Re: HP 11664A Detector - DIY equivalent?

 

开云体育

11664D schematic somewhere? The manual in? files section of grps io is ops only, no ccts.

DaveB, NZ

?

From: [email protected] [mailto:[email protected]] On Behalf Of Paul Bicknell
Sent: Friday, February 04, 2022 22:37
To: [email protected]
Subject: Re: [HP-Agilent-Keysight-equipment] HP 11664A Detector - DIY equivalent?

?

I have done as Jeff described a number of times to get someone a working scaler analyser I use the 11664 C detector adapter for frequencies above 18 Ghz?

so if you are making an adapter without using an old 11664 then the circuit used in the D version is actually better and easier to configure to different detectors ?

Also I have seen possibly on this site a modern board lay out using modern components

Paul B

?

From: [email protected] [mailto:[email protected]] On Behalf Of Jeff Kruth via groups.io
Sent: 04 February 2022 09:15
To: tonycox01@...; [email protected]
Subject: Re: [HP-Agilent-Keysight-equipment] HP 11664A Detector - DIY equivalent?

?

Yes, I have. Polarity of detector doesnt matter, as the signal is chopped at 27.8 KHz.

Best bet is get a blown detector then remove the detector bulkhead and fashion a small aluminum sheet metal plate to hold a BNC, replacing the bulkhead with this. It is the same as an HP11664C detector adapter. Easy-peasy. Works great.

Jeff Kruth

?


In a message dated 2/4/2022 3:52:30 AM Eastern Standard Time, tonycox01@... writes:

?

Hi,

I've been looking for a couple of HP 11664 detectors for a network analyser but was put off by the sky high price - even for second hand items of dubious condition.

I've seen a web page showing a DIY equivalent using a
?33330B?(positive)?detector and an amplifier replicated from the HP service manual.I was wondering if this is a sensible/realistic thing to attempt? Has anybody ever tried this?

thanks

Tony


Re: HP detector

 

That's true .. I don't really play in the areas below 100MHz, I forget the "low end" response ... but I guess for some people it is important.

On Fri, 4 Feb 2022 at 19:54, Jeff Kruth via <kmec=[email protected]> wrote:
A caveat: The guy said he wanted a 1 MHz to X Mhz detector: Most of the 423/420/Sage/FXR,MACOM etc Microwave detectors roll off (some pretty hard) below 10 MHz. This is due often to the DC return, RF Bypass Cap and the input coupling cap (if there is one). Low frequency coaxial detector design is different from microwave ones. I often build my own, just to show students how easy it is.
Jeff Kruth

In a message dated 2/4/2022 2:01:08 PM Eastern Standard Time, robin=[email protected] writes:
?
Just try eBay if you don't find one here.? While the prices of the *unsold* items on eBay are indeed crazy money, if you look at what has actually sold, it is much more reasonable.?? There are some HP 420A for $39 ... good to 12.4GHz
?

On Fri, 4 Feb 2022 at 18:51, Hans Eriksen via <ddtech2000=[email protected]> wrote:
See:


Might have a paper copy of the 8471A manual, which (if I can find it) will scan and make available

/Hans

?

?


Re: HP detector

 

A caveat: The guy said he wanted a 1 MHz to X Mhz detector: Most of the 423/420/Sage/FXR,MACOM etc Microwave detectors roll off (some pretty hard) below 10 MHz. This is due often to the DC return, RF Bypass Cap and the input coupling cap (if there is one). Low frequency coaxial detector design is different from microwave ones. I often build my own, just to show students how easy it is.
Jeff Kruth

In a message dated 2/4/2022 2:01:08 PM Eastern Standard Time, robin@... writes:
?
Just try eBay if you don't find one here.? While the prices of the *unsold* items on eBay are indeed crazy money, if you look at what has actually sold, it is much more reasonable.?? There are some HP 420A for $39 ... good to 12.4GHz
?

On Fri, 4 Feb 2022 at 18:51, Hans Eriksen via <ddtech2000=[email protected]> wrote:

See:


Might have a paper copy of the 8471A manual, which (if I can find it) will scan and make available

/Hans

?

?


Re: HP detector

 

Hans and Eric - thank you, that's very helpful. The Texscan (first time I heard about it) seems to have a very clean spec sheet, sounds like a very good deal for the money.

Robin - that one's pretty good price/spec ratio, but it doesn't seem to quite go as low as I need it (be flat to at least to 1MHz on the low end).?

Thank you all.
Radu.?



