I think it's fixed and running beautifully. The other A19 board was functional but significantly different on top and none of the labeling on the plastic cover matched the board, so I decided I should fix the original. I quickly found that the -12.6V reference was out (only about -1.7V), but it took a while to figure out since there are a number of things that could do it. The culprit was perhaps the absolute last thing I would have expected. I had a couple of usual suspect Ta caps, a couple of high-K ceramic caps, a bunch of negative opamp rail circuitry to make -24V from the -40V supply, and of course the opamp itself that regulates the reference from a TC Zener.
?
After some test point wire tacking and operating measurements I found that the input to the opamp U3 was shorted to ground by C18, showing about 18 ohms. I couldn't believe this high grade plastic cap could have failed - it's not even a high stress spot at all, just extra filtering at the input against 5.11 kR from the Zener. C18 and its counterpart C22 in the +10V reference apparently needed to be high grade (I'm virtually certain these are PPF = polypropylene and foil) types for very low ESR, ESL, and DC leakage. I have some of these salvaged from other gear over the years, so I immediately recognized the type, and I can often tell by the density, which usually "feels" higher than film type caps. This is a relatively fancy and expensive part, and physically big (board space-wise), so there must have been good reason. A cap like this failing in a spot like this has to be an extremely rare random failure.
?
I found an equivalent one in stock, popped it in, and voila, just like new. All the original boards are back in now. I assumed the original setup for A19 is still good enough, so I only did the A20 adjustments, and this time the procedure worked as it was supposed to. It seems to be all working now, but will need some more checkout to be sure. I am happy for now.
?
Ed