VISA problems with ancient test equipment?
Hello,
my HBIB system works with 8568B spectrum anaylyzer, 8970A, 34401A and others. However I fail to connect a 5350B to my system. Diag 80 (HPIB diagnosis) shows no errors. Also tried ddifferent HPIB cables with no change. The counter is not found with Keysight Connection Expert. I get an VISA BFFF0012 error when I try to connect within WinGPIO software. Before digging into the hardware of the HPIB interface within the counter, I assume that VISA is not working with "old" test equipment, as I have same problem with 5343A counter. Is there workaround? Any ideas are appreciated. Regards Ernst |
¿ªÔÆÌåÓý? Hi Ernst, ? Using the 5350B series meters is not simple with GPIB at first glance. But you have to follow a small procedure: ? You must put the meter in HOLD mode to begin. ? - Press the SET key - Press the SAMPLE RATE button - With the LEFT arrow, move the cursor to the HOLD position on the left. - Go to Connection Expert (I used USB-GPIB interface 82357B). - In the menu on the left, choose the interface to which your 5350B is connected. - Choose ADD and GPIB INSTRUMENT - Choose the GPIB address corresponding to your 5350B (example GPIB0 and primary address 14) - Uncheck AUTO IDENTIFY - press on OK - The instrument should appear in the list with the name UNKNOW and the address GPIB0::14::INSTR for this example. - You will not be able to test with Connection Expert, however this will allow using a program to communicate if you add the following two VISA commands, allowing you to complete a reading with the LINEFEED character. In the example, these are two commands used in vb.net. ? errorStatus = visa32.viSetAttribute(vi5351B, VI_ATTR_TERMCHAR_EN, 1) errorStatus = visa32.viSetAttribute(vi5351B, VI_ATTR_TERMCHAR, 10) ? ? ? If you use a GPIB interface from National Instruments, such as GPIB-USB-B or PCI-NI-GPIB for example, it will be possible to communicate with the instrument if you add the two options (Configure EOS): - Enable "Terminate READ on EOS" - EOS byte=10 ? ? ? ? I wrote a small program to test the sensitivity of the 535xB meters a few weeks ago. I modified it to allow only 10 consecutive readings on input 1 or input 2 (terminated 50 Ohms). ? The image shows you the result of a 300 MHz -20dBm signal at input 2. ? If you can't test GPIB communication, I will send you this little program for you to try. ? Yves ? ? De?: [email protected] <[email protected]> De la part de dk1vi via groups.io ? Hello, |
Hello Yves_Tardif,
unfortunately my initial answer bounced from the email address you sent me your program. Thanks for your program. The 5350B is really a weird beast! I made some initial tests on friday following your instructions. At first not working. Had to remove checkmark in Keysight Communicator -> Settings -> "Enable Keysight GPIB cards for 488 programs". Now it worked, but some measurements showed "0". Over the weekend I was busy with other obligations. So I tried again today with no success. Hitting "Stop" did not work so I could not select Input 2. Input 1 not working because of fried sampler. This was the reason the counter was donated to me. Regards Ernst? |
¿ªÔÆÌåÓýHi Ernst, ?
? To follow up on the last email I sent you, here is a basic example in Python. ? However, you must activate the Keysight-NI-488-2 compatibility option in Connection Expert. ?
? Yves ? ? De?: [email protected] <[email protected]> De la part de dk1vi via groups.io ? Hello Yves_Tardif, |
¿ªÔÆÌåÓýHi Ernst, yes, I received your last email, I will return an answer today with a link for modified program¡yesterday was too busy. ? Python 3.10.0 ? Yves ? De?: [email protected] <[email protected]> De la part de dk1vi via groups.io ? Hello Yves, |
Hi Yves,
sent you this message from two different email adresses but both could not be delivered: ============================================ thanks again for your efforts and your detailed message. ? Unfortunately I will be quite busy till mid of this week, so I will report later. ? Ernst ============================================ I`ll keep you informed about my progress. Regards |
Hello Yves,
thanks to your tutorial on how to get rid of "Unknown Instruments" in Keysight Commander I made some progress. Commander works fine now and when I tested your second programm (dated 20.6.2024), all worked fine now - no"0" measurements. Excellent! However -after deinstalling previous programm- your latest program (23.6.2024) could not be installed. I get this message: "The application can not be started. Contact publisher" all in german. Can not include error message here, because there seem to be a limited length for the message. Thanks for your time. Ernst PS maybe our problems are too specific for the group, but email mostly does not work. Sorry |