¿ªÔÆÌåÓý

Date

Re: Measuring the impedance of a lithium ion cell

 

https://static.eleshop.nl/mage/media/downloads/de-5000.jpg


Re: Measuring the impedance of a lithium ion cell

 

And for your series parallel pack, build two sets and measure in same manner.
?
On Tue, Dec 24, 2024 at 08:24 PM, RFI-EMI-GUY wrote:

Would it help to measure two cells in series with polarities reversed to remove the DC offset. And use a common LCR meter Like a DE5000? Obviously both would need to be at same state of charge.


Re: Measuring the impedance of a lithium ion cell

 

Would it help to measure two cells in series with polarities reversed to remove the DC offset. And use a common LCR meter Like a DE5000? Obviously both would need to be at same state of charge.


Re: Measuring the impedance of a lithium ion cell

 

Why sorry to try to help?? I was illustrating the problem,, that the impedance is so low that even 1000, which is clearly impractical, would be needed to even get the impedance into a range that makes any sense.? But even if that would work, it would be so large as to have more artifact than characteristic that I need to find.? And it would annoy the interns to assemble.

As well, the task is also to understand how the design of a battery module is different from a simple calculation of a single cell, so I really need to know the specifics of a cell, then I could measure different module designs.? Also then work up a mathematical model so when a different cell was tested we could somewhat predict the characteristics of a module.

Peter

On 12/24/2024 8:14 PM, Mike Feher via groups.io wrote:

Sorry for trying to help. I did not say 1000.

Mike B. Feher, N4FS

89 Arnold Blvd.

Howell NJ 07731

908-902-3831

-----Original Message-----
From: [email protected] <[email protected]> On Behalf Of Peter Gottlieb via groups.io
Sent: Tuesday, December 24, 2024 8:09 PM
To: [email protected]
Subject: Re: [HP-Agilent-Keysight-equipment] Measuring the impedance of a lithium ion cell

Well a thousand in series would bring the impedance up to 10 ohms, so that's not so practical, especially with the effects of the impedance of the assembly which may exceed what I'm trying to measure.? I could make a straight string of 10 or

15 but that won't really help.? I could boost the excitation signal but was specifically asked to use a small signal.

Peter

On 12/24/2024 8:03 PM, Mike Feher via groups.io wrote:

Peter ¨C
Initial simple thought is to put a bunch in series for the
measurement. Merry Christmas ¨C Mike
Mike B. Feher, N4FS
89 Arnold Blvd.
Howell NJ 07731
908-902-3831
*From:*[email protected]
<[email protected]
<mailto:[email protected]>> *On Behalf Of *Peter

Gottlieb via groups.io
*Sent:* Tuesday, December 24, 2024 6:29 PM
*To:* [email protected]
<mailto:[email protected]>

*Subject:* [HP-Agilent-Keysight-equipment] Measuring the impedance of
a lithium ion cell
As part of an EMI investigation, I was asked if I could find a way to
measure the reactance of a lithium ion cell over the range of 10 kHz
to 20 MHz.? Later I will need to measure a module consisting of a
series-parallel combination of these cells.? The cells are 21700
cylindrical so will have very definite capacitive and inductive components.
The impedances will be low, in the tens of milliohms.
I am not sure how I can do this.? I tried simply hooking up to a 3577A
and test set (knowing the test set only goes down to 100 kHz) but the
cell simply looks like a short (of course).
Does anyone have any thoughts on how I can do this measurement?? I
have other gear to amplify and so forth if a custom setup is needed (likely).
Has anyone been able to make a similar measurement successfully?
Peter
< <>>

?????????? Virus-free.www.avast.com
<
campaign=sig-email&utm_content=emailclient>
<#DAB4FAD8-2DD7-40BB-A1B8-4E2AA1F9FDF2>


Re: Measuring the impedance of a lithium ion cell

 

¿ªÔÆÌåÓý

Sorry for trying to help. I did not say 1000.

?

Mike B. Feher, N4FS

89 Arnold Blvd.

Howell NJ 07731

908-902-3831

?

