8590B low sensitivity problem
Having repaired the PSU and overcome my issue with corrupt NVRAM, I'm now trying to run the self calibration routines. The frequency routines run OK, but the amplitude routines fail very quickly. Testing the analyser with a variety of sources shows that it's very insensitive, unable to discriminate a signal below about 10MHz, improving very slightly as the frequency inceases to 200MHz, but still 70dB too low. I'm having to use the 8590E manual for the fault finding process as I'm unable to locate the service manual for the B; I do have the CLIP for the B version, although there do seem to be differences on my unit, e.g. the backup battery is on the main board and is AA rather than the A123 battery on the memory board indicated in the CLIP. The service manual calls for the HP IF Test Module which isn't something I can lay my hands on, making it rather tricky to fault find signal level issues, especially as the CLIP doesn't have signal levels marked in many places. |
Re: Repairing flat high voltage ribbon used in HP displays
On 2020-12-04 6:13 a.m., Joel Setton wrote:
Hello,Hi Joel, Thanks again for more info. Yes, this is a delicate repair, indeed. Replacing the complete ribbonYes. I'm pretty sure that's the case. I am definitely trying to avoid complete CRT removal from the 1331A. You are right in saying that high-voltage insulation is only requiredMy intuition might be off here, but I would have thought that ordinary heatshrink would be sufficient? It would interrupt the shortest path air gap completely. That looks like an identical cable and junction box. Very neat, you can't tell it has been reglued. I imagine keeping it attached to the CRT while reglueing and holding in a vise was quite a delicate operation? My _biggest_ concern with either regluing or slicing is keeping things intact towards the CRT. Some kind of clamp would seem to be wanted to stop further delamination? At the moment I am favouring the slicing idea. --Toby
|
Re: Repairing flat high voltage ribbon used in HP displays
Hello,
Yes, this is a delicate repair, indeed. Replacing the complete ribbon cable, assuming you could get one, does not look like an option, because the other end of the cable is somehow bonded to the CRT, I think there's no possibility you could remove it in a way which would make it repairable. You are right in saying that high-voltage insulation is only required between the HV trace and the others. Slicing the cable so that the HV conductor is separate from the other ones looks like a reasonable possibility, assuming you could find some nice, flexible tubing which could stand the high voltage (teflon ?). I think it would be more flexible if "not shrunk".? It would not look as nice as the original thing, but there's a good chance it would work. For what it's worth, I am attaching a picture of the flat cable in my 181A. It looks very similar to the one in your 1335A. What is not visible in the picture is that I also installed a piece of plastic sheet inside the cover, so that the ribbon cable would not spring an arc to the sheet metal. Of course, YMMV. I'm staying tuned ! Joel |
Re: Weather resistant, covered Litz wire wanted.
开云体育Hi Donald,
how many meters do you need ? I have AWG24/19 FEP 73, Rainer Am 04.12.2020 um 01:21 schrieb donald collie: Sorry about the off topic post, but can anyone suggest a source of Litz wire that can be used outside in the weather.[PVC, or Teflon, etc covered]. I`m going to build a loop antenna for receive, and need several meters. Thankyou!? <donaldbcollie@...> |
Re: Weather resistant, covered Litz wire wanted.
Thanks for the suggestion George - there`s plenty of cotton covered stuff on eBay [bless em]...................Don C. On Fri, Dec 4, 2020 at 6:08 PM george edmonds via <G6HIG=[email protected]> wrote:
|
Re: Weather resistant, covered Litz wire wanted.
Donald I would suggest that you use standard cotton covered and put several layers of varnish on it. George G6HIG Dover UK On Friday, 4 December 2020, 00:21:54 GMT, donald collie <donaldbcollie@...> wrote: Sorry about the off topic post, but can anyone suggest a source of Litz wire that can be used outside in the weather.[PVC, or Teflon, etc covered]. I`m going to build a loop antenna for receive, and need several meters. Thankyou!? <donaldbcollie@...> |
Re: Performance verification on a HP8510C
One question - are you confusing calibration with verification. If
toggle quoted message
Show quoted text
so, calibration can be done with systm software and a set of standards. The verification needs an external computer and a set of standard air lines. Cheers! Bruce Quoting Job PH4AS <groupsio@...>: Hi Bill, |
Re: 40GHz cal kit.
Pete Manfre
85562a… . On Thu, Dec 3, 2020, 5:20 PM Pete Manfre via <pmanfre=[email protected]> wrote:
|
Re: 40GHz cal kit.
Pete Manfre
Not sure of number off hand but it looks the same.? ?I will look it up ehen i get home.? P On Thu, Dec 3, 2020, 5:04 PM Dr. David Kirkby, Kirkby Microwave Ltd <drkirkby@...> wrote:
|
Re: 40GHz cal kit.
On Thu, 3 Dec 2020 at 04:10, Pete Manfre <pmanfre@...> wrote:
What Keysight kit was it equivalent to? . |
Re: 40GHz cal kit.
