 is 0v on the good A6 but is +2.4v on the bad board. Conclusion: the gain is too high and is tripping U14 in the Bandswitch Comparator.

Change all 4 gain resistors:

R29???????? B0?????????? was 109.73k??????? change to 82.541k (TP7 is now 0v with the above test)

R27???????? B1?????????? was 57.014k??????? change to 42.884k

R25???????? B2?????????? was 40.423k??????? change to 30.423k

R23???????? B3????????? was 40.423k??????? change to 30.423k

Now the plug-in sweeps from 0.01 to 20 GHz and the INSTR PRESET no longer hangs the 8350B.

Conclusion: The A6 board was from an HP 83595. I can now proceed with the calibration bu expect some problems because the resistors I used were 1%, not 0.1%, and are not the exact values. I may not have enough range of adjustment. I may also run into problems in the SRD and PIN diode bias circuits because of component values.

Some of the analog switches are normally closed (PMI SW06) but are shown open on the schematic. U2A pins 1 & 2 are switched, U2A-2 goes to U13-11.

Peter.

?

?

From: hp_agilent_equipment@... [mailto:hp_agilent_equipment@...]
Sent: October-16-15 12:10 PM
To: hp_agilent_equipment@...
Subject: Re: [hp_agilent_equipment] HP 83592B hangs 8350B sweeper

?

?

Peter,

?

I have an 83594A and an 83595A wth boards that have mismatched part numbers so it may not have been anything nefarious but if so its contagious! There are a couple of boards that don't match any of the manuals I've been able to find, and there are switches that open parts of loops, and pots for adjustments, and none are mentioned in any manual. The "joys" of working on older gear!

Steve

WB0DBS

?

?


On Oct 16, 2015, at 10:47 AM, 'Peter' bunge@... [hp_agilent_equipment] <hp_agilent_equipment@...> wrote:

?

Has anyone noticed that boards in HP 8350 plug-ins have part numbers different from the plug-in model?

My 83592 has two boards, A6 and A7, that have part numbers starting with 83595-x. The others start 83592-x.

?

I know that replacing the A6 board (83595-60106) with a good one (83592-60106) gets the plug-in working, even with the 83595-x A7 board.

?

On the bad A6 board I have found that a pot (R63, SRD bias) was missing and the Band 0 jumper was in the B1 position. The board also has two extra pots which are not shown on the schematic or component location. There are seven single turn pots on the top left, viewed from component side. They are shown in the 83595 manual. Did someone try to get it running with the board from a 83595 (0.01 to 26.5 GHz) plug-in?

I need to look carefully at every component. The 60106 in the number is the same (83592-60106) and the boards look identical but I¡¯m sure there are component differences as I have found. I do not believe that this is the original board for the plug-in.

?

I have seen instruments put together with odd parts before. I have an HP8656A chassis that is part 8656A and part 8657A inside. I gave up troubleshooting when I discovered this. This is a very dishonest practice to do this and sell the items as-is, for parts, etc.

?

?


Bill Lauchlan
 

¿ªÔÆÌåÓý

I think A vs B means like an X-Y Plotter. You probably connect a Sweep Output Ramp to ¡°B¡± and then some other Test Point to ¡°A¡±. Rgds Bill Lauchlan

?

From: hp_agilent_equipment@... [mailto:hp_agilent_equipment@...]
Sent: Saturday, October 17, 2015 7:48 PM
To: hp_agilent_equipment@...
Subject: RE: [hp_agilent_equipment] HP 83592B hangs 8350B sweeper

?

?

Has anyone done a calibration on one of these plug-ins?

Step 20 in adjustments: What does A vs B mode on an oscilloscope mean? I think HP means trigger on B and look at A ??(A + B or A ¨C B will not work. The waveforms are completely different time/div and shapes)

A6-TP8 does not have positive sweeps as shown. They go negative from 0 to -10v.

Step 22: A6R37 will not affect TP8 an far as I can see.

Am I missing something?

?

From: hp_agilent_equipment@... [mailto:hp_agilent_equipment@...]
Sent: October-17-15 5:54 PM
To: hp_agilent_equipment@...
Subject: RE: [hp_agilent_equipment] HP 83592B hangs 8350B sweeper

?

?

GREAT PROGRESS!

Thanks Steve, you are quite correct. My A7 board is marked 83595 and works in this 83592 plug-in.

