Vladan- I got the cover off without trouble. By testing I note that when the middle (40 dB) switch is triggered, the output turns off completely. As far as I can determine, the plunger is actuated correctly. Moving it manually also results in the loss of output. The other two switches (10 and 20 dB) work fine. I suppose that either the switch contacts are failing or the 40 dB resistor is open. Apparently both switches and resistors are inside the two-piece machined steel block. All the RF path seems to be inside. It would appear that I need to get this apart, but I am not sure how to. There are quite a few screws attaching the solenoids to the steel block, but I am not seeing what is holding the steel block together. This attenuator has a PC board with several chips on it. I have noticed before that HP seems to have adopted the philosophy that if 2 chips are good, then 20 must be better. It is a surprisingly complicated device. Any advice?? Thanks! Jim At 08:45 PM 12/30/2012, you wrote:
Jim,
You will have to pull out the front panel. There are a few flat-head screws along the front frame (top and bottom, maybe even sides - don't remember any more). Once you remove them, the front panel pulls out of the front frame, though it still has a couple of umbilicals attached. From there, you will see how to get the attenuator out.
The most common failure has to do with the tiny rubber O-rings on the plungers. This is all delicate magnifier work. Do not succumb to the "magnetized housing" myth of repair. The attenuator depends on magnets for latching. It's just that when the O-ring fails and either splits or falls off, the mechanism shifts closer to the magnet and the force of magnetic attraction increases. At that point, the solenoid (which has nothing wrong with it) no longer can pull the parts away from the magnet (which also has nothing wrong with it). This is when some people decide that demagnetizing everything will fix things. If you have the patience, study the design, it's interesting. The goal was to use current only when switching a stage is required and then interrupt all current to the coils as soon as the task is accomplished.
You open the attenuator by removing one of the end caps and sliding the U-shaped steel cover off. It comes off relatively easily. Don't try to remove it by prying with screwdrivers.
There are two versions of the attenuators. The older ones are all electro-mechanical, the newer ones have a p.c. board with some ICs on them.
Good luck, they can be fixed, but the work requires some patience and you may have to order no O-rings from Agilent.
Vladan
--- In <mailto:hp_agilent_equipment%40yahoogroups.com>hp_agilent_equipment@..., Jim Schatzman <james.schatzman@...> wrote:
All-
O.k., I have an 8568B with a bad attenuator (40dB and below are dead).
My big question is - how do you get access to the attenuator to take it apart and clean/repair? It seems to be really buried under a maze of stuff...
Thanks!
Jim
At 01:44 PM 12/27/2012, you wrote:
Thanks a lot of this.
I just managed to get a 8568B that attenuator has some problems with the 20dB step.
These instructions encourage me to open the attenuator and clean the pads with isopropyl alcohol. Now it works fine.
I didn't made a testped for it i just cold moved carefully the solenoids and checked with dg8saq vnwa the attenuator pads. I'm not sure is the solenoid stuck or is it actually a contact failure.
The instructions how to disaasemble the attenuator was very accurate and easy to follow.
BR, Jarmo
[Non-text portions of this message have been removed]
[Non-text portions of this message have been removed]
|
Hi Jim,
Do not unscrew the solenoids on the top, at least not for now. To see what's happening, you will have to open the lower part where the RF action is.
The two SMA connectors have hex-nuts. Remove them. There is thin name plate that you can now remove. It covers the whole area between the SMA connectors. It probably has a few glue spots on its other side, so carefully pry it off without bending the it. Now, you will see a bunch of socket cap screws. When you remove them, you will be able to carefully separate the parts and gain access to the attenuator pads and plungers. The rubber O-rings are on the tiny nylon plungers that are attached to the gold plated spring contacts. I think there are two rings on each plunger, one on the RF side and one on the solenoid side.
Vladan
toggle quoted message
Show quoted text
--- In hp_agilent_equipment@..., Jim Schatzman <james.schatzman@...> wrote: Vladan-
I got the cover off without trouble. By testing I note that when the middle (40 dB) switch is triggered, the output turns off completely. As far as I can determine, the plunger is actuated correctly. Moving it manually also results in the loss of output. The other two switches (10 and 20 dB) work fine.
I suppose that either the switch contacts are failing or the 40 dB resistor is open. Apparently both switches and resistors are inside the two-piece machined steel block. All the RF path seems to be inside. It would appear that I need to get this apart, but I am not sure how to. There are quite a few screws attaching the solenoids to the steel block, but I am not seeing what is holding the steel block together.
This attenuator has a PC board with several chips on it. I have noticed before that HP seems to have adopted the philosophy that if 2 chips are good, then 20 must be better. It is a surprisingly complicated device.
Any advice??
Thanks!
