¿ªÔÆÌåÓý

Date

Re: HP8640B fan

 

¿ªÔÆÌåÓý

Hi, George. ?Your predicament spurred me to research it.


A clear copy of this schematic can be found in 08640-90215. ?That is the newer of two 8640B service manuals?available at KeySight;?I got my copy from KO4BB's fine manuals website.

It is Service Sheet 19 (SS19), Figures 8C-61 and 8C-67, at page 321 of 364 in the PDF document.


This page is the -5.2V regulator as well as the fan driver. ?The regulator is A17U1,?a three-terminal adjustable positive regulator IC?whose output is grounded. ?Its input pulls down the positive pole of a 15V floating unregulated supply until the negative pole is at -5.2V . ?A18R1 trims the regulator IC's adjust pin. ?Downstream circuitry is protected by SCR crowbar A18Q3 controlled by zener A18VR2. ?If the regulated output goes more negative than -6.2V the SCR will fire and blow 2A fuse A18F1. ?2.5-ohm 10W?A18R9 reduces the power dissipated in A17U1.


The 15V floating supply is filtered by A17C5 and rectified by A12CR17 through 20, depicted on SS18.


Fan B1 is powered by 12V, referred to the -5.2V rail. ?12V zener?A18VR1 and A18Q?1/Q2 form a +12V regulator taking input from the positive pole of the?floating 15V. ?Since its negative pole is -5.2V, its positive pole is about +7V.


A18TP1 is raw plus input (about +10V), TP2 is the other side of the fuse, TP3 is -5.2V and fan negative?, and TP4 is fan positive or about +7V.


HTH,

Dave Wise


From: [email protected] <[email protected]> on behalf of george edmonds via groups.io <G6HIG@...>
Sent: Tuesday, October 13, 2020 11:53 AM
To: [email protected]
Subject: Re: [HP-Agilent-Keysight-equipment] HP8640B fan
?
Hi Dave

The schematic I posted is the original HP scan, there is allegedly a full copy of the appropriate HP manual on line but I have never been able to find it. This may be due to Googles regional search policy.

George G6HIG Dover UK
On Tuesday, 13 October 2020, 16:58:40 BST, Dave Wise <david_wise@...> wrote:


?Thank you, George, for this tantalizing information, but I implore you to redraw that schematic and post your sketch even if it's messy freehand.


Regards,

Dave Wise


From: [email protected] <[email protected]> on behalf of george edmonds via groups.io <G6HIG@...>
Sent: Tuesday, October 13, 2020 8:20 AM
To: [email protected]
Subject: Re: [HP-Agilent-Keysight-equipment] HP8640B fan
?
Hi All

Having now had time to find my notes, by serial prefix 2853A (1988) the fan had changed to a two wire one and the -5.2VDC supply and fan board had changed to part number 08640-60392.

This is a vastly simplified board using a three terminal regulator. The only schematic scan of this board that I have been able to find is of poor quality. I have attached a copy.

George G6HIG Dover UK
On Tuesday, 13 October 2020, 14:46:03 BST, alwyn.seeds1 <a.seeds@...> wrote:


Dear All,

It may be a myth, but a colleague told me that on some instruments the fans had controllers that were designed to avoid microphony problems with the instrument- possible that the HP8640B with its cavity oscillator might be one of those?

Regards,

Alwyn

_____________________________________________________

Alwyn Seeds, Director
SynOptika Ltd.,
114 Beaufort Street,
London,
SW3 6BU,
England.


SynOptika Ltd., Registered in England and Wales: No. 04606737
Registered Office: 114 Beaufort Street, London, SW3 6BU, United Kingdom.
_____________________________________________________


HP 70206 Display unit

 

Hi All:
I got a HP 70206 Display unit, on the cheap.
Now, what do I do with it?
Right now, it is keeping the floorboards down.
Is this thing the gateway to an accumulation of HP 70xxx?
Best regards and wishes.
Roy


Re: UNIX files and dd, Re: [HP-Agilent-Keysight-equipment] Agilent 54831M

 

Hi Dave,

Greetings Manuel, I acknowledge your nitpicking and respond with a
bigger nitpicking:
Oh, great ;)

