Re: Looking to connect with other 4396B owners
Hello,
had a lab accident yesterday, so still a bit stressed.
ad 2) nope. But I was told that it uses some kind of transformer,
so the block diagram in the service manual is deceptive. I
furthermore heard RUMORS of the one for the 4195A working on the
4953A if you slinked around with the cables. Sadly, this is all I
know - so far, all managed to elude me.
ad 3) NI is the way to go. Buy a PCIe card used, cry about the
price (around 200EUR), install Visual Studio and smile.
With best regards
Tam Hanna
---
Enjoy electronics? Join 13700 other followers by visiting the Crazy Electronics Lab at
On 21. 3. 2019 20:48, markhaun2000
wrote:
toggle quoted message
Show quoted text
This is my first post to the group, so hello everyone.? You've
been a great resource when I've needed to search for information
on prospective or actual TE purchases; thanks :)
I recently acquired an Agilent 4396B spectrum/network/impedance
analyzer with all of the options (1D5 oven, 1D6 time-gated
measurements, 010 impedance measurement) in non-working
condition.? The main problem was some stressed SMD electrolytics
on the A50 DC-DC converter board, as discussed in service bulletin
11.? There is also something odd in the source assembly, but it
seems to be after the ALC power detector; current theory is the
semi-rigid coax link from the output attenuator to the bulkhead N
connector, or perhaps the N connector itself.? I'm documenting the
repair here:
Looking ahead I have a few questions (and will undoubtedly have
more as I dig into the analyzer's capabilities):
1) The calibration constants are stored in an EEPROM and updated
over GPIB, but as far as I can tell the actual commands and format
of the data are not documented.? The only approved way to update
the calibration is to use an HP-UX program called "ADJ4396B"
running on an ancient HP controller.? Is this likely to be a
problem in the long run?? Should I consider sending the instrument
out for calibration (assuming the cal lab even has access to this
magic software)?? This is for home / hobby use only, so it likely
doesn't matter, but I wonder if a window of opportunity is closing
as these units get older and the last support is dropped.
2) Has anyone actually laid eyes (or camera) on the insides of the
43961A impedance test fixture?? I have no need for it except
curiosity, but it would be fun to leverage the 010 software in my
unit.? It could probably be a worthwhile learning experience to
build something equivalent at lower frequencies.
3) Not a 4396B question per se, but what is the currently
recommended GPIB interface solution in 2019?? I would prefer not
to use any NI stuff or the VISA library, but if that's the price
for being able to run useful third-party software (like KE5FX's
phase-noise tool), I will.? Although in that case I would still
prefer options which *can* work with the VISA library but can also
work without them.? Cross-platform compatibility with Linux and
Windows would be a plus.
Regards,
Mark
|
Re: Understanding the 8753ES frequency loop
All good info.
In this case the output frequency is consistent with the applied current so that is what I¡¯m going to investigate first as it is higher than it should be.
Peter
toggle quoted message
Show quoted text
On Mar 21, 2019, at 5:48 PM, Dave_G0WBX via Groups.Io <g8kbvdave@...> wrote:
Be careful "bench testing" Yig Tuned Ocillators" (YIG = Yttrium Iron Garnet by the way *.)
Older designs that use + and - polarity supplies can be destroyed if you get the power up sequence wrong, the little FET that "is" the oscillator dies. Similar can happen (or at best some damage that makes them produce less signal power) if they are run without a RF load of some sort.
*
Plus Wikipedia has some good references too.
From past experience working professionally servicing Spectrum Analysers to component level that used YIG Oscillators, check the power supply rails that feed the coil drivers, also any degaussing scheme, if such is fitted.
A sudden and permanent step change in the current/frequency relationship, and also a change in the apparent dynamics of the coil driver (stability issues) can be an indication of shorted turns within the tuning coil. If, there are any documented figures for the tuning coil inductance, that can of course be checked with a LCR bridge, at low power.
Take care..
Dave B (G0WBX)
-- Created on and sent from a Unix like PC running and using free and open source software:
|
Re: Looking to connect with other 4396B owners
At 2019-03-21 02:48 PM, markhaun2000 wrote:
3) Not a 4396B question per se,
but what is the currently recommended GPIB interface solution in
2019?? I would prefer not to use any NI stuff or the VISA library,
but if that's the price for being able to run useful third-party software
(like KE5FX's phase-noise tool), I will.? Although in that case I
would still prefer options which *can* work with the VISA library but can
also work without them.? Cross-platform compatibility with Linux and
Windows would be a plus.
