Re: HP 8563E - CRT turns on only after waiting for a while turned off

 

Hi,
?
After disconnecting the 3x 2.45kV cable from the A17 display driver board and by extension from the CRT (heater or cathode supply) itself, the 110V come up normally. When the 8563 is cold i can see that the DCDC tries to start up for about 2 seconds when it finally turns on and produces the correct voltage and the CRT is turned on. When warm (say I turn it off and then a minute later, I turn it on again) the DCDC just tries to start up indefinitely without ever being able to.
It seems something is loading the DCDC in A6, either the CRT itself or something on the A17 display driver board if it is not the HV multiplier/transformer.
?
What is the most likely here? Are there any obvious tests to try and figure this out?
?


Re: Conductive Paint for Microcircuit Repair...

 

On 7/18/24 05:19, Chuck Harris wrote:
All I really need is one tiny little drop, about 0.04mm.
And no flow additives can lower the conductivity when dried/cured.

you've stumped us all as far as keeping it liquid by changing it.

So, what's left is keeping the surrounding very full of the same volatiles in the conductive paste/paint,
or humid with water.

I think a humid atmosphere around the work area is more feasible than anything else dreamed up, so maybe worth testing on non-critical similar parts.


Re: Conductive Paint for Microcircuit Repair...

 

Hmmm, I see that I was thinking of sizes about 10x larger than your requirement. Still, the sharpened tip of a tuberculin needle, with a wick inside, might be small enough.
?
--
Jim Adney
Madison, WI USA


Re: Conductive Paint for Microcircuit Repair...

 

As others have pointed out, a single hair provides no space for capillary action to hold any paint. I suspect that various kinds of hair vary from .002" to .006", so there's room to fit several into a cluster that will hold paint and still be under .014", especially if you can steal bristles from a soft artist's brush. You might even be able to find a VERY small artist's brush.
?
If a pre-made brush is not available, how about slipping a number of hairs into a small hypodermic needle, gluing them in, and then trimming them to extend only slightly past the end of the needle? You might even be able to use the pointed tip of the needle as your fine tip, with the hairs inside acting only as a wick.
?
The fine needles used in what are known as tuberculin syringes come to mind. I think they are 26 gauge and 1/2" long. Feed the hairs all the way in, and fix in place with a tiny drop of glue only at the top, away from the tip.
?
--
Jim Adney
Madison, WI USA


Re: HP 8557A Start-Up Calibration & Linear Mode - Errors in Manual?

 

Frank,
You have a serious problem if you can only get the Log/Lin amplitudes to match at a ref level of -70dBm.
Let¡¯s start from scratch. Operators check pages 3-10 through 3-14 of the 8557A manual.
In Log mode, Center Freq (CF) set to 250MHz Ref Level set to set to -30 dBm, Span set to 10 MHz, Resoluton Bandwith to 1MHz (RBW) Ref level fine set to 0 dBm, Cal signal connected to the input, what do you see on the display? If you have a displayed signal, can you gradually reduce the Span and RBW (30KHz/div. and 10KHz RBW) and keep the signal centered on the display? Has the amplitude of the displayed signal changed much as you gradually stepped the span down? What should happen is the displayed cal signal (250MHz @ -30 dBm), should be at the top graticule line, centered on the display, and as you reduce the Span/RBW switch (coupled) from widest to narrowest should stay at the center frequency of 250MHz and stay at the top graticule line whether you are in 10dB Log / 1dBLog / Lin mode. There are tolerances so you might see 1 vertical division change in amplitude as you switch to. 1dB Log or Lin mode. You might have to slightly adjust the frequency the keep the signal centered, however if you see abrupt changes in frequency and amplitude - a couple of graticule major divisions or more that points to another problem - resolution bandwidth adjustments (for centering and bandwidths). Also as you switch to narrower spans/RBW (mainly at the narrower span sans RBW), the displayed signal may pull/stretch to the right side of the display, if so increase the sweep time/div (SWP) until it looks normal - a bell shaped signal.
Don Bitters


Re: Conductive Paint for Microcircuit Repair...