On Fri, Feb 4, 2022 at 11:01 AM Robin Szemeti via <robin=[email protected]> wrote:
Just try eBay if you don't find one here.? While the prices of the *unsold* items on eBay are indeed crazy money, if you look at what has actually sold, it is much more reasonable.?? There are some HP 420A for $39 ... good to 12.4GHz


On Fri, 4 Feb 2022 at 18:51, Hans Eriksen via <ddtech2000=[email protected]> wrote:
See:


Might have a paper copy of the 8471A manual, which (if I can find it) will scan and make available

/Hans


Re: HP detector

 

Just try eBay if you don't find one here.? While the prices of the *unsold* items on eBay are indeed crazy money, if you look at what has actually sold, it is much more reasonable.?? There are some HP 420A for $39 ... good to 12.4GHz


On Fri, 4 Feb 2022 at 18:51, Hans Eriksen via <ddtech2000=[email protected]> wrote:
See:


Might have a paper copy of the 8471A manual, which (if I can find it) will scan and make available

/Hans


Re: HP detector

 

See:


Might have a paper copy of the 8471A manual, which (if I can find it) will scan and make available

/Hans


Re: HP detector

 

Hans - I've not been able to track down a datasheet or catalog page for the? HP 8471A. The Keysight website is typically very good at having manuals, catalogs, datasheets, etc. for?vintage HP or Agilent equipment, but for this particular one - and it's the first?time I see a broken link on that website - points to catalog pages that are missing this component. The linked catalog goes to? HP 8471D, E etc.?

Greg - it doesn't have to be HP; it's just I've been told the HP is the gold standard on detectors. Which seems to be reflected in price... But really all I need is good linearity. I have a couple of detectors currently - a vintage "Telonic" that doesn't seem to be worth anything, and a AD8307 active one that's pretty good but I still don't have a solid certainty of its linearity or even suitability. So, if another brand is known to be good (linear and reliable) but less glittery than HP and therefore can be had for less, I'd be interested in hearing about that.?

For instance, Wavetek has a detector which may fit the bill (D-152), but I haven't seen one for sale anywhere currently.?

Thank you,
Radu.?

On Fri, Feb 4, 2022 at 8:44 AM Greg Muir via <big_sky_explorer=[email protected]> wrote:

Does it have to be HP?? For the frequency range that you are dealing with HP detectors seem to be a little overkill.? There are plenty of other manufacturers whose offerings would fit your needs at a much lower price especially if found on that auction site.? And many of them offer frequency ranges up to 10 GHz or more.

?

I have even found detectors installed on RF transmission line directional couplers from old broadcast transmitters that were heading for the scrap yard, were of major manufacture and swept out to 8 GHz.? The price was free.

?

Greg


Re: HP detector

 

Does it have to be HP?? For the frequency range that you are dealing with HP detectors seem to be a little overkill.? There are plenty of other manufacturers whose offerings would fit your needs at a much lower price especially if found on that auction site.? And many of them offer frequency ranges up to 10 GHz or more.

?

I have even found detectors installed on RF transmission line directional couplers from old broadcast transmitters that were heading for the scrap yard, were of major manufacture and swept out to 8 GHz.? The price was free.

?

Greg


Re: HP detector

 

Look at the HP 8471A

0.1 to 1200 MHz and BNC

/Hans


Re: HP detector

 

Look for a HP 8471A

0.1 to 1200MHz and BNC


Re: HP 3314A DC Offset issues

 

Hello;

I repaired a 3314A a while back, I did a bunch of waveform and voltage measurements, I've attached my annotated schematics.

The docs I accumulated are here:

https://xdevs.com/doc/HP_Agilent_Keysight/3314A/


Re: HP Test Equipment Logo Plates Wanted.

jmfranke
 

I am down to five spares. How many do you need and what is your shipping address?

John Franke WA4WDL
4500 Ibis Ct
Portsmouth, VA 23703

On February 4, 2022 at 12:28 AM "William Kibler via groups.io" <will.kibler@...> wrote:

Greetings, I have a couple pieces of test equipment missing the HP logo plate. The kind that is held in place with the two bumps.?

One is an HP 5326B the other is the venerable 8640B. I’d really like to complete my restorations with some original logos if possible.?

Please let me know where I might acquire a couple. I’d be very grateful.

Thank you.?


?


Re: Convert monochrome 5454x scope to color

 

Outstanding! That CCFL what I will need to do soon, display is fairly dim.


Re: HP Test Equipment Logo Plates Wanted.

 

Brian, thank you. I'd like to wait a bit and see if anyone closer has any but if not I would be willing to pay shipping along with whatever you think is fair.
I'm in Michigan.?

-William


Re: HP 11664A Detector - DIY equivalent?

 

开云体育

Hi tony there are various pictures of homemade detector adapters in the pictures on this site

Paul B

?

From: [email protected] [mailto:[email protected]] On Behalf Of Tony
Sent: 04 February 2022 09:56
To: [email protected]
Subject: Re: [HP-Agilent-Keysight-equipment] HP 11664A Detector - DIY equivalent?

?

Thanks Jeff - sounds like something I can have a go at. Do you happen to have a picture of your modified detector - be really useful to see the layout.

thanks

Tony


Re: HP 11664A Detector - DIY equivalent?

