Re: Replacement battery for 8566A
The battery is just three AA-sized NiCd cells in series, so you could replace it with anything that you can cobble together from similar cells.? Ed
|
Replacement battery for 8566A
Hi all,
I have an 8566A system that is in need of a new battery so it doesn't lose it's mind all the time. I understand that, unlike the B models, the battery is behind the front panel. Where can I find a replacement that will work?
Thanks!
Sean
|
Hello group,
I am in need of a HP 5086-7592 Yig oscillator.? I have found many on Ebay but figure with the nature of these things I could end up with a part no better than what I have.? Any advice would be greatly appreciated.
Thank you,
Jon
|
New phase lock board for 8753ES - qualitative first test
Today I received a new A11 phase lock board for my misbehaving 8753ES (thank you Pete M!).
A lot of measurements will follow as I have lab time, but I thought I'd run a simple "plug it in and hold ears" test.
Test 48 now passes.? The 1 V/GHz waveform test on the analog bus has the same shape as before, but the starting voltage is below 3.5 volts instead of 3.6 as in the original board.
I still have some intermittent lock issues below 10 MHz but this is major progress.
I have a lot to do, including Root Cause Analysis of the failure of the original phase lock board.? Perhaps the precision 6.2 volt reference zener has gone out of tolerance, it would only need to be a few percent off.? I will see.
What this means is that when the troubleshooting guide says the waveform should be as they show, it means literally in this case.? A waveform starting at 3.6 instead of 3.5 volts may cause failure.
Much more to follow!
Peter
|
Re: HP8753ES versus HP8753D
Thank you pianovt. I had seen the comparison on that web page at
some point and lost track of it. Just what I was looking for!
Steve, K8JQ
On 3/21/2019 7:01 PM, pianovt via
Groups.Io wrote:
toggle quoted message
Show quoted text
|
FWIW I bought an 8648C w/ opt 1E5 for under $1k delivered. Very nice clean instrument which should meet most or all of my needs.
I also received an email from AD today which might be relevant. They have a new part in volume production, the ADF4371. I'm sure there is a significant amount of programming required to turn this into a bench instrument. But the eval board set is under $1K and the specs are *very* impressive. It's clearly worth a close look if you want to go above 4 GHz.
I have been stunned at what you can buy now in the form of an OEM eval board. If you require a broad range of capability, a commercial instrument is generally a better choice. But if you need to do a narrow range of tasks and are competent writing software You can get the same performance for 10% or less of the price. That's pretty cool in my book.
Reg
|
Re: Looking to connect with other 4396B owners
Hello,
well, I am now semi-retired - maybe, if I find another
opportunity when some surplus store closes, I will snipe it.
Until then, I can darkly remember that the manual for the 41951A
contains a very blurry shot. Maybe this can help you...in
principle, I am all game to help out.
Tam
With best regards
Tam Hanna
---
Enjoy electronics? Join 13700 other followers by visiting the Crazy Electronics Lab at
On 21. 3. 2019 23:34, markhaun2000
wrote:
toggle quoted message
Show quoted text
2) nope. But I was told that it uses some kind of
transformer, so the block diagram in the service manual is
deceptive. I furthermore heard RUMORS of the one for the 4195A
working on the 4953A if you slinked around with the cables.
Sadly, this is all I know - so far, all managed to elude me.
Yes, I believe the general shape of the circuit is known or can be
deduced from available literature.? The best summary I have found
is here:
?
?(about the 4291A RF impedance analyzer---similar vintage from the
same design group in Japan)
But it would be nice to see the innards.? The difficult aspects
will be those which cannot be represented in a schematic diagram.
|
Re: NVRAM 28-Pin Encapsulated Package DS1230Y and 6811B
The DS1230Y utilizes an encapsulated internal battery.? The battery is disconnected until external power is applied breaking the "freshness seal" and connecting the battery to the RAM.? So with a new device you are dealing with battery shelf life until it is powered up.? And I assume that nearly all if not all device programmers will supply power to break the seal and allow you to program it and remove it from the socket without losing the newly programmed contents.
Good reading:
Some of the above data reflect older dates.
I see that Digi-Key still handles large quantities of the DS1230Y:
As well as does Mouser:
And others out there.
Greg
|
Re: HP8753ES versus HP8753D
https://www.kirkbymicrowave.co.uk/Everything-you-wanted-to-know-about-the-HP-8753-VNA/
|
Re: Looking to connect with other 4396B owners
2) nope. But I was told that it uses some kind of transformer, so the block diagram in the service manual is deceptive. I furthermore heard RUMORS of the one for the 4195A working on the 4953A if you slinked around with the cables. Sadly, this is all I know - so far, all managed to elude me.