I'll take that as "bait" for a semi--commercial plug for my
KISS-488. It looks like a GPIB plug (actually it's contained inside the
backshell) but has an Ethernet jack on the back end, providing a web
server for your instrument. It can capture HPGL plots, BMPs, or any file
format understood by both the instrument and your browser. Works with any
web browser whether hosted on Windows, Linux, Apple, or your phone. And
it also offers a Telnet interface that allows you to talk directly to the
instrument's command interface. It's listed on eBay right now, but I can
knock a little off the price if you buy it direct and I don't have to pay
eBay fees. I designed it to meet my own need, and I'm selling copies at
just a little over my out-of-pocket costs (doggone GPIB connectors are
EXPENSIVE! as are PCBs when made in low volume). Contact me offline at
SteveHx at HxEngineering dot com for more info without cluttering up the
group's bandwidth.
Steve Hendrix
|
Re: Understanding the 8753ES frequency loop
Be careful "bench testing" Yig Tuned Ocillators" (YIG = Yttrium Iron Garnet by the way *.)
Older designs that use + and - polarity supplies can be destroyed if you get the power up sequence wrong, the little FET that "is" the oscillator dies.? Similar can happen (or at best some damage that makes them produce less signal power) if they are run without a RF load of some sort.
*
Plus Wikipedia has some good references too.
From past experience working professionally servicing Spectrum Analysers to component level that used YIG Oscillators,? check the power supply rails that feed the coil drivers, also any degaussing scheme, if such is fitted.
A sudden and permanent step change in the current/frequency relationship, and also a change in the apparent dynamics of the coil driver (stability issues) can be an indication of shorted turns within the tuning coil.? If, there are any documented figures for the tuning coil inductance, that can of course be checked with a LCR bridge, at low power.
Take care..
Dave B (G0WBX)
-- Created on and sent from a Unix like PC running and using free and open source software:
|
What are main differences between the HP8753D and the HP8753ES? I could maybe figure this out from old HP catalogs but if someone on the list knows I would appreciate your insight.
Thanks.
Steve, K8JQ
|
Re: Understanding the 8753ES frequency loop
On Thu, Mar 21, 2019 at 01:29 PM, Peter Gottlieb wrote:
Should the FM input be zero?
Yes, the numbers I gave for the main coil are strictly that - FM coil disconnected. Incidentally, the main coil has a resistance of approximately 44 Ohms at room temperature. That will give you a way to calculate the voltage across the coil for a static case. Vladan
|
Re: I'm officially lost now with this 8753ES
...who, if they had any influence on file formats and compression algorithms, indicates a serious problem within the company.
Not an uncommon problem, by any stretch, but a serious one.
-Dave
toggle quoted message
Show quoted text
On 3/21/19 4:54 PM, bownes wrote: Yeah, but you have factor in the suits.
On Thu, Mar 21, 2019 at 4:46 PM Dave McGuire <mcguire@... <mailto:mcguire@...>> wrote:
? Pretty sure the 8753D is newer than that, but even if it isn't, JPG was *never* suitable for that sort of imaging.? A good choice (perhaps the only "right" choice) in 1995 would have been GIF.? Heck, even the awful BMP would've been better than JPG!
? ? ? ? ? ? ?-Dave
On 3/21/19 4:44 PM, W2HX wrote: > I don't know the age of these VNAs but PNG was only invented in 1995. Probably needed a few more years before they became popular such that an engineer would consider using it. Maybe that explains it? > ________________________________________ > From: [email protected] <mailto:[email protected]> <[email protected] <mailto:[email protected]>> on behalf of Dave McGuire <mcguire@... <mailto:mcguire@...>> > Sent: Thursday, March 21, 2019 4:36 PM > To: [email protected] <mailto:[email protected]> > Subject: Re: [HP-Agilent-Keysight-equipment] I'm officially lost now with this 8753ES > > On 3/18/19 3:32 PM, pianovt via Groups.Io wrote: >> Yes, PNG is great for screen shots. The problem with the 8753D/ES >> graphics is that the firmware guys didn't do the best job. They added a >> graphics dump, but they only implemented jpg. To make it worse, you >> can't download it via GPIB, it has to be stored to a floppy. At least, I >> don't see a command to download the JPG via GPIB. If anyone here knows >> how, please let me know. > >? ?Wow, I can't imagine that any engineer who was able to get hired by HP > wouldn't understand why JPG isn't appropriate for screen shots or other > line art. (or anything else that isn't a continuous-tone natural scene, > which is what JPG was designed for, and the only thing for which it > works well) > >? ? ? ? ? ? ? -Dave > > -- > Dave McGuire, AK4HZ > New Kensington, PA > > > > > >
-- Dave McGuire, AK4HZ New Kensington, PA
-- Dave McGuire, AK4HZ New Kensington, PA
|
Re: I'm officially lost now with this 8753ES
Yeah, but you have factor in the suits.