 

Hi Chuck,
I've seen you describe all kinds of wondrous techniques over the years and successfully followed some of your advice. Surely, you can grow a (tin) whisker across the gap...
?
Raymond


Re: How to remove the output cable of the isolator in the 8672A

 

Went ahead and decoupled the YTM and shifted it away so I could connect a cable to the low pass filter.
The output power of the power amp was at most +15dBm and dropped around 3 GHz to +5 dBm, way less than the constant input power between +11 and +13 dBm.
As the power amp has a rather low serial number (0365) I'm assuming it is of the "sensitive" type so I ordered a replacement.
I should have measured the amp output power without the isolator and will do that once the replacement amp arrives.


Re: Wanted: 16902B recovery DVD, p/n 16902-14100

 

You're welcome!
?
It looks like there's newer versions up there too.? You can try:
?
?
?
for the full list of files.? Be patient; I think it's un-tarring this gigantic tar file on demand to generate the list.
?
-mark
?
?
?
On Tue, Jul 16, 2024 at 11:38 PM, Eric Smith wrote:

Thank you! I found the image of the media label in my search, but couldn't seem to get from there to the ISO.
?


Re: Conductive Paint for Microcircuit Repair...

 

This is all getting way too complicated for the task. If I
knew how to take a picture with my scope, I should have, but
I don't, and didn't. For that I am truly sorry.

Let me describe it in terms of the macro world we usually work
in:

It is a dual row PC card edge connector where the rows are
staggered, and the conductors connecting to the pads on the outer
row pass in between the pads on the inner row.

The flaw is in one of the conductors that passes in between two
pads on the inner row.

Shrink that model down to sub mm pad widths...

I can't simply cover the gold plated traces, willy-nilly with any
sort of solder paste and blast it with hot air. It will cease to
be a reliable connector if I do.

I can't leave any coating on the surrounding pads to protect my
repair, they are a connector.

I was once a skilled amateur watchmaker, and did lots of work on
ladies watches. This job would have taxed my much younger, skilled
hands back then. Now, at almost 70, I can tell you, things haven't
improved!

As I said originally in the heading, I need a conductive paint
for micro circuit repair.

I need a paint that will allow me to place a single dot over the
top of a flaw that is about 0.04mm in diameter... about 0.0015"
of an inch, or the thickness of one of my hairs.

Thanks for all the help. I think I have enough new ideas to finish
this job.

-Chuck Harris




On Thu, 18 Jul 2024 05:19:55 -0700 "SCMenasian" <scm@...>
wrote:
What I would try (first on a test sample piece) is:
1) Chemically (or thermally) remove the insulation from a piece of
#46 magnet wire. 2) Paint the wire with some Chipquick (T7 or T6
particle size) solder paste. 3) With some tension on the wire (to
hold it straight), heat it to melt the paste and tin the wire. 4) Cut
the tinned wire to length. 5) Paint the entire area of the defective
flexible part with solder resist. 6) With the tip of a scalpel,
scratch a groove (down to the copper conductor but not through it)
where the repair should be made. 7) Apply solder paste so it fills
the groove. 8) Lay the prepared tinned wire in the groove.
9) With a hot air gun, melt the solder paste to complete the
electrical connection. I might experiment with restricting the size
of the hot air stream by using some teflon spaghetti tubing to make a
tiny nozzle.





Re: HP 8557A Start-Up Calibration & Linear Mode - Errors in Manual?

