开云体育

ctrl + shift + ? for shortcuts
© 2025 开云体育

How to I disassemble the HP/Agilent 8495B attenuator ?


 

Hello,
?
I need to disassemble ?the HP/Agilent 8495B ?attenuator, there are no visible screws, how can I do it if I need to service the attenuator ?
?
Thanks in advance!


 

Do you have a manual attenuator or the electronically switched one?
There are 4 screws on each end of the attenuator, if you have SMA/3.5mm connectors, then remove the nuts on those connectors, the cover plate between the connectors, and that will expose the 12-16 screws that hold the top cover with the attenuator pads on it. You will need to loosen at least one end plate to take the
top cover plate/attenuator plate off.
A lot of this information is in the HP/Agilent/Keysight Microwave Accessory catalog and in the HP 8495B Operating and Service manual, yes it does contain theory of operation, schematics, parts breakdown, and parts listing. A .pdf of the manual is attached.

Don Bitters


 

Thanks, I have the manual, mine has no screws on each end, for? a? picture please see
?


 

sorry, mine has N-type connectors and looks exactly as this one
?