Re: 8510C - calibration 3.5/sma
On Tue, 2 Apr 2024 at 13:41, LeonP <leon.pavlovic@...> wrote: Hi Tom, I don't have time to look for it now, but someone created a paper where they modeled the effects of using a 3.5 mm calibration kit with SMA and 2.92 mm connectors. For 2.92 mm, a correction can be applied, but the variability among SMA connectors means that it's pointless using any correction, as the correction is small compared to the variability on the semi-precision SMA connectors. So essentially, don't worry about it.Chances are your DUT is likely to be SMA too. -- Dr David Kirkby Ph.D Kirkby Microwave Ltd Email: drkirkby@... Web: Telephone 07910 441670 (UK) or +44 7910 441670 (international) Registered in England and Wales, company number 08914892.? Registered office: Stokes Hall Lodge, Burnham Rd, Althorne, Essex, CM3 6DT |
Re: 8510C - calibration 3.5/sma
Hi Tom,
thanks for the help. I agree with you. However, my concern is the following: my cal kit is 3.5mm (it is in a fairly good condition) and the calibration assumes I also have the 3.5mm port cables (obviously HP sold you the 8510C with its 26.5GHz 3.5mm cal kit along with the 3.5mm port cables). However, I do not have the 3.5mm port cables, but SMA cables. Therefore there is a 3.5mm/SMA interface. I've seen the adaptor element in the HP cal kit table, what is the use of it? Thanks! |
Re: HP 606B Frequency Dial
¿ªÔÆÌåÓýThank you, Steve.
Your scan I will be able to use as a background in my CAD system to recreate the dial on photo paper.
Thanks,
Mark Sent from From: [email protected] <[email protected]> on behalf of Steve Dench via groups.io <sjdench48@...>
Sent: Tuesday, April 2, 2024 1:36:36 PM To: [email protected] <[email protected]> Subject: Re: [HP-Agilent-Keysight-equipment] HP 606B Frequency Dial ?
Hi Mark
Attached is a scan of the dial from my parts machine, its not perfect as it has black dot marks on it, but it may be a start for you. Its currently in PDF but if you want it scanned in some other format let me know, you may be able to edit the worst marks out. Cheers? Steve.? |
Re: HP 606B Frequency Dial
Hi Mark
Attached is a scan of the dial from my parts machine, its not perfect as it has black dot marks on it, but it may be a start for you. Its currently in PDF but if you want it scanned in some other format let me know, you may be able to edit the worst marks out. Cheers? Steve.? |
Dymec 2460A/M3 plugin manual?
Any chance someone here has a PDF of the manual for these things? I've found some hard copies on eBay so that might be a last resort, but if I can avoid spending $50AUD just for curiosity that'd be nice
I can't even remember when or where I picked this thing up (I suspect in a bulk lot at some point), it's been on the shelf for years but recently I finally found a power cable for it and gave it a test to find it does in fact work.? There really doesn't seem to be much info out there about these things, but it looks pretty handy for a handful of tasks |
HP 606B Frequency Dial
¿ªÔÆÌåÓýHello Group, ? I am currently restoring a HP 606B. Unfortunately, the dial markings are badly scratched/worn. Does anyone have, or would care to take, a high-resolution copy of this dial? The dial scale is 7.5¡± in diameter. ? Thanks for reading, ? Mark |
Re: Service Notes on HPWiki
It's Official!
We now have more than Keysight has on their dedicated Service Notes site () - over 4000! ?
This is not an April Fool's joke - I just finished reconciling the last >100 notes and added them to HPWiki.
?
I am planning to compare HPWiki <-> Keysight every few months and bring any new notes over. Will be interesting how many they publish on?average.
?
And if they get rid of any notes on their site, we will still retain a copy.
?
I am hoping that eventually someone will find and make available a treasure trove of old Service Notes that are no longer available from Keysight.
Best, Wolfgang KI7PFX ? |
Re: 8510C - calibration 3.5/sma
Hi Leon,
it depends on how accurately you know the properties of your cal elements. Do you still have the kit parameters of the manufacturer? If so, do the standards still obey these parameters, i.e. not worn, clean? Cables are another issue. Are your test cables stable when you move them, i.e. amplitude stability, phase stability? As far as I have found out, the 8510C assumes the load standard to be ideal. Thus, the return loss of your load will limit accuracy of measured reflection coefficients near the 50 Ohms point, while short and open will impact high reflection coefficients. Transmission is yet another story. Sorry, no simple answers possible. Best regards, Tom |
8510C - calibration 3.5/sma
Hello group!
