My faithful 3585a has developed a problem. When a BNC cable is connected to either the 50/75 ohm or 1 megohm input and the cable shield is connected to ground contact of another instrument or the ground contact from a wall outlet, spikes walk across the screen. These spikes are about 20dBm above the noise floor.
If I set the instrument to its default settings, I see the spikes walk across the screen. If I save the waveform to B (where it no longer walks), I count 24 spikes. On many start-stop frequency settings I see the same count of spikes. However, on some I see 48 spikes; e.g. a start frequency of 1MHz and a stop of 2MHz.
Anyone have a suggestion of how to diagnose and repair this problem?
Can the work be done without extender boards?
Thanks,
Gordon, KA2NLM
|
Gordon,
Is there any chance that your AC outlet for the 3585 has a defective earth/ground? Is there a bad extension cord or power strip? You can check that with one of those inexpensive outlet testers with three lights.
Can you select "LINE TRIGGER" on the 3585 and see if the spikes stop moving? If they do, it's line related. It could also be something in the 3585 power supply related to grounding.
Vladan
toggle quoted message
Show quoted text
--- In hp_agilent_equipment@..., "gldinnc" wrote: My faithful 3585a has developed a problem. When a BNC cable is connected to either the 50/75 ohm or 1 megohm input and the cable shield is connected to ground contact of another instrument or the ground contact from a wall outlet, spikes walk across the screen. These spikes are about 20dBm above the noise floor.
If I set the instrument to its default settings, I see the spikes walk across the screen. If I save the waveform to B (where it no longer walks), I count 24 spikes. On many start-stop frequency settings I see the same count of spikes. However, on some I see 48 spikes; e.g. a start frequency of 1MHz and a stop of 2MHz.
Anyone have a suggestion of how to diagnose and repair this problem?
Can the work be done without extender boards?
Thanks,
Gordon, KA2NLM
|
Gordon wrote. My faithful 3585a has developed a problem. When a BNC cable is connected to either the 50/75 ohm or 1 megohm input and the cable shield is connected to ground contact of another instrument or the ground contact from a wall outlet, spikes walk across the screen. These spikes are about 20dBm above the noise floor. Maybe your home has been equipped with a smart electric meter. Mine has and it transmits on about 160 kHz and has strong harmonics up to about 10 MHz. It's a very wide band digital signal that is on the power lines. It can easily be induced into any ground loop which is what you are creating when you connect two pieces of grounded equipment together with a coaxial cable. If you can get hold of one of those ancient 2 to 3 prong AC adapters it is good for breaking ground loops. If you have one try it and see if the spikes go away. Regards. Max. K 4 O DS. Email: max@... Transistor site Vacuum tube site: Woodworking site Music site: To subscribe to the fun with transistors group send an email to. funwithtransistors-subscribe@... To subscribe to the fun with tubes group send an email to, funwithtubes-subscribe@... To subscribe to the fun with wood group send a blank email to funwithwood-subscribe@... ----- Original Message ----- From: "gldinnc" <GLDuff@...> To: <hp_agilent_equipment@...> Sent: Wednesday, January 09, 2013 4:24 PM Subject: [hp_agilent_equipment] Failing 3585A Spectrum Analyzer My faithful 3585a has developed a problem. When a BNC cable is connected to either the 50/75 ohm or 1 megohm input and the cable shield is connected to ground contact of another instrument or the ground contact from a wall outlet, spikes walk across the screen. These spikes are about 20dBm above the noise floor. If I set the instrument to its default settings, I see the spikes walk across the screen. If I save the waveform to B (where it no longer walks), I count 24 spikes. On many start-stop frequency settings I see the same count of spikes. However, on some I see 48 spikes; e.g. a start frequency of 1MHz and a stop of 2MHz.
Anyone have a suggestion of how to diagnose and repair this problem?
Can the work be done without extender boards?