 

Thanks Jeff - sounds like something I can have a go at. Do you happen to have a picture of your modified detector - be really useful to see the layout.

thanks

Tony


Re: HP 11664A Detector - DIY equivalent?

 

开云体育

I have done as Jeff described a number of times to get someone a working scaler analyser I use the 11664 C detector adapter for frequencies above 18 Ghz?

so if you are making an adapter without using an old 11664 then the circuit used in the D version is actually better and easier to configure to different detectors ?

Also I have seen possibly on this site a modern board lay out using modern components

Paul B

?

From: [email protected] [mailto:[email protected]] On Behalf Of Jeff Kruth via groups.io
Sent: 04 February 2022 09:15
To: tonycox01@...; [email protected]
Subject: Re: [HP-Agilent-Keysight-equipment] HP 11664A Detector - DIY equivalent?

?

Yes, I have. Polarity of detector doesnt matter, as the signal is chopped at 27.8 KHz.

Best bet is get a blown detector then remove the detector bulkhead and fashion a small aluminum sheet metal plate to hold a BNC, replacing the bulkhead with this. It is the same as an HP11664C detector adapter. Easy-peasy. Works great.

Jeff Kruth

?


In a message dated 2/4/2022 3:52:30 AM Eastern Standard Time, tonycox01@... writes:

?

Hi,

I've been looking for a couple of HP 11664 detectors for a network analyser but was put off by the sky high price - even for second hand items of dubious condition.

I've seen a web page showing a DIY equivalent using a?33330B?(positive)?detector and an amplifier replicated from the HP service manual.I was wondering if this is a sensible/realistic thing to attempt? Has anybody ever tried this?

thanks

Tony


Re: HP 3314A DC Offset issues

 

Hi,

I've just recently fixed poor 3314A, though not with this problem.

But my suggestion is to do a logical sequence of troubleshooting. Study the schematics and work your way along the signal path. If you get an error 34, that means that the signal amplitude is out of limits, obviously because of the offset. The self calibration check the signal right after the output amplifier, before the attenuator. So the problem could be practically anywhere.

However, the offset is injected after the attenuator. So if something is not right with the attenuator, then the controller my be adding offset, that is not suitable for that range. Maybe this is what causes the jumps in your case. I did have problems with the reed relays in the attenuator, the CA3082? driver array was bad and also the actual reed contacts also.

Quite simple to check, look at the output signal with the scope and start from the lowest amplitude and start increasing. You should hear the faint click of the reeds and the signal amplitude should not really change much at that time, when the next attenuator range is switched. If there is a big jump, then you have a problem there. The range of output amplitude is created by the output amplifier creating 0-10V output and various attenuation values create the low level signal. I've documented my repair here:

Otherwise, if this is not the problem (most likely not), then go backwards and follow the signal and you'll find whether the problem is in the output, the preamp, etc. stage. Then check if all waveforms produce this or only one, in that case the waveform shaper may be the issue, etc. Of course as step 0. check all power rails.
The service manual has quite an extensive troubleshooting process, so go though that.

Szabolcs



<andytwgss@...> ezt írta (id?pont: 2022. febr. 3., Cs, 17:15):

Hello folks,

My very first post here.? I have been repairing and trying to do some very basic adjustments to the function generator.? Everything is fine until 5.13 DC Offset Adjustment.

I set up as per instructed, connecting 3314A 's output to my HP 34401A DMM via a 50-ohm dummy load, and set the Offset to .002V.

While Offset @ 0V being 0V, there is a very large offset however, when Offset @ .002 V, the A2-R267 DC Zero range gives -0.67 V ~ -0.12V at the output.? I verified again hooking up to my Tek 2445B scope @ 50-ohm input, offset adjustments appears to be smooth and continuous, except it jumps and always starts from the -ve when adjusting.??

The machine currently has a E34 error code when bootup and adjusting.?

Any ideas?? Thank you and much appreciated.


Re: HP 11664A Detector - DIY equivalent?

 

Yes, I have. Polarity of detector doesnt matter, as the signal is chopped at 27.8 KHz.
Best bet is get a blown detector then remove the detector bulkhead and fashion a small aluminum sheet metal plate to hold a BNC, replacing the bulkhead with this. It is the same as an HP11664C detector adapter. Easy-peasy. Works great.
Jeff Kruth
?

In a message dated 2/4/2022 3:52:30 AM Eastern Standard Time, tonycox01@... writes:
?
Hi,

I've been looking for a couple of HP 11664 detectors for a network analyser but was put off by the sky high price - even for second hand items of dubious condition.

I've seen a web page showing a DIY equivalent using a?33330B?(positive)?detector and an amplifier replicated from the HP service manual.I was wondering if this is a sensible/realistic thing to attempt? Has anybody ever tried this?

thanks

Tony


Re: 8350B *restoration adventure?*

 

PD I'm not in the UK, I live in Madrid, Spain