toggle quoted message
Show quoted text
On Thu, Mar 21, 2019 at 4:46 PM Dave McGuire < mcguire@...> wrote:
? Pretty sure the 8753D is newer than that, but even if it isn't, JPG
was *never* suitable for that sort of imaging.? A good choice (perhaps
the only "right" choice) in 1995 would have been GIF.? Heck, even the
awful BMP would've been better than JPG!
? ? ? ? ? ? ?-Dave
On 3/21/19 4:44 PM, W2HX wrote:
> I don't know the age of these VNAs but PNG was only invented in 1995. Probably needed a few more years before they became popular such that an engineer would consider using it. Maybe that explains it?
> ________________________________________
> From: [email protected] <[email protected]> on behalf of Dave McGuire <mcguire@...>
> Sent: Thursday, March 21, 2019 4:36 PM
> To: [email protected]
> Subject: Re: [HP-Agilent-Keysight-equipment] I'm officially lost now with this 8753ES
>
> On 3/18/19 3:32 PM, pianovt via Groups.Io wrote:
>> Yes, PNG is great for screen shots. The problem with the 8753D/ES
>> graphics is that the firmware guys didn't do the best job. They added a
>> graphics dump, but they only implemented jpg. To make it worse, you
>> can't download it via GPIB, it has to be stored to a floppy. At least, I
>> don't see a command to download the JPG via GPIB. If anyone here knows
>> how, please let me know.
>
>? ?Wow, I can't imagine that any engineer who was able to get hired by HP
> wouldn't understand why JPG isn't appropriate for screen shots or other
> line art. (or anything else that isn't a continuous-tone natural scene,
> which is what JPG was designed for, and the only thing for which it
> works well)
>
>? ? ? ? ? ? ? -Dave
>
> --
> Dave McGuire, AK4HZ
> New Kensington, PA
>
>
>
>
>
>
--
Dave McGuire, AK4HZ
New Kensington, PA
|
Re: NVRAM 28-Pin Encapsulated Package DS1230Y and 6811B
I have a mess of these I got new a few?years ago when I repaired some 2075 eatons.? ?Might still have some around.?
Pete wa2odo
toggle quoted message
Show quoted text
On Thu, Mar 21, 2019, 4:47 PM Reginald Beardsley via Groups.Io <pulaskite= [email protected]> wrote: Battery life on the early Dallas parts was 5-7 years.? I had to replace one in a 5-6 year old Sun workstation.? Dallas stopped making them and for a while Maxim made them.? At present the only NVRAM source appears to be ST Microelectronics.
I am waiting for a pair of chips to arrive from the UK as there is no stock in NA.? They are backordered until June.? I was able to get a pair for my Tek 11801 courtesy of a member of the TekScopes list.? RS sold out 34 chips in 8-10 days while I was waiting for my 11801 to arrive.? I got the last two.
The only demand for these is repairing old instruments, so ST appears to make a small run once a year.? The current ST chips use a lithium cell, so shelf life should be reasonably good.? But the cells are very low capacity (40-120 mAh) and can easily be drained if inserted into conductive foam. They are widely available on eBay, but the freshest I could find were US NOS parts of 2002 manufacture.
Reg
|
Re: NVRAM 28-Pin Encapsulated Package DS1230Y and 6811B
Battery life on the early Dallas parts was 5-7 years. I had to replace one in a 5-6 year old Sun workstation. Dallas stopped making them and for a while Maxim made them. At present the only NVRAM source appears to be ST Microelectronics.
