Repair HP3455A Multimeter
Hi,
my 3325A is driving me crazy and no one could help yet, so I went over to a faulty HP3455A multimeter. It worked many years, but now it comes up with a OL display in DC.
Cal test stops at 11,10,9,8,5,4,3,2,1,0 and a digital value 0.062.
?
All voltages are stable with nearly zero ripple. Vref is 10V, all perfect.
The good service manual (no extrem testing procedure with a signature analyzer as in the 3325A manual)? has many tables with listings of voltages for the individual cal number. The cal feature switches many fets to ground, vref etc.
?
I found suspect voltages at the outputs of U6 a LM339 and replaced it, this fixed some cal errors. The meter now has only cal error #10. Checking the flowchart for #10 and the voltage matrix a found a suspect output on Pin 14 of U5, so I replaced it with no success. The voltage is 8.8V, that seems to be the Gate Bias voltage after 100KOhms, i lifted the pin and supplied shortly -24v, but this was not suceccessful either.
To pin 14 there is Q21 connected, i suppose it does not switch off, Pin 14 should be -24V.?
?
Next I will? replace Q19, the part number is 5070-7074, but I can't find any information and it is not listed in the HP crossreference list. I has 4 pins, may be one pin is can and gate is tied to taht pin as i can measure the gate bias voltage at the can.?
?
Has anyone a spare part or a hint which fet i can use to replace the Q19?
?
Ralf
? |
¿ªÔÆÌåÓý
I don't find a 5070-7074 in the 3455A manual I downloaded, Manual P/N 03455-90003. Looking for Q19, I find A10Q3, 4, 15, 16, 18, 19, 21? which is a 5081-7074 and A15Q19, 20, which is a 2N4391. ? ?You might try swapping Q19 with a different 5070-7074. You might pass diagnostics, or at least find a different failure to troubleshoot. John??? KK6IL On 3/28/2025 4:15 AM, Ralf Rosche via
groups.io wrote:
|
I'd try 2N4119A or PN4119A.? I might have some... Mouser?are mighty proud of them. On Fri, Mar 28, 2025 at 4:15?AM Ralf Rosche via <ralf.rosche=[email protected]> wrote:
|
2N4119.
A bit more investigation this morning.
I have on error 10 ?A10U6 and A10U5 Pin 13 the correct 10 Volts, the drivers U16... are ok.
The inputs are correct as listed in the manual.?
On A10U5 PIN14 I see 8 Volts, should be -24 Volts,?
May be its an error in the manual?
If I open Q21, which is connected to PIN 14, then error 8 and 5 comes up, as expected.
Connecting shortly PIn 14 of A10U5 to -24V clears ?the error 10.?
Hmm very strange.?
I tested Q4 and Q19, they are ok with multimeter, as JFets can be tested with a simple multimeter.
However, I am waiting now for the 2N4119 and will replace them. This are the only involved cal switches mentioned in the flowchart.
Ralf |
Hi have a pdf from the flowcharts, group B schematics and the groupB service excerpt.
HP3455A_groupB_Schematics.pdf
HP3455A_groupB_Schematics.pdf
HP3455A_groupB.pdf
HP3455A_groupB.pdf
HP3455A_flowcharts.pdf
HP3455A_flowcharts.pdf
|
Hi,
yesterday i got the Jfets from digikey, i replaced Q19,Q21,Q4 ,Q22 but the result is the same, still error 10. According to the flowchart the advice is to check Q4,Q19 and components around.
?
There are many components around, so could anyone look at this part of the schematics and give an advice?
?
May be U3??
?
And I have a curious 8.75 V at pin 14 of U5, should be -24V, but not? in focus of cal test 10.? (I replaced already U5,U4,U6)
?
Ralf |
to navigate to use esc to dismiss