Yes, I believe the general shape of the circuit is known or can be deduced from available literature.? The best summary I have found is here: ? ?(about the 4291A RF impedance analyzer---similar vintage from the same design group in Japan) But it would be nice to see the innards.? The difficult aspects will be those which cannot be represented in a schematic diagram.
|
Re: Looking to connect with other 4396B owners
Hello,
had a lab accident yesterday, so still a bit stressed.
ad 2) nope. But I was told that it uses some kind of transformer,
so the block diagram in the service manual is deceptive. I
furthermore heard RUMORS of the one for the 4195A working on the
4953A if you slinked around with the cables. Sadly, this is all I
know - so far, all managed to elude me.
ad 3) NI is the way to go. Buy a PCIe card used, cry about the
price (around 200EUR), install Visual Studio and smile.
With best regards
Tam Hanna
---
Enjoy electronics? Join 13700 other followers by visiting the Crazy Electronics Lab at
On 21. 3. 2019 20:48, markhaun2000
wrote:
toggle quoted message
Show quoted text
This is my first post to the group, so hello everyone.? You've
been a great resource when I've needed to search for information
on prospective or actual TE purchases; thanks :)
I recently acquired an Agilent 4396B spectrum/network/impedance
analyzer with all of the options (1D5 oven, 1D6 time-gated
measurements, 010 impedance measurement) in non-working
condition.? The main problem was some stressed SMD electrolytics
on the A50 DC-DC converter board, as discussed in service bulletin
11.? There is also something odd in the source assembly, but it
seems to be after the ALC power detector; current theory is the
semi-rigid coax link from the output attenuator to the bulkhead N
connector, or perhaps the N connector itself.? I'm documenting the
repair here:
Looking ahead I have a few questions (and will undoubtedly have
more as I dig into the analyzer's capabilities):
1) The calibration constants are stored in an EEPROM and updated
over GPIB, but as far as I can tell the actual commands and format
of the data are not documented.? The only approved way to update
the calibration is to use an HP-UX program called "ADJ4396B"
running on an ancient HP controller.? Is this likely to be a
problem in the long run?? Should I consider sending the instrument
out for calibration (assuming the cal lab even has access to this
magic software)?? This is for home / hobby use only, so it likely
doesn't matter, but I wonder if a window of opportunity is closing
as these units get older and the last support is dropped.
2) Has anyone actually laid eyes (or camera) on the insides of the
43961A impedance test fixture?? I have no need for it except
curiosity, but it would be fun to leverage the 010 software in my
unit.? It could probably be a worthwhile learning experience to
build something equivalent at lower frequencies.
3) Not a 4396B question per se, but what is the currently
recommended GPIB interface solution in 2019?? I would prefer not
to use any NI stuff or the VISA library, but if that's the price
for being able to run useful third-party software (like KE5FX's
phase-noise tool), I will.? Although in that case I would still
prefer options which *can* work with the VISA library but can also
work without them.? Cross-platform compatibility with Linux and
Windows would be a plus.
Regards,
Mark
|
Re: Understanding the 8753ES frequency loop
All good info.
In this case the output frequency is consistent with the applied current so that is what I¡¯m going to investigate first as it is higher than it should be.
Peter
toggle quoted message
Show quoted text
On Mar 21, 2019, at 5:48 PM, Dave_G0WBX via Groups.Io <g8kbvdave@...> wrote:
Be careful "bench testing" Yig Tuned Ocillators" (YIG = Yttrium Iron Garnet by the way *.)
Older designs that use + and - polarity supplies can be destroyed if you get the power up sequence wrong, the little FET that "is" the oscillator dies. Similar can happen (or at best some damage that makes them produce less signal power) if they are run without a RF load of some sort.
*
Plus Wikipedia has some good references too.
From past experience working professionally servicing Spectrum Analysers to component level that used YIG Oscillators, check the power supply rails that feed the coil drivers, also any degaussing scheme, if such is fitted.
A sudden and permanent step change in the current/frequency relationship, and also a change in the apparent dynamics of the coil driver (stability issues) can be an indication of shorted turns within the tuning coil. If, there are any documented figures for the tuning coil inductance, that can of course be checked with a LCR bridge, at low power.
Take care..
Dave B (G0WBX)
-- Created on and sent from a Unix like PC running and using free and open source software:
|
Re: Looking to connect with other 4396B owners
At 2019-03-21 02:48 PM, markhaun2000 wrote:
3) Not a 4396B question per se,
but what is the currently recommended GPIB interface solution in
2019?? I would prefer not to use any NI stuff or the VISA library,
but if that's the price for being able to run useful third-party software
(like KE5FX's phase-noise tool), I will.? Although in that case I
would still prefer options which *can* work with the VISA library but can
also work without them.? Cross-platform compatibility with Linux and
Windows would be a plus.
