Hi Joe,
?
If you just change some parts without knowing why you only make it worse. You give exact 0 information what was your error-picture, what you did or what voltages did you measure. My advice:
print out the schematc, mark the GND reference against which there is the masurement. If you measure, take the DC and AC Voltages and write them in the schematic. Check all voltages from left to right. Only C10, C12 or C14 may have up to 0,1V AC without load.
There may be a defect capacitor. But defect diodes or transistors are more common.
?
before you open the device: unplug it. If open: Isolate ALL life points, no exception. Use an isolation transformer.
?
Are the bridges on the rear according to the schematic?
?
Roland