Keyboard Shortcuts
Likes
- HP-Agilent-Keysight-Equipment
- Messages
Search
HP8592L level issue
Hi, I have an 8592L that has a level discrepancy between high and low bands ... of around 16dB, its the high band that is measuring flat, but low. ... it's not the mixer, that has been swapped for a known worker with exactly the same result (+- 0.5dB) ...? the amplitude, frequency and YTF cals complete without issue ... where to look next? I bought this with a blown mixer, the upper range was down, the lower range was dead.?? Swapping the mixer for a known good one has restored the low range to full function, but the upper range still has the same low level ... Ideas? -- Best regards, |
Re: Re-capping an HP-130 Scopes
I learned allot from his vids.? He does have 8k+ patreaons so I figure he is making at least 20k+ plus a month so he must be doing something right.
He manages to sell his? Cap tester quite well.? I have built is but it is the one thing I rarely use as I figured out 99.9% of all waxies/paper are bad and there is no sense in even testing them or even keeping any of them. I did find a few dud mica domino caps with it but that was from a plie of 200+ NOS I acquired.?? |
Re: HP8620C Fan removal.
On Thu, Oct 28, 2021 at 06:42 PM, Dave_G0WBX wrote:
By the sound of it, non unlike some of the early Tek 465 'scope fans, but probably physically very different.As it happens, some HP 8620 series were equipped with the very same Siemens fan motor that Tek used in their 465-family of 'scopes. A lot has been written about their behaviour and overhaul, see Tekscopes groups.io. Raymond |
HP8620C Fan removal.
Hi.
Anyone have any hard won experience as to how best to tackle the removal of the cooling fan from a HP8620C sweeper mainframe? The fan in mine, is running slow (so the instrument, and especially the plug ins, "run rather warm.") It is also making "grumbling bearing" sounds, so I'd like to remove it for overhaul (clean and lube, as I've done for many others) or replace it with a more modern alternative. I am aware it is an early DC driven device, with external motor position sensor and drivers, plus it's speed is governed by a temperature sensor somewhere. By the sound of it, non unlike some of the early Tek 465 'scope fans, but probably physically very different. I've a "user" manual so far, but that just talks about removing/cleaning/replacing the mesh air filter. Any advice at this stage welcome. Regards to All. Dave B |
Re: Re-capping an HP-130 Scopes
开云体育I’m not a fan.? His assured spew of misinformation rubs me the wrong way. ? Dave Wise ? From: [email protected] [mailto:[email protected]]
On Behalf Of Roy J. Tellason, Sr. via groups.io
Sent: Thursday, October 28, 2021 9:06 AM To: [email protected] Subject: Re: [HP-Agilent-Keysight-equipment] Re-capping an HP-130 Scopes ? On Wednesday 27 October 2021 10:58:33 pm Glen Slick wrote: |
Re: Re-capping an HP-130 Scopes
On Wednesday 27 October 2021 10:58:33 pm Glen Slick wrote:
On Wed, Oct 27, 2021 at 7:53 PM Jim Ford <james.ford@...> wrote:Wow, talk about over-complicating things... You don't need a scope to do this test. Just a wire plugged into any handy audio amplifier will do. One way will get you more hum out of the speaker than the other. An RCA jack allows the use of any handy audio patch cable and it's way cheaper than a BNC. If you do want to build a gizmo, you don't need 3 ICs, either, just a pair of transistors in an astable multivibrator circuit. Put the LEDs in line with the collector of each transistor, and when that one turns on that transistor's collector is going toward ground. If I were going to build something like that a bit of perf board would do, no need to make a circuit board, no surface mount parts either. He mentions the "CD" prefix as being more tolerant of higher power supply voltages than other brands. If he'd gone with the original 4066 rather than the 74HC4066, they can handle power supply voltages of up to 15V no problem. He also mentions wanting longevity in terms of operation, but uses a bipolar 555 rather than the CMOS 7555. Not much of a fan of that channel, I guess... -- Member of the toughest, meanest, deadliest, most unrelenting -- and ablest -- form of life in this section of space, ?a critter that can be killed but can't be tamed. ?--Robert A. Heinlein, "The Puppet Masters" - Information is more dangerous than cannon to a society ruled by lies. --James M Dakin |