Jim
At 08:45 PM 12/30/2012, you wrote:
Jim,
You will have to pull out the front panel. There are a few flat-head screws along the front frame (top and bottom, maybe even sides - don't remember any more). Once you remove them, the front panel pulls out of the front frame, though it still has a couple of umbilicals attached. From there, you will see how to get the attenuator out.
The most common failure has to do with the tiny rubber O-rings on the plungers. This is all delicate magnifier work. Do not succumb to the "magnetized housing" myth of repair. The attenuator depends on magnets for latching. It's just that when the O-ring fails and either splits or falls off, the mechanism shifts closer to the magnet and the force of magnetic attraction increases. At that point, the solenoid (which has nothing wrong with it) no longer can pull the parts away from the magnet (which also has nothing wrong with it). This is when some people decide that demagnetizing everything will fix things. If you have the patience, study the design, it's interesting. The goal was to use current only when switching a stage is required and then interrupt all current to the coils as soon as the task is accomplished.
You open the attenuator by removing one of the end caps and sliding the U-shaped steel cover off. It comes off relatively easily. Don't try to remove it by prying with screwdrivers.
There are two versions of the attenuators. The older ones are all electro-mechanical, the newer ones have a p.c. board with some ICs on them.
Good luck, they can be fixed, but the work requires some patience and you may have to order no O-rings from Agilent.
Vladan
--- In <mailto:hp_agilent_equipment%40yahoogroups.com>hp_agilent_equipment@..., Jim Schatzman <james.schatzman@> wrote:
All-
O.k., I have an 8568B with a bad attenuator (40dB and below are dead).
My big question is - how do you get access to the attenuator to take it apart and clean/repair? It seems to be really buried under a maze of stuff...
Thanks!
Jim
At 01:44 PM 12/27/2012, you wrote:
Thanks a lot of this.
I just managed to get a 8568B that attenuator has some problems with the 20dB step.
These instructions encourage me to open the attenuator and clean the pads with isopropyl alcohol. Now it works fine.
I didn't made a testped for it i just cold moved carefully the solenoids and checked with dg8saq vnwa the attenuator pads. I'm not sure is the solenoid stuck or is it actually a contact failure.
The instructions how to disaasemble the attenuator was very accurate and easy to follow.
BR, Jarmo
|
I've got an attenuator for an 8672A that suffers from failed O-rings. In this case it appears someone used something nasty to try to clean the spring contacts and attenuator pads and whatever was used dripped down the nylon plungers and turned all the O-rings to a mushy black paste. Not a single survivor. Agilent no longer stocks them but provides the dimensions and material type so I just need to research a source.
Steve
toggle quoted message
Show quoted text
On 12/30/2012 10:16 PM, pianovt wrote: Hi Jim,
Do not unscrew the solenoids on the top, at least not for now. To see what's happening, you will have to open the lower part where the RF action is.
The two SMA connectors have hex-nuts. Remove them. There is thin name plate that you can now remove. It covers the whole area between the SMA connectors. It probably has a few glue spots on its other side, so carefully pry it off without bending the it. Now, you will see a bunch of socket cap screws. When you remove them, you will be able to carefully separate the parts and gain access to the attenuator pads and plungers. The rubber O-rings are on the tiny nylon plungers that are attached to the gold plated spring contacts. I think there are two rings on each plunger, one on the RF side and one on the solenoid side.
Vladan
--- In hp_agilent_equipment@... <mailto:hp_agilent_equipment%40yahoogroups.com>, Jim Schatzman <james.schatzman@...> wrote:
Vladan-
I got the cover off without trouble. By testing I note that when the middle (40 dB) switch is triggered, the output turns off completely. As far as I can determine, the plunger is actuated correctly. Moving it manually also results in the loss of output. The other two switches (10 and 20 dB) work fine.
I suppose that either the switch contacts are failing or the 40 dB resistor is open. Apparently both switches and resistors are inside the two-piece machined steel block. All the RF path seems to be inside. It would appear that I need to get this apart, but I am not sure how to. There are quite a few screws attaching the solenoids to the steel block, but I am not seeing what is holding the steel block together.
This attenuator has a PC board with several chips on it. I have noticed before that HP seems to have adopted the philosophy that if 2 chips are good, then 20 must be better. It is a surprisingly complicated device.
Any advice??
Thanks!
Jim
At 08:45 PM 12/30/2012, you wrote:
Jim,
You will have to pull out the front panel. There are a few
flat-head screws along the front frame (top and bottom, maybe even sides - don't remember any more). Once you remove them, the front panel pulls out of the front frame, though it still has a couple of umbilicals attached. From there, you will see how to get the attenuator out.
