Note: the flicker source is not in the ovenized crystal oscillator; the oven is powered even in standby. When powering up from standby, the synthesizer is spot on the frequencies up to 20ghz for about 10 minutes, then the tuning voltage jumps are starting and the frequency offsets are occurring.?

This will be a fun repair, after the nightmare of replacing the mixer (was literally melted) and a blown power supply (wire bonded, so I had to get creative) in the band 0 downconverter.?


Re: Slightly off topic puzzle

 

Meantime the puzzle was solved. The bug is in the 6747B synthesizer, the reference loop (10MHZ), although reported as locked, it ¡°flickers¡±. Easy to see if monitoring the VCO tuning voltage, it quickly changes quite a bit, after the instrument warms up. These flickers create spurious components at the output which in turn falsely trigger the counter gate. Before warming up, the frequencies are dead on.

The root cause of the flickers is not yet known, just in case I ordered the two boards a6 (Vco and loop gain) and a7 (reference divider). I already tried replacing the opamp, still the same. It could be the 500MHz SAW oscillator (unobtanium, but can be replaced by a maxim max2609 Vco plus a gain stage, plus some dc (bias, tuning) scaling,?there¡¯s room enough on the board) or the ECL and schottky counters/flip-flops and gates in the divider and phase detector. My bet goes toward the SAW oscillator hybrid, but fixing the current boards will happen later, I have to fix the synthesizer first.



Re: Hp 70909A vs 70910A

 

Paul

Responding to your older post about the 70910A Receiver Personality software.? I have it loaded on my system and desire to archive it.? Perhaps someone here can offer me some ideas on how to do this/talk me through it?? I'm concerned that I don't corrupt or accidentally delete any of the files while I'm at it.? I have a couple 32KB cards, but the total size of the files is about 700K or so.? I don't have the 9122 series floppy drive.? Any thoughts on copying these files over the GPIB to hard disc so I can post them to KO4BB for posterity (and system safety/longevity)?

Adam


Re: 53131A, 53132A, 53181A Extra Channel option upgrade/downgrade

 

Hi all,

OK - I've stuffed up. my eeprom got erased during this process. I've?done a recal except for the time interval fine.?However my eeprom is pretty much blank now and doesn't have?the header words. Are they at the end of the 8K?


Could I trouble someone for an eeprom image please to compare with what I have now?


I assume the header was at the end of the eeprom as the instructions are to change the start of the eeprom, my eeprom is FF from 0x01 through 0x1F with the firmware revision at 0x28?(3944). It's then blank from 0x200 through 0x1FFF.


My unit has no freq extenders (at the moment), I'm designing my own 12 GHz unit. The 12GHz e*ay units must use the default divide by 128 as?they're not changing the?eeprom so they're trying to get almost 100MHz through the ribbon cable,?which means they are probably pumping a lot of volts down that line to get those 2n3904 transistors to work at that frequency (ft=250Mhz) and given the main channels are divide by 4 before they get to the fpga which has a max clock speed of around 80Mhz,?means the fpga probably isn't expected to work beyond 60Mhz or so.


Thanks

Greg




Re: 8901B Power level zeroisation

 

On 10/07/2017 08:50 AM, Peter Gottlieb hpnpilot@...
[hp_agilent_equipment] wrote:
There was definitely a wave of these going cheap.? That seems to happen with a
lot of stuff, 99.9% of which I miss when looking for something.
Damn. I missed it too. But I bought my 8901B a very long time ago.

The reason I'm not a big fan of this instrument is because although it can do
many things it doesn't seem great at any of them.? Correct me if I'm wrong.? As
a frequency counter a 5334B totally blows it away, as does a Racal 1992.? In
digits, sensitivity, response speed and of course versatility.? I even have a
5335 which I put one of those cheap cell site rubidium oscillators in because
there's so much space inside.? As a power meter the 437/438 and of course the
E4418B seem far superior and are all available, like the 5334 and 5335 counters,
dirt cheap.? That leaves modulation meter, which is why I picked up the 8901B in
the first place.? Unless you're calibrating some piece of gear, a scope (for AM)
or a 53310A (for FM) are far easier and quicker to use and give you more
information.

So, while I have the nice clean 8901B, it mostly serves to keep the rack it's
mounted in from being too easy to move around.? I'm not even incentivized enough
to track down why a good power sensor doesn't work right with the thing.
The 8901B is a fine instrument; it helps to not think of it as a "do
everything" box, though it does do a lot of random stuff. When I want
to measure RF power, I turn on a power meter. When I want to measure
frequency, I turn on a frequency counter. The fact that the 8901B does
these things is almost incidental; they got at least some of that
functionality "for free" in designing a modulation analyzer.

