开云体育

ctrl + shift + ? for shortcuts
© 2025 开云体育
Date

Re: BOONTON 8900B PEAK POWER CALIBRATOR

 

开云体育

To avoid trampling over another group bandwidth i suggest you as this on the Marconi group.io pages
for this question.

And definatly off brand topic, has anyone any details/specs on a Marconi TM9701 VHF/UHF Detector, test accessory. Its an accessory with both a positive and negative output via seperate diodes. Id like to test it but manufacturers specs would be handy.

I suspect it may be possible to help you out.

Regards?


On 3 May 2024, at 22:17, Pete Harrison via groups.io <rockfisher@...> wrote:

?

Hi all, i appreciate this isnt a HPAK instrument but being Boonton it kinda is :)

Has anyone got a full service manual for these please. Ive managed to find one online but it doesnt have a circuit diagram.

I also have a pdf of Boontons no35 notebook? by G.Raymond. Polen who designed the instrument, but id really like the full manual if possible and it will save me trying to draw out the circuit by hand.

?

And definatly off brand topic, has anyone any details/specs on a Marconi TM9701 VHF/UHF Detector, test accessory. Its an accessory with both a positive and negative output via seperate diodes. Id like to test it but manufacturers specs would be handy.

Many thanks in advance Pete Harrison.


HP 3852A firmware update issue

 

I have a 3852A with firmware 3.0 and wanted to use a card that requires firmware 4.2 . I found the instructions and firmware 4.22. I identified my board as 66503 revA? per PCB . Per instructions I only needed to hook pins 1 of U57 and U58 to pin 4 of RP4. I did that but I get a error message that the ROM update is not complete.?
Anyone with revA of 66503 board had this happen?
The suggested remedy on the display is not possible on Rev A boards as there is no jumper wire .?



Re: Attenuator "O" Rings

 

PAI had a minimum order of 100, but when I went up to 500 they started to seem really worth it.? The worst thing about doing a bulk buy was having to count out all those little bags of 50 :^)? I've had no complaints since.
Dan


Re: Attenuator "O" Rings

 

I asked Precision for a quote, I guess we will see Monday.

Wally


Re: BOONTON 8900B PEAK POWER CALIBRATOR

 

Many thanks David, somehow i missed that in my search.

Regards Pete


Re: HP 400H AC VTVM Difficulties

 

开云体育

Since you're okay on the 1mV to 300mV ranges and the trouble starts when you advance to the 1V range, I'm staring at your x1000 range attenuator.? Look for an open there.
What's your serial number?? There were several distinct versions of the circuit as time went on, and the range attenuator was one of the blocks affected.? The early version was known for high noise level.

I've attached repair notes for my 400D which is the same as the 400H except for a lower quality meter movement.? Mine is Serial Prefix 017.? The notes include a partial bibliography of different versions of the manual.

Dave Wise


From: [email protected] <[email protected]> on behalf of SCMenasian <scm@...>
Sent: Saturday, May 4, 2024 5:49 AM
To: [email protected] <[email protected]>
Subject: Re: [HP-Agilent-Keysight-equipment] HP 400H AC VTVM Difficulties
?
Since the problem appears to be a residual reading at the lowest ranges and the grid of V2 helps, you might want to rule out spurious noise getting to V2. Multi MHz oscillation in the V1 stage comes to mind.
I'd check (with a 50 MHz or better 'scope) at the grid of V2. The "phase of the moon" observation supports this hypothesis. Perhaps one or more of the capacitors in the V1 circuit is bad.


Re: Another 3GHz 53132-68003 Chanel 3 option design for the 53131A and 53132A

 

开云体育

Hi Ed,

?

Both 53132A do not count if there is no signal at their CH3 input.

?

I bought these 2 PCBs on eBay, I think they are clones.

?

For 12.4GHz option, the board is identical to this one:

For the 3GHz option, I cannot find the information.

?

Yves

De?: [email protected] <[email protected]> De la part de Ed (scskits)
贰苍惫辞测é?: 4 mai 2024 07:11
??: [email protected]
Objet?: Re: [HP-Agilent-Keysight-equipment] Another 3GHz 53132-68003 Chanel 3 option design for the 53131A and 53132A

?

Yves:

Thanks for the update.

Do the boards count randomly when there is no input?

Were these the OEM options or clones?

ed


Re: Attenuator "O" Rings

 

开云体育

I think precision associates wanted somewhere between 30 and 40 cents each but I don’t recall the minimum order quantity.