-----Original Message-----
From: [email protected] <[email protected]> On Behalf Of Peter Gottlieb via groups.io
Sent: Tuesday, December 24, 2024 8:09 PM
To: [email protected]
Subject: Re: [HP-Agilent-Keysight-equipment] Measuring the impedance of a lithium ion cell

?

Well a thousand in series would bring the impedance up to 10 ohms, so that's not so practical, especially with the effects of the impedance of the assembly which may exceed what I'm trying to measure.? I could make a straight string of 10 or

15 but that won't really help.? I could boost the excitation signal but was specifically asked to use a small signal.

?

Peter

?

?

?

On 12/24/2024 8:03 PM, Mike Feher via groups.io wrote:

>?

> Peter ¨C

>?

> Initial simple thought is to put a bunch in series for the

> measurement. Merry Christmas ¨C Mike

>?

> Mike B. Feher, N4FS

>?

> 89 Arnold Blvd.

>?

> Howell NJ 07731

>?

> 908-902-3831

>?

> *From:*[email protected]

> <[email protected]> *On Behalf Of *Peter

> Gottlieb via groups.io

> *Sent:* Tuesday, December 24, 2024 6:29 PM

> *To:* [email protected]

> *Subject:* [HP-Agilent-Keysight-equipment] Measuring the impedance of

> a lithium ion cell

>?

> As part of an EMI investigation, I was asked if I could find a way to

> measure the reactance of a lithium ion cell over the range of 10 kHz

> to 20 MHz.? Later I will need to measure a module consisting of a

> series-parallel combination of these cells.? The cells are 21700

> cylindrical so will have very definite capacitive and inductive components.

>?

> The impedances will be low, in the tens of milliohms.

>?

> I am not sure how I can do this.? I tried simply hooking up to a 3577A

> and test set (knowing the test set only goes down to 100 kHz) but the

> cell simply looks like a short (of course).

>?

> Does anyone have any thoughts on how I can do this measurement?? I

> have other gear to amplify and so forth if a custom setup is needed (likely).

>?

> Has anyone been able to make a similar measurement successfully?

>?

> Peter

>?

>?

> <>

> ?????????? Virus-free.www.avast.com

> <https://www.avast.com/sig-email?utm_medium=email&utm_source=link&utm_

> campaign=sig-email&utm_content=emailclient>

>?

>?

> <#DAB4FAD8-2DD7-40BB-A1B8-4E2AA1F9FDF2>

>

?

?

?

?


Re: Measuring the impedance of a lithium ion cell

 

Well a thousand in series would bring the impedance up to 10 ohms, so that's not so practical, especially with the effects of the impedance of the assembly which may exceed what I'm trying to measure.? I could make a straight string of 10 or 15 but that won't really help.? I could boost the excitation signal but was specifically asked to use a small signal.

Peter

On 12/24/2024 8:03 PM, Mike Feher via groups.io wrote:

Peter ¨C

Initial simple thought is to put a bunch in series for the measurement. Merry Christmas ¨C Mike

Mike B. Feher, N4FS

89 Arnold Blvd.

Howell NJ 07731

908-902-3831

*From:*[email protected] <[email protected]> *On Behalf Of *Peter Gottlieb via groups.io
*Sent:* Tuesday, December 24, 2024 6:29 PM
*To:* [email protected]
*Subject:* [HP-Agilent-Keysight-equipment] Measuring the impedance of a lithium ion cell

As part of an EMI investigation, I was asked if I could find a way to measure the reactance of a lithium ion cell over the range of 10 kHz to 20 MHz.? Later I will need to measure a module consisting of a series-parallel combination of these cells.? The cells are 21700 cylindrical so will have very definite capacitive and inductive components.

The impedances will be low, in the tens of milliohms.

I am not sure how I can do this.? I tried simply hooking up to a 3577A and test set (knowing the test set only goes down to 100 kHz) but the cell simply looks like a short (of course).

Does anyone have any thoughts on how I can do this measurement?? I have other gear to amplify and so forth if a custom setup is needed (likely).

Has anyone been able to make a similar measurement successfully?

