Re: HP 141T Power Supply Help Needed
While I wait for a new socket, I went ahead and soldered the transistor in the relevant socket.? ? The other sockets look ok.? ? Like I say, I now have all 4 supplies running and adjusted.? ? ?Starting to muse over the A5 board presently.? ?I guess flood gun damage is possible,? but I'll check out the pulse and blank portions first.? ? ?Will report back as that progresses.? ? On Sat, Sep 14, 2024 at 11:16?AM Michael Bafaro via <m.bafaro=[email protected]> wrote:
|
Re: HP 141T Power Supply Help Needed
¿ªÔÆÌåÓýBill, ? My friend Pete brought over his 141T with plugins and it was intermittent and I found that the pass devices were not always making contact.? The sockets have been thermal cycled enough that the brass contacts have been annealed.? I had it all working and all of a sudden the -100 was missing.? I tapped the pass device and it started working again!? I removed the collector screws and if it were not for the sill pad the transistor would have fallen out!!? It had ZERO insertion force.? I ordered some new sockets Keystone Electronics P/N? 4608 and installed them and now it is solid.? The sockets are riveted in and that was a real pain as well as having to unsolder all of the wires and working in that small space to get the screws out and back in.? UGH. I installed the new sockets with 4-40 X ? screws lock washers and nuts and incase the phenolic shrinks with age I put some Locktite on the threads. ? Mike ? From: [email protected] <[email protected]> On Behalf Of Bill Berzinskas
Sent: Saturday, September 14, 2024 9:37 AM To: [email protected] Subject: Re: [HP-Agilent-Keysight-equipment] HP 141T Power Supply Help Needed ? All 4 voltages look good now.?? I popped the plugins into the case.?? 3 lights - scanning, log ref 0 and an X.??? No life on the crt however.?? Going to take a break and Crack open the manual some more.????? Not super pumped that it just started working, especially without a root cause.? But I'll take it for today man... ? On Sat, Sep 14, 2024, 10:23?AM Bill Berzinskas via <bberzinskas=[email protected]> wrote:
|
Re: HP 141T Power Supply Help Needed
¿ªÔÆÌåÓýBill, ? Wow!? Bad diodes I would not have expected that.? Now you have to check the pass device and the rest? of the regulator. ? Mike ? From: [email protected] <[email protected]> On Behalf Of Bill Berzinskas
Sent: Saturday, September 14, 2024 12:19 AM To: [email protected] Subject: Re: [HP-Agilent-Keysight-equipment] HP 141T Power Supply Help Needed ? For S&Gs, I jumpered in a set of SB520 and that brought the -12.6v output up to ~23v!? ? I went ahead and ordered 1N3209s.? ? On Sat, Sep 14, 2024 at 12:38?AM Bill Berzinskas via <bberzinskas=[email protected]> wrote:
|
Re: HP 141T Power Supply Help Needed
¿ªÔÆÌåÓýBill, ? See that there is continuity from the ?negative side of C6 to the anodes of the diodes.?? If that is good use the diode scale on the DMM and see that? you have about .7V forward on each diode. ?You may have to disconnect the common cathode connection which is on the outside of the unit. ?These are stud mounted diodes and should be almost indestructible. ? Mike ? From: [email protected] <[email protected]> On Behalf Of Bill Berzinskas
Sent: Friday, September 13, 2024 11:39 PM To: [email protected] Subject: Re: [HP-Agilent-Keysight-equipment] HP 141T Power Supply Help Needed ? Took both side panels off, dug the transistor holder out a bit more and proceeded to trace the wiring.? ?It all checks out, but I now have strong reservations about the status of CR1 and CR2.? ?I believe I am seeing?