However I suspected that the A6 board that was causing all the trouble was actually from an 83595 not an 83592. I removed the 4L and 4H pots, installed the R63 pot, moved the Band jumper, ?and concentrated on the Sweep Control and Bandswitch circuits.

The A3 SW1 forces the signals to go through the variable gain amplifier for troubleshooting. Using this setting I could sweep band 0 but was finding that the TP7 signal was going positive by 2.4 volts. It should start at 0v and go negative to -10v max. There was an offset in the variable gain amplifier! I had also noticed that the gain seemed high. The schematics showed the gain resistors to be the same as the ones on the bad A6 board but they were different from the ones on the good A6 board. The parts list agreed with the resistors on the good A6 board so I assumed them to be correct. This was not the first error on the schematics I had found.

Troubleshooting: Set sweep from 1 MHz to 1.1 MHz so TP5 is essentially at 0v. Check TP1 (BandSw DAC O/P) is 1.195v on good A6 and 1.194 on bad A6. So far so good. Check U5-6 is -10.05v on good A6 but is? -12.73v on bad board. TP7 is 0v on the good A6 but is +2.4v on the bad board. Conclusion: the gain is too high and is tripping U14 in the Bandswitch Comparator.

Change all 4 gain resistors:

R29???????? B0?????????? was 109.73k??????? change to 82.541k (TP7 is now 0v with the above test)

R27???????? B1?????????? was 57.014k??????? change to 42.884k

R25???????? B2?????????? was 40.423k??????? change to 30.423k

R23???????? B3????????? was 40.423k??????? change to 30.423k

Now the plug-in sweeps from 0.01 to 20 GHz and the INSTR PRESET no longer hangs the 8350B.

Conclusion: The A6 board was from an HP 83595. I can now proceed with the calibration bu expect some problems because the resistors I used were 1%, not 0.1%, and are not the exact values. I may not have enough range of adjustment. I may also run into problems in the SRD and PIN diode bias circuits because of component values.

Some of the analog switches are normally closed (PMI SW06) but are shown open on the schematic. U2A pins 1 & 2 are switched, U2A-2 goes to U13-11.

Peter.

?

?

From: hp_agilent_equipment@... [mailto:hp_agilent_equipment@...]
Sent: October-16-15 12:10 PM
To: hp_agilent_equipment@...
Subject: Re: [hp_agilent_equipment] HP 83592B hangs 8350B sweeper

?

?

Peter,

?

I have an 83594A and an 83595A wth boards that have mismatched part numbers so it may not have been anything nefarious but if so its contagious! There are a couple of boards that don't match any of the manuals I've been able to find, and there are switches that open parts of loops, and pots for adjustments, and none are mentioned in any manual. The "joys" of working on older gear!

Steve

WB0DBS

?

?


On Oct 16, 2015, at 10:47 AM, 'Peter' bunge@... [hp_agilent_equipment] <hp_agilent_equipment@...> wrote:

?

Has anyone noticed that boards in HP 8350 plug-ins have part numbers different from the plug-in model?

My 83592 has two boards, A6 and A7, that have part numbers starting with 83595-x. The others start 83592-x.

?

I know that replacing the A6 board (83595-60106) with a good one (83592-60106) gets the plug-in working, even with the 83595-x A7 board.

?

On the bad A6 board I have found that a pot (R63, SRD bias) was missing and the Band 0 jumper was in the B1 position. The board also has two extra pots which are not shown on the schematic or component location. There are seven single turn pots on the top left, viewed from component side. They are shown in the 83595 manual. Did someone try to get it running with the board from a 83595 (0.01 to 26.5 GHz) plug-in?

I need to look carefully at every component. The 60106 in the number is the same (83592-60106) and the boards look identical but I¡¯m sure there are component differences as I have found. I do not believe that this is the original board for the plug-in.

?

I have seen instruments put together with odd parts before. I have an HP8656A chassis that is part 8656A and part 8657A inside. I gave up troubleshooting when I discovered this. This is a very dishonest practice to do this and sell the items as-is, for parts, etc.

?

?


 

X = horizontal axis
Y = vertical
There is no sweep, no trigger : the "sweep" results from the ramp applied to the horizontal deflection amplifier. X v. Y exits almost only on analog scopes (1741A on the 83592X manual)
--------------------------------------------

On Sun, 10/18/15, 'Bill Lauchlan' bill.lauchlan@... [hp_agilent_equipment] <hp_agilent_equipment@...> wrote:

Subject: RE: [hp_agilent_equipment] HP 83592B hangs 8350B sweeper
To: hp_agilent_equipment@...
Date: Sunday, October 18, 2015, 3:03 AM


?