The most common failure has to do with the tiny rubber O-rings on
the plungers. This is all delicate magnifier work. Do not succumb to the "magnetized housing" myth of repair. The attenuator depends on magnets for latching. It's just that when the O-ring fails and either splits or falls off, the mechanism shifts closer to the magnet and the force of magnetic attraction increases. At that point, the solenoid (which has nothing wrong with it) no longer can pull the parts away from the magnet (which also has nothing wrong with it). This is when some people decide that demagnetizing everything will fix things. If you have the patience, study the design, it's interesting. The goal was to use current only when switching a stage is required and then interrupt all current to the coils as soon as the task is accomplished.
You open the attenuator by removing one of the end caps and sliding
the U-shaped steel cover off. It comes off relatively easily. Don't try to remove it by prying with screwdrivers.
There are two versions of the attenuators. The older ones are all
electro-mechanical, the newer ones have a p.c. board with some ICs on them.
Good luck, they can be fixed, but the work requires some patience
and you may have to order no O-rings from Agilent.
Vladan
--- In
<mailto:hp_agilent_equipment%40yahoogroups.com>hp_agilent_equipment@... <mailto:hp_agilent_equipment%40yahoogroups.com>, Jim Schatzman <james.schatzman@> wrote:
All-
O.k., I have an 8568B with a bad attenuator (40dB and below are
dead).
My big question is - how do you get access to the attenuator to
take it apart and clean/repair? It seems to be really buried under a maze of stuff...
Thanks!
Jim
At 01:44 PM 12/27/2012, you wrote:
Thanks a lot of this.
I just managed to get a 8568B that attenuator has some problems
with the 20dB step.
These instructions encourage me to open the attenuator and clean
the pads with isopropyl alcohol. Now it works fine.
I didn't made a testped for it i just cold moved carefully the
solenoids and checked with dg8saq vnwa the attenuator pads. I'm not sure is the solenoid stuck or is it actually a contact failure.
The instructions how to disaasemble the attenuator was very
accurate and easy to follow.
BR, Jarmo
|
Ran across these guys while designing a gas analyzer. They make all kinds of 0-Rings.
Karl
toggle quoted message
Show quoted text
--- In hp_agilent_equipment@..., Steve Krull <Steve-Krull@...> wrote: I've got an attenuator for an 8672A that suffers from failed O-rings. In this case it appears someone used something nasty to try to clean the spring contacts and attenuator pads and whatever was used dripped down the nylon plungers and turned all the O-rings to a mushy black paste. Not a single survivor. Agilent no longer stocks them but provides the dimensions and material type so I just need to research a source.
Steve
On 12/30/2012 10:16 PM, pianovt wrote:
Hi Jim,
Do not unscrew the solenoids on the top, at least not for now. To see what's happening, you will have to open the lower part where the RF action is.
The two SMA connectors have hex-nuts. Remove them. There is thin name plate that you can now remove. It covers the whole area between the SMA connectors. It probably has a few glue spots on its other side, so carefully pry it off without bending the it. Now, you will see a bunch of socket cap screws. When you remove them, you will be able to carefully separate the parts and gain access to the attenuator pads and plungers. The rubber O-rings are on the tiny nylon plungers that are attached to the gold plated spring contacts. I think there are two rings on each plunger, one on the RF side and one on the solenoid side.
Vladan
--- In hp_agilent_equipment@... <mailto:hp_agilent_equipment%40yahoogroups.com>, Jim Schatzman <james.schatzman@> wrote:
Vladan-
I got the cover off without trouble. By testing I note that when the middle (40 dB) switch is triggered, the output turns off completely. As far as I can determine, the plunger is actuated correctly. Moving it manually also results in the loss of output. The other two switches (10 and 20 dB) work fine.
I suppose that either the switch contacts are failing or the 40 dB resistor is open. Apparently both switches and resistors are inside the two-piece machined steel block. All the RF path seems to be inside. It would appear that I need to get this apart, but I am not sure how to. There are quite a few screws attaching the solenoids to the steel block, but I am not seeing what is holding the steel block together.
This attenuator has a PC board with several chips on it. I have noticed before that HP seems to have adopted the philosophy that if 2 chips are good, then 20 must be better. It is a surprisingly complicated device.
Any advice??
Thanks!
Jim
At 08:45 PM 12/30/2012, you wrote:
Jim,
You will have to pull out the front panel. There are a few
flat-head screws along the front frame (top and bottom, maybe even sides - don't remember any more). Once you remove them, the front panel pulls out of the front frame, though it still has a couple of umbilicals attached. From there, you will see how to get the attenuator out.
The most common failure has to do with the tiny rubber O-rings on
the plungers. This is all delicate magnifier work. Do not succumb to the "magnetized housing" myth of repair. The attenuator depends on magnets for latching. It's just that when the O-ring fails and either splits or falls off, the mechanism shifts closer to the magnet and the force of magnetic attraction increases. At that point, the solenoid (which has nothing wrong with it) no longer can pull the parts away from the magnet (which also has nothing wrong with it). This is when some people decide that demagnetizing everything will fix things. If you have the patience, study the design, it's interesting. The goal was to use current only when switching a stage is required and then interrupt all current to the coils as soon as the task is accomplished.