Is there anyone here using the 8510C with a mechanical 3.5 kit? I have a mechanical 3.5mm 26.5GHz cal kit, but my port cables are 18GHz SMA. What is the proper calibration sequence for such a setup? And what kind of performance can I expect (Sxx levels)? Many thanks, Leon PS: I've sent the same message to the svna group weeks ago, but the moderator probably denied it, since it never came through, no help there. |
Re: Another draft: 8160A
Unfortunately, the gold plating on the many boards in these makes them a target of the gold scrappers.
toggle quoted message
Show quoted text
I am lucky to have one of these generators in excellent condition with the 020 option and use it from time to time. Peter On 4/1/2024 9:41 AM, Gianni Becattini via groups.io wrote:
|
Re: HP8560a smoke came out
I think the problem is not really correctable given the
circumstances of a system that injects power back into the grid. The problem is that losses are against you. When you draw from the grid, your transformer and wiring all drops the voltage down from the voltage generated by the power plant, down to the nominal voltage that you expect to see at your house. But, when you inject power back into the grid, you have to overcome the losses of your transformer, wire, etc. before you can actually feed the grid with your power. That necessarily requires you to pump more voltage into the grid than the nominal power line voltage that you would expect to see in your house. Further, because of other issues in the grid (automatic station voltage regulators...), your solar power can only feed other subscribers on your local transformer. That is too small of a population to allow you to avoid a race to the maximum voltage and beyond. You see, your inverter wants to feed excess power to the grid, so it has to raise its voltage until it starts to take up load from the grid. But, your neighbor's solar inverter wants to do the same thing, so it has to raise its voltage until it starts to take up load from the grid, which causes your inverter to shed load, so it cranks the voltage up... and you are in an arms race. If you set your inverter so that it cannot raise your local (house) voltage until it is high enough to allow you to actually supply power to the grid, you will never supply power to the grid. Your best bet is to use a variac, or other step down auto transformer to drop your workbench voltage down to a safe value. -Chuck Harris On Sun, 31 Mar 2024 19:15:20 -0400 "Harvey White" <madyn@...> wrote: Someplace, somewhere, you might be able to find a voltage regulator. |
Re: HP8560a smoke came out
Someplace, somewhere, you might be able to find a voltage regulator.? I suspect that they can be either the saturable reactor kind, or perhaps an old servo/motor driven autotransformer.? It's possible you might just want to knock down the 240-250 to 230-240, which might be more in line to what the equipment expects.
toggle quoted message
Show quoted text
Harvey On 3/31/2024 7:01 PM, glenn percy via groups.io wrote:
Hi Harvey |
Re: HP8560a smoke came out
Hi Harvey
Afraid not.?? Solar panels are very popular in Australia with at least 1/3rd of homes having roof top solar panels.? In my street of about 30 homes more than half have solar. What this tends to do on sunny days, with so much solar energy in a street is to push up the nominal 230v mains voltage to sometimes in excess of 250v.? Glenn vk3pe |
Re: Mounting HP Iron in 19" Racks
This is not directly related to the topic, but our local HAM club has at least three full size racks we need to get rid of.? This is the Anne Arundel Amateur Radio club in Davidsonville, MD. Drop me a note if you're interested and I'll put you in touch with the right person. Jim Simpson |
Re: Mounting HP Iron in 19" Racks
All of my 19" racks have mounting holes front and back, as usual,
but also have mounting holes on the sides as well. I drill pieces of 1-1/2" aluminum angle to fit the mounting holes on the side when the 19" instrument lines up with an appropriate set of holes on the front. Then, you can just slide the instrument onto the full depth angle slides, and you can also use the screw holes in the front panel for added security. The aluminum angles easily handle a couple of hundred pounds each pair. The heaviest thing I have mounted that way is a FLUKE linear amplifier, 5205A. It weighs a cool 120lbs. -Chuck Harris On Sat, 30 Mar 2024 17:28:24 -0400 "Peter Gottlieb" <hpnpilot@...> wrote: I use equipment shelves that only mount from the front, but the one |