Peter


<> Virus-free.www.avast.com <>

<#DAB4FAD8-2DD7-40BB-A1B8-4E2AA1F9FDF2>


Re: Measuring the impedance of a lithium ion cell

 

¿ªÔÆÌåÓý

Peter ¨C

?

Initial simple thought is to put a bunch in series for the measurement. Merry Christmas ¨C Mike

?

Mike B. Feher, N4FS

89 Arnold Blvd.

Howell NJ 07731

908-902-3831

?

From: [email protected] <[email protected]> On Behalf Of Peter Gottlieb via groups.io
Sent: Tuesday, December 24, 2024 6:29 PM
To: [email protected]
Subject: [HP-Agilent-Keysight-equipment] Measuring the impedance of a lithium ion cell

?

As part of an EMI investigation, I was asked if I could find a way to measure the reactance of a lithium ion cell over the range of 10 kHz to 20 MHz.? Later I will need to measure a module consisting of a series-parallel combination of these cells.? The cells are 21700 cylindrical so will have very definite capacitive and inductive components.

?

The impedances will be low, in the tens of milliohms.

?

I am not sure how I can do this.? I tried simply hooking up to a 3577A and test set (knowing the test set only goes down to 100 kHz) but the cell simply looks like a short (of course).

?

Does anyone have any thoughts on how I can do this measurement?? I have other gear to amplify and so forth if a custom setup is needed (likely).

?

Has anyone been able to make a similar measurement successfully?

?

Peter


Virus-free.


Re: 8714ES - Network Analyzer - Can it be converted from 75ohm to 50?

 

Thanks!

-----Message d'origine-----
De : [email protected] <[email protected]> De la part de Peter Gottlieb via groups.io
Envoy¨¦ : 24 d¨¦cembre 2024 19:49
? : [email protected]
Objet : Re: [HP-Agilent-Keysight-equipment] 8714ES - Network Analyzer - Can it be converted from 75ohm to 50?

Nice!


Re: 8714ES - Network Analyzer - Can it be converted from 75ohm to 50?

 

Nice!

On 12/24/2024 7:36 PM, Yves Tardif via groups.io wrote:
This is a software I made for my VNA 8753ES and 8753D.


-----Message d'origine-----
De : [email protected] <[email protected]> De la part de Peter Gottlieb via groups.io
Envoy¨¦ : 24 d¨¦cembre 2024 18:58
? : [email protected]
Objet : Re: [HP-Agilent-Keysight-equipment] 8714ES - Network Analyzer - Can it be converted from 75ohm to 50?

How do you capture the screen data?






Re: 8714ES - Network Analyzer - Can it be converted from 75ohm to 50?

 

This is a software I made for my VNA 8753ES and 8753D.

-----Message d'origine-----
De : [email protected] <[email protected]> De la part de Peter Gottlieb via groups.io
Envoy¨¦ : 24 d¨¦cembre 2024 18:58
? : [email protected]
Objet : Re: [HP-Agilent-Keysight-equipment] 8714ES - Network Analyzer - Can it be converted from 75ohm to 50?

How do you capture the screen data?


Re: 8714ES - Network Analyzer - Can it be converted from 75ohm to 50?

 

How do you capture the screen data?

On 12/24/2024 6:13 PM, Yves Tardif via groups.io wrote:

I just made a comparison between an 8753D 75 Ohms and an 8753ES while testing a filter.

Same cables, same calibration type and same calibration kit obviously.

I superimposed the two images as best I could.

The 8753D 75 Ohms uses mechanical adapters for an impedance of 50 Ohms.

It is not absolute perfection, but still very reasonable.

Red curve: ????S21 8753ES

Green curve: S11 8753ES

Blue curve: ???S21 8753D 75 Ohms

Cyan curve: ??S11 8753D 75 Ohms

The image of the 8753D 75 ohms has been superimposed on that of the 8753ES. We can barely see a few green pixels that are not hidden by the cyan curve of the 8753D 75 Ohms.

Now for a 75 Ohm 8714ET and another at 50 Ohms, I don't know if the performances will be similar...I have never made this comparison.