You open the attenuator by removing one of the end caps and sliding
the U-shaped steel cover off. It comes off relatively easily. Don't try to remove it by prying with screwdrivers.
There are two versions of the attenuators. The older ones are all
electro-mechanical, the newer ones have a p.c. board with some ICs on them.
Good luck, they can be fixed, but the work requires some patience
and you may have to order no O-rings from Agilent.
Vladan
--- In
<mailto:hp_agilent_equipment%40yahoogroups.com>hp_agilent_equipment@... <mailto:hp_agilent_equipment%40yahoogroups.com>, Jim Schatzman <james.schatzman@> wrote:
All-
O.k., I have an 8568B with a bad attenuator (40dB and below are
dead).
My big question is - how do you get access to the attenuator to
take it apart and clean/repair? It seems to be really buried under a maze of stuff...
Thanks!
Jim
At 01:44 PM 12/27/2012, you wrote:
Thanks a lot of this.
I just managed to get a 8568B that attenuator has some problems
with the 20dB step.
These instructions encourage me to open the attenuator and clean
the pads with isopropyl alcohol. Now it works fine.
I didn't made a testped for it i just cold moved carefully the
solenoids and checked with dg8saq vnwa the attenuator pads. I'm not sure is the solenoid stuck or is it actually a contact failure.
The instructions how to disaasemble the attenuator was very
accurate and easy to follow.
BR, Jarmo
[Non-text portions of this message have been removed]
[Non-text portions of this message have been removed]
|
Thanks Karl - I've emailed them to see if they will sell in small quantities.
Steve K.
toggle quoted message
Show quoted text
On 12/31/2012 12:46 PM, Karl wrote:
Ran across these guys while designing a gas analyzer. They make all kinds of 0-Rings.
Karl
--- In hp_agilent_equipment@... <mailto:hp_agilent_equipment%40yahoogroups.com>, Steve Krull <Steve-Krull@...> wrote:
I've got an attenuator for an 8672A that suffers from failed O-rings. In
this case it appears someone used something nasty to try to clean the spring contacts and attenuator pads and whatever was used dripped down the nylon plungers and turned all the O-rings to a mushy black paste. Not a single survivor. Agilent no longer stocks them but provides the dimensions and material type so I just need to research a source.
Steve
On 12/30/2012 10:16 PM, pianovt wrote:
Hi Jim,
Do not unscrew the solenoids on the top, at least not for now. To see what's happening, you will have to open the lower part where the RF action is.
The two SMA connectors have hex-nuts. Remove them. There is thin name plate that you can now remove. It covers the whole area between the SMA connectors. It probably has a few glue spots on its other
side, so
carefully pry it off without bending the it. Now, you will see a bunch
of socket cap screws. When you remove them, you will be able to carefully separate the parts and gain access to the attenuator pads and plungers. The rubber O-rings are on the tiny nylon plungers that are attached to the gold plated spring contacts. I think there are two
rings on each plunger, one on the RF side and one on the solenoid side.
Vladan
--- In hp_agilent_equipment@...
<mailto:hp_agilent_equipment%40yahoogroups.com>
<mailto:hp_agilent_equipment%40yahoogroups.com>, Jim Schatzman <james.schatzman@> wrote:
Vladan-
I got the cover off without trouble. By testing I note that when
the
middle (40 dB) switch is triggered, the output turns off completely. As far as I can determine, the plunger is actuated correctly. Moving it manually also results in the loss of output. The other two switches
(10 and 20 dB) work fine.
I suppose that either the switch contacts are failing or the 40 dB resistor is open. Apparently both switches and resistors are inside the two-piece machined steel block. All the RF path seems to be inside. It would appear that I need to get this apart, but I am not sure how to. There are quite a few screws attaching the solenoids to the steel block, but I am not seeing what is holding the steel block together.
This attenuator has a PC board with several chips on it. I have noticed before that HP seems to have adopted the philosophy that if 2 chips are good, then 20 must be better. It is a surprisingly complicated device.
Any advice??
Thanks!
Jim
At 08:45 PM 12/30/2012, you wrote:
Jim,
You will have to pull out the front panel. There are a few
flat-head screws along the front frame (top and bottom, maybe even sides - don't remember any more). Once you remove them, the front panel pulls out of the front frame, though it still has a couple of umbilicals attached. From there, you will see how to get the attenuator out.
