开云体育

Date

8596E adjustment SAMPLING MATHC

 

I have a faulty 8596E, on page 42 the sampler match adjustment, I don't see the signal I should have when carrying out the procedure. I ask if the anomaly may reside in the sampling oscillator circuit, it reports the name of three signals, SD-RELDCK-SAMP5W-SAMPDRY, in this circuit I don't see any signal present, what should I look for?


Re: 8516A cable question

Lothar baier
 

开云体育

First just because the source works with a 8515 doesn’t mean its ok as the 8516 requires different source parameters as the 15 ,? generally no if found or phaselock error comes about if the source power is too low to generate a meaningful IF signal from the reference samplers , for the 8516 the power into the testset must be between 0 and 10dBm for the 45MHz to 20GHz range and 10dBm for the 20-40GHz range , you mentioned that you have a flashing light on your source when the error occurs so more than likely your source goes unleveled when you use the doubler !

?

?

?

This is the link to the service manual ,? check the unratioed and sampler test procedure to make sure your samplers are ok !

?

?

?

From: [email protected] <[email protected]> On Behalf Of Willy via groups.io
Sent: Sunday, March 20, 2022 6:04 PM
To: [email protected]
Subject: Re: [HP-Agilent-Keysight-equipment] 8516A cable question

?

07 APR 92 shows up on the right display of my 8340B, SN is 3050A. Is this new enough?


Re: Protecting Equipment

 

开云体育

I made two of these to use with older communications receivers. Power transformers in the day often used paper that becomes acidic and crackely now that it's seventy years old, National recivers are especially vulnerable. I wonder if vacuum impregnating them with silicon oil or winding varnish would help keep them from? developing the intra-winding shorts they are so prone to. I can't recall ever seeing one of them have an inter-winding short, an open winding, or a short to ground.

?? Bruce Gentry, KA2IVY







On 3/21/22 10:59, Roy J. Tellason, Sr. wrote:

On Sunday 20 March 2022 02:02:21 pm greenboxmaven via groups.io wrote:
An older Uninterruptable Power Supply makes a very convenient way to 
reduce line voltage. Find one of the old "loaf of bread" sized units 
that uses the large 60 cycle transformer, rather than the newer ones 
that use a switching inverter. Remove everything except the transformer, 
fuse holder, outlets and switch. Use one half of the? primary winding 
along with the 120 volt secondary to make an autotranformer. It will 
drop 120 volts to about 109, which is much easier on older gear designed 
for 110 or 115 volts.

 ???? Bruce Gentry, KA2IVY
Interesting!  I happen to have a pile of those...


Re: Protecting Equipment

 

On Sunday 20 March 2022 02:02:21 pm greenboxmaven via groups.io wrote:
An older Uninterruptable Power Supply makes a very convenient way to
reduce line voltage. Find one of the old "loaf of bread" sized units
that uses the large 60 cycle transformer, rather than the newer ones
that use a switching inverter. Remove everything except the transformer,
fuse holder, outlets and switch. Use one half of the? primary winding
along with the 120 volt secondary to make an autotranformer. It will
drop 120 volts to about 109, which is much easier on older gear designed
for 110 or 115 volts.

???? Bruce Gentry, KA2IVY
Interesting! I happen to have a pile of those...

--
Member of the toughest, meanest, deadliest, most unrelenting -- and
ablest -- form of life in this section of space, ?a critter that can
be killed but can't be tamed. ?--Robert A. Heinlein, "The Puppet Masters"
-
Information is more dangerous than cannon to a society ruled by lies. --James
M Dakin


Re: BS, was Re: [HP-Agilent-Keysight-equipment] Spread the word -- Swap Meet Returns!

 

On Monday 21 March 2022 01:41:19 am Michael A. Terrell wrote:
My first real soldering tool was a Weller 8200 soldering gun ,
I still have (and occasionally use) one of those.

in its carry case.
They had a carry case? Never saw one of those...

