419A Chopper Replacement
Good morning,
I have a 419A Nulling Voltmeter that I just finished replacing all the electrolytic capacitors in. This cleared up my 'after warming up no operation' problem. However, I also had to really have my way with both the chopper freq adj and neon current adj pots to get them to fire _something_ like reasonably. I can't get the frequency to be correct (too slow by close to half), and the waveform is pretty distorted over most of the range, and just _somewhat_ distorted over the rest. This makes me think I probably need new choppers. I assume these are unobtainum. 1. is my logic sound, or is there more likely a problem with the driver instead? 2. what will a somewhat slow, distorted chopper waveform do to the correct operation of this otherwise-excellent meter? Thanks very much! |
J Forster
Many Ne bulbs have minute amounts of radioactive material in them to
toggle quoted message
Show quoted text
stabilize the firing point. As the radioactive emission rate decays, the bulbs become erratic. Put in new bulbs, if you can. Best, -John Christian A Weagle wrote: Good morning, |
--- In hp_agilent_equipment@..., J Forster <jfor@...> wrote:
That was my first impulse; however, the manual (paragraph 5-27) states that "physical alignment and selection of the neon bulbs are critical" and that replacement of the entire assy is the only thing to do. Can you speak to this at all? If I _do_ replace the neons, are standard NE-2 bulbs the right thing to use? Thanks for your help! |
J Forster
I don't know about the alignment. My experience was with the opto-chopper in
toggle quoted message
Show quoted text
the 410C. Mine was replaced as a unit by the company cal lab. There are only a very few wire pigtail Ne's made. I think there was a means of ID'ing which is in there by looking very carefully at the glass pinch area. There is something stamped into the glass as I remember. You might ask someone like BulbDirect that specializes in lamps. Best, -John Christian A Weagle wrote: --- In hp_agilent_equipment@..., J Forster <jfor@...> wrote:That was my first impulse; however, the manual (paragraph 5-27) states |
John Day
At 04:22 PM 11/3/2007, you wrote:
I don't know about the alignment. My experience was with the opto-chopper inOr ask on the NEON-NIXIE Yahoo group, a lot of expertise and experience there too. John Best, |
Harvey White
On Sat, 03 Nov 2007 15:22:26 -0500, you wrote:
I don't know about the alignment. My experience was with the opto-chopper inI remember NE2-D and NE2-J as well as the NE2; different wattages, IIRC. Harvey
|
deane kidd who has tektronix parts might have the neon bulbs you need
toggle quoted message
Show quoted text
----- Original Message -----
From: Harvey White To: hp_agilent_equipment@... Sent: Saturday, November 03, 2007 9:07 PM Subject: Re: [hp_agilent_equipment] Re: 419A Chopper Replacement On Sat, 03 Nov 2007 15:22:26 -0500, you wrote: >I don't know about the alignment. My experience was with the opto-chopper in >the 410C. Mine was replaced as a unit by the company cal lab. > >There are only a very few wire pigtail Ne's made. I think there was a means of >ID'ing which is in there by looking very carefully at the glass pinch area. >There is something stamped into the glass as I remember. You might ask someone >like BulbDirect that specializes in lamps. I remember NE2-D and NE2-J as well as the NE2; different wattages, IIRC. Harvey > >Best, >-John > > > >Christian A Weagle wrote: > >> --- In hp_agilent_equipment@..., J Forster <jfor@...> wrote: >> > >> > Many Ne bulbs have minute amounts of radioactive material in them to >> > stabilize the firing point. As the radioactive emission rate decays, the >> > bulbs become erratic. Put in new bulbs, if you can. >> >> That was my first impulse; however, the manual (paragraph 5-27) states >> that "physical alignment and selection of the neon bulbs are critical" >> and that replacement of the entire assy is the only thing to do. >> >> Can you speak to this at all? If I _do_ replace the neons, are >> standard NE-2 bulbs the right thing to use? >> >> Thanks for your help! |
toggle quoted message
Show quoted text
-----Original Message-----I don't know the 419A. For what it's worth, my Fluke 895A differential voltmeter had a distorted chopper waveform that was about half normal frequency. Turned out to be a short on one of the power supplies that the chop oscillator drives in addition to the choppers. Dave Wise |
Well, here I am answering my own question(s). This one was sort of my
fault. To wit: When I got this unit, I tried to turn it on, no luck. Seeing the battery charge pushbuttons, I feared the worst (leaky cells corroding the PCBs), and opened it up. Found some wacky NiCds (basically, four 'cells', each the size of sub-Cs, each labelled as 6V 225mAh (so probably 5 big fat coil cells in fancy wrapping), chained up to provide +/- 12V, and dated in the early 80's) that were truly dead - no leaks though. I removed them as some stickers on the back of the unit implied that they were trickle charged whenever it was plugged in, and they probably wouldn't like that in their current state! Changed a bunch of the electrolytic caps too. I then proceeded to operate the unit, sort of. It was all drifty, lots of random brief full-scale deflections, couldn't cal the photochopper, etc... Changed the rest of the caps. No help. After a close perusal of the power supply page of the schematic, though, I saw what was wrong. The only place that this meter gets its' internal ground is from the center of that battery stack! It basically runs from, and continuously charges them from, the top and bottom of a full-bridge rectifier. Ground is sort of virtual - if there are no batteries, there is no fixed ground. Hacking in a pair of 12-cell NiMH packs makes everything better - no drifts or abrupt swings, and I could cal the photochopper. The packs are somewhat higher voltage, but I think they should regulate down to the correct 12V after a while of running off them to discharge them, before plugging the unit in. What an unusual way of doing business! |
they can be replaced with zener diodes and bypass caps
toggle quoted message
Show quoted text
----- Original Message -----
From: Christian A Weagle To: hp_agilent_equipment@... Sent: Thursday, November 08, 2007 9:20 PM Subject: [hp_agilent_equipment] Re: 419A Chopper Replacement Well, here I am answering my own question(s). This one was sort of my fault. To wit: When I got this unit, I tried to turn it on, no luck. Seeing the battery charge pushbuttons, I feared the worst (leaky cells corroding the PCBs), and opened it up. Found some wacky NiCds (basically, four 'cells', each the size of sub-Cs, each labelled as 6V 225mAh (so probably 5 big fat coil cells in fancy wrapping), chained up to provide +/- 12V, and dated in the early 80's) that were truly dead - no leaks though. I removed them as some stickers on the back of the unit implied that they were trickle charged whenever it was plugged in, and they probably wouldn't like that in their current state! Changed a bunch of the electrolytic caps too. I then proceeded to operate the unit, sort of. It was all drifty, lots of random brief full-scale deflections, couldn't cal the photochopper, etc... Changed the rest of the caps. No help. After a close perusal of the power supply page of the schematic, though, I saw what was wrong. The only place that this meter gets its' internal ground is from the center of that battery stack! It basically runs from, and continuously charges them from, the top and bottom of a full-bridge rectifier. Ground is sort of virtual - if there are no batteries, there is no fixed ground. Hacking in a pair of 12-cell NiMH packs makes everything better - no drifts or abrupt swings, and I could cal the photochopper. The packs are somewhat higher voltage, but I think they should regulate down to the correct 12V after a while of running off them to discharge them, before plugging the unit in. What an unusual way of doing business! |