Pete Manfre
??? P On Wed, Dec 2, 2020, 10:14 PM Mike Vande Voort <mike@...> wrote:
|
Re: Hp 8664A
Just to set the record straight, the 8644, 8645 and 8665 were all developed at the same time as part of what we called the Performance Signal Generator family. They share hardware but there are differences between them. The 8643 is an 8644 with a different power supply and without the 140ns frequency discriminator option. The 8664 is an 8665 without the output section to operate above 3 GHz.
toggle quoted message
Show quoted text
The 8645 was designed for fast frequency hopping applications. The 8644 was a general purpose generator covering the same frequency range. The 8665 with its YIG oscillator extended frequency coverage to 4.2GHz (A version) and 6 GHz (B version). All three utilized a single loop fractional N synthesizer with frequency discriminators to enhance phase noise performance. These generators were designed at the Spokane Division of HP. One other generator followed which was the 8647. It was a low cost generator that used the signal generator hardware developed for the 8920 test set. Around 1992 the RF signal generator charter moved from Spokane to Santa Rosa so Spokane could concentrate on cellular test. The 8648 used some of the 8647 design but the RF hardware was different and was designed in Santa Rosa. To be complete, the 8640 and 8660 were designed at the Stanford Park division and their production was moved to Spokane when Spokane split off from Stanford Park. The 8642 was designed in Spokane to be a programmable replacement for the 8640. The 8642 was a multi-loop synthesizer design that used fractional N synthesis in a low frequency loop. The 8662 was designed at Stanford Park and the 8663 was in development when it was transferred to Spokane. These generators had very good close in phase noise performance but were worse than the 8640 or 8642 at higher offsets. Jim KD7F On Dec 2, 2020, at 11:20 PM, Don Bitters via groups.io <donbitters@...> wrote: |
Re: Repairing flat high voltage ribbon used in HP displays
开云体育Toby wrote on 12/3/2020 9:27 AM:<snip>Hello, Toby and the group--I can see some very difficult aspects to this repair. If it's delaminating along the whole length then the whole length would need to be reglued, but a perfect seal would only be needed in the gap between the 7.5kV conductor and the others? I can offer the following high-voltage NOS wire which may provide another option to repairing the ribbon cable: High-voltage wire: Brim Electronics p/n CHV22-6.5, AWG#22 (19 strands of AWG#34) insulated with Teflon ? tape overwrap, plus special anticorona treatment of the conductor. Rated for 32.5 KVDC maximum. or 10 KVAC maximum. Outer diameter is 0.141 inch. ?Limited
qty. available at $1.50 per foot, plus USPS postage. Note that overseas shipment may be extremely
expensive! Questions
welcomed, PayPal honored. 73-- Brad? AA1iP |
Re: Repairing flat high voltage ribbon used in HP displays
On 2020-12-03 3:27 a.m., Joel Setton wrote:
Hello Toby,Hi Joel, Yes it does, thanks for the reply. I'd be interested to know what your ribbon cable looks like. Was it removable at the CRT end, or integrally connected? I can see some very difficult aspects to this repair. If it's delaminating along the whole length then the whole length would need to be reglued, but a perfect seal would only be needed in the gap between the 7.5kV conductor and the others? I am also thinking that slicing the 7.5kV part away from the other two and heatshrinking it along its length would also work. I might even be able to avoid removing the entire CRT this way (it took me a few _hours_ to remove a broken CRT from a 1335A plus I'd be very worried about damaging this one. These CRTs now seem to be unobtainable, period). Never attempted anything like this myself so it is great to hear from someone who has resolved it before! --Toby
|
Re: Performance verification on a HP8510C
开云体育Hi Job, You are welcome, verification process is basically having VNA to measuring all components in a calibration kit and then compare the obtained S parameter data to the HP factory supplied data (with verification kit) to see how much it is off. I think by measuring the verification kit component and look on the displayed S11, S21 you should already know if your cal kit or system is ok or not. You don’t really need to run the verification program, unless you have other serious need. I strongly suggest you to download full manual in my previous link and read section 8 for more detail. Good luck. ? ? Best regards,?? Steve Cheng? ??????????????? North American EMC Certification Services Inc. ? |
Re: Performance verification on a HP8510C
开云体育Job Please contact me off list Hardy ? Fra:
[email protected]
[mailto:[email protected]] P? vegne af Job PH4AS ? Hi Steve, Thanks for highlighting the requirements. I guess I was hoping for a more modern solution. Unfortunately it seems the commercial HTBasic package is the only option in that sense. At least I have not found indications otherwise. I better first get hold of the software disks and see if I can figure out what they do. Best regards, Job ? |
Re: Hp 8664A- HP8643A
开云体育Dear All,The HP8664A is in the same family as the HP8643A, with many modules common. On my HP8643A, the battery only seems to charge when the generator is switched on, not when in standby. I, too, had calibration errors when I bought it. There were two causes: 1. An internal connector that was not properly seated. 2. The battery issue. These days, I switch it on for a few hours every couple of weeks- that ensures it holds last settings and does not get stuck in the cal. errors problems. I’m sure the battery could do with replacement- does anyone know where the battery is and what is a suitable replacement type? Other than this issue, this is my chosen generator for reciprocal mixing measurements, as it has very low amplitude and phase noise for a synthesised generator of that era- I believe it was made for testing aircraft landing systems (ground segment- it weighs so much, an aircraft might struggle to take off, unless it was a B52). Regards, Alwyn ? _____________________________________________________ Alwyn Seeds, Director SynOptika Ltd., 114 Beaufort Street, London, SW3 6BU, England. SynOptika Ltd., Registered in England and Wales: No. 04606737 Registered Office: 114 Beaufort Street, London, SW3 6BU, United Kingdom. _____________________________________________________ |