(snip)

I used 99.5% tin, .5% antimony solder.
I just acquired a roll of that stuff. I'm not sure what I'm gonna do with it...

--
Member of the toughest, meanest, deadliest, most unrelenting -- and
ablest -- form of life in this section of space, ?a critter that can
be killed but can't be tamed. ?--Robert A. Heinlein, "The Puppet Masters"
-
Information is more dangerous than cannon to a society ruled by lies. --James
M Dakin


Re: Agilent 3458A Troubleshooting

 

Hi George,

Well the good news is that it is something basic rather than something (as yet) complicated.... :)

Another vote here for xDevs site & EEVblog?forum threads. I assume you've done a complete check for power & toasted components. Double check that everything is plugged in where it should be (say, by comparison with say xDevs images) - and that all the cables and sockets seem good. Particularly those to the digital board and display (and check the display voltage too) - it sounds like it's not even starting its boot sequence properly.

If not socketed already, I agree with your intuition, and would desolder, read and then socket the Dallas rams and ROMs. Compare ROMs against known good and also see if you can get any data out of the Dallas rams - I would think you have more than a 50/50 chance. You may want to change the electrolytics while you're there - a list is on my site:?

You don't say much about the provenance - maybe there isn't any, but maybe you can get a few clues? Re: 'but someone has maintained over the years' - hmmm, I think I might take all the boards out and give both sides a good look under a magnifier to see if you can see whether any detail work has been done already. Great care with fingerprint contamination of the analogue board of course. A swap with a known good digital board would be a helpful and quick diagnosis aid if you can borrow one. All the manuals, including schematics, are out there as pdfs, of course.

Ah, you're in Sydney! - I think there's a good chance EEVblog Dave could help a lot in person :) From (hazy) memory I think he has a 3458a or two to compare with / test against.

Alan


Re: BS, was Re: [HP-Agilent-Keysight-equipment] Spread the word -- Swap Meet Returns!

 

While I am younger, I also used a wood burning tool for soldering a few times.? My soldering iron had been taken away as punishment for something (I no longer recall), and I had a project I wanted to finish...? it didn't have a proper tinned tip, of course, but worked well and became my backup iron.? I still don't understand the appeal of charring wood!

-Dave

On Sunday, March 20, 2022, 09:57:23 PM PDT, John kolb <jlkolb@...> wrote:



Young punk kids :)? In 1957 I was soldering using the tool from my wood
burning set.

John? 1942


On 3/20/2022 8:33 AM, John Griessen wrote:
>
> On 3/20/22 04:51, Paul Bicknell wrote:
>> Hi
>> wasn't the EE20 the ad on set with speaker and a lot more components
>> wish I still had it as it was an inspiration for lots my age into
>> electronics
>
> I had one of the beginner sets from Philips also when I was 7 or 8 in
> Beaumont Texas.? I'm 1957 vintage also.
>
>
>
>
>
>






Re: BS, was Re: [HP-Agilent-Keysight-equipment] Spread the word -- Swap Meet Returns!

 

My first real soldering tool was a Weller 8200 soldering gun , in its carry case. I bought it before I turned 13 to repair raids and TVs in the mid '60s. I had a wood burner but the tips were too big to use on most electronics because you would burn wires. It was followed by an ENDEVCO desoldering iron a few months later. I still have the gun.
I next bought a Weller SP23, and a SP175 to solder the mounting rings of tube sockets to a brass chassis for an RF amp. Since then I've used many other irons, and an oxyacetylene torch to solder the 3" copper pipes in an RCA TV transmitter. Some idiot tried to fix a leak by brazing the pipe to the brass fittings. I had fun fixing that, since they were custom made and nearly 40 years old. I used 99.5% tin, .5% antimony solder. I tinned the ends of the pipe and inside the brass first, to ensure no leaks. That solder had a mirror finish. :)

I couldn't get spare tips for the SP23 while stationed in Alaska, so I asked the Base engineers for some scrap welding rod and to borrow a die and the handle to thread it. A very large pair of dykes gave a decent chisel tip. That was almost 50 years ago and I still have a couple of those tips. Solder doesn't eat brass like it does pure copper.


