What is bandwidth?
The classic definition is when the measured voltage is at the
half power point, which is 0.707 x V (to be measured).
So, your 350MHz scope should show at least 0.707V for a 1V signal
at 350MHz. Manufacturers typically try to do a little better than
what they specify...
500MHz is nearly an octave above the scope's rated bandwidth. What
you measure there could be zero, and it would not be out of spec.
More probably, it will have peaks and valleys all over the map above
the rated bandwidth. But probably would be about 0.25V, with most
of the scopes I have used.
The probe is 10M, and it a real work of art to get such a high impedance
to pass anything at 500MHz. It can't be done without treating the
1M input of the scope and the 10M probe as a unit as far as adjusting
the scope goes. Typically, any probe above 100MHz has to be designed
and adjusted to work with the host scope.
Clearly these probes were made to some other standard than your
350MHz scope.
My rule for scope measurements is don't use a 10M 10x probe for
any measurements above 50MHz, and then only if the probe is grounded
right at the circuit with a short and direct connection to the sleeve
at the probe tip... Never with the ground clip lead!
-Chuck Harris
On Tue, 09 Apr 2024 11:44:41 -0700 "Jinxie" <paul666@...>
wrote:
Hi all,
There's some seller on Ebay offering Agilent 10073C oscilloscope
probes at a knock-down price and naturally I bought one like a
complete idiot would. Anyway, this is supposed to be a 500Mhz passive
probe so I decided to compare it against my existing selection of
passive probes just to get some idea if that bandwidth really was as
advertised. So I set up my Marconi 5.4Ghz RF signal generator for
50-500Mhz output and tested the probes against it one by one.
Obviously the lower BW probes I have couldn't make the grade and fell
flat on their faces well before the half-gig mark and some before
100Mhz even. That was of course expected, however, the Ebay one's
performance was no better and possibly a bit worse than an old 150Mhz
Tek probe I had lying around and it simply could not cut it even at
300Mhz, let alone 500. However, before I shoot my mouth off and
complain about this seller sending me a fake Agilent probe, I thought
I'd best check with the Panel here to make sure I haven't forgotten
to allow for anything in my testing. The scope I am using for this
test is a 350Mhz Tek analog one and it can 'see' waveforms perfectly
well at half gig when connected with a high bandwidth probe (my
3.5Ghz one) so I know the scope's not at fault. Any suggestions?