Thanks,
Gordon, KA2NLM
------------------------------------
Yahoo! Groups Links
|
I would not do that, even if it's just an experiment. Defeating the safety ground on most HP/Agilent equipment can damage it and other equipment that is attached. If the instrument has a line filter (and almost all of them do), defeating the earth/ground wire will float the chassis up to some significant portion of the line voltage. Though the impedance will be relatively high, chances are good that you will blow up something like its mixer or a preamp that might be attached to the analyzer as soon as you complete the ground by touching the equipment. This used to be a common way to damage spectrum analyzer front ends.
Vladan
toggle quoted message
Show quoted text
--- In hp_agilent_equipment@..., "Max Robinson" wrote: If you can get hold of one of those ancient 2 to 3 prong AC adapters it is good for breaking ground loops. If you have one try it and see if the spikes go away.
Regards.
Max. K 4 O DS.
Email: max@...
|
You need to determine whether the spikes are actual input signals from ground loops, or artifacts in the IF or display system. Can you measure the frequency spacing to see what the apparent interfering frequency is? Also, try different sweep rates and spans to see if the frequency measurement is consistent, or if it's dependent on the display function. Also, does the 3585A use a switching power supply, or a linear type?
Knowing the frequency will help to diagnose the problem and source.
Ed
toggle quoted message
Show quoted text
--- In hp_agilent_equipment@..., "Max Robinson" wrote: Gordon wrote.
My faithful 3585a has developed a problem. When a BNC cable is connected to either the 50/75 ohm or 1 megohm input and the cable shield is connected to ground contact of another instrument or the ground contact from a wall outlet, spikes walk across the screen. These spikes are about 20dBm above the noise floor.
Maybe your home has been equipped with a smart electric meter. Mine has and it transmits on about 160 kHz and has strong harmonics up to about 10 MHz. It's a very wide band digital signal that is on the power lines. It can easily be induced into any ground loop which is what you are creating when you connect two pieces of grounded equipment together with a coaxial cable. If you can get hold of one of those ancient 2 to 3 prong AC adapters it is good for breaking ground loops. If you have one try it and see if the spikes go away.
Regards.
Max. K 4 O DS.
Email: max@...
Transistor site Vacuum tube site: Woodworking site
Music site:
To subscribe to the fun with transistors group send an email to. funwithtransistors-subscribe@...
To subscribe to the fun with tubes group send an email to, funwithtubes-subscribe@...
To subscribe to the fun with wood group send a blank email to funwithwood-subscribe@...
----- Original Message ----- From: "gldinnc" To: Sent: Wednesday, January 09, 2013 4:24 PM Subject: [hp_agilent_equipment] Failing 3585A Spectrum Analyzer
My faithful 3585a has developed a problem. When a BNC cable is connected to either the 50/75 ohm or 1 megohm input and the cable shield is connected to ground contact of another instrument or the ground contact from a wall outlet, spikes walk across the screen. These spikes are about 20dBm above the noise floor.
If I set the instrument to its default settings, I see the spikes walk across the screen. If I save the waveform to B (where it no longer walks), I count 24 spikes. On many start-stop frequency settings I see the same count of spikes. However, on some I see 48 spikes; e.g. a start frequency of 1MHz and a stop of 2MHz.
Anyone have a suggestion of how to diagnose and repair this problem?
Can the work be done without extender boards?
Thanks,
Gordon, KA2NLM
------------------------------------
Yahoo! Groups Links
|
In response to the replies to date:
1. The spikes stop moving if I trigger off the line. 2. The number of spikes is not depended on frequency; therefore the frequency between spikes varies. 3. The 3585a worked fine until recently. No wiring changes have been made. Two different runs back to the breaker box have been used with the same result.
More suggestions please,
Gordon, KA2NLM
toggle quoted message
Show quoted text
--- In hp_agilent_equipment@..., "gldinnc" wrote: My faithful 3585a has developed a problem. When a BNC cable is connected to either the 50/75 ohm or 1 megohm input and the cable shield is connected to ground contact of another instrument or the ground contact from a wall outlet, spikes walk across the screen. These spikes are about 20dBm above the noise floor.
If I set the instrument to its default settings, I see the spikes walk across the screen. If I save the waveform to B (where it no longer walks), I count 24 spikes. On many start-stop frequency settings I see the same count of spikes. However, on some I see 48 spikes; e.g. a start frequency of 1MHz and a stop of 2MHz.