Yves

*De?:*[email protected] <[email protected]> *De la part de* Patricio A. Greco via groups.io
*·¡²Ô±¹´Ç²â¨¦?:* 24 d¨¦cembre 2024 17:17
*??:* [email protected]
*Objet?:* Re: [HP-Agilent-Keysight-equipment] 8714ES - Network Analyzer - Can it be converted from 75ohm to 50?

I agree

Ing. Patricio A. Greco

Gral. Mart¨ªn Rodr¨ªguez 2159
San Miguel (1663)
Buenos Aires
T: +5411-4455-2557 <tel:+5411-4455-2557>
F: +5411-4032-0072 <tel:+5411-4032-0072>
www.servicios-electronicos.com <>



On 24 Dec 2024, at 18:52, Peter Gottlieb via groups.io
<hpnpilot@...> wrote:

?While not ideal, for most practical applications the accuracy and dynamic
range will be fine, and certainly far better than not having the measurements.

Peter



On 12/24/2024 4:47 PM, Patricio A. Greco via groups.io wrote:

There are an important loose in measurement acuraccy¡­

Ing. Patricio A. Greco

Gral. Mart¨ªn Rodr¨ªguez 2159

San Miguel (1663)

Buenos Aires

T: +5411-4455-2557 <tel:+5411-4455-2557>

F: +5411-4032-0072 <tel:+5411-4032-0072>

www.servicios-electronicos.com <>
<>

On 24 Dec 2024, at 17:58, Yves Tardif via groups.io
<yves_tardif@...> wrote:

?

Hi,

Your 8714ES can be calibrated in 50 Ohms, if you have a 50 Ohm kit
to do it.

But be very careful, the N-type connectors are not compatible with
50 Ohm N connectors.

You will damage the VNA connectors if you try to make a connection
of 50 Ohm connectors in 75 Ohm ones...

I have an 8712 here, and I looked for a conversion to 50 Ohms, but
the input of the couplers are really at 75 Ohms, so you can not do
a conversion without changing the directional couplers.

But another solution is to use a 75 to 50 mechanical adapter.

Expensive adapter, but it exists.

I have already done this exercise and it works very well.

There is also the minimum loss pad 75 to 50 Ohms, but you lose
dynamic range, but it works too.



Yves

*De?:*[email protected]
<[email protected]> *De la part de* Anthony
Kuzub via groups.io

*·¡²Ô±¹´Ç²â¨¦?:* 24 d¨¦cembre 2024 15:17

*??:* [email protected]

*Objet?:* [HP-Agilent-Keysight-equipment] 8714ES - Network
Analyzer - Can it be converted from 75ohm to 50?

Good day

I hope you are Ohm for the holidays!

I recently got an 8714ES - Network Analyzer - I plan on using it
for Antenna design / filter tuning and Amplifierd qualifications
for RF microphone work.

But on closer in spection it is setup and has the calibration kit
for 75ohm- (it mainly did video equipment testing)

1.Can it be converted from 75ohm to 50ohm?

2.IS there anything I can do externally to make it behave like a
50ohm device?

Thanks and Regards - I really apprecite the hivemind here.

Anthony







Measuring the impedance of a lithium ion cell

 

As part of an EMI investigation, I was asked if I could find a way to measure the reactance of a lithium ion cell over the range of 10 kHz to 20 MHz.? Later I will need to measure a module consisting of a series-parallel combination of these cells.? The cells are 21700 cylindrical so will have very definite capacitive and inductive components.
?
The impedances will be low, in the tens of milliohms.
?
I am not sure how I can do this.? I tried simply hooking up to a 3577A and test set (knowing the test set only goes down to 100 kHz) but the cell simply looks like a short (of course).
?
Does anyone have any thoughts on how I can do this measurement?? I have other gear to amplify and so forth if a custom setup is needed (likely).
?
Has anyone been able to make a similar measurement successfully?
?
Peter


Re: 8714ES - Network Analyzer - Can it be converted from 75ohm to 50?

 

¿ªÔÆÌåÓý

I just made a comparison between an 8753D 75 Ohms and an 8753ES while testing a filter.

Same cables, same calibration type and same calibration kit obviously.

I superimposed the two images as best I could.

The 8753D 75 Ohms uses mechanical adapters for an impedance of 50 Ohms.