Re: OT: NI GPIB-USB-HS+ repair question
<>
toggle quoted message
Show quoted text
-----Original Message-----
From: [email protected] <[email protected]> On Behalf Of nigel adams via groups.io Sent: 28 October 2021 15:04 To: [email protected] Subject: Re: [HP-Agilent-Keysight-equipment] OT: NI GPIB-USB-HS+ repair question Please can you describe how you rewrote the memory chip and where did you find the code and routine to perform it. -----Original Message----- From: [email protected] <[email protected]> On Behalf Of David C. Partridge Sent: 28 October 2021 14:00 To: [email protected] Subject: Re: [HP-Agilent-Keysight-equipment] OT: NI GPIB-USB-HS+ repair question You have to pry it open with a metal spudger. Good luck with it ... I'd be very interested to know what the problem is. I managed to fix a GPIB-USB-HS (the older version) by re-writing its memory IC. David -----Original Message----- From: [email protected] <[email protected]> On Behalf Of Tom B Sent: 28 October 2021 12:57 To: [email protected] Subject: [HP-Agilent-Keysight-equipment] OT: NI GPIB-USB-HS+ repair question Hello All, I have a National Instruments GPIB-USB-HS+ adapter that appears to be dead. It does not respond at all when plugged into a USB port. What I need to know is how to open it up. I took out the two visible screws but would not come apart. There do not appear to be any additional screws under the labels. Could someone tell me how to open this device without damaging it? Tom Bryan N3AJA |
Re: OT: NI GPIB-USB-HS+ repair question
Please can you describe how you rewrote the memory chip and where did you find the code and routine to perform it.
toggle quoted message
Show quoted text
-----Original Message-----
From: [email protected] <[email protected]> On Behalf Of David C. Partridge Sent: 28 October 2021 14:00 To: [email protected] Subject: Re: [HP-Agilent-Keysight-equipment] OT: NI GPIB-USB-HS+ repair question You have to pry it open with a metal spudger. Good luck with it ... I'd be very interested to know what the problem is. I managed to fix a GPIB-USB-HS (the older version) by re-writing its memory IC. David -----Original Message----- From: [email protected] <[email protected]> On Behalf Of Tom B Sent: 28 October 2021 12:57 To: [email protected] Subject: [HP-Agilent-Keysight-equipment] OT: NI GPIB-USB-HS+ repair question Hello All, I have a National Instruments GPIB-USB-HS+ adapter that appears to be dead. It does not respond at all when plugged into a USB port. What I need to know is how to open it up. I took out the two visible screws but would not come apart. There do not appear to be any additional screws under the labels. Could someone tell me how to open this device without damaging it? Tom Bryan N3AJA |
Re: Looking to buy HP6268B 40V 30A Power Supply
开云体育Dear Wally,I have the 60V version, 6274B. Under J3E syllabic loading the noises from the mains transformer were loud and a little worrying, so I decided not to use it for transmitter powering. Perhaps the experts on here could comment whether such use is likely to cause problems. Other than that, the Rifa capacitors went up with a sizzle on 240V switch-on after a couple of years unused. Replaced those with more generously voltage rated modern alternatives and has worked well since. Regards, Alwyn _____________________________________________________ Alwyn Seeds, Director SynOptika Ltd., 114 Beaufort Street, London, SW3 6BU, England. SynOptika Ltd., Registered in England and Wales: No. 04606737 Registered Office: 114 Beaufort Street, London, SW3 6BU, United Kingdom. _____________________________________________________ |
Re: OT: NI GPIB-USB-HS+ repair question
You have to pry it open with a metal spudger.