So, it helps to think of it as just that...a modulation analyzer. For
that, it does very well. But don't be tempted by its "I'll replace
everything on your bench" feature list. Right tool for the job, etc.

That said, though, for my applications, I can't tell you the last time
I modulated a carrier "analog". ;) For that reason, my pristine 8901B
sees almost no use.

-Dave

--
Dave McGuire, AK4HZ
New Kensington, PA


Re: RF Voltmeter?

 

On Fri, Oct 6, 2017 at 11:35 PM, Jeremy Nichols <jn6wfo@...> wrote:
The 411A is the older, vacuum tube version of the 3406, vintage the same as the 400D/H/L AC and the 412 DC voltmeters.
They are really very different. The 411A uses a traditional diode
detector, that is supposedly linearized through the use of a matched
feedback diode. The 3406A uses a fast sampler to convert the RF input
to an IF frequency for measurement, which is a much more accurate
technique.


Re: Affordable VNA Cal Kit in 12~15Ghz

 

On 10/07/2017 08:13 PM, Mel Swanberg wa6jbd@...
[hp_agilent_equipment] wrote:
I'll make a third recommendation for Kirkby Microwave cal kits. I've purchased two of them, and the proof is in the verification tests. When my measurements closely match the supplied verification device, it just doesn't get better than that.

One of my kits even came with custom coefficients for a number of different frequencies that I specified. When compared to what else is out there, you'd be hard pressed to find a better option.
I have two as well, good stuff.

-Dave

--
Dave McGuire, AK4HZ
New Kensington, PA


Re: Affordable VNA Cal Kit in 12~15Ghz

Mel Swanberg
 

I'll make a third recommendation for Kirkby Microwave cal kits. I've purchased two of them, and the proof is in the verification tests. When my measurements closely match the supplied verification device, it just doesn't get better than that.

One of my kits even came with custom coefficients for a number of different frequencies that I specified. When compared to what else is out there, you'd be hard pressed to find a better option.

Mel - WA6JBD


--------------------------------------------

On Sat, 10/7/17, robert8rpi@... [hp_agilent_equipment] <hp_agilent_equipment@...> wrote:

Subject: Re: [hp_agilent_equipment] Affordable VNA Cal Kit in 12~15Ghz
To: hp_agilent_equipment@...
Date: Saturday, October 7, 2017, 7:55 AM

Hi,
It's not a simple job to
measure a cal kit, most cal labs will not
"calibrate" an unknown set of components.
I second the recommendation for Kirby
Microwave. Dave's kits are excellent and a much better
option than some unkown overpriced kit bought on ebay.
Search the group for more information, it's all been
said before.

Robert
G8RPI.


Re: Affordable VNA Cal Kit in 12~15Ghz

 

¿ªÔÆÌåÓý

I'm interested.

I think I can get access to the a cal kit through work.

6GHz would be plenty since I don't have equipment that goes past 3GHz.

Ed

On 10/7/2017 3:52 AM, Tobias Pluess tobias.pluess@... [hp_agilent_equipment] wrote:
???

Hi Calvin


depending on your frequency range of interest, you can make your own cal kit. I made a SMA kit which works fine up to 6GHz, and a N kit which works to approx. 7GHz. However, to calibrate your own cal kit, you need to have access once to a 'real' cal kit, then measure your own standards.
If you are interested, I can tell you more details about this.

Tobias HB9FSX


-------- Original message --------
Date: 10/7/17 05:03 (GMT+01:00)
Subject: [hp_agilent_equipment] Affordable VNA Cal Kit in 12~15Ghz

???

Hi there,


I need to measure and tune some of the RF passive components i.e. filters, DC, ring resonators etc I made for my hobby projects. I would like to get some more precision but I don't have $5000+ laying around for a calibration kit.


Looking at ebay, there are some DC-26.5g HP calkit offered at $2500 range and claiming "fully working". How would I know if it is "working" without sending it to Keysight for a verification?


Is there any Cal Kit in the 12~15 ghz range that doesn't cost an arm and a leg, and provide definition files that I can load to a 8510C VNA?


Have anyone tried to make their own open, short, load, through, then rent a calibrated VNA to characterize them to make their own poorman's cal kit?


Thanks,

Calvin



Virus-free.


Re: Affordable VNA Cal Kit in 12~15Ghz

 

I hope this won't be taken personally. But I think the OP is being done a disservice by being told he needs to buy something that he is clearly skilled enough to make himself.

It was said is that someone like Transcat won't measure S parameters for a set of passive components over a specified frequency range. That may be true, but I find it rather amazing if it is. NIST certainly will, but I expect that they're a lot more expensive. This is vastly less work than an instrument calibration, so it should be cheaper. There are no adjustments to be made.