Anyone have a suggestion of how to diagnose and repair this problem?
Can the work be done without extender boards?
Thanks,
Gordon, KA2NLM
|
By the way, I have scanned the power supply trouble shooting section and it seems like most of the likely steps call for the use of extender boards. Is it realistic to attempt repair without extender boards?
Gordon, KA2NLM
toggle quoted message
Show quoted text
--- In hp_agilent_equipment@..., "gldinnc" wrote: In response to the replies to date:
1. The spikes stop moving if I trigger off the line. 2. The number of spikes is not depended on frequency; therefore the frequency between spikes varies. 3. The 3585a worked fine until recently. No wiring changes have been made. Two different runs back to the breaker box have been used with the same result.
More suggestions please,
Gordon, KA2NLM
--- In hp_agilent_equipment@..., "gldinnc" wrote:
My faithful 3585a has developed a problem. When a BNC cable is connected to either the 50/75 ohm or 1 megohm input and the cable shield is connected to ground contact of another instrument or the ground contact from a wall outlet, spikes walk across the screen. These spikes are about 20dBm above the noise floor.
If I set the instrument to its default settings, I see the spikes walk across the screen. If I save the waveform to B (where it no longer walks), I count 24 spikes. On many start-stop frequency settings I see the same count of spikes. However, on some I see 48 spikes; e.g. a start frequency of 1MHz and a stop of 2MHz.
Anyone have a suggestion of how to diagnose and repair this problem?
Can the work be done without extender boards?
Thanks,
Gordon, KA2NLM
|
Hi Gordon,
Since you managed to stop the spikes from moving with LINE TRIGGER, I would say that you have a power supply filtering problem. I don't remember the inside of the 3585 any more, but I suspect that you could easily use a scope to check all the power supplies for ripple. That would be the first thing I would do and you may not need the extenders for that.
Vladan
toggle quoted message
Show quoted text
--- In hp_agilent_equipment@..., "gldinnc" wrote: By the way, I have scanned the power supply trouble shooting section and it seems like most of the likely steps call for the use of extender boards. Is it realistic to attempt repair without extender boards?
Gordon, KA2NLM
--- In hp_agilent_equipment@..., "gldinnc" wrote:
In response to the replies to date:
1. The spikes stop moving if I trigger off the line. 2. The number of spikes is not depended on frequency; therefore the frequency between spikes varies. 3. The 3585a worked fine until recently. No wiring changes have been made. Two different runs back to the breaker box have been used with the same result.
More suggestions please,
Gordon, KA2NLM
|
To repair without extender boards on boards where frequencies are not critical [ ie, power supply] you can solder short stiff wires to points you want to monitor, the wire long enough to be accessable above the board. Can be very time consuming but quite practical. Obviously you can't use this approach where the inductance / capacitance of the wire significantly effects the circuit. Ken g8beq
|
Since line trigger seems to hold it, and, if I understand what you mean by it being "not depend on frequency" to mean the span has no effect, then it must be an internal problem of line ripple getting into the IF or display. But what about at different sweep rates? If it is ripple getting in, the apparent spacing should change with sweep rate, but still be steady as long as it's line-triggered.
Once this is determined, the remaining mystery is why it's affected by external ground loops. It could be an internal cap or other failure (even mechanical), causing excessive ripple currents flowing within the chassis or ground structure - maybe even something in the line filter could do it.
I believe you mentioned the spikes were down near the noise floor, so the actual interfering signal amplitudes involved may be quite small and hard to spot, depending on where and how it's actually getting in.
A large amount of ripple on a power supply could do that, so as Vladan suggested, check those first. But, even if that's the cause, it doesn't seem to explain the mystery part, unless it's somehow causing internal ground loops as I mentioned above. Otherwise, it should be present all the time, regardless of external connections.
