¿ªÔÆÌåÓý

Date

HP 6632B PSU showing CV and CC at the same time

 

Hi!
?
I recently repaired a 663120A where the crow-bar just kicked in and did it's job. The trouble was to find the right soldered in mini-fuse that was half way hidden under the interface board.
?
So I hoped for an easy fix on my 6632B too but no. I tried to follow the failure diagrams in the service manual and just got the info to replace the interface board. But the signals on the interface to that board do show a lot of good and only a few out of range values. Especially the signals that should make the display show CV and CC are active, so the interface board is right.
?
The fault itself is, that I can enable the output of the unit as long as I program 0V. As soon as I try to program more then about 0.7V the system shows CV + CC at the same time and no output voltage.
?
The measured voltage on the VFD is close to my multimeter, so the ADC seems to work.
?
One thing I recognized is, that the input bus voltages are all high. But I double-checked that the transformer ist configured for 230V. But the (+) rail has 45V compared to the service manual saying it should have 39V. The (-) 9V rail has -12V. Is there anything known about a faulty series of transformers or a bug in the wiring diagram?
?
73 de Ulrich
?


Re: Tantalum Capacitors

 

I think Adrian is doing a good job in going slowly through the analysis. Sometimes you get the wrong way and that happens to all of us. Making interesting episodes is one thing, but having to repair something in time as the customer is waiting outside is another one.
?
Like Usagi and others channels, as an old guy sitting in front of the screen, there are these moments where you scream at them, telling "There is the fault right in front of you!" but the guy on the screen is going the whole wrong way. But that is part of the fun.
?
Back to the capacitors I have successfully replaced Tantalum types by Aluminum Polymer. These combine the benefit of being dry and having a low ESR. Even dry electrolyte sounds great and you want to replace all caps by this type, there are problems. You can replace Tantalum bei Al-Poly as the PSU is prepared for low ESR caps or there are small resistors in front of them (sometimes there are coils in front that filter but also add enough serial resistance)?
?
Replacing Al-Electrolyte with Al-Poly will result in a big punch on the PSU on power-on but not as hard as if you replace low value AL-Electrolyte by MLCC ceramics. So the PSU needs to be prepared for that. And there is no way to replace Al-Electrolyte with AL-Poly in signal related places except you adapt surrounding inductance or resistance values.
?
In any smaller test equipment like power-meters and such, especially when the supply is linear regulated, Al-Poly is good. But if you replace all more than 300 Al-Electrolytes by Al-Poly in a Stabilock 4031, you might trigger overload on power on of the PSU.?
But again, if you replace the 10uF caps on the 4031 CPU and secondary CPU board by 1uF Al-Poly or MLCC, that should work absolutely fine, as the series resistance is much lower and the energy required is transferred much faster. But the punch on the PSU is shorter and lower because of the reduced capacity. The Al-Poly will not go up in flames like to original ones and not peeing on the board like wet caps and eating the traces.
?
And then again, replacing all SMD electrolytes in the small Stabilock 4015 by Al-Poly will cost over 170€ and may introduce different problems. Using Automotive or Aerospace certified Al-Electrolytes from the best brands you can get, will cost less then 80€ and make the unit work again for the next 40 years. And there is still hope, that modern Al-Electrolytes do not leak and eat your traces from the PCB, as the capacitors from the 2000'ds do when the capacitor plague was a thing.?
?
73 de Ulrich
DC3AX / AG1U / OE3UPR
?


Re: Shielded probe cables?

 

We are talking about something like the
HP/Agilent 54620-61601 Logic Analyzer Probe Cable, right?


Re: Refurbished E4407B 26.5GHz spectrum analyzer for sale with nice options price 2500US/2300Euro

 

¿ªÔÆÌåÓý

New price unit is still availalable (Black friday sale Loool)

?Hello here are your ultimate christmas gift.
?
I have for sale a nice Agilent E4407B spectrum analyzer for sale. It has been repaired and works perfectly.

Range 9KHz to 26.5GHz?

It has a modern design and utilize same menu structure as the expensive PSA series just more compact.
Fast measurements.
It measures up to specs all up to 26.5Ghz.

Nice options installed
1DR : Narrow band resolution bandwidth
AYZ : External mixer option interface hardware
1DS: 3GHz Preamplifier option?
B72: Memory expansion
1D5: High stability Frequency reference

The Preamp option makes the dynamic range down to -165dBm@1hz bw in the range 9khz to 3Ghz.?
Above 3GHz yig filter makes the spectrum clean.?

Can be shipped for 50Euro in europe or 75US for overseas.?


Unit is used and is in good condition with some marks here and there .
Look closely on the pics for actual condition


ask any questions

Price 2500us/2300euro plus shipping

73 Peter OZ1LPR


Re: HP141T with multiple sections

 

Many thanks David and Gary, so I guess best not take any risk, I'm going to power up my HP141T with one full set of section: 8552B and 8553B plus the great active probe 1121A which I've forgot to mention. Let's hope its stoarge for more than 2 years will not affect any PSU, electrolytic power cap and so.
?
In the mean time, I was looking in our forum for HP141T key words then found this very interesting old comment on LCD retrofit kit
?
Looked briefly to what was done on the ? it is suddenly mind blowing how the 70's HP141T has been colorized, other display possibilities, this is awesome.
?
Suppose I remove the CRT unit and electronics from my HP141T to replace with a LCD kit, do you know how much weight in kg would be saved and power consumption gained ?
?
I saw on the Rohde&Schwarz forum someone who used the same LCD-FPGA demo board kit to replace a CRT unit, not sure where I saw this but it is the same situation with TD500/700 oscilloscope and Advantest R3261/R3361/R365/R3365/R3271/R3371, the cost or impossible to find old CRT => the challenge to LCD retrofit which permits to colorize, add new function if proper hacking of the legacy board.
?
?