The leading dot making the file "hidden" is a convention of usage,
and is not handled by the filesystem or the core of the OS (the kernel)
itself in any way. It is interpreted and treated that way by the "ls"
program, which lists a directory. When a program opens a directory (via
the OS system calls) and walks down the list of files it contains, they
will all be present, with or without leading dots.
I absolutely second this. Back at the university some distributed filesystems
were even mounted at "..." in the user's home dir, so "~/.../remote_file" was
just a regular file path. Indeed, I continued to use this method on my private
box, just because it's so easy to type and looks so clean... but I'm getting
off topic ;)

I blame the late hour and the soldering fumes (...)
Haha, that's a good one! My favorite data sheet for some solder wire (Alpha
Fluitin 1532) even states as "benefit" the "pleasant pine smell". Mhmmm...


Greetings,
Manuel.


Re: HP8640B fan

 

Hi Dave

The schematic I posted is the original HP scan, there is allegedly a full copy of the appropriate HP manual on line but I have never been able to find it. This may be due to Googles regional search policy.

George G6HIG Dover UK
On Tuesday, 13 October 2020, 16:58:40 BST, Dave Wise <david_wise@...> wrote:


?Thank you, George, for this tantalizing information, but I implore you to redraw that schematic and post your sketch even if it's messy freehand.


Regards,

Dave Wise


From: [email protected] <[email protected]> on behalf of george edmonds via groups.io <G6HIG@...>
Sent: Tuesday, October 13, 2020 8:20 AM
To: [email protected]
Subject: Re: [HP-Agilent-Keysight-equipment] HP8640B fan
?
Hi All

Having now had time to find my notes, by serial prefix 2853A (1988) the fan had changed to a two wire one and the -5.2VDC supply and fan board had changed to part number 08640-60392.

This is a vastly simplified board using a three terminal regulator. The only schematic scan of this board that I have been able to find is of poor quality. I have attached a copy.

George G6HIG Dover UK
On Tuesday, 13 October 2020, 14:46:03 BST, alwyn.seeds1 <a.seeds@...> wrote:


Dear All,

It may be a myth, but a colleague told me that on some instruments the fans had controllers that were designed to avoid microphony problems with the instrument- possible that the HP8640B with its cavity oscillator might be one of those?

Regards,

Alwyn

_____________________________________________________

Alwyn Seeds, Director
SynOptika Ltd.,
114 Beaufort Street,
London,
SW3 6BU,
England.


SynOptika Ltd., Registered in England and Wales: No. 04606737
Registered Office: 114 Beaufort Street, London, SW3 6BU, United Kingdom.
_____________________________________________________


Re: HP-3456A - Question Regarding Resistance Measurement

Bob Albert
 

I suspect my GR 1644 might come pretty close but even so it's only rated for 1% if I remember correctly.

Bob

On Tuesday, October 13, 2020, 11:25:38 AM PDT, Robert G8RPI via groups.io <robert8rpi@...> wrote:


Adrian gave the answer but I think it was lost in the noise.

The 3456A 90 day accuracy on >10M range is only 1.8% _+? count. Even 24h is 1.3%
This is clearly not adequate to measure a 1% resistor even if the 3456A was recently calibrated.

You need a different meter to check a 90M 1% resistor.

Robert G8RPI.


Re: HP-3456A - Question Regarding Resistance Measurement

 

Adrian gave the answer but I think it was lost in the noise.

The 3456A 90 day accuracy on >10M range is only 1.8% _+? count. Even 24h is 1.3%
This is clearly not adequate to measure a 1% resistor even if the 3456A was recently calibrated.

You need a different meter to check a 90M 1% resistor.

Robert G8RPI.


Re: HP-3456A - Question Regarding Resistance Measurement

 

HI George,

Thanks for the information. I knew you'd have insight into this one.

I don't have a precision 10M resistor, but I did find a 10M in my parts drawer (I think it's metal film 5% but not sure about that tolerance band color). In any case, the 3456A measured it as 9.9172M.

I placed the questionable 90M and the 9.9172M in parallel and the 3456A showed the resulting resistance as 8.9441M. That makes the resulting value of the questionable 90M to be 91.1524M. Given that 90M a 1% could still be in tolerance at 90.9M, that's not too far off and given that my 3456A is not calibrated, then I'm guessing the 90M is probably within spec and the rest of that difference is attributable to lack of calibration.