 

Thank you so much Don for your detailed reply.? So for the first adjustment (log/lin relationship cal) I can get the 250Mhz cal peak to the top graticule line in both linear and 10db/div modes.? The setting is ~-70dB when you factor in both coarse and fine knob positions.? But when I try the next adjustment which is the absolute gain cal (Step 33 in the operator's check), the step states to bring the coarse ref level down to -30dB and the fine to 0dB; then by utilizing the 'reference level cal' trimpot, I should be able to bring the 250MHz cal peak all the way to the top graticule in linear mode.? But I can't even get it to first graticule line with these settings and the 'reference level cal' pot turned to max.? Here's the link to my photo album of CRT traces and 8557A settings for each step I'm referring to if that helps to understand my questions (/g/HP-Agilent-Keysight-equipment/album?id=296447).? I numbered them and included the Operator's Check description from the manaul.??
?
Is there a way I can calculate where the height of the 250MHz cal peak should be based on its output spec? Manual states it is -30dBm. I apologize if I am asking rudimentary questions.? I think I need to brush up on my theory regarding signal amplitude and decibels.? This repair is my first introduction to using spectrum analyzers as well.
?
I will get a copy of the 8558A manual as well!
?
-Frank?


Re: Conductive Paint for Microcircuit Repair...

 

What I would try (first on a test sample piece) is:
1) Chemically (or thermally) remove the insulation from a piece of #46 magnet wire.
2) Paint the wire with some Chipquick (T7 or T6 particle size) solder paste.
3) With some tension on the wire (to hold it straight), heat it to melt the paste and tin the wire.
4) Cut the tinned wire to length.
5) Paint the entire area of the defective flexible part with solder resist.
6) With the tip of a scalpel, scratch a groove (down to the copper conductor but not through it) where
? ?the repair should be made.
7) Apply solder paste so it fills the groove.
8) Lay the prepared tinned wire in the groove.
9) With a hot air gun, melt the solder paste to complete the electrical connection.
I might experiment with restricting the size of the hot air stream by using some teflon spaghetti
tubing to make a tiny nozzle.


Re: Conductive Paint for Microcircuit Repair...

 

Solder would have been my first choice, but this is too
small for my smallest soldering iron, and the substrate
isn't meant for solder temperatures. My smallest iron is
a Weller MT1000, which is meant for soldering surface mount
IC's with about a 1mm pitch.

Everything I do at this level is under an Olympus stereo zoom
scope. Ladies wrist watches are easy compared to this stuff.
(I used ladies wrist watches to train myself for SMD work some
30 years ago.)

I tried using watch oilers, the problem is the viscosity of
watch oils and greases is much higher than this silver paint
that I have been using. No drop is formed when I dip into
the paint. All that happens is I put a thin coat on the
oiler, and it dries while I move from the paint to the flex
board.

All I really need is one tiny little drop, about 0.04mm.

-Chuck Harris


On Wed, 17 Jul 2024 16:30:49 -0700 "Scott Vance via groups.io"
<scott_l_vance@...> wrote:
For flexible connections, I've had the best luck with really thin
wire.? You can use solder on a soldering iron at lowish heat to very
carefully burn off the coating and prep it.? Next cut it to length
and then solder in place (or conductive 2-part silver epoxy or
whatever).

For putting down fine lines, in the watchmaking trade we use small
oilers.? They have a tiny tip which holds just a bit of liquid.
Cousins UK is generally the best source for this sort of thing.? ?If
I was putting down a conductive adhesive and it had to be precise it
would be this under a good scope.? Good luck!


Re: HP 8557A Start-Up Calibration & Linear Mode - Errors in Manual?

 

The object of the adjustment is to get no relative change in displayed signal amplitude when you change between Log mode and Lin mode. In the adjustment you are trying to do one adjust affects the DC offset more and the other affects the overall gain of the Log amp. Start the procedure with the fine adjust centered in the adjustment range, then use the coarse adjust to get no shift in amplitude when switching between Log and Lin modes on the display. Then use the overall gain to bring the signal amplitude to the top graticule of the display while decreasing the coarse adjust to 0. Adjust the fine to maintain no relative change when switching between Log mode and Lin mode, if you still run out of fine adjustment range, then note where the coarse adjustment is set (the knob is scaled 0 to -11). If you are off by a small amount on the coarse knob (less than 1dB) then you may have a slight DC offset when switching into Lin mode, if you are off by about -2.5 dB or more on the coarse knob, you have a gain issue from the front end through the log amp section. I have seen this caused by ¡°wounded¡± (leaky) input mixer diodes. Another symptom of this will be marginal or out-of-tolerance frequency response. More typically, you may have excessive DC offset in the Lin mode (caused typically by a leaky transistor in the Log amp Lin mode).
You generally will not not notice a DC offset in Log mode, but it will definitely be noticed in Lin mode.
For whatever reason the 8557A procedure is a little harder to follow than the same procedure in the 8558A manual. Download that portion from the 8558A manual and compare notes with the 8557A procedure that you have. This may help you understand the adjustment procedure.
I am recalling this from memory so there may be a glitch in what I am citing.
Feel free to post more questions.
Don Bitters