Ed
toggle quoted message
Show quoted text
--- In hp_agilent_equipment@..., "pianovt" wrote: Hi Gordon,
Since you managed to stop the spikes from moving with LINE TRIGGER, I would say that you have a power supply filtering problem. I don't remember the inside of the 3585 any more, but I suspect that you could easily use a scope to check all the power supplies for ripple. That would be the first thing I would do and you may not need the extenders for that.
Vladan
--- In hp_agilent_equipment@..., "gldinnc" wrote:
By the way, I have scanned the power supply trouble shooting section and it seems like most of the likely steps call for the use of extender boards. Is it realistic to attempt repair without extender boards?
Gordon, KA2NLM
--- In hp_agilent_equipment@..., "gldinnc" wrote:
In response to the replies to date:
1. The spikes stop moving if I trigger off the line. 2. The number of spikes is not depended on frequency; therefore the frequency between spikes varies. 3. The 3585a worked fine until recently. No wiring changes have been made. Two different runs back to the breaker box have been used with the same result.
More suggestions please,
Gordon, KA2NLM
|
Thanks to Vladan, Ken and Ed for their suggestions.
I did a little more testing. Here is what I found:
1. With a battery operated source, everything seems fine. 2. If an external ground is connected to the ground of the tracking generator BNC or the external trigger BNC, no spikes appear. 3. Connecting external ground to the 1meg or 50/75 ground generates spikes. 4. If the span is less than 500kHz the spikes disappear (or seem minor), but if the span goes from 499,999 to 500,000 they hop up to the reported level.
While I can pull the covers off and begin to test power supply ripple, I would hope to get a little more directed.
Thoughts?
Gordon, KA2NLM
|
Hi Gordon,
I assume the battery operated source is a battery with an inverter that creates 120Vac/60Hz, right? Does the inverter output have a safety ground that is connected to earth?
If not, please do the following experiment. Disconnect everything from the 3585 except for the power cord and plug it into your usual ac line socket. Turn the 3585 on. Take a multimeter and put it on AC Volts . Attach one lead of the meter to something that you know is grounded (earth) and touch the other meter lead to the BNC shells and any conductive chassis parts on the 3585. Is there any measurable voltage?
Vladan
toggle quoted message
Show quoted text
--- In hp_agilent_equipment@..., "gldinnc" wrote: Thanks to Vladan, Ken and Ed for their suggestions.
I did a little more testing. Here is what I found:
1. With a battery operated source, everything seems fine. 2. If an external ground is connected to the ground of the tracking generator BNC or the external trigger BNC, no spikes appear. 3. Connecting external ground to the 1meg or 50/75 ground generates spikes. 4. If the span is less than 500kHz the spikes disappear (or seem minor), but if the span goes from 499,999 to 500,000 they hop up to the reported level.
While I can pull the covers off and begin to test power supply ripple, I would hope to get a little more directed.
Thoughts?
Gordon, KA2NLM
|
Vladan,
I was a little too brief (confusing) in my first point. I was using a battery operated signal source to put a test signal on the 3585a, not to power the 3585a.
I have conducted the suggested test using a Fluke 189. I get no voltage across any of the BNC shells or the 3585a chassis.
Thanks for sticking with me on this. I really would like to get it to work.
Gordon.
toggle quoted message
Show quoted text
--- In hp_agilent_equipment@..., "pianovt" wrote: Hi Gordon,
I assume the battery operated source is a battery with an inverter that creates 120Vac/60Hz, right? Does the inverter output have a safety ground that is connected to earth?
If not, please do the following experiment. Disconnect everything from the 3585 except for the power cord and plug it into your usual ac line socket. Turn the 3585 on. Take a multimeter and put it on AC Volts . Attach one lead of the meter to something that you know is grounded (earth) and touch the other meter lead to the BNC shells and any conductive chassis parts on the 3585. Is there any measurable voltage?
Vladan
--- In hp_agilent_equipment@..., "gldinnc" wrote:
Thanks to Vladan, Ken and Ed for their suggestions.
