Re: HP 83590A - Help understanding frequency band switching
Probably a little late in the game for this but, I would have suggested walking through the adjustments without adjusting anything first to identify where your problems are, then use that in the troubleshooting.
Can you adjust the baseband (1) properly, low end 2.0GHz (offset), high end 7.7GHz (gain)?, if yes, then you have said that band (2) low end 7.0GHz adjusts correctly (band 2 offset), but 13.5GHz (band 2 gain) runs out of adjustment range, high or low? Here you want to check the tuning voltage on all bands to make sure that you are getting the correct tuning voltages, then you check for the correct frequency outputs vs. the tuning voltages.? You do need a pretty good DMM (ie HP 3456A, 34401A, or equivalent) because IIRC, the YTO, YTM tuning is 1V/GHz ( 1mV/MHz), and if IIRC, the adjust tolerance is 5mV.? Each band adjustment depends on the the lower band adjust being correct. Other things to check for are the bands being switched correctly (here is where you had your band switch question), I think there is a band switch table that gives a truth table (1's and 0's; on and off), for all of the band nos. Typically they will be 5Vdc, 0Vdc for TTL, and 12-15Vdc, 0Vdc for CMOS devices. This will be in either the theory or the troubleshooting. If for instance band 1 and 2 are on at the same time it will give you your symptom, but the adjust ranges will be off at both high and low ends, probably less effect at the low end and much worse on the high end (I have seen this problem once or twice). Consider this, all of the adjustments in an ideal situation, would be at or near the center of their adjustment range. Also each adjustment should be paired with a fixed resistor to set the offset and gain for each band, the adjustment is basically a fine tune for the fixed resistor. Don Bitters? |
Re: Conductive Paint for Microcircuit Repair...
I dont know composition of your conductive "ink" but if only problem is fast solvent evaporation, you can try to make solvent (acetone or MEK) saturated suronding atmosphere... Put piece of cotton toped with acetone on the top of working areas with hole in the middle to access your area of interest.? This will slow down ink solvent evaporation. On Tue, 16 Jul 2024, 19:59 Peter Gottlieb via , <hpnpilot=[email protected]> wrote: Solder paste looks like boulders at this magnification.? There must be a finer |
Re: Conductive Paint for Microcircuit Repair...
Solder paste looks like boulders at this magnification.? There must be a finer grade than I have.
toggle quoted message
Show quoted text
On 7/16/2024 1:48 PM, SCMenasian via groups.io wrote:
This is a long shot - and a project in itself. |
Re: Conductive Paint for Microcircuit Repair...
This is a long shot - and a project in itself.
If you can get some really fine wire (for example, a single strand from a Litz wire, with the insulation chemically removed), coat it with solder paste and lay it in the region you want to repair and heat it until the paste melts, you might be able to make the repair. By my estimate #46 wire should be about right. A quick search turned up magnet wire all the way to #56!!! |
Re: Conductive Paint for Microcircuit Repair...
I have a bunch of highly conductive 2 part silver epoxy but it is past expiration date and a bit hard to mix.? It does solidify very well after mixing though.? It might work?? I have a whole bunch of the small sets of part A and B if you are inclined to want to play with some.
toggle quoted message
Show quoted text
On 7/16/2024 1:14 PM, Chuck Harris via groups.io wrote:
Hi Peter, |
Re: Conductive Paint for Microcircuit Repair...
Years ago when I used to paint cars with laquer, thinner was available
toggle quoted message
Show quoted text
in several grades, depending on application temperature. You wanted the solvent to flash off in a reasonable time, not too fast and not too slow. The grades were generally fast, medium and slow. Well, you can still get slow laquer thinner, although I don't think you'll like the price. However, looking up the mds shows it a near equal combination of acetone, methanol and toluene with the remainder mostly naptha and a teeny bit of heptane. You might try thinning that with toluene and/or naptha and see how it behaves. Paul On Tue, Jul 16, 2024 at 12:24:56PM -0400, Chuck Harris wrote:
A form of silver print is what I used before, it came in --
Paul Amaranth, GCIH | Manchester MI, USA Aurora Group of Michigan, LLC | Security, Systems & Software paul@... | Unix/Linux - We don't do windows |
Re: Conductive Paint for Microcircuit Repair...