As for the feasibility, that is not an issue as was noted by others. For bleeding edge work there are references that get sent from lab to lab. The measurements are then reviewed and discussed at length.

I'm quite certain that David's kits are a much better choice than name brand stuff bought on eBay in unknown condition. I would expect he provides a set of calibration sweeps to compensate for the unavoidable errors in the standards. He certainly is no more able to make perfect standards than anyone else. Are David's calibration sweeps as good as Keysight's? Quite unlikely. They can afford to get in house references measured by NIST. Keysight's price for their references reflects the costs and the sales volume.

If the test fixture is designed to probe microstrip devices, then the article I cited is clearly the way to go. I was driving most of the day and doing algebra in my head without paper and pencil is unreliable at best. So I've not worked out the algebra for multiple devices with multiple connectors. An assumption of connector equivalence may be required to account for connector terms. At the moment I don't know.

There is a vast literature on the problem of quantifying instrument and reference errors. All you need is as many measurements as there are errors in the references. And a willingness to let yourself get beat up for awhile by the algebra.

It is certainly true, that in a commercial setting buying a set of references from Keysight is the prudent approach. But for a hobbyist it's silly.

A story from a different realm.

A salesman was calling on a company. While chatting with one of the employees he asked what they made. "Gears of all types." "Really? I thought you had to buy those."

Yeah, I'm a reflection seismologist, not an EE. But 1D (transmission line) problems are the first thing we have to learn. After that we spend most of our time on 2D and 3D issues because the earth is not 1D. It's not EM, but the algebra is the same. I've been quite surprised at how intimidated people are by matrix algebra.

Have Fun!
Reg

-


Re: Need HP transformer IDs

 

Thanks Vladan. Being audio types would make sense, since some of the windings have very large L, that is even out of range of my 4276A. They may also be inductors - one has only two terminals (out of six) that had actually been soldered to. I don't know if simple inductors could be included in this part prefix - maybe they used a transformer winding as a big choke. I thought that these may have been from an old 740A that I junked out years ago, but can't find a manual for that. The 740B manual seems to show use of one of this style package for a dual 500 H choke in the front-end filter, but it's a different part number (9110-0197).

I've been doing some large scale garage cleanout, and rediscovering stuff that I hadn't seen in years, or didn't even know I had. Nowadays I mark most big parts that I keep with what they came from, whatever other info is evident, and the date. Back then I didn't, except for a few things that were very unusual.

Ed


Re: Need HP transformer IDs

 

This is not 100% accurate, but most 9120- transformers were classified as "audio". Maybe it was used in an audio oscillator (vacuum tube).

Vladan


Need HP transformer IDs

 

I have some old HP canned transformers to ID - probably signal type - and I can't recall where they came from. Does anyone recognize these part numbers, and what they may have went to? They are hermetically sealed cans about 1.5" D x 2" H. They have 1965 date codes, so had to be from some older instruments.

9120-0096
9120-0097

Matching them to an instrument application usually tells enough about the application and characteristics. If there happens to be spec sheets somewhere, even better, but I doubt it. My searching so far has found nothing.

Ed


HP AN/USM 281A Oscilliscope needs repair (San Antonio TX)

 

Hello all,


I am completely new to oscilliscopes, and I just acquired one of these for free, since it was not working. Cool thing is that the technical manual is downloadable and the user manual as well, since this is a military unit.


Now the rub: I have never used an oscilliscope, and might not ever use one, however I'm learning electronics repair and my Dad is telling me that this could be an invaluable piece of troubleshooting equipment. Another problem is that you normally need a couple things to troubleshoot an oscilliscope, such as a frequency generator and - another oscilliscope.


I've used the flow chart on the technical manual and it appears that the horizontal signal board needs to be repaired. The CRT projects a beam only when I press the TEST button and only vertical adjustments are possible. Horizontal adjustments while holding the test? button down do nothing.


Anyone in the San Antonio TX area willing to help a newbie out? I work long hours (3am - 8pm mostly) so the time I get off is usually weekends.


Re: Affordable VNA Cal Kit in 12~15Ghz

 

Hi,
It's not a simple job to measure a cal kit, most cal labs will not "calibrate" an unknown set of components.
I second the recommendation for Kirby Microwave. Dave's kits are excellent and a much better option than some unkown overpriced kit bought on ebay. Search the group for more information, it's all been said before.

Robert G8RPI.


Re: RF Voltmeter?

 

¿ªÔÆÌåÓý

Most people I know who repair their boatanchors have relied on the HP 410C.? These are readily seen for sale with the RF probe and these have been converted to solid state.? See

?

