What happened to HP/Agilent detailed circuit schematics
HI HP experts,
Maybe I am trapped in the mists of time, but what happened to 'proper' schematic diagrams in HP service manuals?
I have been trying to work on an HP ESG-D3000 and scoured the web for service information.
I found what appeared to be both operators and service manuals but the service manual only has block schematics.
Whereas every nut/bolt/clip/washer/rivet and aluminium panel is detailed in the minutest detail!
Am I missing a schematic supplement, or is servicing these days expected to just be a 'board swapping' exercise?
I look forward to your comments.
Dave |
Hello Dave,
toggle quoted message
Show quoted text
You are not missing the schematic supplement, it is not published or sold anymore. I am not sure if this is because of IP inside or because they don't want you to repair the product to force you to buy a new one. Or maybe both?! Another topic related to the new designs is that they use a lot of micro-controllers and FPGAs and even if you would have the schematic and let's say you can get the exact part number to replace the micro-controller or the FPGA you will not get the microcode/software to write it back. So it is pretty useless for the new designs to get a schematic if you cannot get the complete firmware for individual components. Regards, Razvan On 26/08/2024 11:57, Dave B via groups.io wrote:
HI HP experts, |
HI, well I suspected that may be the case and in the digital side of the instruments, (as you say) one wouldn't have a chance of a repair.
However I would have imagined that just about every possibility of RF circuitry would have been explored by now so wouldn't particularly be top secret?
And, even in my HP 1725A 'scope they used 'hybrid' (secret?) modules, and just depicted them as a block but showed the rest of the analogue workings as a 'proper' schematic.
It's frustrating as I'm developing a tiny solid state (plug and play) LED backlight 'module' to replace the CCFL inverter and tube.?
The backlight is borne to fail as the CCFL lamp dies after a while. The inverter and lamp were (I believe) made easy to replace as they are pretty accessible without having to
access any of the electronics within the sig-gen. (It's just behind the front panel).
However, without knowing what is supplying the +5v and the dimming control voltage could easily be answered if a schematic was available.
Questions like:-
Does the control come from a DAC (I think it does) but what drive capability does this have. Is the output of the DAC protected by a series resistor?
Whilst I could start to take the instrument internal covers off to trace this out...........sigh!
Dave |
Consider that we live in an amoral litigious society, where
companies have patented just about every conceivable combination of thoughts, algorithms, or circuits, that have happened in the past, present, or future, and covetous lawyers roam the planet sniffing out shares of any monies that can be taken from the legally weak. In such a society, it is impossible to design circuitry when you have to challenge your every thought, and trace the origin of your every idea just to see who owes what and to whom... So why, in such a society, would you put your company in the sights of marauding lawyers by freely revealing exactly what you have done inside of your circuits, ASICS, ROMS, and FPGA's? Make the b**tard lawyers work for their plunder! Component level servicing manuals are a relic from a much friendlier time; a time when most people had moral values, and lawyers were few, and traded their craft for livestock and produce. -Chuck Harris On Mon, 26 Aug 2024 04:59:29 -0700 "Dave B" <davebullockmbe@...> wrote: HI, well I suspected that may be the case and in the digital side of |
I've considered such replacements as I do design microprocessor boards with (replaced by LED) CCFL displays.
toggle quoted message
Show quoted text
Many of the inverters have a DC control voltage as you know.? I'd be tempted to have the LED string capable of working off the dc supply voltage for the inverter, and I'd bag the whole idea of worrying about the DAC.? Feed a very small (8 pin or so) microprocessor's A/D input through a divider and let the processor generate the PWM.? That, with a transistor level shifter allows you to use a P channel power FET to control the LEDS. IF you had access to the software and IF it had an I2C interface you could control on the main board, something like a PCA9634 FM LED controller would be overkill, but would work.? I use that LED controller on my generic boards and just reserve one channel of 8 for the display backlight. Harvey On 8/26/2024 7:59 AM, Dave B wrote:
HI, well I suspected that may be the case and in the digital side of the instruments, (as you say) one wouldn't have a chance of a repair. |
Dave,
?