About a year ago, they were around 1/3 of the price of apple rubber (NY).


Re: Attenuator "O" Rings

 

Interesting idea, Adrian. I've used Amazon as source of o-rings when I repaired my 8657B. But to your pertinent point, I think I'd go McMaster-Carr or similar today:?. Their products are very good and well classified, spec'ed, and extremely finely organized for searching.?
I'd definitely go this route should I have to do it again.?
Radu.?


On Sat, May 4, 2024 at 2:29?AM Adrian Nicol via <Adrian=[email protected]> wrote:
In metric-land I got a bunch of 1mm ID 2mm OD watch stem seals, eBay or Amazon I seem to recall. That was about 10 years back on an 8566A - still working fine - done several other attenuators since.
However while they seem OK at the moment, nowadays having had cheap plumbing 'rubber' seals go gooey after about 5 years, I have little faith in unbranded stuff and I'll bet it's just a matter of time until they start to return to their original liquid hydrocarbon form.
Probably worth finding some of that size from a known (and trusted) manufacturer?

On 04/05/2024 00:08, Wallace Gasiewicz via wrote:

Does anyone know what size and where I could get "O" rings for my HP 8714 VNA.Attenuator??
One is not there and attenuator is not working properly.

wally KC9INK


Re: Another 3GHz 53132-68003 Chanel 3 option design for the 53131A and 53132A

 

I have the latest version of my PCB's ready to build. I added a trimpot for the bias on the prescaler chip to do away with the need for selecting a resistor, which is accessible through a convenient spare hole in the side of the chassis.
This way you can easily dial in the bias to optimise the tradeoff between stopping random counts with no input signal and sensitivity of the prescaler.

Here is the sensitivity of my previous version (pdf), once I have my latest version constructed, I'll make a better measurement of its performance.
It exceeds the HP spec by a bit up the top of the frequency range, so I'll have to figure out how to fix that.


And a 3D render of the design:


Re: HP 400H AC VTVM Difficulties

 

Since the problem appears to be a residual reading at the lowest ranges and the grid of V2 helps, you might want to rule out spurious noise getting to V2. Multi MHz oscillation in the V1 stage comes to mind.
I'd check (with a 50 MHz or better 'scope) at the grid of V2. The "phase of the moon" observation supports this hypothesis. Perhaps one or more of the capacitors in the V1 circuit is bad.


Re: HP 400H AC VTVM Difficulties

 

Have a look for the 400D & 400L on here too, they are almost identical to the 400H.

Looking at the manual, it lists a number of paper capacitors, these will almost certainly be leaky, some are grid coupling capacitors.
Also from experience I've had a number of the 1500uF low voltage capacitors fail open circuit, these were single cans rather than the dual can used in this, they were in a pair of 412A and another in a 5245L.

David


HP 400H AC VTVM Difficulties

 

Greetings to the Group:

?? I may have botched the search function, as I can't find any actual discussions of repairs to this instrument.?? If so, I apologize for taking up the bandwidth.

?? However, I do seem to need a bit of help with this instrument.?? The unit will not zero on the 1 volt / 0 dB scale; it reads anywhere from about 1/3 scale to 2/3 scale, depending on phases of the moon or whatever.?? The reading decreases proportionally on the higher voltage scales.?? On the .3 volt / -10 dB scale, the zero is pretty close, and remains so on the more sensitive scales.

?? On the .3 volt / -10 dB scale, the meter cannot be calibrated; with R29 at the maximum position and .3 volts applied, the reading is .18 volts.?? R29 seems to have little effect on the meter level.?? It can be used to bring the meter to read .3 volts on the 1 volt scale as per the manual. but the meter reads only slightly less with no signal applied and the input terminals shorted.

?? The manual suggests that the trouble is in the power supply, and indeed, I see a 60 Hz signal in the last two stages of gain (V4 and V5).?? Unfortunately, the power supply looks very clean with my oscilloscope.?? I note that there is only 3 mV of ripple permitted, so I had to use a series .1 uF capacitor to protect my scope from the B+ when using a X1 probe.?? At 5 mV per division, I see only about 1 mV (barely discernible).?? I have two of these instruments and have them opened up side by side on the bench.?? One works well and the other is failed as described.?? The two instruments have very similar power supply ripple, with the good instrument possibly being a bit worse than the failed one.