I'll take that as "bait" for a semi--commercial plug for my
KISS-488. It looks like a GPIB plug (actually it's contained inside the
backshell) but has an Ethernet jack on the back end, providing a web
server for your instrument. It can capture HPGL plots, BMPs, or any file
format understood by both the instrument and your browser. Works with any
web browser whether hosted on Windows, Linux, Apple, or your phone. And
it also offers a Telnet interface that allows you to talk directly to the
instrument's command interface. It's listed on eBay right now, but I can
knock a little off the price if you buy it direct and I don't have to pay
eBay fees. I designed it to meet my own need, and I'm selling copies at
just a little over my out-of-pocket costs (doggone GPIB connectors are
EXPENSIVE! as are PCBs when made in low volume). Contact me offline at
SteveHx at HxEngineering dot com for more info without cluttering up the
group's bandwidth.
Steve Hendrix
|
Re: Understanding the 8753ES frequency loop
Be careful "bench testing" Yig Tuned Ocillators" (YIG = Yttrium Iron Garnet by the way *.)
Older designs that use + and - polarity supplies can be destroyed if you get the power up sequence wrong, the little FET that "is" the oscillator dies.? Similar can happen (or at best some damage that makes them produce less signal power) if they are run without a RF load of some sort.
*
Plus Wikipedia has some good references too.
From past experience working professionally servicing Spectrum Analysers to component level that used YIG Oscillators,? check the power supply rails that feed the coil drivers, also any degaussing scheme, if such is fitted.
A sudden and permanent step change in the current/frequency relationship, and also a change in the apparent dynamics of the coil driver (stability issues) can be an indication of shorted turns within the tuning coil.? If, there are any documented figures for the tuning coil inductance, that can of course be checked with a LCR bridge, at low power.
Take care..
Dave B (G0WBX)
-- Created on and sent from a Unix like PC running and using free and open source software:
|
What are main differences between the HP8753D and the HP8753ES? I could maybe figure this out from old HP catalogs but if someone on the list knows I would appreciate your insight.
Thanks.
Steve, K8JQ
|
Re: Understanding the 8753ES frequency loop
On Thu, Mar 21, 2019 at 01:29 PM, Peter Gottlieb wrote:
Should the FM input be zero?
Yes, the numbers I gave for the main coil are strictly that - FM coil disconnected. Incidentally, the main coil has a resistance of approximately 44 Ohms at room temperature. That will give you a way to calculate the voltage across the coil for a static case. Vladan
|
Re: I'm officially lost now with this 8753ES
...who, if they had any influence on file formats and compression algorithms, indicates a serious problem within the company.
Not an uncommon problem, by any stretch, but a serious one.
-Dave
toggle quoted message
Show quoted text
On 3/21/19 4:54 PM, bownes wrote: Yeah, but you have factor in the suits.
On Thu, Mar 21, 2019 at 4:46 PM Dave McGuire <mcguire@... <mailto:mcguire@...>> wrote:
? Pretty sure the 8753D is newer than that, but even if it isn't, JPG was *never* suitable for that sort of imaging.? A good choice (perhaps the only "right" choice) in 1995 would have been GIF.? Heck, even the awful BMP would've been better than JPG!
? ? ? ? ? ? ?-Dave
On 3/21/19 4:44 PM, W2HX wrote: > I don't know the age of these VNAs but PNG was only invented in 1995. Probably needed a few more years before they became popular such that an engineer would consider using it. Maybe that explains it? > ________________________________________ > From: [email protected] <mailto:[email protected]> <[email protected] <mailto:[email protected]>> on behalf of Dave McGuire <mcguire@... <mailto:mcguire@...>> > Sent: Thursday, March 21, 2019 4:36 PM > To: [email protected] <mailto:[email protected]> > Subject: Re: [HP-Agilent-Keysight-equipment] I'm officially lost now with this 8753ES > > On 3/18/19 3:32 PM, pianovt via Groups.Io wrote: >> Yes, PNG is great for screen shots. The problem with the 8753D/ES >> graphics is that the firmware guys didn't do the best job. They added a >> graphics dump, but they only implemented jpg. To make it worse, you >> can't download it via GPIB, it has to be stored to a floppy. At least, I >> don't see a command to download the JPG via GPIB. If anyone here knows >> how, please let me know. > >? ?Wow, I can't imagine that any engineer who was able to get hired by HP > wouldn't understand why JPG isn't appropriate for screen shots or other > line art. (or anything else that isn't a continuous-tone natural scene, > which is what JPG was designed for, and the only thing for which it > works well) > >? ? ? ? ? ? ? -Dave > > -- > Dave McGuire, AK4HZ > New Kensington, PA > > > > > >
-- Dave McGuire, AK4HZ New Kensington, PA
-- Dave McGuire, AK4HZ New Kensington, PA
|
Re: I'm officially lost now with this 8753ES
Yeah, but you have factor in the suits.