Not to be rude, but it does appear that you are trapped in the mists of time as you put it.? As a current lab instrumentation engineer and self-taught electronics engineer/repair specialist, there is no company that I am aware of today (at least in the biotech industry manufacturing lab equipment) that is providing schematics for their internal circuitry.? They don't even provide service manuals anymore (no I don't mean user manuals).? Service manuals as in manuals with troubleshooting, circuit descriptions, calibration procedures, and PM procedures.??
?
I tend to agree with Chuck.? And this is a topic that could be discussed for hours.? But this is what the whole 'Right to Repair' movement is about which I am huge proponent of.? It is one of the reasons I started my own YouTube channel to thwart the manufacturers and share with scientists how they can repair their equipment on their own.
?
Also this comment below is probably the most short-sighted and completely ridiculous.? There is still much repair you can do without having access to the firmware.? Although I believe companies should offer that, too.??
?
"Another topic related to the new designs is that they use a lot of
micro-controllers and FPGAs and even if you would have the schematic and let's say you can get the exact part number to replace the micro-controller or the FPGA you will not get the microcode/software to write it back. So it is pretty useless for the new designs to get a schematic if you cannot get the complete firmware for individual components." ?
-Frank |
HI Harvey,
Your creations sound far superior to my KISS approach to the issue.
I have been designing my replacement 'module' with the idea it just 'drops in' in place of the existing inverter.
I have watched an interesting Youtube video of a complicated attempt to solve this problem and gleaned the control voltages and supply data from this.
I have not ventured inside the instrument I have on my bench yet, as I only want to go inside it once!
It looks like there is a 3 pin connector with +5V, GND and a variable 'Dim' control voltage.
According to the video, the control voltage varies from 0.7V (display off) to 4.7V (full brightness) set by rotating the front panel adjustment.
So...
I'm assuming that the HV inverter would be taking a much greater current from the 5V supply than a few LED's to generate the same light as a CCFL, so that's the supply sorted?
My concern was that without a schematic I don't know what the source impedance of the control voltage was, leading to my 'missing schematics' discovery!
As you hinted, I am powering the LEDs off the 5V and using a high impedance analogue input device to exacerbate any loading issues.
Assuming the voltages are the same on this instrument as measured on the video, then I should have a CCFL replacement backlight module for a great deal less
than an OEM module.
It all works on the breadboard, I'm just waiting for the pcb's to arrive to start manufacturing and carry out a real life test.
Dave |
inline.
On 8/26/2024 11:48 AM, Dave B wrote: HI Harvey,Sometimes I think more may be less... I have been designing my replacement 'module' with the idea it just 'drops in' in place of the existing inverter.this would fit in the same space as the inverter board.? An 8 or 14 pin microprocessor (depending on innards), a very tiny regulator, a few SMT resistors and two transistors (one a DPACK). I have watched an interesting Youtube video of a complicated attempt to solve this problem and gleaned the control voltages and supply data from this.Then the 5 volt can be a bit of a problem. Think 20 ma/LED, and 9 LEDS fit in a strip (I have such strips, pc board wise).? So maximum non-PWM controlled current is about 180 ma/strip, and that, for 2 strips, is 360 ma.? The problem is that the LEDS may not have enough voltage for 2 in series.? Hence parallel. What's done in some LED backlit displays is to put the LEDS in series (say needing about 18 to 20 volts), and then run them with a small inverter.? That inverter might just have brightness control.... They make chips for that.? So perhaps that chip, a few transistors, a small inductor might just do the job for you. According to the video, the control voltage varies from 0.7V (display off) to 4.7V (full brightness) set by rotating the front panel adjustment.Depends on number of LEDS, and hookup.? So not quite so easily sorted. My concern was that without a schematic I don't know what the source impedance of the control voltage was, leading to my 'missing schematics' discovery!That works. Assuming the voltages are the same on this instrument as measured on the video, then I should have a CCFL replacement backlight module for a great deal lessWhich is a good idea.? Multiple ways of getting there. It all works on the breadboard, I'm just waiting for the pcb's to arrive to start manufacturing and carry out a real life test.Ah, then you've gone further in this design than I thought. Best of luck, then, and let me know if I can contribute more ideas. Harvey Dave |
Today’s corporate focus on product manufacture and service is many fold.