On Mon, Mar 21, 2022 at 12:57 AM John kolb <jlkolb@...> wrote:

Young punk kids :)? In 1957 I was soldering using the tool from my wood
burning set.

John? 1942

On 3/20/2022 8:33 AM, John Griessen wrote:
>
> On 3/20/22 04:51, Paul Bicknell wrote:
>> Hi
>> wasn't the EE20 the ad on set with speaker and a lot more components
>> wish I still had it as it was an inspiration for lots my age into
>> electronics
>
> I had one of the beginner sets from Philips also when I was 7 or 8 in
> Beaumont Texas.? I'm 1957 vintage also.
>
>
>
>
>
>






Re: BS, was Re: [HP-Agilent-Keysight-equipment] Spread the word -- Swap Meet Returns!

 

Young punk kids :) In 1957 I was soldering using the tool from my wood burning set.

John 1942

On 3/20/2022 8:33 AM, John Griessen wrote:
On 3/20/22 04:51, Paul Bicknell wrote:
Hi
wasn't the EE20 the ad on set with speaker and a lot more components
wish I still had it as it was an inspiration for lots my age into electronics
I had one of the beginner sets from Philips also when I was 7 or 8 in Beaumont Texas.? I'm 1957 vintage also.


Re: Protecting Equipment

 

开云体育

Yes, they are because the inductors in the LISN can generate some nasty spikes when device power is cycled.

?

Most conducted EMC specs cut off at 30 MHz, or at least did up to the last time I worked in that area about 10 years ago.

?

Tom, N8ZM

?

From: [email protected] <[email protected]> On Behalf Of Lothar baier
Sent: Sunday, March 20, 2022 10:14 PM
To: [email protected]
Subject: Re: [HP-Agilent-Keysight-equipment] Protecting Equipment

?

Those are transient limiters that are especially designed for this purpose , usually they only cover to 30 or 100MHz t match the frequency range of the LISN ( I only have seen the ones made by ETS that went to 100MHz not sure if they make higher frequency ones )

As far as the limiter itself concerns usually one attenuator of 3Db is used on the input followed by a diode limiter which is followed by another attenuator and then another ?limiter .

?

?

From: [email protected] <[email protected]> On Behalf Of n8zmTWH via groups.io
Sent: Sunday, March 20, 2022 8:22 PM
To: [email protected]
Subject: Re: [HP-Agilent-Keysight-equipment] Protecting Equipment

?

Limiters are very commonly recommended for Conducted Immunity testing in EMC labs where the Spec An input comes from the Line Impedance Stabilization Network (LISN) because the turn on/off transients can be quite high. The levels expected to be measured typically should be much less than the distortion level of the limiter. The recommended limiter in this case actually has an intentional 10 dB insertion loss.

?

Tom, N8ZM

?

From: [email protected] <[email protected]> On Behalf Of Lothar baier
Sent: Sunday, March 20, 2022 9:03 PM
To: [email protected]
Subject: Re: [HP-Agilent-Keysight-equipment] Protecting Equipment

?

A few words about limiters ,

Generally limiters are used in systems to protect receivers usually in RADAR or EW applications ,? in test systems the use of limiters s usually frowned upon ( with very few exceptions)? and for good reason !

First of there is a cost penalty that is not insignificant , the cost for a connectorized limiter usually starts at about $600 and can go up into the $2K range .

Limiters also degrade VSWR and add additional system loss .

But the by far biggest problem is that limiters are non linear devices and as such excellent harmonics generators , harmonics will start to occur before the limiter hits the hard threshold and worsens as the power increases so if you put one in front a spectrum analyzer you have to be cautious not to end up with tst results that are worse than they actually are

?