? ? |
Re: HP 141T Power Supply Help Needed
All 4 voltages look good now.?? I popped the plugins into the case.?? 3 lights - scanning, log ref 0 and an X.??? No life on the crt however.?? Going to take a break and Crack open the manual some more.????? Not super pumped that it just started working, especially without a root cause.? But I'll take it for today man... On Sat, Sep 14, 2024, 10:23?AM Bill Berzinskas via <bberzinskas=[email protected]> wrote:
|
Re: HP 141T Power Supply Help Needed
Try this again....?? I've been marking this schematic for to long to keep it straight... I am looking at it though Q9 Q10 Q11 Across cr1 and cr2 is 23v.? I measure 98.7 and -99v on other supplies currently.? Now this is absolutely wild....? doing nothing other than measuring around, I now have a very adjustable negative supply.???? Started on -15 and adjusted to -12.6 with no problem.??? Uggh.? I don't understand what just happened.?? On Sat, Sep 14, 2024, 9:50?AM Frank Mashockie via <fmashockie=[email protected]> wrote:
|
Re: N9010A 32-bit XSA Software for Windows 7
Hi Tony,
?
It's been quite a while since I did this, so I don't recall if the upgrade to Win7 required a 64-bit capable CPU, but ISTR that it did. I do remember having to buy a new processor board for my N9020A for the OS upgrade.
?
If I am wrong about the 64-bit requirement, I have several older Win7 versions of the XSA software and I'd be happy to Dropbox one of them to you. What version were you running under XP?
?
BTW, I looked through several different release notes, but in none of them could I find any indications that the Win7 OS was required to be 64-bit. That may just have been assumed. I think the determination of compatibility of the OS and the XSA software was left up to the installer software.
?
Regards,
Stan |
Re: HP 141T Power Supply Help Needed
Also, as a sanity check, when is the last time you checked the other rails?? Is your -100V rail still in spec?? Without a temp compensated zener, these 140/180 power supplies can get unstable.? I've seen them intermittently drop all the rails due to similar issues.
?
-Frank |
Re: HP 141T Power Supply Help Needed
Bill,
?
Your measurements aren't making sense.? Also, this is being compounded by not providing enough info as to what point you are referencing when you take your measurements.? For example, Mike asked for voltage measurements across anodes and cathodes of CR1 and CR2.??
?
Not sure how you have 1.5V at the emitter and 1.3V at base of Q9.? There should be 0.6V drop from base to emitter there.? Unless you've transcribed your measurements wrong, popped in a transistor backwards, etc.? Also, you state you have 2.2V at Cs of Q10 and Q11.? Well those are tied to the base of Q9 - which you reported as = 1.3V.? They should all be at the same potential.
?
Are you referencing the schematic?? The service manual is widely available online.? I think if you were, you'd be seeing some of these discrepencies.??
?
-Frank |
Re: HP 141T Power Supply Help Needed
Diode to chassis is how I was measuring it.??? I do see expected voltage at the cathode side of my makeshift rectifier.? 17.5v. The 24v last night was the fuse..?? here's the most recent checks. Q9 Q10 Q11 On Sat, Sep 14, 2024, 5:15?AM Dave_G0WBX via <g8kbvdave=[email protected]> wrote:
|
HP attenuator O-rings
The usual problem - degraded O rings!
They seem around 1mm ID 0.5mm CS (so 2mm OD) or maybe .039" ID, .024" Cross-section, which can readily be found in Butyl-rubber as watch crown gaskets, but there are various comments that they need to be Viton/FKM in order to last, but so far I've not found a supplier willing to sell me a sensible number (in UK). Does anyone know of one or will watch crown gaskets do? Alan |
Re: HP 141T Power Supply Help Needed
¿ªÔÆÌåÓýBill.For the sake of clarity, what exact points in the circuit are you measuring between? Across each diode (anode to cathode) or between one side of each diode vs Chassis Ground? If the latter, you have a big problem.? The diode Cathodes, "should" be some 6 (or 9) Volts Positive wrt chassis. What is shown during a Diode Test on a DMM, depends on the DMM, and what else is still connected to the diode in question.? Such tests usually are only valid on a diode "out of circuit" (or with one end disconnected at least.)? Even then, a faulty diode (that breaks down / leaks above some voltage well under it's spec') will often seem OK on a DMM. Regards. Dave B. From: Bill Berzinskas Date: Sat, 14 Sep 2024 04:38:53 GMT Took both side panels off, dug the transistor holder out a bit
more and proceeded to trace the wiring.? ?It all checks out, but I
now have strong reservations about the status of CR1 and CR2.? ?I
believe I am seeing?