The most common failure has to do with the tiny rubber O-rings on
the plungers. This is all delicate magnifier work. Do not succumb to the "magnetized housing" myth of repair. The attenuator depends on magnets for latching. It's just that when the O-ring fails and either splits or falls off, the mechanism shifts closer to the magnet and the
force of magnetic attraction increases. At that point, the solenoid (which has nothing wrong with it) no longer can pull the parts away from the magnet (which also has nothing wrong with it). This is when some people decide that demagnetizing everything will fix things. If you have the patience, study the design, it's interesting. The goal was to use current only when switching a stage is required and then interrupt all current to the coils as soon as the task is accomplished.
You open the attenuator by removing one of the end caps and
sliding
the U-shaped steel cover off. It comes off relatively easily. Don't try to remove it by prying with screwdrivers.
There are two versions of the attenuators. The older ones are all
electro-mechanical, the newer ones have a p.c. board with some ICs on them.
Good luck, they can be fixed, but the work requires some patience
and you may have to order no O-rings from Agilent.
Vladan
--- In
<mailto:hp_agilent_equipment%40yahoogroups.com>hp_agilent_equipment@... <mailto:hp_agilent_equipment%40yahoogroups.com>
<mailto:hp_agilent_equipment%40yahoogroups.com>, Jim Schatzman <james.schatzman@> wrote:
All-
O.k., I have an 8568B with a bad attenuator (40dB and below are
dead).
My big question is - how do you get access to the attenuator to
take it apart and clean/repair? It seems to be really buried under a maze of stuff...
Thanks!
Jim
At 01:44 PM 12/27/2012, you wrote:
Thanks a lot of this.
I just managed to get a 8568B that attenuator has some problems
with the 20dB step.
These instructions encourage me to open the attenuator and
clean
the pads with isopropyl alcohol. Now it works fine.
I didn't made a testped for it i just cold moved carefully the
solenoids and checked with dg8saq vnwa the attenuator pads. I'm not sure is the solenoid stuck or is it actually a contact failure.
The instructions how to disaasemble the attenuator was very
accurate and easy to follow.
BR, Jarmo
|
Do you guys have the dimensions or a suggested part number for replacements? Since these are for the 8568, they would most likely be used in other units. If they have a minimum quantity, maybe a group buy would be in order. Distribution would be easy with 1st class mail. Also, a better material might be a good change, maybe silicone, etc.
Thanks, Tom
toggle quoted message
Show quoted text
----- Original Message ----- From: Steve Krull To: hp_agilent_equipment@... Sent: Monday, December 31, 2012 3:21 PM Subject: Re: [hp_agilent_equipment] Re: DIY: Repair of HP 8568B Step Attenuators - another question...
Thanks Karl - I've emailed them to see if they will sell in small quantities.
Steve K.
On 12/31/2012 12:46 PM, Karl wrote: > > > Ran across these guys while designing a gas analyzer. They make > all kinds of 0-Rings. > > > > Karl > > --- In hp_agilent_equipment@... > <mailto:hp_agilent_equipment%40yahoogroups.com>, Steve Krull > <Steve-Krull@...> wrote: > > > > I've got an attenuator for an 8672A that suffers from failed > O-rings. In > > this case it appears someone used something nasty to try to clean the > > spring contacts and attenuator pads and whatever was used dripped down > > the nylon plungers and turned all the O-rings to a mushy black paste. > > Not a single survivor. Agilent no longer stocks them but provides the > > dimensions and material type so I just need to research a source. > > > > Steve > > > > > > On 12/30/2012 10:16 PM, pianovt wrote: > > > > > > Hi Jim, > > > > > > Do not unscrew the solenoids on the top, at least not for now. To see > > > what's happening, you will have to open the lower part where the RF > > > action is. > > > > > > The two SMA connectors have hex-nuts. Remove them. There is thin name > > > plate that you can now remove. It covers the whole area between the > > > SMA connectors. It probably has a few glue spots on its other > side, so > > > carefully pry it off without bending the it. Now, you will see a > bunch > > > of socket cap screws. When you remove them, you will be able to > > > carefully separate the parts and gain access to the attenuator pads > > > and plungers. The rubber O-rings are on the tiny nylon plungers that > > > are attached to the gold plated spring contacts. I think there are > two > > > rings on each plunger, one on the RF side and one on the solenoid > side. > > > > > > Vladan > > > > > > --- In hp_agilent_equipment@... > <mailto:hp_agilent_equipment%40yahoogroups.com> > > > <mailto:hp_agilent_equipment%40yahoogroups.com>, Jim Schatzman > > > <james.schatzman@> wrote: > > > > > > > > Vladan- > > > > > > > > I got the cover off without trouble. By testing I note that when > the > > > middle (40 dB) switch is triggered, the output turns off completely. > > > As far as I can determine, the plunger is actuated correctly. Moving > > > it manually also results in the