?

It is not absolute perfection, but still very reasonable.

?

Red curve: ????S21 8753ES

Green curve: S11 8753ES

Blue curve: ???S21 8753D 75 Ohms

Cyan curve: ??S11 8753D 75 Ohms

?

The image of the 8753D 75 ohms has been superimposed on that of the 8753ES. We can barely see a few green pixels that are not hidden by the cyan curve of the 8753D 75 Ohms.

Now for a 75 Ohm 8714ET and another at 50 Ohms, I don't know if the performances will be similar...I have never made this comparison.

?

Yves

?

?

De?: [email protected] <[email protected]> De la part de Patricio A. Greco via groups.io
·¡²Ô±¹´Ç²â¨¦?: 24 d¨¦cembre 2024 17:17
??: [email protected]
Objet?: Re: [HP-Agilent-Keysight-equipment] 8714ES - Network Analyzer - Can it be converted from 75ohm to 50?

?

I agree?

Ing. Patricio A. Greco

Gral. Mart¨ªn Rodr¨ªguez 2159
San Miguel (1663)
Buenos Aires
T:?+5411-4455-2557
F:?+5411-4032-0072



On 24 Dec 2024, at 18:52, Peter Gottlieb via groups.io <hpnpilot@...> wrote:

?While not ideal, for most practical applications the accuracy and dynamic range will be fine, and certainly far better than not having the measurements.

Peter



On 12/24/2024 4:47 PM, Patricio A. Greco via groups.io wrote:

There are an important loose in measurement acuraccy¡­

Ing. Patricio A. Greco

Gral. Mart¨ªn Rodr¨ªguez 2159

San Miguel (1663)

Buenos Aires

T: +5411-4455-2557 <tel:+5411-4455-2557>

F: +5411-4032-0072 <tel:+5411-4032-0072>

<>

?

On 24 Dec 2024, at 17:58, Yves Tardif via groups.io <yves_tardif@...> wrote:

?

?

?

Hi,

?

Your 8714ES can be calibrated in 50 Ohms, if you have a 50 Ohm kit to do it.

?

But be very careful, the N-type connectors are not compatible with 50 Ohm N connectors.

?

You will damage the VNA connectors if you try to make a connection of 50 Ohm connectors in 75 Ohm ones...

?

I have an 8712 here, and I looked for a conversion to 50 Ohms, but the input of the couplers are really at 75 Ohms, so you can not do a conversion without changing the directional couplers.

?

But another solution is to use a 75 to 50 mechanical adapter.

?

Expensive adapter, but it exists.

?

I have already done this exercise and it works very well.

?

There is also the minimum loss pad 75 to 50 Ohms, but you lose dynamic range, but it works too.

?

?

Yves

?

*De?:*[email protected] <[email protected]> *De la part de* Anthony Kuzub via groups.io

*·¡²Ô±¹´Ç²â¨¦?:* 24 d¨¦cembre 2024 15:17

*??:* [email protected]

*Objet?:* [HP-Agilent-Keysight-equipment] 8714ES - Network Analyzer - Can it be converted from 75ohm to 50?

?

Good day

?

I hope you are Ohm for the holidays!

?

I recently got an 8714ES - Network Analyzer - I plan on using it for Antenna design / filter tuning and Amplifierd qualifications for RF microphone work.

?

But on closer in spection it is setup and has the calibration kit for 75ohm- (it mainly did video equipment testing)

?

1.Can it be converted from 75ohm to 50ohm?

?

2.IS there anything I can do externally to make it behave like a 50ohm device?

?

Thanks and Regards - I really apprecite the hivemind here.

?

Anthony

?







Re: 8714ES - Network Analyzer - Can it be converted from 75ohm to 50?

 

¿ªÔÆÌåÓý

I agree?
Ing. Patricio A. Greco

Gral. Mart¨ªn Rodr¨ªguez 2159
San Miguel (1663)
Buenos Aires
T:?+5411-4455-2557
F:?+5411-4032-0072

On 24 Dec 2024, at 18:52, Peter Gottlieb via groups.io <hpnpilot@...> wrote:

?While not ideal, for most practical applications the accuracy and dynamic range will be fine, and certainly far better than not having the measurements.