?? Filament supply DC ripple is more pronounced, but is no worse than (and may be slightly better if I recall correctly) than the working unit.

?? Grounding the grid of V2 gets rid of the noise and allows the instrument to zero.?? V1 and V2 have both been replaced with no improvement.?? All of the tubes have been pulled from the instrument and checked with a TV-7 B/U tube tester.?? All check good for shorts, mutual conductance and gas.

?? I am looking for any suggestions at this point.... I seem to be completely stumped.

? Thanks for reading and for nany suggestions you may have.

73,
--
Jim Thorusen
KB6GM


Re: BOONTON 8900B PEAK POWER CALIBRATOR

 

The HP 8900B manual with circuits diagrams is available from BAMA;

David


Re: HP 3852a extended memory modules info

 

The CPU board slides out from the back. The ram slot is available through a cutout on the CPU board. The CPU is a 68000 and HP made only 256k and 1M cards for these while other vendors made 2M and 4M versions. I think other card where not available due to ram cost.? Not sure that the system can handle more then 4M but the CPU is able to address up to 16M of address space.?


Re: Another 3GHz 53132-68003 Chanel 3 option design for the 53131A and 53132A

 

Razvan:

The 3GHz board is a divide by 128, with a max output of 23.4375MHz at 3GHz.
That has to get to the main board through the ribbon cable, not the best way to connect.
At 12GHz the output would be almost 100MHz so I think that the higher frequency option boards likely have an additional divider to keep the output at a reasonable frequency.

There are two pins (CH3CODE0 and CH3CODE1) on the connector that are used to tell the main board that one of the options is installed.

I would think that a combination of those pins would select one of the three options or no option.

I will try to see what effect they have later this weekend.

ed


Re: Another 3GHz 53132-68003 Chanel 3 option design for the 53131A and 53132A

 

Yves:

Thanks for the update.

Do the boards count randomly when there is no input?

Were these the OEM options or clones?

ed


Re: Attenuator "O" Rings

 

开云体育

In metric-land I got a bunch of 1mm ID 2mm OD watch stem seals, eBay or Amazon I seem to recall. That was about 10 years back on an 8566A - still working fine - done several other attenuators since.
However while they seem OK at the moment, nowadays having had cheap plumbing 'rubber' seals go gooey after about 5 years, I have little faith in unbranded stuff and I'll bet it's just a matter of time until they start to return to their original liquid hydrocarbon form.
Probably worth finding some of that size from a known (and trusted) manufacturer?

On 04/05/2024 00:08, Wallace Gasiewicz via groups.io wrote:


Does anyone know what size and where I could get "O" rings for my HP 8714 VNA.Attenuator??
One is not there and attenuator is not working properly.

wally KC9INK


8593A display problem

 

hello friends,

I got a 8593A with a unknown problem,
to me it looks to be a failes display board. Any tips to short the search ?
thank you.

see the short video I have made from that :-)


greetings from germany
Martin


Re: Another 3GHz 53132-68003 Chanel 3 option design for the 53131A and 53132A

 

开云体育

Hi,

?

I bought options 030 and 124 (3GHz and 12.4GHz) on eBay a few years ago.

I tested them again today and here is the result on their respective sensitivity.

?

Option 124 doesn't pass specs below 1.8GHz, but the rest is ok.

?

his gives you an overview of the performance of these modules.

?

Yves

?

?

De?: [email protected] <[email protected]> De la part de Ed (scskits)
贰苍惫辞测é?: 3 mai 2024 13:23
??: [email protected]
Objet?: Re: [HP-Agilent-Keysight-equipment] Another 3GHz 53132-68003 Chanel 3 option design for the 53131A and 53132A

?

Tony:

The 53131A had three options for the channel 3 input, 3GHz, 5GHz, and 12GHz.

I have never seen anything on EBAY over 3GHz installed in a unit.

I have seen Chinese Channel 3 boards for 3GHz (a lot of them) and 6GHz (just 1) currently listed.
I do not see the 12GHz board listed at this time, although it was listed in the past.

I did a board a few years back using the HMC432 (divide by 2) and two HMC434 (divide by 8) that could get to 6GHz or higher.

A less expensive version would use the HMC432 driving an MC12079 programmed to divide by 64, have the layout for it but never made it.

Never did any follow up on these because the Chinese clone boards were so inexpensive it was not worth pursuing.

ed