I did a little more testing. Here is what I found:
1. With a battery operated source, everything seems fine. 2. If an external ground is connected to the ground of the tracking generator BNC or the external trigger BNC, no spikes appear. 3. Connecting external ground to the 1meg or 50/75 ground generates spikes. 4. If the span is less than 500kHz the spikes disappear (or seem minor), but if the span goes from 499,999 to 500,000 they hop up to the reported level.
While I can pull the covers off and begin to test power supply ripple, I would hope to get a little more directed.
Thoughts?
Gordon, KA2NLM
|
Hi Gordon,
I will have to find a manual for the 3585A before I suggest anything else.
Vladan
toggle quoted message
Show quoted text
--- In hp_agilent_equipment@..., "gldinnc" wrote: Vladan,
I was a little too brief (confusing) in my first point. I was using a battery operated signal source to put a test signal on the 3585a, not to power the 3585a.
I have conducted the suggested test using a Fluke 189. I get no voltage across any of the BNC shells or the 3585a chassis.
Thanks for sticking with me on this. I really would like to get it to work.
Gordon.
--- In hp_agilent_equipment@..., "pianovt" wrote:
Hi Gordon,
I assume the battery operated source is a battery with an inverter that creates 120Vac/60Hz, right? Does the inverter output have a safety ground that is connected to earth?
If not, please do the following experiment. Disconnect everything from the 3585 except for the power cord and plug it into your usual ac line socket. Turn the 3585 on. Take a multimeter and put it on AC Volts . Attach one lead of the meter to something that you know is grounded (earth) and touch the other meter lead to the BNC shells and any conductive chassis parts on the 3585. Is there any measurable voltage?
Vladan
--- In hp_agilent_equipment@..., "gldinnc" wrote:
Thanks to Vladan, Ken and Ed for their suggestions.
I did a little more testing. Here is what I found:
1. With a battery operated source, everything seems fine. 2. If an external ground is connected to the ground of the tracking generator BNC or the external trigger BNC, no spikes appear. 3. Connecting external ground to the 1meg or 50/75 ground generates spikes. 4. If the span is less than 500kHz the spikes disappear (or seem minor), but if the span goes from 499,999 to 500,000 they hop up to the reported level.
While I can pull the covers off and begin to test power supply ripple, I would hope to get a little more directed.
Thoughts?
Gordon, KA2NLM
|
Hi Gordon,
If I were troubleshooting this problem, I would proceed as follows:
1. Open the top cover and check the power supplies for ripple. I would do this first, because it's the easiest thing to do and a bad filter capacitor might be the problem. On page 8-10 of Service Manual Vol. 2, you will see all the test points where you can probe with the scope and measure D.C.
2. If that all looks fine (no ripple on the supplies), my next guess would be to look at Figure 10-38 (service manual Vol. 2) "Power Distribution". In the upper right corner is what looks like a common mode filter T2. This filter is hooked up to the U1 and U3 regulators. Check the output of the regulators for ripple. Look at T2 to see if something is wrong with it. The reason I zoomed in on that circuit is because terminals 1 and 4 of T2 isolate the power supply ground from the input board ground. On the schematics, the power supply ground is called GND, and the ground for the front end is called GND1. In other words, T2 is where the BNC shells of your input connectors get their ground (GND1) and that ground is not the same as the power supply ground (GND). Using an Ohm meter, you should however see continuity (probably 0 Ohms) between the shells of the BNCs (GND1) and the power supply ground (GND).
Other than that, I don't know what to suggest. Best luck,
Vladan
toggle quoted message
Show quoted text
--- In hp_agilent_equipment@..., "gldinnc" wrote: Vladan,
I was a little too brief (confusing) in my first point. I was using a battery operated signal source to put a test signal on the 3585a, not to power the 3585a.
I have conducted the suggested test using a Fluke 189. I get no voltage across any of the BNC shells or the 3585a chassis.
Thanks for sticking with me on this. I really would like to get it to work.