I think A vs B means
like an X-Y Plotter. You probably connect a Sweep Output
Ramp to ¡°B¡± and then some other Test Point to ¡°A¡±.
Rgds Bill Lauchlan ?From:
hp_agilent_equipment@...
[mailto:hp_agilent_equipment@...]
Sent: Saturday, October 17, 2015 7:48
PM
To:
hp_agilent_equipment@...
Subject: RE: [hp_agilent_equipment] HP
83592B hangs 8350B sweeper ?? Has anyone done a
calibration on one of these plug-ins?Step 20 in
adjustments: What does A vs B mode on an oscilloscope mean?
I think HP means trigger on B and look at A ??(A + B or A
¨C B will not work. The waveforms are completely different
time/div and shapes)A6-TP8 does not have
positive sweeps as shown. They go negative from 0 to -10v.
Step 22: A6R37 will
not affect TP8 an far as I can see.Am I missing
something??From: hp_agilent_equipment@...
[mailto:hp_agilent_equipment@...]

Sent: October-17-15 5:54 PM
To: hp_agilent_equipment@...
Subject: RE: [hp_agilent_equipment] HP
83592B hangs 8350B sweeper?? GREAT PROGRESS!Thanks Steve, you are
quite correct. My A7 board is marked 83595 and works in this
83592 plug-in.However I suspected that
the A6 board that was causing all the trouble was actually
from an 83595 not an 83592. I removed the 4L and 4H pots,
installed the R63 pot, moved the Band jumper, ?and
concentrated on the Sweep Control and Bandswitch
circuits.The A3 SW1
forces the signals to go through the variable gain amplifier
for troubleshooting. Using this setting I could sweep band 0
but was finding that the TP7 signal was going positive by
2.4 volts. It should start at 0v and go negative to -10v
max. There was an offset in the variable gain amplifier! I
had also noticed that the gain seemed high. The schematics
showed the gain resistors to be the same as the ones on the
bad A6 board but they were different from the ones on the
good A6 board. The parts list agreed with the resistors on
the good A6 board so I assumed them to be correct. This was
not the first error on the schematics I had found.Troubleshooting: Set sweep
from 1 MHz to 1.1 MHz so TP5 is essentially at 0v. Check TP1
(BandSw DAC O/P) is 1.195v on good A6 and 1.194 on bad A6.
So far so good. Check U5-6 is -10.05v on good A6 but is?
-12.73v on bad board. TP7 is 0v on the good A6 but is +2.4v
on the bad board. Conclusion: the gain is too high and is
tripping U14 in the Bandswitch Comparator.Change all 4 gain
resistors:R29????????
B0?????????? was 109.73k??????? change to
82.541k (TP7 is now 0v with the above test)R27????????
B1?????????? was 57.014k??????? change to
42.884kR25????????
B2?????????? was 40.423k??????? change to
30.423kR23????????
B3????????? was 40.423k??????? change to
30.423kNow the
plug-in sweeps from 0.01 to 20 GHz and the INSTR PRESET no
longer hangs the 8350B.Conclusion: The A6 board
was from an HP 83595. I can now proceed with the calibration
bu expect some problems because the resistors I used were
1%, not 0.1%, and are not the exact values. I may not have
enough range of adjustment. I may also run into problems in
the SRD and PIN diode bias circuits because of component
values.Some of the
analog switches are normally closed (PMI SW06) but are shown
open on the schematic. U2A pins 1 & 2 are switched,
U2A-2 goes to U13-11.Peter.??From: hp_agilent_equipment@...
[mailto:hp_agilent_equipment@...]

Sent: October-16-15 12:10 PM
To: hp_agilent_equipment@...
Subject: Re: [hp_agilent_equipment] HP
83592B hangs 8350B sweeper?? Peter,?I have an 83594A and an
83595A wth boards that have mismatched part numbers so it
may not have been anything nefarious but if so its
contagious! There are a couple of boards that don't
match any of the manuals I've been able to find, and
there are switches that open parts of loops, and pots for
adjustments, and none are mentioned in any manual. The
"joys" of working on older gear!