Yes, I know, this is all very crude due to the lack of proper calibration but it at least gives me something that makes just slightly better sense than the directly-measured 90M value I was seeing.

BTW, did you ever find those Red Devil resistors? We discussed this very early this year and I think you said you thought you had some of those that could be used as a somewhat better resistance "standard" than what I happen to have but you were going to wait until your move was complete, etc., to locate them. If you do happen across those, I would still like to get one if that's still possible.

Thanks again,
Barry - N4BUQ

----- Original Message -----
From: ghnatiuk@...
To: [email protected]
Sent: Tuesday, October 13, 2020 10:55:23 AM
Subject: Re: [HP-Agilent-Keysight-equipment] HP-3456A - Question Regarding Resistance Measurement

Barry,
You really need to calibrate your 3456A.? ?The Hewlett Packard / Agilent /
Keysight? DMM meters that were not specifically designed for high resistance
or low current applications as the 4329A or 34420A measure high resistances
by placing the 10 Meg Ohm voltage attenuator (used in the 100V and 1KV
range) in parallel with the unknown and computing the unknown value? from

1/ Rx =? ?1/ Rt? - 1/ 10 Meg

Rt = parallel combination of unknown and internal 10 Meg attenuator
Rx = unknown value

This method keeps the total resistance within the 10 Meg range that can be
measured with reasonable accuracy by the meter but the computation becomes
more and more inaccurate as the unknown resistance goes up in value due to
the meter's ability to resolve the difference between a 10 Meg resistor and
something in parallel with it.? ?Any inaccuracies in the 10 Meg range are
amplified using the extended ohms function.

The 10 Meg attenuator resistor is a? low TEMPCO? tantalum nitride? fine-line
resistor that is not trimmed to precision accuracy? ( 0.5% ) which will
limit the overall accuracy with which a resistor in parallel with it can be
measured by the computation method I described above.

The reason this is done in this manner for a multi-function DMM instrument is
that it is very difficult to make a 100nA current source that would be
needed to inject into a 100 Meg Ohm resistor to limit the voltage across it
so as not to overload the A/D converter input at 10V.? The Ohm's converter
circuitry? would require an internal range resistor on the order of 40 Meg
Ohm in the fine-line resistor network which then would take up a huge amount
of real estate in that network,? not to mention cost.

This extended Ohms function is also available thru the I/O ports on the 3468A
and 3478A meters that puts the 10 Meg attenuator resistance across the puts
terminals and supplies the value of the parallel combination of this with
the unknown over the bus.? Then with a calculator using this value? and the
equation above you can compute the value of the unknown.? There was not
enough room in the memory to code it into the firmware so we left it as an
extended ohms function that can only be accessed thru the I/O? bus.

Conclusion:? get your meter calibrated so the? attenuator 10 Meg resistor is
characterized.
OR in the alternative, obtain a precision 10 MEG or other resistor ( 0.1% or
better ) and put your unknown in parallel with it and measure the parallel
combination using the 10 Meg range of the 3456A and use the equation above
as does the 3456A to compute the unknown value - or use the voltage divider
method suggested earlier by? Stephen? Bell.? ?Good luck with this.

George Hnatiuk, PE






Re: HP8640B fan

 

¿ªÔÆÌåÓý

?Thank you, George, for this tantalizing information, but I implore you to redraw that schematic and post your sketch even if it's messy freehand.


Regards,

Dave Wise


From: [email protected] <[email protected]> on behalf of george edmonds via groups.io <G6HIG@...>
Sent: Tuesday, October 13, 2020 8:20 AM
To: [email protected]
Subject: Re: [HP-Agilent-Keysight-equipment] HP8640B fan
?
Hi All

Having now had time to find my notes, by serial prefix 2853A (1988) the fan had changed to a two wire one and the -5.2VDC supply and fan board had changed to part number 08640-60392.

This is a vastly simplified board using a three terminal regulator. The only schematic scan of this board that I have been able to find is of poor quality. I have attached a copy.