Gordon.
|
Gordon,
Another possibility is that W2 which connects A70 Power Supply to A76 Power Distribution isn't making a good ground connection between the two. (Fig. 10-38, Vol. 2)
Vladan
toggle quoted message
Show quoted text
--- In hp_agilent_equipment@..., "pianovt" wrote:
Hi Gordon,
If I were troubleshooting this problem, I would proceed as follows:
1. Open the top cover and check the power supplies for ripple. I would do this first, because it's the easiest thing to do and a bad filter capacitor might be the problem. On page 8-10 of Service Manual Vol. 2, you will see all the test points where you can probe with the scope and measure D.C.
2. If that all looks fine (no ripple on the supplies), my next guess would be to look at Figure 10-38 (service manual Vol. 2) "Power Distribution". In the upper right corner is what looks like a common mode filter T2. This filter is hooked up to the U1 and U3 regulators. Check the output of the regulators for ripple. Look at T2 to see if something is wrong with it. The reason I zoomed in on that circuit is because terminals 1 and 4 of T2 isolate the power supply ground from the input board ground. On the schematics, the power supply ground is called GND, and the ground for the front end is called GND1. In other words, T2 is where the BNC shells of your input connectors get their ground (GND1) and that ground is not the same as the power supply ground (GND). Using an Ohm meter, you should however see continuity (probably 0 Ohms) between the shells of the BNCs (GND1) and the power supply ground (GND).
Other than that, I don't know what to suggest. Best luck,
Vladan
--- In hp_agilent_equipment@..., "gldinnc" wrote:
Vladan,
I was a little too brief (confusing) in my first point. I was using a battery operated signal source to put a test signal on the 3585a, not to power the 3585a.
I have conducted the suggested test using a Fluke 189. I get no voltage across any of the BNC shells or the 3585a chassis.
Thanks for sticking with me on this. I really would like to get it to work.
Gordon.
|
Vladan,
Here are the results of my testing :(
1. The test points on A71-A74 all are well within tolerance and show no ripple. There is some pulse noise in the 20KHz range, but I would guess that is normal (I have a 200 Meg Scope).
2. I checked the power out of A76 and there is no ripple on J3, J7, J9 and J10. Again there is a little high frequency pulse noise.
3. Resistance is nominally 0 between A76 J10 and the BNC shell.
What I find curious, is that the spikes on the display only happen when the frequency span is 500,000 Hz or higher. Thoughts on this?
Thanks,
Gordon.
toggle quoted message
Show quoted text
--- In hp_agilent_equipment@..., "pianovt" wrote:
Hi Gordon,
If I were troubleshooting this problem, I would proceed as follows:
1. Open the top cover and check the power supplies for ripple. I would do this first, because it's the easiest thing to do and a bad filter capacitor might be the problem. On page 8-10 of Service Manual Vol. 2, you will see all the test points where you can probe with the scope and measure D.C.
2. If that all looks fine (no ripple on the supplies), my next guess would be to look at Figure 10-38 (service manual Vol. 2) "Power Distribution". In the upper right corner is what looks like a common mode filter T2. This filter is hooked up to the U1 and U3 regulators. Check the output of the regulators for ripple. Look at T2 to see if something is wrong with it. The reason I zoomed in on that circuit is because terminals 1 and 4 of T2 isolate the power supply ground from the input board ground. On the schematics, the power supply ground is called GND, and the ground for the front end is called GND1. In other words, T2 is where the BNC shells of your input connectors get their ground (GND1) and that ground is not the same as the power supply ground (GND). Using an Ohm meter, you should however see continuity (probably 0 Ohms) between the shells of the BNCs (GND1) and the power supply ground (GND).
Other than that, I don't know what to suggest. Best luck,
Vladan
--- In hp_agilent_equipment@..., "gldinnc" wrote:
Vladan,
I was a little too brief (confusing) in my first point. I was using a battery operated signal source to put a test signal on the 3585a, not to power the 3585a.
I have conducted the suggested test using a Fluke 189. I get no voltage across any of the BNC shells or the 3585a chassis.
Thanks for sticking with me on this. I really would like to get it to work.
Gordon.
|
Hi Gordon,
This is almost certainly related to the way the LO is structured. Generally speaking, the LO has to make some changes as the span increases. In your case, for spans that are 500 kHz and larger, the LO uses a different method of generating the signal. I will look at the manual, but this may not be the easiest way to find a bad part, especially since you see the problem (to a lesser extent) in narrower spans as well.