Anode - 24v dc / 17v ac Cathode - 0.5v DC Assuming it should register something on my DMM, since its a diode and all..? ?Nothing.? ?? I take it these are the source of all my woes?? ? Not sure what the heck I was measuring before..? -- Created on and sent from a Unix like PC running and using open source software: |
Re: HP 141T Power Supply Help Needed
Re:-
"That discrepancy between A2Q9.C and Q4.C is certainly a typo, or perhaps the two points were measured at different times and the mains voltage changed, and no-one noticed that they're by definition the same." Under different load conditions perhaps?? (Plug-ins present, or not.) I see there is also mention of 2V ripple on the +ve unregulated side (Q4.Q9 collector.)? Presumably when under some unspecified load. (I don't have a copy of the full manual.) Regards. Dave B. -- Created on and sent from a Unix like PC running and using open source software: |
N9010A 32-bit XSA Software for Windows 7
Hey All,
?
I'm just wondering if anyone may have a copy saved somewhere of this software.
?
I'm trying to bring an old N9010A with a 32-bit x86 processor onto a later OS than Windows XP (Windows 7) but it seems that Keysight either never made a Win7 x86 version of the software or has let it fall off their website and the internet.
?
The N9010A will run Windows 7 but the Installers for the older software are gated to check the Windows Version so I can't just re-run the WinXP installer and when you move from Windows XP to Windows Vista (as part of the OS upgrade process) it breaks the app install (it expects to run the installer again which then errors out).
?
The right way to address this is to take the N9010A from WinXP to Win7 and then reinstall the app but all the versions hosted on the Keysight website expect that you have an x64 process board in the N9010A.
?
Any help is appreciated.
?
Thanks,
?
TonyG |
Re: HP 141T Power Supply Help Needed
For S&Gs, I jumpered in a set of SB520 and that brought the -12.6v output up to ~23v!? ? I went ahead and ordered 1N3209s.? -23v is a bit concerning, but I will check the pass trans one more time and remeasure everything in the morning.? ? On Sat, Sep 14, 2024 at 12:38?AM Bill Berzinskas via <bberzinskas=[email protected]> wrote:
|
Re: HP 141T Power Supply Help Needed
Took both side panels off, dug the transistor holder out a bit more and proceeded to trace the wiring.? ?It all checks out, but I now have strong reservations about the status of CR1 and CR2.? ?I believe I am seeing?
Anode - 24v dc / 17v ac Cathode - 0.5v DC Assuming it should register something on my DMM, since its a diode and all..? ?Nothing.? ?? I take it these are the source of all my woes?? ? Not sure what the heck I was measuring before..?? ?
? |
Re: HP 5086-7906 YTO
¿ªÔÆÌåÓý? Hi Aldri, ? ? For your second converter, when you do debugging it's easy to make a mistake, there is so much manipulation. You have to keep the positive side, this part is set and you have a second spare converter. ? For the YTF, it's not easy, centering the spheres is not enough, don't forget that it's a bandpass filter. So you have to align the filter for a certain reference frequency, but which one, we could assume that it's in the center, but not necessarily, it could be at the beginning for example (lowest frequency). The filter will deform according to the frequency, it's about finding a happy medium. The YTF calibration will find the coefficients of a polynomial trying to have the best for all frequencies. The level differences will be compensated during the frequency flatness calibration. However, don't forget that if you have too much loss to compensate, the DANL test could become out of spec. ? The fact that you use a precise external power supply is not a problem for the alignment of your YTF. The important parameter of this power supply is its stability. ? I had sent you a picture of my 8595E, here it is again, it represents a cumulative test of the frequencies: yellow curve -> test at 3GHz cyan curve -> test at 5GHz magenta curve -> test at 6.5GHz We can see that on the left, there could have been an improvement by re-adjusting one of the spheres, but nothing is certain. The insertion loss varied from 2 to 3 dB between the minimum and the maximum of these 3 frequencies. You have similar results, I think your YTF has good results...nothing is perfect :) ? You say that you have problems in the high band, what are the frequencies in play. from 2.9GHz to 6.5GHz? (highband). For the second converter, the