toggle quoted message
Show quoted text
On Thu, Mar 21, 2019 at 4:46 PM Dave McGuire < mcguire@...> wrote:
? Pretty sure the 8753D is newer than that, but even if it isn't, JPG
was *never* suitable for that sort of imaging.? A good choice (perhaps
the only "right" choice) in 1995 would have been GIF.? Heck, even the
awful BMP would've been better than JPG!
? ? ? ? ? ? ?-Dave
On 3/21/19 4:44 PM, W2HX wrote:
> I don't know the age of these VNAs but PNG was only invented in 1995. Probably needed a few more years before they became popular such that an engineer would consider using it. Maybe that explains it?
> ________________________________________
> From: [email protected] <[email protected]> on behalf of Dave McGuire <mcguire@...>
> Sent: Thursday, March 21, 2019 4:36 PM
> To: [email protected]
> Subject: Re: [HP-Agilent-Keysight-equipment] I'm officially lost now with this 8753ES
>
> On 3/18/19 3:32 PM, pianovt via Groups.Io wrote:
>> Yes, PNG is great for screen shots. The problem with the 8753D/ES
>> graphics is that the firmware guys didn't do the best job. They added a
>> graphics dump, but they only implemented jpg. To make it worse, you
>> can't download it via GPIB, it has to be stored to a floppy. At least, I
>> don't see a command to download the JPG via GPIB. If anyone here knows
>> how, please let me know.
>
>? ?Wow, I can't imagine that any engineer who was able to get hired by HP
> wouldn't understand why JPG isn't appropriate for screen shots or other
> line art. (or anything else that isn't a continuous-tone natural scene,
> which is what JPG was designed for, and the only thing for which it
> works well)
>
>? ? ? ? ? ? ? -Dave
>
> --
> Dave McGuire, AK4HZ
> New Kensington, PA
>
>
>
>
>
>
--
Dave McGuire, AK4HZ
New Kensington, PA
|
Re: NVRAM 28-Pin Encapsulated Package DS1230Y and 6811B
I have a mess of these I got new a few?years ago when I repaired some 2075 eatons.? ?Might still have some around.?
Pete wa2odo
toggle quoted message
Show quoted text
On Thu, Mar 21, 2019, 4:47 PM Reginald Beardsley via Groups.Io <pulaskite= [email protected]> wrote: Battery life on the early Dallas parts was 5-7 years.? I had to replace one in a 5-6 year old Sun workstation.? Dallas stopped making them and for a while Maxim made them.? At present the only NVRAM source appears to be ST Microelectronics.
I am waiting for a pair of chips to arrive from the UK as there is no stock in NA.? They are backordered until June.? I was able to get a pair for my Tek 11801 courtesy of a member of the TekScopes list.? RS sold out 34 chips in 8-10 days while I was waiting for my 11801 to arrive.? I got the last two.
The only demand for these is repairing old instruments, so ST appears to make a small run once a year.? The current ST chips use a lithium cell, so shelf life should be reasonably good.? But the cells are very low capacity (40-120 mAh) and can easily be drained if inserted into conductive foam. They are widely available on eBay, but the freshest I could find were US NOS parts of 2002 manufacture.
Reg
|
Re: NVRAM 28-Pin Encapsulated Package DS1230Y and 6811B
Battery life on the early Dallas parts was 5-7 years. I had to replace one in a 5-6 year old Sun workstation. Dallas stopped making them and for a while Maxim made them. At present the only NVRAM source appears to be ST Microelectronics.
I am waiting for a pair of chips to arrive from the UK as there is no stock in NA. They are backordered until June. I was able to get a pair for my Tek 11801 courtesy of a member of the TekScopes list. RS sold out 34 chips in 8-10 days while I was waiting for my 11801 to arrive. I got the last two.
The only demand for these is repairing old instruments, so ST appears to make a small run once a year. The current ST chips use a lithium cell, so shelf life should be reasonably good. But the cells are very low capacity (40-120 mAh) and can easily be drained if inserted into conductive foam. They are widely available on eBay, but the freshest I could find were US NOS parts of 2002 manufacture.
Reg
|