From: [email protected] <[email protected]> On Behalf Of Jim Ford via groups.io
Sent: Sunday, March 20, 2022 2:53 PM
To: [email protected]
Subject: Re: [HP-Agilent-Keysight-equipment] Protecting Equipment

?

Mini-Circuits sells limiters as well as DC blocks, but they only go up to 8 GHz.? Not sure about the price, but they are most likely competitive.? Assuming you actually need one.

?

Jim Ford

?

------ Original Message ------

From: "Matt Huszagh" <huszaghmatt@...>

Sent: 3/20/2022 6:52:14 AM

Subject: Re: [HP-Agilent-Keysight-equipment] Protecting Equipment

?

It's worth adding that a DC block only provides some measure of safety. The ramp up voltage of a DC signal is an AC signal and can pass right through your series cap. If you look at DC blocking filter designs like the HP 3048A option K23, there's a limiter after the series cap. It seems like the "right" devices for this sort of thing are RF power limiters. Unfortunately, the cost of many of these devices new from reputable manufacturers probably matches or exceeds what many people paid for their second-hand spectrum analyzer. Maybe there are good second-hand market options, but I haven't looked into to it. If you want something to reliably protect a low frequency SA, that shouldn't be too hard to design and build. But, if you want to get the full 22 GHz range out of your 8566, I expect that wouldn't be so easy.


Re: Youtube video HP service center in the 90th

 

If you look at the bottom edge of the video, it shows a timing error that flashes in time with the vertical jitter.

On Thu, Mar 17, 2022 at 4:42 PM Dave_G0WBX via <g8kbvdave=[email protected]> wrote:
Amazing how the video bounces up/down in time with the music! Almost as
if it was supposed to do that.

Regards.

???? Dave B.

--
Created on and sent from a Unix like PC running and using free and open source software:







Re: Protecting Equipment

Lothar baier
 

开云体育

Those are transient limiters that are especially designed for this purpose , usually they only cover to 30 or 100MHz t match the frequency range of the LISN ( I only have seen the ones made by ETS that went to 100MHz not sure if they make higher frequency ones )

As far as the limiter itself concerns usually one attenuator of 3Db is used on the input followed by a diode limiter which is followed by another attenuator and then another ?limiter .

?

?

From: [email protected] <[email protected]> On Behalf Of n8zmTWH via groups.io
Sent: Sunday, March 20, 2022 8:22 PM
To: [email protected]
Subject: Re: [HP-Agilent-Keysight-equipment] Protecting Equipment

?

Limiters are very commonly recommended for Conducted Immunity testing in EMC labs where the Spec An input comes from the Line Impedance Stabilization Network (LISN) because the turn on/off transients can be quite high. The levels expected to be measured typically should be much less than the distortion level of the limiter. The recommended limiter in this case actually has an intentional 10 dB insertion loss.

?

Tom, N8ZM

?

From: [email protected] <[email protected]> On Behalf Of Lothar baier
Sent: Sunday, March 20, 2022 9:03 PM
To: [email protected]
Subject: Re: [HP-Agilent-Keysight-equipment] Protecting Equipment

?

A few words about limiters ,

Generally limiters are used in systems to protect receivers usually in RADAR or EW applications ,? in test systems the use of limiters s usually frowned upon ( with very few exceptions)? and for good reason !

First of there is a cost penalty that is not insignificant , the cost for a connectorized limiter usually starts at about $600 and can go up into the $2K range .

Limiters also degrade VSWR and add additional system loss .

But the by far biggest problem is that limiters are non linear devices and as such excellent harmonics generators , harmonics will start to occur before the limiter hits the hard threshold and worsens as the power increases so if you put one in front a spectrum analyzer you have to be cautious not to end up with tst results that are worse than they actually are

?

From: [email protected] <[email protected]> On Behalf Of Jim Ford via groups.io
Sent: Sunday, March 20, 2022 2:53 PM
To: [email protected]
Subject: Re: [HP-Agilent-Keysight-equipment] Protecting Equipment

?