loss of output. The other two > switches > > > (10 and 20 dB) work fine. > > > > > > > > I suppose that either the switch contacts are failing or the 40 dB > > > resistor is open. Apparently both switches and resistors are inside > > > the two-piece machined steel block. All the RF path seems to be > > > inside. It would appear that I need to get this apart, but I am not > > > sure how to. There are quite a few screws attaching the solenoids to > > > the steel block, but I am not seeing what is holding the steel block > > > together. > > > > > > > > This attenuator has a PC board with several chips on it. I have > > > noticed before that HP seems to have adopted the philosophy that if 2 > > > chips are good, then 20 must be better. It is a surprisingly > > > complicated device. > > > > > > > > Any advice?? > > > > > > > > Thanks! > > > > > > > > Jim > > > > > > > > > > > > > > > > At 08:45 PM 12/30/2012, you wrote: > > > > > > > > > > > > > > >Jim, > > > > > > > > > >You will have to pull out the front panel. There are a few > > > flat-head screws along the front frame (top and bottom, maybe even > > > sides - don't remember any more). Once you remove them, the front > > > panel pulls out of the front frame, though it still has a couple of > > > umbilicals attached. From there, you will see how to get the > > > attenuator out. > > > > > > > > > >The most common failure has to do with the tiny rubber O-rings on > > > the plungers. This is all delicate magnifier work. Do not succumb to > > > the "magnetized housing" myth of repair. The attenuator depends on > > > magnets for latching. It's just that when the O-ring fails and either > > > splits or falls off, the mechanism shifts closer to the magnet and > the > > > force of magnetic attraction increases. At that point, the solenoid > > > (which has nothing wrong with it) no longer can pull the parts away > > > from the magnet (which also has nothing wrong with it). This is when > > > some people decide that demagnetizing everything will fix things. If > > > you have the patience, study the design, it's interesting. The goal > > > was to use current only when switching a stage is required and then > > > interrupt all current to the coils as soon as the task is > accomplished. > > > > > > > > > >You open the attenuator by removing one of the end caps and > sliding > > > the U-shaped steel cover off. It comes off relatively easily. Don't > > > try to remove it by prying with screwdrivers. > > > > > > > > > >There are two versions of the attenuators. The older ones are all > > > electro-mechanical, the newer ones have a p.c. board with some ICs on > > > them. > > > > > > > > > >Good luck, they can be fixed, but the work requires some patience > > > and you may have to order no O-rings from Agilent. > > > > > > > > > >Vladan > > > > > > > > > >--- In > > > > <mailto:hp_agilent_equipment%40yahoogroups.com>hp_agilent_equipment@... > <mailto:hp_agilent_equipment%40yahoogroups.com> > > > <mailto:hp_agilent_equipment%40yahoogroups.com>, Jim Schatzman > > > <james.schatzman@> wrote: > > > > >> > > > > >> All- > > > > >> > > > > >> O.k., I have an 8568B with a bad attenuator (40dB and below are > > > dead). > > > > >> > > > > >> My big question is - how do you get access to the attenuator to > > > take it apart and clean/repair? It seems to be really buried under a > > > maze of stuff... > > > > >> > > > > >> Thanks! > > > > >> > > > > >> Jim > > > > >> > > > > >> > > > > >> > > > > >> At 01:44 PM 12/27/2012, you wrote: > > > > >> > > > > > >> > > > > > >> >Thanks a lot of this. > > > > >> > > > > > >> >I just managed to get a 8568B that attenuator has some problems > > > with the 20dB step. > > > > >> > > > > > >> >These instructions encourage me to open the attenuator and > clean > > > the pads with isopropyl alcohol. Now it works fine. > > > > >> > > > > > >> >I didn't made a testped for it i just cold moved carefully the > > > solenoids and checked with dg8saq vnwa the attenuator pads. I'm not > > > sure is the solenoid stuck or is it actually a contact failure. > > > > >> > > > > > >> >The instructions how to disaasemble the attenuator was very > > > accurate and easy to follow. > > > > >> > > > > > >> >BR, Jarmo > > > > >> > > > > > >> > > > > > >> > > > > >> > > > > >> > > > > >> > > > > > > > > > > > > > > > > > > > > > > > > > > > > > > > > > > > > > > > > > > > >
|
Tom,
The dimensions are .039" ID, .024" xsection diameter, and the originals were made of nitrile. I haven't had an 8568 attenuator apart but maybe one of the list members can verify if HP used the same o-rings in all the programmable attenuators. I'll see what the supplier says about cost and quantity and report to the group.
Happy New Year!
Steve
toggle quoted message
Show quoted text
On 12/31/2012 2:49 PM, Tom Miller wrote: Do you guys have the dimensions or a suggested part number for replacements? Since these are for the 8568, they would most likely be used in other units. If they have a minimum quantity, maybe a group buy would be in order. Distribution would be easy with 1st class mail. Also, a better material might be a good change, maybe silicone, etc.