toggle quoted message
Show quoted text
Good luck with it ... I'd be very interested to know what the problem is. I managed to fix a GPIB-USB-HS (the older version) by re-writing its memory IC. David -----Original Message-----
From: [email protected] <[email protected]> On Behalf Of Tom B Sent: 28 October 2021 12:57 To: [email protected] Subject: [HP-Agilent-Keysight-equipment] OT: NI GPIB-USB-HS+ repair question Hello All, I have a National Instruments GPIB-USB-HS+ adapter that appears to be dead. It does not respond at all when plugged into a USB port. What I need to know is how to open it up. I took out the two visible screws but would not come apart. There do not appear to be any additional screws under the labels. Could someone tell me how to open this device without damaging it? Tom Bryan N3AJA |
Re: Looking to buy HP6268B 40V 30A Power Supply
On Wed, Oct 27, 2021 at 12:36 PM, kc0wox Leeper wrote:
Where are you located? I may have one but I'm near Denver and won't ship it.Would have to ship to Miami FL. I know these can be a bear to pack and ship. Thank you. |
OT: NI GPIB-USB-HS+ repair question
Hello All,
I have a National Instruments GPIB-USB-HS+ adapter that appears to be dead.? It does not respond at all when plugged into a USB port.?? What I need to know is how to open it up.? I took out the two visible screws but would not come apart.? There do not appear to be any additional screws under the labels.? Could someone tell me how to open this device without damaging it? Tom Bryan N3AJA |
Re: Looking to buy HP6268B 40V 30A Power Supply
On Wed, Oct 27, 2021 at 02:59 PM, Wallace Gasiewicz wrote:
I bought the PS because I was sick and tired of dealing with birdies on SMPSThere you go, just another -very- good reason to stay with a good old linear PS. I have enough birdies already just dealing with the Ethernet cables and routers, don't need any more :-) |
Re: HP54720D Service manual
Hello Dave, |
Re: HP200CD oscillator
X-rays don't happen at such low voltages. The blue glow,
if on the envelope, is fluorescence caused by electron bombardment. If you get a glow that is inter-electrode, it is ionized gas. -Chuck Harris On Wed, 27 Oct 2021 20:15:24 -0700 "Richard Knoppow" <dickburk@...> wrote: ?? Some tubes like the ones you mentioned, have a deep blue glow |
Re: HP200CD oscillator
Some tubes like the ones you mentioned, have a deep blue glow on the inside of the envelope due to x-ray bombardment. This is different than a gas glow, which is inside the tube, usually between the plate and heater. This is usually a violet glow. This is from gas that has either evolved from the elements or has gotten in due to a leak. If a tube glows brightly its not going to last long.
toggle quoted message
Show quoted text
On 10/27/2021 5:24 PM, Mark Bielman wrote:
Most power amp output tubes (6AQ5, 6L6, EL34, etc) will glow a beautiful blue. --
Richard Knoppow dickburk@... WB6KBL |
Re: Re-capping an HP-130 Scopes
On Wed, Oct 27, 2021 at 7:53 PM Jim Ford <james.ford@...> wrote:
This might have been the video which was relevant to capacitor discussion in this thread: Are Your Capacitors Installed Backwards? Build this and find out |
Re: Re-capping an HP-130 Scopes
On 2021-10-27 10:40 p.m., Glen Slick wrote:
On Wed, Oct 27, 2021, 7:31 PM Jim Ford <james.ford@... <mailto:james.ford@...>> wrote:+1 to Mr C. I get a lot of value from his streams for just $2 a month to Patreon. --Toby |
Re: Re-capping an HP-130 Scopes
Thanks, Glen. I've missed the last few days due to exceeding my email storage limit at my ISP.? Friends sending me too many 10-15 MB videos! Jim
------ Original Message ------
From: "greenboxmaven via groups.io" <ka2ivy@...>
Sent: 10/27/2021 9:41:03 PM
Subject: Re: [HP-Agilent-Keysight-equipment] Re-capping an HP-130 Scopes
|