608 output attenuator.
You may remember that I was concerned about the integrity of the output attenuator in my 608D. Today I checked it again the way I had calibrated it with a short cable and a pass through 50 ohm termination at the scope end. I set the frequency to 10 MHz and the attenuator to 100 mV to be sure there would be no proximity effect between the pickup coil and the output tuned circuit. With the scope reading 100 mV within accuracy spects I then removed the termination and measured 200 mV within accuracy spects. From this I conclude that my output attenuator has not been damaged by someone transmitting into it.
Regards. Max. K 4 O DS. Email: max@... Transistor site Vacuum tube site: Woodworking site Music site: To subscribe to the fun with transistors group send an email to. funwithtransistors-subscribe@... To subscribe to the fun with tubes group send an email to, funwithtubes-subscribe@... To subscribe to the fun with wood group send a blank email to funwithwood-subscribe@... |
Bob Albert
Ah you were luckier than I. My Marconi TF1066/4 not only had a broken attenuator rack (from removing from cabinet without first turning it inward) but also a burned out 47 Ohm resistor at the probe.
toggle quoted message
Show quoted text
It required complete removal of the probe assembly, including some gear work and springs and so on. A lot of work but now the generator output is accurate and the attenuator works smoothly. The resistor is 1/4 Watt so it doesn't take much to burn it out. Maybe 5 Volts of rf for most of a minute. A 100 Watt transmitter delivers 71 Volts; that should do it in a second or less. Bob --- On Tue, 12/25/12, Max Robinson <max@...> wrote:
From: Max Robinson <max@...> |
I don't have any antenna or RF impedance bridge that I trust. So the 50 ohm termination and "infinite" impedance test is all I have.
toggle quoted message
Show quoted text
Regards. Max. K 4 O DS. Email: max@... Transistor site Vacuum tube site: Woodworking site Music site: To subscribe to the fun with transistors group send an email to. funwithtransistors-subscribe@... To subscribe to the fun with tubes group send an email to, funwithtubes-subscribe@... To subscribe to the fun with wood group send a blank email to funwithwood-subscribe@... ----- Original Message -----
From: "Patrick Wong" <patwong3@...> To: <hp_agilent_equipment@...> Sent: Tuesday, December 25, 2012 9:16 PM Subject: [hp_agilent_equipment] Re: 608 output attenuator. Hi Max, |
On 12/26/2012 12:01 AM, Max Robinson wrote:
It would be better if nobody top-posts. That having been said; As you place and remove the 50 Ohm load, you are mismatching THE GENERATOR, such that the signal may vary by 2:1. If you want the test to make sense, you need to connect the generator output thru a fairly large series resistor--say 470 ohms--to the 608 input port. Connect your scope thru a 10:1 probe to that same input port. Now, with that connected, add a 51 ohm resistor in parallel with the 608 input. If the scope reading drops by about 50%, then the 608 has a 50 ohm impedance at the port. If it drops by quite a bit more than 50%, then the 608 has a higher impedance than 50 Ohms, and there would appear to be a problem with the attenuator circuit. BTW, if you don't have a scope, a high-impedance VTVM that will read at 10 MHz (it doesn't have to be accurate) will work similarly. Doug, WA2SAY |