Re: Conductive Paint for Microcircuit Repair...

 

Chipquick makes grade T7 (2-11 microns) and T6 (5-15 microns) as well as
coarser grades.


Re: 8702A service manual [SEC=UNOFFICIAL]

 

¿ªÔÆÌåÓý

I¡¯ve changed some of my HP test equipment over to the Simmconn Labs LCD display. Even the ones with a colour display and a RGB output. They are not cheap, but the display quality and the slightly larger viewing area is great, and it makes the instrument look more modern. I don¡¯t know of anyone else that has made a video output capability or LCD conversion for those early HP8753A mono displays. To upgrade the instrument, I think there were several boards had to be changed, as well as the display.

James Goodwin


N9020A MXA programming Python

 

Hello Group,
?
I am writing some small scripts to control the N9020A signal analyzer. Basically as a learning experience, when needed for "serious work" I will use the scripts as a starting point. The Spectrum Analysis and Phase Noise modes are well defined. The Noise Figure mode quite less so. I have the N9020A programming manual (abt 3300 pages) but it does not really cover the Noise Figure mode. I can operate the SA manually of course but that does not really give a clue how to program the instrument. Is anyone familiar with the NF mode and is there a programming manual available that I have not found (on Keysight's website)?
Thanks in advance,
Harke, PA0HRK
?


Re: HP 8970A Noise Figure Meter - Searching for an A6 Input Assembly board

 

This is the A6 Input Assembly (picture from the Internet)
?


HP 8970A Noise Figure Meter - Searching for an A6 Input Assembly board

 

Hello all,
Two years ago I bought an HP 8970A Noise Figure Meter for parts, that did not come complete. Recently I started working on it, whichs means I try to make it like new again. That means taking it appart, clean, check, repair and reassemble. A lot of work, but I like doing it. I did find a problem in the A3 20MHz IF Assembly. +15V and -15V had been swapped and serveral inductors in the -15V line were missing. Somebody had been trying to fix this meter and stopped with it. I reinstalled the inductors and tested the assembly. I found one broken amplifier transistor and then the assembly tested well.
For the rest all other assemblies that are there seem to function, BUT the A6 Input Assembly board is missing.
So my question is, is there a member that has an HP 8970A for parts lying around and has an A6 board for me? It would make me very glad to get this meter going again.
I do have a friends US domestic shipping address, just in case.
?
Jaap PA0JRK
Netherlands


Re: Conductive Paint for Microcircuit Repair...

 

Chuck said
"There is a one part epoxy available, that is heat cured, but
it requires 2-300C for 15 seconds, which I think will melt
the kapton substrate."
?
A Kapton substrate will easilly survive 300C for 15 seconds. There are grades that will work at 400.? If it is amber / brown translucent it is probably Kapton. If colourless or opaque yellow it may be polyester. Test the edge with a soldering iron.
?
Robert.


Re: Conductive Paint for Microcircuit Repair...

 

Dia 0.04mm is #46 AWG. I have repaired/rewired flex boards by supplanting the entire trace with magnet wire 40 AWG soldered to entry-exit points of the flex and secured with flexible adhesive mix.? If the flex is just a part of assembly, this works (provided some larger pads are available). If it is a part of disk head flex assembly flailing about, it will not work very long. It is possible to make the wire "wavy" to allow for expansion-contraction, perhaps.