On Tue, Jun 25, 2024 at 06:45 PM, W4PJB wrote:

......

One thing I did discover this evening is the red LED on my A6A7 YTX Current Driver is lit. Not very brightly, which is probably how I didn't see it until this evening, but it is lit. I haven't had any time to dive into that yet but I suspect it isn't helping. But I'm also getting a proper, full amplitude signal to the LO OUT, perhaps this isn't affecting anything regardless. To be determined another evening.?


=====
Red LED says heater is working hard to bring up temperature of YTX. It should turn on bright at power up, and turn off when YTX temperature is within target. I just checked on my 8566B, red LED was on at power up and was out in less than 1 minute. On your unit does it start bright and get dim later or is it always dim?

As Jeff wrote LO output signal doesn't need YTX. However, if YTX is not working high band will not down convert input RF to IF.?

In case it helps your debug: I measured resistance of YTX pins to ground on a spare YTX of unknown state. ?

pins 2 and 5 are two coils in series ~ 60-ohms across
pin 3 heater drive: 238-ohm to ground
pin 4 heater sensor: 455-ohm ground

Ozan


Re: 8566B - Low signal on 2-22 GHz band

 

I have been reading with interest. At the LO output, you are measuring the YIG local oscillator, correct? However, this tells you nothing about the YIG tuned mixer, which, horror of horrors, I hope is not bad as a big problem if it is. You want to trace the signal up to and after the tuned mixer. After will tell you all the IF stuff is OK and before will let you know the RF input signal is getting to the mixer. A low level 321 MHz signal input to the correct port on the front panel (jumper ports) will tell you the signal goes thru the IFs and all the way to the screen. You may have to disconnect some hardline going to the YTM to see if the RF path is good. If the YTM is not tuning correctly, Bob's your uncle!
Jeff Kruth


Re: 8566B - Low signal on 2-22 GHz band

 

Hi All,?

Thanks again for the suggestions. I'm calling it a night, but I wanted to update this thread before turning in.?

I have already run the test on the LO OUTPUT, and I do get a mixed 321.4 MHz s + my center tuning frequency at the proper amplitude. For example, a 3 GHz input signal, with center tuning at 3 GHz, yields a 3.3214 GHz output on an SA. I did just quickly try putting a short 3 inch SMA jumper to the IF In/Out and there was no change in performance.?I will run the other checks tomorrow.?

One thing I did discover this evening is the red LED on my A6A7 YTX Current Driver is lit. Not very brightly, which is probably how I didn't see it until this evening, but it is lit. I haven't had any time to dive into that yet but I suspect it isn't helping. But I'm also getting a proper, full amplitude signal to the LO OUT, perhaps this isn't affecting anything regardless. To be determined another evening.?

I really appreciate all of the ideas, and I am testing every theory as time allows. Thank you!?


Artek Manuals - vacation announcement

 

A message posting requested by Artek Manuals:
Artek Manuals will on vacation from June 27th through July 5th, returning to answer questions, process orders on Thursday July 6th.

regards, Ron
-


Re: 8566B - Low signal on 2-22 GHz band

 

Follow this sequence:

Right next to the 1st LO OUTPUT connector (front panel) is the IF jumper. Remove the jumper and check it for continuity. The one in the picture on EEVBLOG does not look like the factory original, so maybe someone made a replacement and it's open.
Connect your troubleshooting SA to the "IF OUTPUT" SMA connector. Is the 321.4 MHz signal present?
If you still have a good 321.4 MHz signal there, reconnect the jumper. Does the IF signal make it through the 1.5 GHz low pass?
If it does, and you are sure the cable between A16 LPF and A4 SECOND CONVERTER is good, then the problem is in or around A4 Second Converter. That microcircuit contains a switch. Check the d.c. drive signal to the switch by changing between low band (0-2.5) and high band (2-22). Every time you change bands, the control voltage should do something.
If the control voltage is working, and you have a 321.4 signal going into the second converter microcircuit, then I would suspect the switch inside.

Vladan


Re: 8566B - Low signal on 2-22 GHz band

 

Hi Pete,
you can check:
if there is IF out (321.4MHz) from 1.5GHz LPF A6A16 and from ACLU A6A5 to LPF (jumper A6W22).
if there is YTX A6A8 drive from current driver A6A7
if there is resistance on YTX coil
if there is diode bias to YTX from A6A10

Be careful as YTX diode is very sensitive to static discharge.