Thanks, Tom
----- Original Message ----- From: Steve Krull To: hp_agilent_equipment@... <mailto:hp_agilent_equipment%40yahoogroups.com> Sent: Monday, December 31, 2012 3:21 PM Subject: Re: [hp_agilent_equipment] Re: DIY: Repair of HP 8568B Step Attenuators - another question...
Thanks Karl - I've emailed them to see if they will sell in small quantities.
Steve K.
On 12/31/2012 12:46 PM, Karl wrote:
Ran across these guys while designing a gas analyzer. They make all kinds of 0-Rings.
Karl
--- In hp_agilent_equipment@... <mailto:hp_agilent_equipment%40yahoogroups.com>
<mailto:hp_agilent_equipment%40yahoogroups.com>, Steve Krull <Steve-Krull@...> wrote:
I've got an attenuator for an 8672A that suffers from failed O-rings. In
this case it appears someone used something nasty to try to clean the spring contacts and attenuator pads and whatever was used dripped down the nylon plungers and turned all the O-rings to a mushy black paste. Not a single survivor. Agilent no longer stocks them but provides the dimensions and material type so I just need to research a source.
Steve
On 12/30/2012 10:16 PM, pianovt wrote:
Hi Jim,
Do not unscrew the solenoids on the top, at least not for now.
To see
what's happening, you will have to open the lower part where the RF action is.
The two SMA connectors have hex-nuts. Remove them. There is thin
name
plate that you can now remove. It covers the whole area between the SMA connectors. It probably has a few glue spots on its other side, so
carefully pry it off without bending the it. Now, you will see a bunch
of socket cap screws. When you remove them, you will be able to carefully separate the parts and gain access to the attenuator pads and plungers. The rubber O-rings are on the tiny nylon plungers that are attached to the gold plated spring contacts. I think there are two
rings on each plunger, one on the RF side and one on the solenoid side.
Vladan
--- In hp_agilent_equipment@...
<mailto:hp_agilent_equipment%40yahoogroups.com>
<mailto:hp_agilent_equipment%40yahoogroups.com>
<mailto:hp_agilent_equipment%40yahoogroups.com>, Jim Schatzman <james.schatzman@> wrote:
Vladan-
I got the cover off without trouble. By testing I note that when
the
middle (40 dB) switch is triggered, the output turns off completely. As far as I can determine, the plunger is actuated correctly. Moving it manually also results in the loss of output. The other two switches
(10 and 20 dB) work fine.
I suppose that either the switch contacts are failing or the 40 dB resistor is open. Apparently both switches and resistors are inside the two-piece machined steel block. All the RF path seems to be inside. It would appear that I need to get this apart, but I am not sure how to. There are quite a few screws attaching the solenoids to the steel block, but I am not seeing what is holding the steel block together.
This attenuator has a PC board with several chips on it. I have noticed before that HP seems to have adopted the philosophy that
if 2
chips are good, then 20 must be better. It is a surprisingly complicated device.
Any advice??
Thanks!
Jim
At 08:45 PM 12/30/2012, you wrote:
Jim,
You will have to pull out the front panel. There are a few
flat-head screws along the front frame (top and bottom, maybe even sides - don't remember any more). Once you remove them, the front panel pulls out of the front frame, though it still has a couple of umbilicals attached. From there, you will see how to get the attenuator out.
The most common failure has to do with the tiny rubber O-rings on
the plungers. This is all delicate magnifier work. Do not succumb to the "magnetized housing" myth of repair. The attenuator depends on magnets for latching. It's just that when the O-ring fails and
either
splits or falls off, the mechanism shifts closer to the magnet and the
force of magnetic attraction increases. At that point, the solenoid (which has nothing wrong with it) no longer can pull the parts away from the magnet (which also has nothing wrong with it). This is when some people decide that demagnetizing everything will fix things. If you have the patience, study the design, it's interesting. The goal was to use current only when switching a stage is required and then interrupt all current to the coils as soon as the task is accomplished.
You open the attenuator by removing one of the end caps and
sliding
the U-shaped steel cover off. It comes off relatively easily. Don't try to remove it by prying with screwdrivers.
There are two versions of the attenuators. The older ones are all
electro-mechanical, the newer ones have a p.c. board with some
ICs on
them.
Good luck, they can be fixed, but the work requires some patience
and you may have to order no O-rings from Agilent.
Vladan
--- In
<mailto:hp_agilent_equipment%40yahoogroups.com>hp_agilent_equipment@... <mailto:hp_agilent_equipment%40yahoogroups.com>
<mailto:hp_agilent_equipment%40yahoogroups.com>
<mailto:hp_agilent_equipment%40yahoogroups.com>, Jim Schatzman <james.schatzman@> wrote:
All-
O.k., I have an 8568B with a bad attenuator (40dB and below are
dead).
My big question is - how do you get access to the attenuator to
take it apart and clean/repair? It seems to be really buried under a maze of stuff...