George G6HIG Dover UK
On Tuesday, 13 October 2020, 14:46:03 BST, alwyn.seeds1 <a.seeds@...> wrote:


Dear All,

It may be a myth, but a colleague told me that on some instruments the fans had controllers that were designed to avoid microphony problems with the instrument- possible that the HP8640B with its cavity oscillator might be one of those?

Regards,

Alwyn

_____________________________________________________

Alwyn Seeds, Director
SynOptika Ltd.,
114 Beaufort Street,
London,
SW3 6BU,
England.


SynOptika Ltd., Registered in England and Wales: No. 04606737
Registered Office: 114 Beaufort Street, London, SW3 6BU, United Kingdom.
_____________________________________________________


Re: HP-3456A - Question Regarding Resistance Measurement

 

Barry,
You really need to calibrate your 3456A.? ?The Hewlett Packard / Agilent / Keysight? DMM meters that were not specifically designed for high resistance or low current applications as the 4329A or 34420A measure high resistances by placing the 10 Meg Ohm voltage attenuator (used in the 100V and 1KV range) in parallel with the unknown and computing the unknown value? from? ?

? ? ? ? ? ? ? ? ? ? ? ? ? ? ? ? ? ? ?1/ Rx =? ?1/ Rt? - 1/ 10 Meg

Rt = parallel combination of unknown and internal 10 Meg attenuator
Rx = unknown value

This method keeps the total resistance within the 10 Meg range that can be measured with reasonable accuracy by the meter but the computation becomes more and more inaccurate as the unknown resistance goes up in value due to the meter's ability to resolve the difference between a 10 Meg resistor and something in parallel with it.? ?Any inaccuracies in the 10 Meg range are amplified using the extended ohms function.

The 10 Meg attenuator resistor is a? low TEMPCO? tantalum nitride? fine-line resistor that is not trimmed to precision accuracy? ( 0.5% ) which will limit the overall accuracy with which a resistor in parallel with it can be measured by the computation method I described above.

The reason this is done in this manner for a multi-function DMM instrument is that it is very difficult to make a 100nA current source that would be needed to inject into a 100 Meg Ohm resistor to limit the voltage across it so as not to overload the A/D converter input at 10V.? The Ohm's converter circuitry? would require an internal range resistor on the order of 40 Meg Ohm in the fine-line resistor network which then would take up a huge amount of real estate in that network,? not to mention cost.

This extended Ohms function is also available thru the I/O ports on the 3468A and 3478A meters that puts the 10 Meg attenuator resistance across the puts terminals and supplies the value of the parallel combination of this with the unknown over the bus.? Then with a calculator using this value? and the equation above you can compute the value of the unknown.? There was not enough room in the memory to code it into the firmware so we left it as an extended ohms function that can only be accessed thru the I/O? bus.

Conclusion:? get your meter calibrated so the? attenuator 10 Meg resistor is characterized.
OR in the alternative, obtain a precision 10 MEG or other resistor ( 0.1% or better ) and put your unknown in parallel with it and measure the parallel combination using the 10 Meg range of the 3456A and use the equation above as does the 3456A to compute the unknown value - or use the voltage divider method suggested earlier by? Stephen? Bell.? ?Good luck with this.

George Hnatiuk, PE


Re: HP8640B fan

 

On my 8640B the fan mechanical noise is clearly visible in the
close-in spectrum. Unplugging the fan makes the two little AM spikes
disappear from the screen.

Frank IZ8DWF

On Tue, Oct 13, 2020 at 5:32 PM Dave Wise <david_wise@...> wrote:

Anecdote. I was doing some phase-locked FM work and saw ripple in my discriminator. The fix was lifting the 8640B off the Tektronix 500-series scope it was balanced atop. So yes it's worth an experiment or two.


Dave Wise

________________________________
From: [email protected] <[email protected]> on behalf of alwyn.seeds1 via groups.io <a.seeds@...>
Sent: Tuesday, October 13, 2020 6:45 AM
To: [email protected]
Subject: Re: [HP-Agilent-Keysight-equipment] HP8640B fan

Dear All,

It may be a myth, but a colleague told me that on some instruments the fans had controllers that were designed to avoid microphony problems with the instrument- possible that the HP8640B with its cavity oscillator might be one of those?