73s
Konstantinos

¦²¦Ó¦É? ¦³¦Å¦Ó?¦Ñ¦Ó¦Ç 26 ¦©¦Ï¦Ô¦Í?¦Ï¦Ô 2024 ¦Ò¦Ó¦É? 01:13:54 ¦Ð.¦Ì. EEST, ¦Ï ¦Ö¦Ñ?¦Ò¦Ó¦Ç? W4PJB <petebibbyjr@...> ?¦Ã¦Ñ¦Á¦×¦Å:


Thanks everyone, I tried running the Preselector Peak Routine, no success. Also, just for kicks, I swapped out the YIG with a supposed "tested good" unit. Also, no change, although I didn't expect this to be the issue.?

The SA works perfectly on the low band. There are no errors such as YTO unlock, PL1/PL2, REF Unlock (except when the Oven light is on), M/N unlock, HET unlock, etc. It simply gets a very, very weak signal on the high band. Like -94 dB weak. So it's something in this path for sure. My signal generator only goes to 6 GHz, but now that I think of it, I have an Anritsu SA with TG that goes to 7.1 GHz. Perhaps I can try injecting a 7 GHz signal with the Anritsu to see if it works on the next band up.?

I tested the RF switch with another SA with TG to 7 GHz, there is no attenuation in the switch on either band. I should also add, I did put another SA on the "1ST LO OUTPUT", set a really narrow sweep span on the 8566B (might have been 0), set attenuation to "0", and was able to measure my input signal (3 GHz) + the 321.4 MHz signal, at full strength from the LO OUT. So we've got? great signal at least to there.?



Re: 8566B - Low signal on 2-22 GHz band

 

Thanks everyone, I tried running the Preselector Peak Routine, no success. Also, just for kicks, I swapped out the YIG with a supposed "tested good" unit. Also, no change, although I didn't expect this to be the issue.?

The SA works perfectly on the low band. There are no errors such as YTO unlock, PL1/PL2, REF Unlock (except when the Oven light is on), M/N unlock, HET unlock, etc. It simply gets a very, very weak signal on the high band. Like -94 dB weak. So it's something in this path for sure. My signal generator only goes to 6 GHz, but now that I think of it, I have an Anritsu SA with TG that goes to 7.1 GHz. Perhaps I can try injecting a 7 GHz signal with the Anritsu to see if it works on the next band up.?

I tested the RF switch with another SA with TG to 7 GHz, there is no attenuation in the switch on either band. I should also add, I did put another SA on the "1ST LO OUTPUT", set a really narrow sweep span on the 8566B (might have been 0), set attenuation to "0", and was able to measure my input signal (3 GHz) + the 321.4 MHz signal, at full strength from the LO OUT. So we've got? great signal at least to there.?



Re: k&s 1470 Wire bonding reconstruction

 

Hello Massimo,

Thanks for the tip with the patents!

As for the heater I already have some left over power resistors so I can
build something similar to yours.

I will most probably come to you with more questions regarding the wire
bonding machine.

73!
Razvan

On 25/06/2024 22:09, iv3hed via groups.io wrote:
Razvan, to get an idea of the components of the machine try looking up
the patent for that bonding model.

I did the same thing for the 1470 and found these two patents:
US4202482
US4239144

I suggest you try looking them up at this site:
<>

As for the heater, you can build it yourself with a power resistor.
In my heater I used a 50W resistor to get maximum 150 ¡ãC .
I attach some photos to give you some ideas.

Regards,
Massimo


Re: k&s 1470 Wire bonding reconstruction

 

Razvan, to get an idea of the components of the machine try looking up the patent for that bonding model.

I did the same thing for the 1470 and found these two patents:
US4202482
US4239144

I suggest you try looking them up at this site:

As for the heater, you can build it yourself with a power resistor.
In my heater I used a 50W resistor to get maximum 150 ¡ãC .
I attach some photos to give you some ideas.

Regards,
Massimo


Re: HP 3852a extended memory modules info

 


Great job !
I would like 1 bare board,

thanks,
rick


Re: HP 8566B - Can my high voltage transformer be faulty?

 

¿ªÔÆÌåÓý

Saevar,

?

I used to troubleshoot Solid State CRT power supplies. Replace the fuse and any shorted components that drive the HV transformer. Use a variable

power supply and bring up the voltage slowly to the HV output semiconductor. Monitor the output waveform of the semiconductor(usually the collector) with

an oscilloscope (max range setting). ?The waveform should have *no* small spikes or otherwise called ringing observed in between the higher amplitude

ones. If you have ringing I would suspect the transformer as having shorted turns. With a bad transformer, as the voltage increases the unwanted spikes

continue to grow. This condition usually causes the final semiconductor to fail(short) and then blow the fuse.