SteveWB0DBS??
On Oct 16,
2015, at 10:47 AM, 'Peter' bunge@...
[hp_agilent_equipment] <hp_agilent_equipment@...>
wrote:? Has anyone noticed that boards in HP
8350 plug-ins have part numbers different from the plug-in
model?My 83592 has two boards, A6 and A7,
that have part numbers starting with 83595-x. The others
start 83592-x.?I know that replacing the A6 board
(83595-60106) with a good one (83592-60106) gets the plug-in
working, even with the 83595-x A7 board.?On the bad A6 board I have found that
a pot (R63, SRD bias) was missing and the Band 0 jumper was
in the B1 position. The board also has two extra pots which
are not shown on the schematic or component location. There
are seven single turn pots on the top left, viewed from
component side. They are shown in the 83595 manual. Did
someone try to get it running with the board from a 83595
(0.01 to 26.5 GHz) plug-in? I need to look carefully at every
component. The 60106 in the number is the same (83592-60106)
and the boards look identical but I¡¯m sure there are
component differences as I have found. I do not believe that
this is the original board for the plug-in.?I have seen instruments put together
with odd parts before. I have an HP8656A chassis that is
part 8656A and part 8657A inside. I gave up troubleshooting
when I discovered this. This is a very dishonest practice to
do this and sell the items as-is, for parts,
etc.??










#yiv0923979647 #yiv0923979647 --
#yiv0923979647ygrp-mkp {
border:1px solid #d8d8d8;font-family:Arial;margin:10px
0;padding:0 10px;}

#yiv0923979647 #yiv0923979647ygrp-mkp hr {
border:1px solid #d8d8d8;}

#yiv0923979647 #yiv0923979647ygrp-mkp #yiv0923979647hd {
color:#628c2a;font-size:85%;font-weight:700;line-height:122%;margin:10px
0;}

#yiv0923979647 #yiv0923979647ygrp-mkp #yiv0923979647ads {
margin-bottom:10px;}

#yiv0923979647 #yiv0923979647ygrp-mkp .yiv0923979647ad {
padding:0 0;}

#yiv0923979647 #yiv0923979647ygrp-mkp .yiv0923979647ad p {
margin:0;}

#yiv0923979647 #yiv0923979647ygrp-mkp .yiv0923979647ad a {
color:#0000ff;text-decoration:none;}
#yiv0923979647 #yiv0923979647ygrp-sponsor
#yiv0923979647ygrp-lc {
font-family:Arial;}

#yiv0923979647 #yiv0923979647ygrp-sponsor
#yiv0923979647ygrp-lc #yiv0923979647hd {
margin:10px
0px;font-weight:700;font-size:78%;line-height:122%;}

#yiv0923979647 #yiv0923979647ygrp-sponsor
#yiv0923979647ygrp-lc .yiv0923979647ad {
margin-bottom:10px;padding:0 0;}

#yiv0923979647 #yiv0923979647actions {
font-family:Verdana;font-size:11px;padding:10px 0;}

#yiv0923979647 #yiv0923979647activity {
background-color:#e0ecee;float:left;font-family:Verdana;font-size:10px;padding:10px;}

#yiv0923979647 #yiv0923979647activity span {
font-weight:700;}

#yiv0923979647 #yiv0923979647activity span:first-child {
text-transform:uppercase;}

#yiv0923979647 #yiv0923979647activity span a {
color:#5085b6;text-decoration:none;}

#yiv0923979647 #yiv0923979647activity span span {
color:#ff7900;}

#yiv0923979647 #yiv0923979647activity span
.yiv0923979647underline {
text-decoration:underline;}

#yiv0923979647 .yiv0923979647attach {
clear:both;display:table;font-family:Arial;font-size:12px;padding:10px
0;width:400px;}

#yiv0923979647 .yiv0923979647attach div a {
text-decoration:none;}

#yiv0923979647 .yiv0923979647attach img {
border:none;padding-right:5px;}

#yiv0923979647 .yiv0923979647attach label {
display:block;margin-bottom:5px;}

#yiv0923979647 .yiv0923979647attach label a {
text-decoration:none;}

#yiv0923979647 blockquote {
margin:0 0 0 4px;}

#yiv0923979647 .yiv0923979647bold {
font-family:Arial;font-size:13px;font-weight:700;}

#yiv0923979647 .yiv0923979647bold a {
text-decoration:none;}

#yiv0923979647 dd.yiv0923979647last p a {
font-family:Verdana;font-weight:700;}

#yiv0923979647 dd.yiv0923979647last p span {
margin-right:10px;font-family:Verdana;font-weight:700;}