Peter



On 12/24/2024 4:47 PM, Patricio A. Greco via groups.io wrote:
There are an important loose in measurement acuraccy¡­
Ing. Patricio A. Greco
Gral. Mart¨ªn Rodr¨ªguez 2159
San Miguel (1663)
Buenos Aires
T: +5411-4455-2557 <tel:+5411-4455-2557>
F: +5411-4032-0072 <tel:+5411-4032-0072>
www.servicios-electronicos.com <http://www.servicios-electronicos.com/>

On 24 Dec 2024, at 17:58, Yves Tardif via groups.io <yves_tardif@...> wrote:

?

Hi,

Your 8714ES can be calibrated in 50 Ohms, if you have a 50 Ohm kit to do it.

But be very careful, the N-type connectors are not compatible with 50 Ohm N connectors.

You will damage the VNA connectors if you try to make a connection of 50 Ohm connectors in 75 Ohm ones...

I have an 8712 here, and I looked for a conversion to 50 Ohms, but the input of the couplers are really at 75 Ohms, so you can not do a conversion without changing the directional couplers.

But another solution is to use a 75 to 50 mechanical adapter.

Expensive adapter, but it exists.

I have already done this exercise and it works very well.

There is also the minimum loss pad 75 to 50 Ohms, but you lose dynamic range, but it works too.

https://www.pasternack.com/n-female-n-male-straight-adapter-pe9350-p.aspx

Yves

*De?:*[email protected] <[email protected]> *De la part de* Anthony Kuzub via groups.io
*·¡²Ô±¹´Ç²â¨¦?:* 24 d¨¦cembre 2024 15:17
*??:* [email protected]
*Objet?:* [HP-Agilent-Keysight-equipment] 8714ES - Network Analyzer - Can it be converted from 75ohm to 50?

Good day

I hope you are Ohm for the holidays!

I recently got an 8714ES - Network Analyzer - I plan on using it for Antenna design / filter tuning and Amplifierd qualifications for RF microphone work.

But on closer in spection it is setup and has the calibration kit for 75ohm- (it mainly did video equipment testing)

1.Can it be converted from 75ohm to 50ohm?

2.IS there anything I can do externally to make it behave like a 50ohm device?

Thanks and Regards - I really apprecite the hivemind here.

Anthony








Re: HP 6645A Fan Noise

 

RIFA caps are brownish rectangles. And had RIFA stamped on them ?



One RIFA X cap ready to commit suicide..

Wilko


Re: 8714ES - Network Analyzer - Can it be converted from 75ohm to 50?

 

You could also make your own minimum loss (about 5.7 dB, IIRC) 75 to 50 ohm matching pads.? It's 3 resistors in a pi configuration.? I built one a few years back, with an edge-mount SMA female on the 50 ohm side and an F connector female salvaged from some 75 ohm TV device on the 75 ohm side.? It worked well for a while, and I was able to see the carriers from my DirecTV box on my 8566A spectrum analyzer without ripple from impedance mismatch.? But one of the resistors went open circuit at some point, and the ripple was just as bad as with a direct SMA to F connector thru (which I also happened to have lying around the bench, from God knows where!).

At some point, I'll try again and since now I have a VNA (8 GHz Advantest R3767CG), I'll build 2 and sweep them back to back to characterize the match, loss, and symmetry (S parameters).? Assuming I don't just buy one or two from Mini-Circuits.? Hmmm, at $104 a pop for an SFFF-5075+, maybe not!? So Mini-Circuits uses 2 resistors only for the SFFF-5075+; I assume that's why the loss is a bit higher than the minimum at ~5.9 dB.

Anyway, if you want to build your own, search for minimum loss pad on the Internet.? I found a calculator that gave me the appropriate values.? Good luck!

Jim Ford
Laguna Hills, California, USA

On Tuesday, December 24, 2024 at 02:58:07 PM CST, Yves Tardif via groups.io <yves_tardif@...> wrote:


Hi,

?

Your 8714ES can be calibrated in 50 Ohms, if you have a 50 Ohm kit to do it.