Thanks!
Jim
At 01:44 PM 12/27/2012, you wrote:
Thanks a lot of this.
I just managed to get a 8568B that attenuator has some
problems
with the 20dB step.
These instructions encourage me to open the attenuator and
clean
the pads with isopropyl alcohol. Now it works fine.
I didn't made a testped for it i just cold moved carefully the
solenoids and checked with dg8saq vnwa the attenuator pads. I'm not sure is the solenoid stuck or is it actually a contact failure.
The instructions how to disaasemble the attenuator was very
accurate and easy to follow.
BR, Jarmo
|
I recall Lothar telling me several years ago that these O-rings were available for watch repair. -Eric
toggle quoted message
Show quoted text
--- In hp_agilent_equipment@..., Steve Krull <Steve-Krull@...> wrote: I've got an attenuator for an 8672A that suffers from failed O-rings. In this case it appears someone used something nasty to try to clean the spring contacts and attenuator pads and whatever was used dripped down the nylon plungers and turned all the O-rings to a mushy black paste. Not a single survivor. Agilent no longer stocks them but provides the dimensions and material type so I just need to research a source.
Steve
|
For a picture of the plastic rod with attached o-rings, lookup 5060-0342 at parts.agilent.com. I bought some back in September, but they are apparently only available to Agilent inside operations now. Unfortunately, it is difficult (for me, anyway) to remove the rods for o-ring replacement without damaging them. It made sense that they were sold as a combined unit.
--John Gord
toggle quoted message
Show quoted text
--- In hp_agilent_equipment@..., "Eric" <eric_haskell@...> wrote: I recall Lothar telling me several years ago that these O-rings were available for watch repair. -Eric
--- In hp_agilent_equipment@..., Steve Krull <Steve-Krull@> wrote:
I've got an attenuator for an 8672A that suffers from failed O-rings. In this case it appears someone used something nasty to try to clean the spring contacts and attenuator pads and whatever was used dripped down the nylon plungers and turned all the O-rings to a mushy black paste. Not a single survivor. Agilent no longer stocks them but provides the dimensions and material type so I just need to research a source.
Steve
|
All the products that use this part.
a.. 11758V a.. 33320G a.. 33320H a.. 33321H a.. 33322Q a.. 3708A a.. 70904A a.. 70907B-H10 a.. 70908A-H10 a.. 70908A-H25 a.. 71100C a.. 71100P a.. 71210P a.. 71400C a.. 71401C a.. 8340B a.. 83810B a.. 84100EM a.. 84110EM a.. 84125A a.. 84125B a.. 84300A-E53 a.. 84300A-E57 a.. 84300A-E95 a.. 8494G a.. 8494H a.. 8495H a.. 8496G a.. 85070M a.. 8514B a.. 8515A a.. 85301B a.. 85301C a.. 8542E a.. 8546A a.. 8590L a.. 8590L-R13 a.. 8591EM a.. 8593E a.. 86030A a.. 86037A a.. 86037B a.. 86037C a.. 8642B a.. 8662A a.. 8664A a.. 8665A a.. 8665B a.. 8672A a.. 86794B a.. 8752C a.. 8753D a.. 8753E a.. 8753ES a.. 8902A a.. E2500B a.. E2500B-508 a.. E2505A a.. E2507B a.. E2508A a.. E2747A-003 a.. E4901B a.. E4902B a.. E5501A a.. E5501B a.. E5502B a.. E5503A a.. E5503B a.. E5504A a.. E5504B a.. E7350A a.. Z5200FB
toggle quoted message
Show quoted text
----- Original Message ----- From: johncharlesgord To: hp_agilent_equipment@... Sent: Tuesday, January 01, 2013 1:20 AM Subject: [hp_agilent_equipment] Re: DIY: Repair of HP 8568B Step Attenuators - another question...
For a picture of the plastic rod with attached o-rings, lookup 5060-0342 at parts.agilent.com. I bought some back in September, but they are apparently only available to Agilent inside operations now. Unfortunately, it is difficult (for me, anyway) to remove the rods for o-ring replacement without damaging them. It made sense that they were sold as a combined unit.
--John Gord
--- In hp_agilent_equipment@..., "Eric" <eric_haskell@...> wrote: > > I recall Lothar telling me several years ago that these O-rings were available for watch repair. -Eric > > --- In hp_agilent_equipment@..., Steve Krull <Steve-Krull@> wrote: > > > > I've got an attenuator for an 8672A that suffers from failed O-rings. In > > this case it appears someone used something nasty to try to clean the > > spring contacts and attenuator pads and whatever was used dripped down > > the nylon plungers and turned all the O-rings to a mushy black paste. > > Not a single survivor. Agilent no longer stocks them but provides the > > dimensions and material type so I just need to research a source. > > > > Steve > > >
|