Hi Peter,
The first time I tried, I used it at factory dilution. However, over the 10 years since I did it last, the syringes hardened... So, I reliquified them with acetone... The acetone did a nice job of dissolving the paint, but now it dries even faster. So, I let (Hah! I let!) the acetone evaporate a bit, and then replaced it with MEK, which is like acetone with a slower evaporation rate, but, it didn't help enough. I even thinned the stuff with MEK to 3x the volume, but the MEK still lets the paint skin over too quickly. I could buy new, but I would be back where I started. I was hoping that there was a better form of the silver print. There is a one part epoxy available, that is heat cured, but it requires 2-300C for 15 seconds, which I think will melt the kapton substrate. -Chuck Harris On Tue, 16 Jul 2024 12:44:08 -0400 "Peter Gottlieb" <hpnpilot@...> wrote: I have the same kind of repair to do and have not found an acceptable |
Re: 8702A service manual
Hello Wayne, I have an 8702A which?IIRC is an 8753C with firmware to support the optical function by using the optical S parameter test set (which I do not have). Apparently one could mix optical and electrical measurements with the proper test set but I just select?electrical as I only have the S Parameter test set for electrical.? AFAIK the base electronics hardware on the 8702A and the 8753 are identical only the firmware is different so if you can find a manual for the 8753 you should be good to go. As far as the external display goes I don't know but I believe some members on the EEVblog forum site have converted the 8753's to an LCD display so that might be an option for you. Sam W3OHM Sam Reaves
ARS W3OHM
Owner / Moderator of: LeCroy Owners Group on Groups.io Sencore Owners Group on Groups.io Sprint Layout Group on Groups.io?? Pulsonix EDA Group on Groups.io LPKF Owners Group on Groups.io
Electronics and Mechanical Hardware Design Engineering Manager Staff Scientist Andritz Rolls Global Research Center (RETIRED) |
Re: Conductive Paint for Microcircuit Repair...
I have the same kind of repair to do and have not found an acceptable solution.
toggle quoted message
Show quoted text
Would your paint work if a little extra solvent was added and mixed in? Peter On 7/16/2024 12:24 PM, Chuck Harris via groups.io wrote:
A form of silver print is what I used before, it came in |
Re: Conductive Paint for Microcircuit Repair...
A form of silver print is what I used before, it came in
little 1ml syringes. The problem is the solvent evaporates quickly enough that between dipping the one hair brush into the paint, and touching the damaged spot, the paint has skinned over, and won't transfer. The last time I did it, I put the paint right next to the damage, and dipped, moved 0.1", and painted. Even with that small distance, the paint laid down like a mixture of straw and thick mud, as the crust that formed on the wet paint on the brush laid down in sheets and strings on the repair. Got to be a better way! -Chuck Harris On Tue, 16 Jul 2024 10:33:52 -0400 "Michael A. Terrell" <terrell.michael.a@...> wrote: I used to use General Cement Silver Print to repair conductive |
Re: Conductive Paint for Microcircuit Repair...
I used to use General Cement Silver Print? to repair conductive traces. I believe it was a lacquer with very fine silver particles in suspension. I don't know if it is still available, because my? satellite internet can take an hour and a dozen tries to open most web pages. during the day. It is to be replaced by a fiber optic service, but that hasn't? happened yet. On Tue, Jul 16, 2024 at 8:55?AM Chuck Harris via <cfharris=[email protected]> wrote: Hi Gang, |
Re: Conductive Paint for Microcircuit Repair...
Ok. I saw the arm-hair statement, but not the fact that that was the width of the trace.
toggle quoted message
Show quoted text
Thanks. DaveD KC0WJN ============================== All spelling mistakes are the responsibilty of the reader (Rick Renz, STK, ca. 1994) ============================== On Jul 16, 2024, at 09:50, Chuck Harris via groups.io <cfharris@...> wrote: |
Re: Conductive Paint for Microcircuit Repair...