Regards,

Alwyn

_____________________________________________________

Alwyn Seeds, Director
SynOptika Ltd.,
114 Beaufort Street,
London,
SW3 6BU,
England.


SynOptika Ltd., Registered in England and Wales: No. 04606737
Registered Office: 114 Beaufort Street, London, SW3 6BU, United Kingdom.
_____________________________________________________


Re: HP8640B fan

 

¿ªÔÆÌåÓý

?Anecdote. ?I was doing some phase-locked FM work and saw?ripple in my discriminator. ?The fix was lifting?the 8640B off the Tektronix 500-series scope it was balanced atop. ?So yes it's?worth an experiment or two.


Dave Wise


From: [email protected] <[email protected]> on behalf of alwyn.seeds1 via groups.io <a.seeds@...>
Sent: Tuesday, October 13, 2020 6:45 AM
To: [email protected]
Subject: Re: [HP-Agilent-Keysight-equipment] HP8640B fan
?
Dear All,

It may be a myth, but a colleague told me that on some instruments the fans had controllers that were designed to avoid microphony problems with the instrument- possible that the HP8640B with its cavity oscillator might be one of those?

Regards,

Alwyn

_____________________________________________________

Alwyn Seeds, Director
SynOptika Ltd.,
114 Beaufort Street,
London,
SW3 6BU,
England.


SynOptika Ltd., Registered in England and Wales: No. 04606737
Registered Office: 114 Beaufort Street, London, SW3 6BU, United Kingdom.
_____________________________________________________


Re: HP8640B fan

 

Hi All

Having now had time to find my notes, by serial prefix 2853A (1988) the fan had changed to a two wire one and the -5.2VDC supply and fan board had changed to part number 08640-60392.

This is a vastly simplified board using a three terminal regulator. The only schematic scan of this board that I have been able to find is of poor quality. I have attached a copy.

George G6HIG Dover UK
On Tuesday, 13 October 2020, 14:46:03 BST, alwyn.seeds1 <a.seeds@...> wrote:


Dear All,

It may be a myth, but a colleague told me that on some instruments the fans had controllers that were designed to avoid microphony problems with the instrument- possible that the HP8640B with its cavity oscillator might be one of those?

Regards,

Alwyn

_____________________________________________________

Alwyn Seeds, Director
SynOptika Ltd.,
114 Beaufort Street,
London,
SW3 6BU,
England.


SynOptika Ltd., Registered in England and Wales: No. 04606737
Registered Office: 114 Beaufort Street, London, SW3 6BU, United Kingdom.
_____________________________________________________


Re: HP8640B fan

 

¿ªÔÆÌåÓý

Dear All,

It may be a myth, but a colleague told me that on some instruments the fans had controllers that were designed to avoid microphony problems with the instrument- possible that the HP8640B with its cavity oscillator might be one of those?

Regards,

Alwyn

_____________________________________________________

Alwyn Seeds, Director
SynOptika Ltd.,
114 Beaufort Street,
London,
SW3 6BU,
England.


SynOptika Ltd., Registered in England and Wales: No. 04606737
Registered Office: 114 Beaufort Street, London, SW3 6BU, United Kingdom.
_____________________________________________________


Re: UNIX files and dd, Re: [HP-Agilent-Keysight-equipment] Agilent 54831M

 

No need to be so derisive towards windows users.

I'm pretty sure that dd gives final transfer statistics at the
end of a good or bad data transfer. (viz: status=noxfer to suppress
the final transfer statistics, and status=progress to give periodic
transfer statistics updates during the transfer...)

Also, using dd to transfer images to memory cards (CF, SD, thumb drives)
will not always work out like you may think it should. For instance
dd'ing a copy of a bootable CF card onto a data storage type CF card
will not result result in a bootable CF card.

-Chuck Harris

Dave McGuire wrote:

? Another thing is silence in the case of success.? Assume that the computer did what
you told it to do...there is no need for it to say "Hey, you told me to do something,
so I did it!"? That's dumb.? The computer should only say something in an unexpected
situation, i.e. when there was an error.? If you type a command at a UNIX shell and
it responds with another shell prompt, it did what it told you to do.