Profits made from equipment sales are only part of the equation. There is (sometimes much more) money to be made through support. If you don’t provide good documentation the customer is basically faced with having to send the item in for repair. Second, in order to produce a very detailed document describing every nut and bolt on how to fix the item requires a gaggle of tech documentation people which costs $$$ to employ. Make manuals simple and you have fewer related employees to deal with. Third, it has been shown that if users do choose to repair their own equipment that fixing things at the higher level (i.e. PC board, etc.) proves to be easier both for the customer and the factory service department where it reduces time to fix. Neither the customer nor the factory needs to dig deep to find a problem if they follow the simple diagnostic charts provided in the manuals. And factory service effort is reduced by simply plugging boards into test fixtures to localize the fault. Or, if more economical, they may opt to simply scrap the boards and present a new board fee for the user when they purchase through “exchange.” There is more but I will not continue. One must remember that we now live in a “disposable society” where people need instant everything which includes immediate replacement if something breaks. As for those of us legacy equipment collectors/users down here in the trenches we gradually see our inventory becoming older and older in age as we either cannot or opt not to try to fix “mystery boxes” now being manufactured. And as it has been said by others in previous posts, opening a box only to see a PC board with only one non-repairable module in the center of it certainly adds to the issue. And even if that module can be replaced by ordering another from the manufacture time is of a limit given the shorter and shorter time replacement part inventories are held after the manufacture is discontinued for that device. It’s expensive to store parts for old equipment since everyone wants the “latest and greatest” products and will achieve that goal by perpetually buying more new ones. From that the importance of supplying parts for legacy equipment becomes less and less. It’s the corporate “bottom line” speaking here. A note on sending equipment out for repair. Yes, there are those third party agencies that will offer service long after the factory does not. But what I have seen even through actual visits to these places is that they, too often have scanty data to use to fix stuff. From that many of them will tell the customer that the item cannot be repaired and if the customer opts to simply abandon the item with the repair agency that agency will relinquish it to a “bone yard” where they can use parts out of it to fix other units. Greg |
Had to laugh.? Chuck is so right about this.? As someone who managed an engineering team for a larger media company we were always dealing with legal intellectual property nonsense.? My all time favorite was Antor Media suing just about everyone with a claim that their patent US5734961 for transmitting information recorded on information storage from a central server via a high data rate digital telecommunications network to subscribers covered anyone serving content to a customer.? Our company, which would not be considered legally weak, decided to cave and just strike a deal just because it didn't want to waste time dealing with this stupidity.
?
?
The patent (You will laugh):
?
I highly recommend listening to the podcast "When Patents Attack".? After all Why would a company rent an office in a tiny town in East Texas, put a nameplate on the door, and leave it completely empty for a year?
?
?
?
On Mon, Aug 26, 2024 at 06:36 AM, Chuck Harris wrote:
Consider that we live in an amoral litigious society, where |
Hello Frank,
toggle quoted message
Show quoted text
From your point of view maybe it is short-sighted but I encountered issues in the past where we had some Altera and Atmel ICs that was faulty including other passive components around them and we replaced all of them with new parts but the system wouldn't work. Contacting manufacturer didn't help. They don't share the files to rebuild the equipment and since we repaired it they don't want to accept it for repair also. In theory you can repair everything but in reality you cannot do it. Most test equipment now is running some custom PC board that in theory can be repaired but if some NAND flash is damaged and you cannot get the full firmware how you will fix it? Most vendors will say it is "proprietary" code or what ever and they will not provide it to you. What I would like to have is the PSU schematics at least. I saw a lot of new test equipment with PSUs broken/fried and many of those PSUs were very complex full with SMD components. Maybe you can share more details related to fixing modern test equipment to understand the complexity. Regards, Razvan On 26.08.24 17:45, Frank Mashockie via groups.io wrote:
Dave, Also this comment below is probably the most short-sighted and |
I've got three of the ESGD series sig gens here and so far, none of the backlights have failed. However, it is only a matter of time before each one fails. The oldest one is an ESGD4000A. This was an ex-rental I purchased about 10 years ago. It already had a lot of run time hours and it is up to 78,000 hours now. It may be that the display backlight has already been repaired in the past as it has been run for so many hours.