Vladan
toggle quoted message
Show quoted text
--- In hp_agilent_equipment@..., "gldinnc" wrote: What I find curious, is that the spikes on the display only happen when the frequency span is 500,000 Hz or higher. Thoughts on this?
|
Many SAs that I know of change operating modes somewhere in their span range, from the LO being fully and continuously synthesized, to using open-loop linear sweep tuning. Once the span is large enough, there's no need for precise frequency control, since you couldn't resolve it on screen anyway.
The 500 kHz point is probably where this happens on the 3585A, so look for what things happen internally at this mode change.
Ed
toggle quoted message
Show quoted text
--- In hp_agilent_equipment@..., "gldinnc" wrote: Vladan,
Here are the results of my testing :(
1. The test points on A71-A74 all are well within tolerance and show no ripple. There is some pulse noise in the 20KHz range, but I would guess that is normal (I have a 200 Meg Scope).
2. I checked the power out of A76 and there is no ripple on J3, J7, J9 and J10. Again there is a little high frequency pulse noise.
3. Resistance is nominally 0 between A76 J10 and the BNC shell.
What I find curious, is that the spikes on the display only happen when the frequency span is 500,000 Hz or higher. Thoughts on this?
Thanks,
Gordon.
--- In hp_agilent_equipment@..., "pianovt" wrote:
Hi Gordon,
If I were troubleshooting this problem, I would proceed as follows:
1. Open the top cover and check the power supplies for ripple. I would do this first, because it's the easiest thing to do and a bad filter capacitor might be the problem. On page 8-10 of Service Manual Vol. 2, you will see all the test points where you can probe with the scope and measure D.C.
2. If that all looks fine (no ripple on the supplies), my next guess would be to look at Figure 10-38 (service manual Vol. 2) "Power Distribution". In the upper right corner is what looks like a common mode filter T2. This filter is hooked up to the U1 and U3 regulators. Check the output of the regulators for ripple. Look at T2 to see if something is wrong with it. The reason I zoomed in on that circuit is because terminals 1 and 4 of T2 isolate the power supply ground from the input board ground. On the schematics, the power supply ground is called GND, and the ground for the front end is called GND1. In other words, T2 is where the BNC shells of your input connectors get their ground (GND1) and that ground is not the same as the power supply ground (GND). Using an Ohm meter, you should however see continuity (probably 0 Ohms) between the shells of the BNCs (GND1) and the power supply ground (GND).
Other than that, I don't know what to suggest. Best luck,
Vladan
--- In hp_agilent_equipment@..., "gldinnc" wrote:
Vladan,
I was a little too brief (confusing) in my first point. I was using a battery operated signal source to put a test signal on the 3585a, not to power the 3585a.
I have conducted the suggested test using a Fluke 189. I get no voltage across any of the BNC shells or the 3585a chassis.
Thanks for sticking with me on this. I really would like to get it to work.
Gordon.
|
Hi Gordon,
When you change the span, other things in the analyzer change too. The resolution bandwidth and video bandwidth probably change at the 500kHz span setting as well. That means you can do another experiment:
Set the span to 500 kHz so you can see the spikes and then defeat the automatic coupling for the IF resolution bandwidth. Step the resolution bandwidth up and down and see if the spikes change as a function of res. bw. alone. Repeat for the video bandwidths.
Vladan
toggle quoted message
Show quoted text
--- In hp_agilent_equipment@..., "pianovt" wrote: Hi Gordon,
This is almost certainly related to the way the LO is structured. Generally speaking, the LO has to make some changes as the span increases. In your case, for spans that are 500 kHz and larger, the LO uses a different method of generating the signal. I will look at the manual, but this may not be the easiest way to find a bad part, especially since you see the problem (to a lesser extent) in narrower spans as well.
Vladan
--- In hp_agilent_equipment@..., "gldinnc" wrote:
What I find curious, is that the spikes on the display only happen when the frequency span is 500,000 Hz or higher. Thoughts on this?
|