But be very careful, the N-type connectors are not compatible with 50 Ohm N connectors.

You will damage the VNA connectors if you try to make a connection of 50 Ohm connectors in 75 Ohm ones...

I have an 8712 here, and I looked for a conversion to 50 Ohms, but the input of the couplers are really at 75 Ohms, so you can not do a conversion without changing the directional couplers.

?

But another solution is to use a 75 to 50 mechanical adapter.

Expensive adapter, but it exists.

I have already done this exercise and it works very well.

?

There is also the minimum loss pad 75 to 50 Ohms, but you lose dynamic range, but it works too.

?

?

?

Yves

De?: [email protected] <[email protected]> De la part de Anthony Kuzub via groups.io
·¡²Ô±¹´Ç²â¨¦?: 24 d¨¦cembre 2024 15:17
??: [email protected]
Objet?: [HP-Agilent-Keysight-equipment] 8714ES - Network Analyzer - Can it be converted from 75ohm to 50?

?

Good day

I hope you are Ohm for the holidays!

I recently got an 8714ES - Network Analyzer - I plan on using it for Antenna design / filter tuning and Amplifierd qualifications for RF microphone work.

But on closer in spection it is setup and has the calibration kit for 75ohm- (it mainly did video equipment testing)

1.??? Can it be converted from 75ohm to 50ohm?

2.??? IS there anything I can do externally to make it behave like a 50ohm device?

?

Thanks and Regards - I really apprecite the hivemind here.?

?

Anthony?


Re: 8714ES - Network Analyzer - Can it be converted from 75ohm to 50?

 

While not ideal, for most practical applications the accuracy and dynamic range will be fine, and certainly far better than not having the measurements.

Peter

On 12/24/2024 4:47 PM, Patricio A. Greco via groups.io wrote:
There are an important loose in measurement acuraccy¡­
Ing. Patricio A. Greco
Gral. Mart¨ªn Rodr¨ªguez 2159
San Miguel (1663)
Buenos Aires
T: +5411-4455-2557 <tel:+5411-4455-2557>
F: +5411-4032-0072 <tel:+5411-4032-0072>
www.servicios-electronicos.com <>

On 24 Dec 2024, at 17:58, Yves Tardif via groups.io <yves_tardif@...> wrote:

?

Hi,

Your 8714ES can be calibrated in 50 Ohms, if you have a 50 Ohm kit to do it.

But be very careful, the N-type connectors are not compatible with 50 Ohm N connectors.

You will damage the VNA connectors if you try to make a connection of 50 Ohm connectors in 75 Ohm ones...

I have an 8712 here, and I looked for a conversion to 50 Ohms, but the input of the couplers are really at 75 Ohms, so you can not do a conversion without changing the directional couplers.

But another solution is to use a 75 to 50 mechanical adapter.

Expensive adapter, but it exists.

I have already done this exercise and it works very well.

There is also the minimum loss pad 75 to 50 Ohms, but you lose dynamic range, but it works too.



Yves

*De?:*[email protected] <[email protected]> *De la part de* Anthony Kuzub via groups.io
*·¡²Ô±¹´Ç²â¨¦?:* 24 d¨¦cembre 2024 15:17
*??:* [email protected]
*Objet?:* [HP-Agilent-Keysight-equipment] 8714ES - Network Analyzer - Can it be converted from 75ohm to 50?

Good day

I hope you are Ohm for the holidays!

I recently got an 8714ES - Network Analyzer - I plan on using it for Antenna design / filter tuning and Amplifierd qualifications for RF microphone work.

But on closer in spection it is setup and has the calibration kit for 75ohm- (it mainly did video equipment testing)

1.Can it be converted from 75ohm to 50ohm?

2.IS there anything I can do externally to make it behave like a 50ohm device?

Thanks and Regards - I really apprecite the hivemind here.

Anthony


Re: 8714ES - Network Analyzer - Can it be converted from 75ohm to 50?

