开云体育

E4421B Display Pfroblem


 

Am i to believe that there are no schematics available for the ESG signal generators or are they available from some
source unknown to me,
cheers
Geoff
GI0GDP


 

Hi Geoff,

What is the problem with the display?

You can get schematics from milan.vasic@...

Tom
N3AJA

On 8/7/2024 12:56 PM, geoffrey pike via groups.io wrote:
Am i to believe that there are no schematics available for the ESG signal generators or are they available from some
source unknown to me,
cheers
Geoff
GI0GDP


 

开云体育

I have the schematics but nothing about the display itself.
The controller stuff is on the motherboard:

There's also the backlight display driver there.

Mark
KJ7CYN


From: [email protected] <[email protected]> on behalf of Tom B <tbryan@...>
Sent: Wednesday, August 7, 2024 2:26 PM
To: [email protected] <[email protected]>
Subject: Re: [HP-Agilent-Keysight-equipment] E4421B Display Pfroblem
?
Hi Geoff,

What is the problem with the display?

You can get schematics from milan.vasic@...

Tom
N3AJA


On 8/7/2024 12:56 PM, geoffrey pike via groups.io wrote:
> Am i to believe that there are no schematics available for the ESG
> signal generators or are they available from some
> source unknown to me,
> cheers
> Geoff
> GI0GDP
>






 

Hi Tom,
The display problem is that the front panel controls don't work and obviously you can't see anything!
The back light seems OK, early days as i haven't opened it up yet to look but i always like to have CLIP data
Thanks for the drawing Mark
cheers
Geoff
GI0GDP


 

开云体育

Hi Geoff,

If there is no change when you adjust the contrast controls, then the problem is likely the contrast circuit.? There is a test point for that on the bottom of the circuit board.?? The voltage should change when you adjust the contrast controls.

Also, check where the ribbon cable from the display plugs in.? Make sure it is seated correctly.

Tom


On 8/8/2024 12:40 PM, geoffrey pike via groups.io wrote:

Hi Tom,
The display problem is that the front panel controls don't work and obviously you can't see anything!
The back light seems OK, early days as i haven't opened it up yet to look but i always like to have CLIP data
Thanks for the drawing Mark
cheers
Geoff
GI0GDP


 

开云体育

Looks to me like it's on the top. TP602 by U614 and L603


From: [email protected] <[email protected]> on behalf of Tom B <tbryan@...>
Sent: Thursday, August 8, 2024 11:26 AM
To: [email protected] <[email protected]>
Subject: Re: [HP-Agilent-Keysight-equipment] E4421B Display Pfroblem
?

Hi Geoff,

If there is no change when you adjust the contrast controls, then the problem is likely the contrast circuit.? There is a test point for that on the bottom of the circuit board.?? The voltage should change when you adjust the contrast controls.

Also, check where the ribbon cable from the display plugs in.? Make sure it is seated correctly.

Tom


On 8/8/2024 12:40 PM, geoffrey pike via groups.io wrote:
Hi Tom,
The display problem is that the front panel controls don't work and obviously you can't see anything!
The back light seems OK, early days as i haven't opened it up yet to look but i always like to have CLIP data
Thanks for the drawing Mark
cheers
Geoff
GI0GDP