Mini-Circuits sells limiters as well as DC blocks, but they only go up to 8 GHz.? Not sure about the price, but they are most likely competitive.? Assuming you actually need one.

?

Jim Ford

?

------ Original Message ------

From: "Matt Huszagh" <huszaghmatt@...>

Sent: 3/20/2022 6:52:14 AM

Subject: Re: [HP-Agilent-Keysight-equipment] Protecting Equipment

?

It's worth adding that a DC block only provides some measure of safety. The ramp up voltage of a DC signal is an AC signal and can pass right through your series cap. If you look at DC blocking filter designs like the HP 3048A option K23, there's a limiter after the series cap. It seems like the "right" devices for this sort of thing are RF power limiters. Unfortunately, the cost of many of these devices new from reputable manufacturers probably matches or exceeds what many people paid for their second-hand spectrum analyzer. Maybe there are good second-hand market options, but I haven't looked into to it. If you want something to reliably protect a low frequency SA, that shouldn't be too hard to design and build. But, if you want to get the full 22 GHz range out of your 8566, I expect that wouldn't be so easy.


Re: Protecting Equipment

 

开云体育

I mis-spoke. I meant Conducted Emissions. Sorry.

?

Tom, N8ZM

?

From: [email protected] <[email protected]> On Behalf Of n8zmTWH@...
Sent: Sunday, March 20, 2022 9:22 PM
To: [email protected]
Subject: Re: [HP-Agilent-Keysight-equipment] Protecting Equipment

?

Limiters are very commonly recommended for Conducted Immunity testing in EMC labs where the Spec An input comes from the Line Impedance Stabilization Network (LISN) because the turn on/off transients can be quite high. The levels expected to be measured typically should be much less than the distortion level of the limiter. The recommended limiter in this case actually has an intentional 10 dB insertion loss.

?

Tom, N8ZM

?

From: [email protected] <[email protected]> On Behalf Of Lothar baier
Sent: Sunday, March 20, 2022 9:03 PM
To: [email protected]
Subject: Re: [HP-Agilent-Keysight-equipment] Protecting Equipment

?

A few words about limiters ,

Generally limiters are used in systems to protect receivers usually in RADAR or EW applications ,? in test systems the use of limiters s usually frowned upon ( with very few exceptions)? and for good reason !

First of there is a cost penalty that is not insignificant , the cost for a connectorized limiter usually starts at about $600 and can go up into the $2K range .

Limiters also degrade VSWR and add additional system loss .

But the by far biggest problem is that limiters are non linear devices and as such excellent harmonics generators , harmonics will start to occur before the limiter hits the hard threshold and worsens as the power increases so if you put one in front a spectrum analyzer you have to be cautious not to end up with tst results that are worse than they actually are

?

From: [email protected] <[email protected]> On Behalf Of Jim Ford via groups.io
Sent: Sunday, March 20, 2022 2:53 PM
To: [email protected]
Subject: Re: [HP-Agilent-Keysight-equipment] Protecting Equipment

?

Mini-Circuits sells limiters as well as DC blocks, but they only go up to 8 GHz.? Not sure about the price, but they are most likely competitive.? Assuming you actually need one.

?

Jim Ford

?

------ Original Message ------

From: "Matt Huszagh" <huszaghmatt@...>

Sent: 3/20/2022 6:52:14 AM

Subject: Re: [HP-Agilent-Keysight-equipment] Protecting Equipment

?

It's worth adding that a DC block only provides some measure of safety. The ramp up voltage of a DC signal is an AC signal and can pass right through your series cap. If you look at DC blocking filter designs like the HP 3048A option K23, there's a limiter after the series cap. It seems like the "right" devices for this sort of thing are RF power limiters. Unfortunately, the cost of many of these devices new from reputable manufacturers probably matches or exceeds what many people paid for their second-hand spectrum analyzer. Maybe there are good second-hand market options, but I haven't looked into to it. If you want something to reliably protect a low frequency SA, that shouldn't be too hard to design and build. But, if you want to get the full 22 GHz range out of your 8566, I expect that wouldn't be so easy.