?

Some HV output sections have a damper diode in this area that also fails.

?

I don¡¯t have the schematic in front of me but you can use this for a general approach. ?

?

If this test appears to look fine, then possibly on the output of the transformer you will find a HV Tripler circuit that could be faulty. The Tripler

semiconductor device could be self-contained or an arrangement of HV diodes. As the name implies the Tripler multiplies the DC voltage that

connects to the CRT. Having a high voltage probe is the key to measuring the anode voltage on the CRT.

?

Dave

?


>Have you had a high voltage transformer be damaged or do you have any good idea how I can verify whether it is working or not? I can't see any induction >values nor turn ratio of the transformer in the troubleshooting manual so I don't know what are normal values.


Re: 3562A Help/advice

 

A goofed up character generator (ROM) or maybe a dodgy address line to the character generator would be my first guesses. Address buffer chip gone poof maybe?
I do not expect this problem to be caught by POST.

Wilko


HP 8566B - Can my high voltage transformer be faulty?

 

I have a HP 8566B as the title implies. The CRT tube does not turn on and I believe I have traced (at least one) fault to A1A3 high voltage regulator board. The HV oscillator does not oscillate and it just blows the 1A fuse. The Q7 bjt's collector is connected to the high voltage transformer making up the inductor part of the oscillator. The base is also connected to the transformer in a feedback configuration.

The display worked for the first 30 minutes or so of operation after I bought it and then when I inserted low (-10dBm) power at the input the display suddenly turned off, no smoke, no smell but the fuse on A1A6 was blown.

Have you had a high voltage transformer be damaged or do you have any good idea how I can verify whether it is working or not? I can't see any induction values nor turn ratio of the transformer in the troubleshooting manual so I don't know what are normal values.


Re: HP 8340B sweeper and 8757A Scalar Network Analyzer. Do they work together?

 

¿ªÔÆÌåÓý

My pleasure, Marco! Let me know if I can provide any other information!

Steve



On Jun 25, 2024, at 10:01?AM, Marco Marchesi via groups.io <marco.marchesi@...> wrote:

?Great!
So it's time to retire the 8350B and hire the 8340B!
Thank you for your time.
All the best Steve


Re: HP 8405A Vector Voltmeter Probes

 

Hi Jeremy and group,

any updates on the HP8405A probe repair ?

I've bought now a bunch of MMBFJ176 from Mouser and found a source for the 5082-2835 diode here in Europe. In browsing thru Avago Schottky diode data sheets, I found an interesting pair of common anode (HSMS-2823) and cathode (HSMS-2824) double diodes. From the data sheet their electrical characteristics look identical to the single 282Z you tested. Maybe they could make an equivalent circuit to the 4 separate diodes of the HP8405A sampler. What do you think ?
From the HSMS-282x series there is also a bridge and a ring quad (-P & -R) with identical (?!?) pin-out. But they have "wrong" orientation compared to the HP8405A sampler.
Any other suggestions ?

Regards?????? Andreas


Re: HP 8340B sweeper and 8757A Scalar Network Analyzer. Do they work together?

 

Great!
So it's time to retire the 8350B and hire the 8340B!
Thank you for your time.
All the best Steve


Re: HP 8340B sweeper and 8757A Scalar Network Analyzer. Do they work together?

 

¿ªÔÆÌåÓý

Hi Marco,

Yes, let¡¯s see if the picture comes through:

IMG_5365

Steve



On Jun 25, 2024, at 7:34?AM, Marco Marchesi via groups.io <marco.marchesi@...> wrote:

?Thank you Steve,
so you can see the 8757A display annotated with the start-stop frequency set on 8340B?


Re: Frequency References

 

In message <[email protected]>, maxwelloau
<maxwelloau@...> writes
Just use a Leo Bodnar precise reference locked to GPS into a FA-3
Frequency counter with OCXO Vectron oven. That thing is ultra
stable and after a? tweak on receiving from new, its not moved
after a few months? of use. Yes both are on 24hrs? a day.?
Very economical to get extreme accuracy.?
_._,_._,_

Yes second that, good units:)...

--
Tony Sayer


Re: HP 8340B sweeper and 8757A Scalar Network Analyzer. Do they work together?

 

Thank you Steve,
so you can see the 8757A display annotated with the start-stop frequency set on 8340B?


Re: HP 3852a extended memory modules info

 

I will try to make the bodge wire? invisible.?