Ashley repairs and makes custom boards for the 410C,? I have 2 from her and they are well worth it .? There is also a SO-239 coax interface adapter available where you can use the RF probe to directly measure transmitter output.

?

?

Jerry NY2KW


Re: Affordable VNA Cal Kit in 12~15Ghz

 

Can't you send a DIY cal kit to a calibration lab for measurement? I don't have a VNA yet, but that was what I was planning to do unless I manage to stumble across someone who will do it for me for free.

I was reading papers on VNA cal last night trying to find copies of Engen & Hoer, 1979 and Eul & Schiek, 1988 that didn't cost an arm and a leg from IEEE. Sadly now joy there though I did find some good papers.

Take a look at:

Advances in Radio Science (2003) 1: 21¨C25
Calibration of vector network analyzers on the basis of the LRR-method
Rolfe and Schiek,

which is available via google scholar. They show 4 microstrip calibration standards that should be very amenable to DIY construction.

There's a huge amount of algebra in the paper , but it's all trivial. It's just presentation of all the details of least squares (L2). One can do better using least summed absolute error (L1). In either case, reading the math is far worse than actually doing the computation.

Have Fun!
Reg


Re: 8901B Power level zeroisation

 

There was definitely a wave of these going cheap.? That seems to happen with a lot of stuff, 99.9% of which I miss when looking for something.

Interesting that the 8901B takes 8480 series power sensors.? I tried plugging one in last night.? I got "uncal" and "recal" lights.? The sensor zeroed but cal failed with an error 15 and error 14.? I tried the sensor and cable on a regular power meter and it worked perfectly so I guess there's a problem in my 8901B.? Perhaps that is why I got it so cheap.

The reason I'm not a big fan of this instrument is because although it can do many things it doesn't seem great at any of them.? Correct me if I'm wrong.? As a frequency counter a 5334B totally blows it away, as does a Racal 1992.? In digits, sensitivity, response speed and of course versatility.? I even have a 5335 which I put one of those cheap cell site rubidium oscillators in because there's so much space inside.? As a power meter the 437/438 and of course the E4418B seem far superior and are all available, like the 5334 and 5335 counters, dirt cheap.? That leaves modulation meter, which is why I picked up the 8901B in the first place.? Unless you're calibrating some piece of gear, a scope (for AM) or a 53310A (for FM) are far easier and quicker to use and give you more information.

So, while I have the nice clean 8901B, it mostly serves to keep the rack it's mounted in from being too easy to move around.? I'm not even incentivized enough to track down why a good power sensor doesn't work right with the thing.

Peter

On 10/6/2017 11:40 PM, Dave McGuire Mcguire@... [hp_agilent_equipment] wrote:

On 10/06/2017 09:43 PM, Dennis Czelusniak czelusniakd@...
[hp_agilent_equipment] wrote:
I just had a look at ebay listings and do not believe the current asking
prices. A few years back it seemed like no one could give them away. I found mine at a local machinery recycle warehouse. On the last Thursday of the month, any electrical left on the shelves is drastically slashed and open to offers.

I dunno man, I think you got really lucky. I've never seen them go cheap.

-Dave

--
Dave McGuire, AK4HZ
New Kensington, PA


Re: Slightly off topic puzzle

 

You can open the counter and check the different loops, are they locked or not? You may need
the service manual for this.

G?ran


Re: Affordable VNA Cal Kit in 12~15Ghz

 

¿ªÔÆÌåÓý

Hi Calvin

depending on your frequency range of interest, you can make your own cal kit. I made a SMA kit which works fine up to 6GHz, and a N kit which works to approx. 7GHz. However, to calibrate your own cal kit, you need to have access once to a 'real' cal kit, then measure your own standards.
If you are interested, I can tell you more details about this.

Tobias HB9FSX


-------- Original message --------
From: "guancalvin@... [hp_agilent_equipment]"
Date: 10/7/17 05:03 (GMT+01:00)
To: hp_agilent_equipment@...
Subject: [hp_agilent_equipment] Affordable VNA Cal Kit in 12~15Ghz

?

Hi there,


I need to measure and tune some of the RF passive components i.e. filters, DC, ring resonators etc I made for my hobby projects. I would like to get some more precision but I don't have $5000+ laying around for a calibration kit.


Looking at ebay, there are some DC-26.5g HP calkit offered at $2500 range and claiming "fully working". How would I know if it is "working" without sending it to Keysight for a verification?


Is there any Cal Kit in the 12~15 ghz range that doesn't cost an arm and a leg, and provide definition files that I can load to a 8510C VNA?


Have anyone tried to make their own open, short, load, through, then rent a calibrated VNA to characterize them to make their own poorman's cal kit?


Thanks,

Calvin