path loss for the highband is -1.8dB. After this point two bands go through the same path, so if the YTF is correct, either there is a problem with the YTF control source, the first converter or the second converter (highband section). You should be able to easily find where the excess loss is. ? You are at the end of your troubles, don't give up, you will find it. ? ? --------------------? Here is the verification of this second converter after reassembly. The flatness is good, only 0.03dB for 5MHz width at output IF 321.4MHz (lowBand). Multiplier x6 for 3600 MHz, adjusted for best match (lower loss at 321.4MHz), very sensitive adjustments. Path for 321.4MHz highBand was verified too. ? One sweep with 50 MHz span, another with 2950MHz span¡to see the stop band. The loss for IF highband include loss of filter and RF switch, around -1.8dB. ? Yves De?: [email protected] <[email protected]> De la part de alfa beta ? Yves, thanks for the info about the 2nd converter? After having installed the one bought on the bay?I checked?the original one?on the bench and it?seems to work normally ? I'm starting to suspect the blindness on the LOW BAND was due to a bad connection I did when the repair was suspended years ago ¡ ouch ! ? Meanwhile I'm investigating?a certain deafness on the HIGH BAND (from 2.4 to 11 dB) ? I measured the insertion loss of the SYTF I had realigned and found figures?from 5.2??to? 8.5 dB on the range? 3 to 6 GHz ? These figures?seem not too bad?and?the AMP CAL routine should be able to recover them ...unless there is some hardware problem in the generation of the tuning current (I performed the insertion loss measurements by manually tuning? with an external precision power supply) Will see ... Adri ? ? ? ?
![]()
Second Converter HiBand 321.4 MHz filter.png
![]()
Second Converter HiBand 321.4 MHz.png
![]()
Lowband Second Converter IF out 321.4MHz.png
![]()
Test YTF.png
![]()
Test YTF_.png
|
Re: HP 5086-7906 YTO
Hi Razvan,
toggle quoted message
Show quoted text
thanks also for the service note. I have an E4407B, I had not noticed that it was the same second converter. Yves -----Message d'origine-----
De?: [email protected] <[email protected]> De la part de Razvan Popescu via groups.io ·¡²Ô±¹´Ç²â¨¦?: 13 septembre 2024 03:49 ??: [email protected] Objet?: Re: [HP-Agilent-Keysight-equipment] HP 5086-7906 YTO Hello Yves, This second converter assembly is also used in the 8563E and also ESA E4407B spectrum analyzers. In fact I think it is common for all 6.5/13.5/26.5GHz versions. Mine had failed in the E4407B and it looks like it is related to an incorrectly manufactured transformer inside the assembly. I attached the Agilent Service Note related to this issue. Part number affected is 5086-7958. Thanks for the nice PDF with all explanations. Regards, Razvan On 13/09/2024 07:51, Yves Tardif via groups.io wrote: Hi Adri, |
Re: HP 141T Power Supply Help Needed
¿ªÔÆÌåÓý
Comments are based on a scan of 00141-90914, 141T mainframe S/P 1615A, printed June 1976, file downloaded from BAMA.
The positive side of C6 varies with mains voltage.? If the regulator is working, it's whatever it takes to make the negative end be -12.6V .
There are complications when you try to compare winding AC voltage with DC output.? Capacitance, ESR, winding resistance, diode characteristics, load.? You can calculate a rough approximation only.
That discrepancy between A2Q9.C and Q4.C is certainly a typo, or perhaps the two points were measured at different times and the mains voltage changed, and no-one noticed that they're by definition the same.
Dave Wise
From: [email protected] <[email protected]> on behalf of Michael Bafaro via groups.io <m.bafaro@...>
Sent: Friday, September 13, 2024 3:13 PM To: [email protected] <[email protected]> Subject: Re: [HP-Agilent-Keysight-equipment] HP 141T Power Supply Help Needed ?
Sounds good to me. ? Mike ? From: [email protected] <[email protected]>
On Behalf Of Frank Mashockie ? Dave, ? The schematic for the HP 141T shows all voltages referenced to ground.? If that is the case, you know the voltage at the positive side of C6.? It is listed as 5.9VDC.? So you can infer that the voltage across C6 is 18.6V.? From there, you can find what the AC voltage across the winding should be, correct??? ? Also, in the schematic they show collector of Q9 at 6.61VDC and collector of Q4 at 5.9VDC.? But apparently Q4 and Q9 are tied together at the collector by wire 968.? What is the reason for the 0.6V drop?? Or is this an error??? ? -Frank |