? And no dancing paperclips.

? Computing for grownups, when they need to get work done.

???????????? -Dave


Re: HP HIL keyboard and mouse wanted - in Australia

 

¿ªÔÆÌåÓý

Hello,

lol - we are working on something similar!


Tam

With best regards
Tam HANNA 

Enjoy electronics? Join 15k7 other followers by visiting the Crazy Electronics Lab at  
On 2020. 10. 13. 14:57, Anders wrote:

I am working on a PS/2 and/or USB converter to HIL. I have had the prototype in use with my HP9000/340 since this spring. Contact me off-list and we shall see what we can do. I guess there is no shortage of PS/2 or USB keyboards "down under"?

Yes, I developed this because I bought a 9000/340 without keyboard and spare keyboards were as expensive as the original computer...


Re: HP8640B fan

 

Some of the 8640Bs had the multi-speed fan and some had single-speed standard square fans. HP changed the design somewhere along the way so there's a mix of both out there.

Thanks,
Barry - N4BUQ

----- Original Message -----
From: "Szabolcs Szigeti" <szigiszabolcs@...>
To: [email protected]
Sent: Tuesday, October 13, 2020 1:42:35 AM
Subject: Re: [HP-Agilent-Keysight-equipment] HP8640B fan

Hi,

As far as I know, the 8640B has a 3 phase motor for the fan, and one of the
power supply boards contains the driver and speed regulator circuits for
it. Unless, of course in your unit it has already been replaced by
something else.

Szabolcs


Henry Newton <EequalsIR@...> ezt ¨ªrta (id?pont: 2020. okt. 12., H,
23:04):

Hello all,
I've just ordered the brass gears from the eBay fellow in India. Many
thanks to all who presented the gear information.
While waiting, I need to replace my noisy fan in my 8640B.
From the red & black leads going to a nearby board, it appears it is a 5V
fan. My fan has a black colored motor.
Any ideas on the CFM? Or a fan source?

Thanks







Re: HP HIL keyboard and mouse wanted - in Australia

 

I am working on a PS/2 and/or USB converter to HIL. I have had the prototype in use with my HP9000/340 since this spring. Contact me off-list and we shall see what we can do. I guess there is no shortage of PS/2 or USB keyboards "down under"?

Yes, I developed this because I bought a 9000/340 without keyboard and spare keyboards were as expensive as the original computer...


Re: Agilent 54831M

 

I'm also not a Linux expert, so I'll let other people speak on that.

For the power, I know there is no Molex cable anywhere. They uses industrial PSU and not standard computer one. So of course, you won't have a ready to use cable to plug in any adapter. But, the rear CD-Rom and floppy must have power since it's a 3.5inch IDE cable (without power) that goes to it. I found an old picture of mine where you can see the power cable. It is possible that the last user lost the adapter board to plug the floppy and drive, but the power cable should still be somewhere in the scope.


Re: HP8640B fan

 

Hi

Towards the end of production of the HP8640B the fan was changed to a standard two wire one and the power supply board was modified.

George G6HIG Dover UK


On Tuesday, 13 October 2020, 07:43:27 BST, Szabolcs Szigeti <szigiszabolcs@...> wrote:


Hi,

As far as I know, the 8640B has a 3 phase motor for the fan, and one of the power supply boards contains the driver and speed regulator circuits for it. Unless, of course in your unit it has already been replaced by something else.

Szabolcs


Henry Newton <EequalsIR@...> ezt ¨ªrta (id?pont: 2020. okt. 12., H, 23:04):

Hello all,
I've just ordered the brass gears from the eBay fellow in India.? Many thanks to all who presented the gear information.
While waiting, I need to replace my noisy fan in my 8640B.?
From the red & black leads going to a nearby board, it appears it is a 5V fan.? My fan has a black colored motor.
Any ideas on the CFM?? Or a fan source?

Thanks


Re: HP-3456A - Question Regarding Resistance Measurement

 

Hi, may I add one line?

1. get whatever resistors in the 5-10M range you have so they sum up to approx. 100M.

2. solder them all together.

3. measure resistance of each of the resistors and sum up the results.

4. measure resistance of all the resistors connected in series... and tell us what you found out.

G?ran