?
So far it has proved to be completely reliable. I don't recall having to repair it. The other two ESGD sig gens have had problems, the worst being an E4433B. I've had to repair this numerous times. The MMIC devices used in the signal path are prone to failure and also some of the memory chips in the UND dual arb board needed to be replaced. The PSU failed in the E4438C. The PSU used in the E4438C is huge and very complicated, but it proved easy to repair. I don't have schematics for any of these sig gens but managed to repair them quickly and easily for little cost.
?
I converted an old Aeroflex 3414 sig gen display to an LED backlight recently with great success. Hopefully, the ESGD sig gens will be as easy to convert?
?
? |
Razvan,
?
Your points are valid, but my point is don't give the manufacturers the benefit of the doubt.? The customer deserves this information.? I apologize as I came off a bit strong with my wording.?
?
Unless the MCU has failed completely, I've ran into many cases where only a GPIO pin or two has gone bad.? In these cases, flashing the MCU is still an option.? As long as it isn't read protected (you'd be surprised how many manufacturers don't, especially on older equipment).? Or by sharing firmware of MCUs you've flashed with communities like this (or EEVBlog), all you need is a new MCU to perform the programming.? Sure it may take some additional tools, but it is pretty easy in most cases.??
?
As consumers, repairs, or designers we shouldn't allow manufacturers to continue these repair prevention strategies that ultimately stifle the sharing of information and community knowledge.? We shouldn't give them the benefit of the doubt that it is acceptable to forgo sharing of schematics just because the equipment has an MCU or FPGA.? That doesn't stop manufacturers from protecting their creations either.? For example, in pharma, a drug can be patented, yet we still know the chemical composition.? It is not kept secret; the entire public knows.? It should be no different for electronics.?
?
-Frank
?
? |
Does the customer deserve information on how to
repair an incandescent light bulb? Perhaps the customer should be given access to cheap replacement tungsten filaments? Or, if that isn't practical, instructions on how they mixed and forged tungsten? Should the manufacturer be required to include a service port on each light bulb to make the filament reachable without damage? -Chuck Harris On Mon, 26 Aug 2024 13:00:01 -0700 "Frank Mashockie" <fmashockie@...> wrote: Razvan, |
This entire group is only possible because HP/Agilent for a time used to provide schematics.? They still provide service manuals as far as I am aware.? That is all proponents of right to repair are asking for.? To prevent the manufacturers from monopolizing the repair market.? And to allow consumers to repair the things they bought themselves however they would like.? If you don't agree with that, then why are you in this group Chuck???
?
? |
As usual, something between the ridiculous extremes would be the best situation.? Between say, manufacturers being forced by the government to divulge all their secrets and consumers forced to have every single product returned to the mfr for repair.? ? ? ? ? Jim Ford, Laguna Hills, California, USA?
toggle quoted message
Show quoted text
|
I agree there Jim.? While I do agree with some legislation, I have not been too keen on what I've seen passed so far.? Manufacturers are already circumventing some of these laws passed in CA from what I've heard.? I should have clarified that the best way to thwart manufacturers is groups like this, forums, YouTube content creators and bloggers.? If anything, the lack of service manuals and schematics has made me a better problem solver and critical thinker in my own work (or at least I think lol).? ? |
That assumes you have the time to reverse engineer the faulty product, at least to some degree.? These days I'm commuting about two and a half hours each day, so not having at least a schematic usually prevents me from being able to repair much of anything.? ? ? ? ?Jim
toggle quoted message
Show quoted text
|
to navigate to use esc to dismiss