Re: Protecting Equipment

 

开云体育

Limiters are very commonly recommended for Conducted Immunity testing in EMC labs where the Spec An input comes from the Line Impedance Stabilization Network (LISN) because the turn on/off transients can be quite high. The levels expected to be measured typically should be much less than the distortion level of the limiter. The recommended limiter in this case actually has an intentional 10 dB insertion loss.

?

Tom, N8ZM

?

From: [email protected] <[email protected]> On Behalf Of Lothar baier
Sent: Sunday, March 20, 2022 9:03 PM
To: [email protected]
Subject: Re: [HP-Agilent-Keysight-equipment] Protecting Equipment

?

A few words about limiters ,

Generally limiters are used in systems to protect receivers usually in RADAR or EW applications ,? in test systems the use of limiters s usually frowned upon ( with very few exceptions)? and for good reason !

First of there is a cost penalty that is not insignificant , the cost for a connectorized limiter usually starts at about $600 and can go up into the $2K range .

Limiters also degrade VSWR and add additional system loss .

But the by far biggest problem is that limiters are non linear devices and as such excellent harmonics generators , harmonics will start to occur before the limiter hits the hard threshold and worsens as the power increases so if you put one in front a spectrum analyzer you have to be cautious not to end up with tst results that are worse than they actually are

?

From: [email protected] <[email protected]> On Behalf Of Jim Ford via groups.io
Sent: Sunday, March 20, 2022 2:53 PM
To: [email protected]
Subject: Re: [HP-Agilent-Keysight-equipment] Protecting Equipment

?

Mini-Circuits sells limiters as well as DC blocks, but they only go up to 8 GHz.? Not sure about the price, but they are most likely competitive.? Assuming you actually need one.

?

Jim Ford

?

------ Original Message ------

From: "Matt Huszagh" <huszaghmatt@...>

Sent: 3/20/2022 6:52:14 AM

Subject: Re: [HP-Agilent-Keysight-equipment] Protecting Equipment

?

It's worth adding that a DC block only provides some measure of safety. The ramp up voltage of a DC signal is an AC signal and can pass right through your series cap. If you look at DC blocking filter designs like the HP 3048A option K23, there's a limiter after the series cap. It seems like the "right" devices for this sort of thing are RF power limiters. Unfortunately, the cost of many of these devices new from reputable manufacturers probably matches or exceeds what many people paid for their second-hand spectrum analyzer. Maybe there are good second-hand market options, but I haven't looked into to it. If you want something to reliably protect a low frequency SA, that shouldn't be too hard to design and build. But, if you want to get the full 22 GHz range out of your 8566, I expect that wouldn't be so easy.


Re: Protecting Equipment

Lothar baier
 

开云体育

A few words about limiters ,

Generally limiters are used in systems to protect receivers usually in RADAR or EW applications ,? in test systems the use of limiters s usually frowned upon ( with very few exceptions)? and for good reason !

First of there is a cost penalty that is not insignificant , the cost for a connectorized limiter usually starts at about $600 and can go up into the $2K range .

Limiters also degrade VSWR and add additional system loss .

But the by far biggest problem is that limiters are non linear devices and as such excellent harmonics generators , harmonics will start to occur before the limiter hits the hard threshold and worsens as the power increases so if you put one in front a spectrum analyzer you have to be cautious not to end up with tst results that are worse than they actually are

?

From: [email protected] <[email protected]> On Behalf Of Jim Ford via groups.io
Sent: Sunday, March 20, 2022 2:53 PM
To: [email protected]
Subject: Re: [HP-Agilent-Keysight-equipment] Protecting Equipment

?