Re: HP 8341B 1992 firmware version

 

OK, now I understand, so anyway the cheapest option would be probably buy dedicated display for 8530A with already loaded software for 8530A.


Re: HP 8341B 1992 firmware version

 

08530-80004 is part you need


?


Re: HP 8341B 1992 firmware version

 

Key is just programmable DIP IC (PAL20L8B-2CNS), inserted in empty socket on small board behind floppy.
But there is no file for programmer


Re: HP 8341B 1992 firmware version

 

Hi Dimitrij,
Oh, I see, so it won't be easy to convert "normal" 8510C display to 8530A display because display for 8530A has additional board with PAL circuit, am I right? So probably only solution is to buy dual identity display (OPT 011) or display specified only for 8530A with already loaded software, do I understand correctly?
Best regards, Wojciech
?


Re: HP 8341B 1992 firmware version

 

Read this topic:
/g/HP-Agilent-Keysight-equipment/topic/101428782


Re: HP 8341B 1992 firmware version

 

8530A firmware already there:
/g/HP-Agilent-Keysight-equipment/files/HP%20&%20Agilent%20Firmware%20&%20Programs/HP%208530%20firm%20A0166%20.exe

But you need appropriate PAL key 08530-80004 to run it


Re: HP 8341B 1992 firmware version

 

Hi Ed,
I have never done this before but of course I can try to save it. Is 1.44MB floppy disk OK for that?
BTW I'm also looking for software for 8530A microwave receiver (for antenna measurements) which I could load to normal display from 8510C. Do you maybe know if I could find it somewhere?
Best regards, Wojciech


Re: HP141T with multiple sections

 

As I recall, the manual warns that the power supplies may not regulate without plug-ins, but I'm relying on my memory here, and it ain't what it used to be.

Gary Appel

On 12/3/2024 1:22 PM, Don Bitters via groups.io wrote:
You can turn on the 141 display by itself, but will only have a dot display which you can move with the horiz/vert controls. I would not leave it on for any length of time with no plug-ins installed. The plug-ins can be verified to be operable by the set up and operation verifications in the manuals.
In the right hands this is still a usable system. You could sell it (guesstimate 200-600€ for all - if working). It is still pretty repairable, unlike some of the modern test equipment. Local pickup is very much recommended, the total system is not lightweight.
You could donate it to a tech school. If donated to a tech school there are many good articles and app notes still available on usage, theory, and applications.

Don Bitters




Re: Fan replacement 8341B

 

Sunon A1175HBL-T


Re: HP141T with multiple sections

 

You can turn on the 141 display by itself, but will only have a dot display which you can move with the horiz/vert controls. I would not leave it on for any length of time with no plug-ins installed. The plug-ins can be verified to be operable by the set up and operation verifications in the manuals.
In the right hands this is still a usable system. You could sell it (guesstimate 200-600€ for all - if working). It is still pretty repairable, unlike some of the modern test equipment. Local pickup is very much recommended, the total system is not lightweight.
You could donate it to a tech school. If donated to a tech school there are many good articles and app notes still available on usage, theory, and applications.

Don Bitters


Re: Fan replacement 8341B

 

¿ªÔÆÌåÓý

Yes, it is quiet, hard to tell it is on.? I should have noted, I used a fan that has slightly less air flow than the original.? This instrument will be used on the bench, at room temperature, not cabinet mounted.

Spinning the old fan by hand it is easy to hear the worn bearings.

After being on for a while, nothing warm was noted.

One more thing, after removing the covers, and moving it around, the RF gaskets fell? out of the rails in the side covers. Keep track of them, easy to lay down and re-install covers.

?

Gary

?

?

?

?

From: [email protected] <[email protected]> On Behalf Of saevar via groups.io
Sent: Tuesday, December 3, 2024 4:11 PM
To: [email protected]
Subject: Re: [HP-Agilent-Keysight-equipment] Fan replacement 8341B

?

Good job,

Is it any quieter than the original?

?


Re: Fan replacement 8341B

 

Good job,
Is it any quieter than the original?
?


Re: HP141T with multiple sections

 

This depends on the age of the mainframe, late ones have a interlock connection & relay in the power supply, to prevent operation with no plug-ins fitted.
But for testing something in unknown condition, I prefer to make sure the power supply rails are correct before fitting plug-ins, this also applies to the 18x series mainframes.
?
David
?
On Tue, Dec 3, 2024 at 09:42 AM, tantratron wrote:

My first question: since the HP141T was not run for more than 2 years, is it possible to run the HP141T alone with no section added.


Re: Looking for Noise Source Driver Assembly E4401-60123 for E4407B Spectrum Analyzer

 

I'm afraid I've never looked inside any of the NFA units. They do look like an ESA on the outside but I think the internal signal path will have quite a few changes.
?
The noise figure of the NFA will be about 4dB up to 500 MHz and the ESA will be more like 7dB. So there will be differences in the front end design.
I'd also expect the NFA to have a dedicated noise detector that will have very good linearity. I think that it would also have to have accurate attenuators or at least some way to calibrate the attenuators for accuracy.
?
I tried to find a service manual for the N8973A but had no luck.
?
?


Re: 8350A + 83525A mechanical issues

 

I guess I should say IF there's a battery problem. The message flow is a little confusing to me - not sure about any battery issues.?? Ed