#yiv0923979647 dd.yiv0923979647last p
span.yiv0923979647yshortcuts {
margin-right:0;}

#yiv0923979647 div.yiv0923979647attach-table div div a {
text-decoration:none;}

#yiv0923979647 div.yiv0923979647attach-table {
width:400px;}

#yiv0923979647 div.yiv0923979647file-title a, #yiv0923979647
div.yiv0923979647file-title a:active, #yiv0923979647
div.yiv0923979647file-title a:hover, #yiv0923979647
div.yiv0923979647file-title a:visited {
text-decoration:none;}

#yiv0923979647 div.yiv0923979647photo-title a,
#yiv0923979647 div.yiv0923979647photo-title a:active,
#yiv0923979647 div.yiv0923979647photo-title a:hover,
#yiv0923979647 div.yiv0923979647photo-title a:visited {
text-decoration:none;}

#yiv0923979647 div#yiv0923979647ygrp-mlmsg
#yiv0923979647ygrp-msg p a span.yiv0923979647yshortcuts {
font-family:Verdana;font-size:10px;font-weight:normal;}

#yiv0923979647 .yiv0923979647green {
color:#628c2a;}

#yiv0923979647 .yiv0923979647MsoNormal {
margin:0 0 0 0;}

#yiv0923979647 o {
font-size:0;}

#yiv0923979647 #yiv0923979647photos div {
float:left;width:72px;}

#yiv0923979647 #yiv0923979647photos div div {
border:1px solid
#666666;height:62px;overflow:hidden;width:62px;}

#yiv0923979647 #yiv0923979647photos div label {
color:#666666;font-size:10px;overflow:hidden;text-align:center;white-space:nowrap;width:64px;}

#yiv0923979647 #yiv0923979647reco-category {
font-size:77%;}

#yiv0923979647 #yiv0923979647reco-desc {
font-size:77%;}

#yiv0923979647 .yiv0923979647replbq {
margin:4px;}

#yiv0923979647 #yiv0923979647ygrp-actbar div a:first-child {
margin-right:2px;padding-right:5px;}

#yiv0923979647 #yiv0923979647ygrp-mlmsg {
font-size:13px;font-family:Arial, helvetica, clean,
sans-serif;}

#yiv0923979647 #yiv0923979647ygrp-mlmsg table {
font-size:inherit;font:100%;}

#yiv0923979647 #yiv0923979647ygrp-mlmsg select,
#yiv0923979647 input, #yiv0923979647 textarea {
font:99% Arial, Helvetica, clean, sans-serif;}

#yiv0923979647 #yiv0923979647ygrp-mlmsg pre, #yiv0923979647
code {
font:115% monospace;}

#yiv0923979647 #yiv0923979647ygrp-mlmsg * {
line-height:1.22em;}

#yiv0923979647 #yiv0923979647ygrp-mlmsg #yiv0923979647logo {
padding-bottom:10px;}


#yiv0923979647 #yiv0923979647ygrp-msg p a {
font-family:Verdana;}

#yiv0923979647 #yiv0923979647ygrp-msg
p#yiv0923979647attach-count span {
color:#1E66AE;font-weight:700;}

#yiv0923979647 #yiv0923979647ygrp-reco
#yiv0923979647reco-head {
color:#ff7900;font-weight:700;}

#yiv0923979647 #yiv0923979647ygrp-reco {
margin-bottom:20px;padding:0px;}

#yiv0923979647 #yiv0923979647ygrp-sponsor #yiv0923979647ov
li a {
font-size:130%;text-decoration:none;}

#yiv0923979647 #yiv0923979647ygrp-sponsor #yiv0923979647ov
li {
font-size:77%;list-style-type:square;padding:6px 0;}

#yiv0923979647 #yiv0923979647ygrp-sponsor #yiv0923979647ov
ul {
margin:0;padding:0 0 0 8px;}

#yiv0923979647 #yiv0923979647ygrp-text {
font-family:Georgia;}

#yiv0923979647 #yiv0923979647ygrp-text p {
margin:0 0 1em 0;}

#yiv0923979647 #yiv0923979647ygrp-text tt {
font-size:120%;}

#yiv0923979647 #yiv0923979647ygrp-vital ul li:last-child {
border-right:none !important;
}
#yiv0923979647