As I stated in my original post, about 0.04mm, or one of
my arm hairs. -Chuck Harris On Tue, 16 Jul 2024 09:47:12 -0400 "Dave Daniel" <kc0wjn@...> wrote: What is the width of the flex cable trace under discussion? |
Re: Conductive Paint for Microcircuit Repair...
What is the width of the flex cable trace under discussion?
toggle quoted message
Show quoted text
DaveD KC0WJN ============================== All spelling mistakes are the responsibilty of the reader (Rick Renz, STK, ca. 1994) ============================== On Jul 16, 2024, at 09:45, Chuck Harris via groups.io <cfharris@...> wrote: |
Re: Conductive Paint for Microcircuit Repair...
Hmmm?
"Masking" tape is generally made of crepe paper that has crenulations that are bigger than my entire trace width. I suppose some kapton tape pieces could be arranged to make some sort of mask... Thanks for the idea... I will think on it. -Chuck Harris On Tue, 16 Jul 2024 08:07:06 -0500 "Dave Casey" <polara413@...> wrote:
|
HP 8557A Start-Up Calibration & Linear Mode - Errors in Manual?
Hey y'all!? I was hoping to get some help or clarification on the start-up calibration procedure for the HP 8557A/8558B spectrum analyzer plug-in.? First, I did not get this unit in working order and I have completed a lot of restoration on it. Also, I've never used/owned a spectrum analyzer before this unit, which likely contributes to the trouble I'm having.?
?
The problem I'm having is with the linear/10 dB/div calibration.? In this step, the manual states that you should be able to adjust both the Linear and 10dB/div by adjusting the 'Reference Level Fine' knob to get the 250MHz calibration signal peak all the way to the top graticule line.? But I can't do that without adjusting the 'Coarse reference level' knob (the manual does allude to using the coarse reference level knob for this adjustment in the troubleshooting section of the manual).? The manual states if this adjustment cannot be done, then check the linear/log amplifier board A11, but I can't find any issues there.
?
My question is, can anyone who owns one of these make sense of the manual?? Or possibly have they confirmed an issue with the way it is described?? It seems a blog on repairing this instrument (pretty much the same repair I had to perform) also seems to allude to issues with the manual.? My cal signal and amplitude for both linear and log modes looks similar to his (blog link here; scroll to the bottom )
?
Thanks in advance for your help!
?
-Frank |
Re: Conductive Paint for Microcircuit Repair...
Use masking tape to define your trace width. Dave Casey? On Tue, Jul 16, 2024, 7:55 AM Chuck Harris via <cfharris=[email protected]> wrote: Hi Gang, |
Conductive Paint for Microcircuit Repair...
Hi Gang,
I need a workable conductive paint for use on repairing a really tiny flexible circuit board trace. Due to a flaw in manufacturing, a small bit (about the diameter of a human hair) of a conductive trace is once again missing. About 10 years ago, I fixed the problem with a conductive silver paint, and a brush made from one of my arm hairs. The whole operation was right at the limits of my abilities to manipulate things by hand. Now, it is 10 years later... The first time, I was vexed by the quickness with which the paint's solvent evaporated, and skinned over. I got the job done before, but it looked like a clotted mess. What I would like is a one part conductive paint that flows, and works like water or ink, but has a working time of 30 seconds, or so... while making a hair thick trace... which isn't quite the same thing as making a 1/16th in brush stroke... Conductive thermosetting epoxies look promising, but I don't know how they were meant to be applied, or what working with them is like. The board is orange, like kapton (polyamide), but I don't really know what its temperature limits are. The trace is exactly the same width as one of my descended from northern European diaspora arm hairs... about 0.04mm. Ideas? -Chuck Harris |
Re: 8702A service manual [SEC=UNOFFICIAL]
Yes I am getting that feeling. The 8702A uses a vector display and not RGB? like the later models. one of the Simmconn kits looks like it is the only way to go unless there are boards I can swap out to upgrade the 8702A to a later model, which I doubt. Is there any benefit in changing firmware from 8702A to 8753 as far as functionality? for electrical testing?
I did find the 8753 manuals , but not much info on teh video side. worked out? that the
8702A uses 1349A display unit, but not an easy or cheap upgrade to an LCD :( |