Mini-Circuits sells limiters as well as DC blocks, but they only go up to 8 GHz.? Not sure about the price, but they are most likely competitive.? Assuming you actually need one.

?

Jim Ford

?

------ Original Message ------

From: "Matt Huszagh" <huszaghmatt@...>

Sent: 3/20/2022 6:52:14 AM

Subject: Re: [HP-Agilent-Keysight-equipment] Protecting Equipment

?

It's worth adding that a DC block only provides some measure of safety. The ramp up voltage of a DC signal is an AC signal and can pass right through your series cap. If you look at DC blocking filter designs like the HP 3048A option K23, there's a limiter after the series cap. It seems like the "right" devices for this sort of thing are RF power limiters. Unfortunately, the cost of many of these devices new from reputable manufacturers probably matches or exceeds what many people paid for their second-hand spectrum analyzer. Maybe there are good second-hand market options, but I haven't looked into to it. If you want something to reliably protect a low frequency SA, that shouldn't be too hard to design and build. But, if you want to get the full 22 GHz range out of your 8566, I expect that wouldn't be so easy.


Re: BS, was Re: [HP-Agilent-Keysight-equipment] Spread the word -- Swap Meet Returns!

 

I had one of these from International Rectifier back in the mid 60's.
https://www.worthpoint.com/worthopedia/ir-international-rectifier-solar-1810736590


Re: 8516A cable question

Lothar baier
 

开云体育

Yes that should work fine !

?

From: [email protected] <[email protected]> On Behalf Of Willy via groups.io
Sent: Sunday, March 20, 2022 6:04 PM
To: [email protected]
Subject: Re: [HP-Agilent-Keysight-equipment] 8516A cable question

?

07 APR 92 shows up on the right display of my 8340B, SN is 3050A. Is this new enough?


Re: HP 54522A - risetime measurement vs. graticule readings

 

Here is the attachment.?


On Sun, Mar 20, 2022 at 5:09 PM Radu Bogdan Dicher via <vondicher=[email protected]> wrote:
Hi all,
Working on aligning another instrument last night, my scope - which I typically find extremely reliable for all the readings it's able to perform (from voltage/level to time-based and everything in between) - has been giving me some confusing readings. I could use some community "sanity checks" on my thinking here, please.?

Per enclosed scope screenshot (coming up), the rising segment of this triangular wave (which triggers a VCOs period) is in my book (per graticules indication) 120ms. And yet, the scope sees that as a 32.35ms risetime wave. The scope reading was rather fluctuating, but at no point coming even close to the 120ms I'd expect. Please note the period reading seems to be correct.?

Given the information presented on the image, when I was doing this last night, I decided to ignore the "risetime" reading and execute the adjustment I was required (120ms sweep time) as the image shows it.?

What could be going on here?.... Am I looking at this wrong? Or shall I start worrying that my scope is beginning to exhibit some sort of illness?

Thank you.?
Radu.?


HP 54522A - risetime measurement vs. graticule readings

 

Hi all,
Working on aligning another instrument last night, my scope - which I typically find extremely reliable for all the readings it's able to perform (from voltage/level to time-based and everything in between) - has been giving me some confusing readings. I could use some community "sanity checks" on my thinking here, please.?

Per enclosed scope screenshot (coming up), the rising segment of this triangular wave (which triggers a VCOs period) is in my book (per graticules indication) 120ms. And yet, the scope sees that as a 32.35ms risetime wave. The scope reading was rather fluctuating, but at no point coming even close to the 120ms I'd expect. Please note the period reading seems to be correct.?

Given the information presented on the image, when I was doing this last night, I decided to ignore the "risetime" reading and execute the adjustment I was required (120ms sweep time) as the image shows it.?

What could be going on here?.... Am I looking at this wrong? Or shall I start worrying that my scope is beginning to exhibit some sort of illness?

Thank you.?
Radu.?


Re: 8516A cable question

 

07 APR 92 shows up on the right display of my 8340B, SN is 3050A. Is this new enough?