 

¿ªÔÆÌåÓý

There are an important loose in measurement acuraccy¡­?
Ing. Patricio A. Greco
Gral. Mart¨ªn Rodr¨ªguez 2159
San Miguel (1663)
Buenos Aires
T:?+5411-4455-2557
F:?+5411-4032-0072

On 24 Dec 2024, at 17:58, Yves Tardif via groups.io <yves_tardif@...> wrote:

?

Hi,

?

Your 8714ES can be calibrated in 50 Ohms, if you have a 50 Ohm kit to do it.

But be very careful, the N-type connectors are not compatible with 50 Ohm N connectors.

You will damage the VNA connectors if you try to make a connection of 50 Ohm connectors in 75 Ohm ones...

I have an 8712 here, and I looked for a conversion to 50 Ohms, but the input of the couplers are really at 75 Ohms, so you can not do a conversion without changing the directional couplers.

?

But another solution is to use a 75 to 50 mechanical adapter.

Expensive adapter, but it exists.

I have already done this exercise and it works very well.

?

There is also the minimum loss pad 75 to 50 Ohms, but you lose dynamic range, but it works too.

?

?

?

Yves

De?: [email protected] <[email protected]> De la part de Anthony Kuzub via groups.io
·¡²Ô±¹´Ç²â¨¦?: 24 d¨¦cembre 2024 15:17
??: [email protected]
Objet?: [HP-Agilent-Keysight-equipment] 8714ES - Network Analyzer - Can it be converted from 75ohm to 50?

?

Good day

I hope you are Ohm for the holidays!

I recently got an 8714ES - Network Analyzer - I plan on using it for Antenna design / filter tuning and Amplifierd qualifications for RF microphone work.

But on closer in spection it is setup and has the calibration kit for 75ohm- (it mainly did video equipment testing)

1.??? Can it be converted from 75ohm to 50ohm?

2.??? IS there anything I can do externally to make it behave like a 50ohm device?

?

Thanks and Regards - I really apprecite the hivemind here.?

?

Anthony?


Re: 8714ES - Network Analyzer - Can it be converted from 75ohm to 50?

 

¿ªÔÆÌåÓý

This is like to convert a cat in a dog?
Ing. Patricio A. Greco

Gral. Mart¨ªn Rodr¨ªguez 2159
San Miguel (1663)
Buenos Aires
T:?+5411-4455-2557
F:?+5411-4032-0072

On 24 Dec 2024, at 17:46, Robert G8RPI via groups.io <robert8rpi@...> wrote:

?
To convert to 50R you would have to replace the couplers, front panel connectors the cables between them.
?
Robert.


Re: 8714ES - Network Analyzer - Can it be converted from 75ohm to 50?

 

¿ªÔÆÌåÓý

Hi,

?

Your 8714ES can be calibrated in 50 Ohms, if you have a 50 Ohm kit to do it.

But be very careful, the N-type connectors are not compatible with 50 Ohm N connectors.

You will damage the VNA connectors if you try to make a connection of 50 Ohm connectors in 75 Ohm ones...

I have an 8712 here, and I looked for a conversion to 50 Ohms, but the input of the couplers are really at 75 Ohms, so you can not do a conversion without changing the directional couplers.

?

But another solution is to use a 75 to 50 mechanical adapter.

Expensive adapter, but it exists.

I have already done this exercise and it works very well.

?

There is also the minimum loss pad 75 to 50 Ohms, but you lose dynamic range, but it works too.

?

?

?

Yves

De?: [email protected] <[email protected]> De la part de Anthony Kuzub via groups.io
·¡²Ô±¹´Ç²â¨¦?: 24 d¨¦cembre 2024 15:17
??: [email protected]
Objet?: [HP-Agilent-Keysight-equipment] 8714ES - Network Analyzer - Can it be converted from 75ohm to 50?

?

Good day

I hope you are Ohm for the holidays!

I recently got an 8714ES - Network Analyzer - I plan on using it for Antenna design / filter tuning and Amplifierd qualifications for RF microphone work.

But on closer in spection it is setup and has the calibration kit for 75ohm- (it mainly did video equipment testing)

1.??? Can it be converted from 75ohm to 50ohm?

2.??? IS there anything I can do externally to make it behave like a 50ohm device?

?

Thanks and Regards - I really apprecite the hivemind here.?

?