I am waiting for a pair of chips to arrive from the UK as there is no stock in NA. They are backordered until June. I was able to get a pair for my Tek 11801 courtesy of a member of the TekScopes list. RS sold out 34 chips in 8-10 days while I was waiting for my 11801 to arrive. I got the last two.
The only demand for these is repairing old instruments, so ST appears to make a small run once a year. The current ST chips use a lithium cell, so shelf life should be reasonably good. But the cells are very low capacity (40-120 mAh) and can easily be drained if inserted into conductive foam. They are widely available on eBay, but the freshest I could find were US NOS parts of 2002 manufacture.
Reg
|
Re: I'm officially lost now with this 8753ES
I was thinking along the same lines. Barely even heard of PNG initially, it might have been considered risky to use a format which might have died out.
Peter
toggle quoted message
Show quoted text
On Mar 21, 2019, at 4:44 PM, W2HX <w2hx@...> wrote:
I don't know the age of these VNAs but PNG was only invented in 1995. Probably needed a few more years before they became popular such that an engineer would consider using it. Maybe that explains it? ________________________________________ From: [email protected] <[email protected]> on behalf of Dave McGuire <mcguire@...> Sent: Thursday, March 21, 2019 4:36 PM To: [email protected] Subject: Re: [HP-Agilent-Keysight-equipment] I'm officially lost now with this 8753ES
On 3/18/19 3:32 PM, pianovt via Groups.Io wrote: Yes, PNG is great for screen shots. The problem with the 8753D/ES graphics is that the firmware guys didn't do the best job. They added a graphics dump, but they only implemented jpg. To make it worse, you can't download it via GPIB, it has to be stored to a floppy. At least, I don't see a command to download the JPG via GPIB. If anyone here knows how, please let me know. Wow, I can't imagine that any engineer who was able to get hired by HP wouldn't understand why JPG isn't appropriate for screen shots or other line art. (or anything else that isn't a continuous-tone natural scene, which is what JPG was designed for, and the only thing for which it works well)
-Dave
-- Dave McGuire, AK4HZ New Kensington, PA
|
Re: I'm officially lost now with this 8753ES
Pretty sure the 8753D is newer than that, but even if it isn't, JPG was *never* suitable for that sort of imaging. A good choice (perhaps the only "right" choice) in 1995 would have been GIF. Heck, even the awful BMP would've been better than JPG!
-Dave
toggle quoted message
Show quoted text
On 3/21/19 4:44 PM, W2HX wrote: I don't know the age of these VNAs but PNG was only invented in 1995. Probably needed a few more years before they became popular such that an engineer would consider using it. Maybe that explains it? ________________________________________ From: [email protected] <[email protected]> on behalf of Dave McGuire <mcguire@...> Sent: Thursday, March 21, 2019 4:36 PM To: [email protected] Subject: Re: [HP-Agilent-Keysight-equipment] I'm officially lost now with this 8753ES
On 3/18/19 3:32 PM, pianovt via Groups.Io wrote:
Yes, PNG is great for screen shots. The problem with the 8753D/ES graphics is that the firmware guys didn't do the best job. They added a graphics dump, but they only implemented jpg. To make it worse, you can't download it via GPIB, it has to be stored to a floppy. At least, I don't see a command to download the JPG via GPIB. If anyone here knows how, please let me know. Wow, I can't imagine that any engineer who was able to get hired by HP wouldn't understand why JPG isn't appropriate for screen shots or other line art. (or anything else that isn't a continuous-tone natural scene, which is what JPG was designed for, and the only thing for which it works well)
-Dave
-- Dave McGuire, AK4HZ New Kensington, PA
-- Dave McGuire, AK4HZ New Kensington, PA
|
Re: I'm officially lost now with this 8753ES
I don't know the age of these VNAs but PNG was only invented in 1995. Probably needed a few more years before they became popular such that an engineer would consider using it. Maybe that explains it? ________________________________________ From: [email protected] < [email protected]> on behalf of Dave McGuire <mcguire@...> Sent: Thursday, March 21, 2019 4:36 PM To: [email protected]Subject: Re: [HP-Agilent-Keysight-equipment] I'm officially lost now with this 8753ES On 3/18/19 3:32 PM, pianovt via Groups.Io wrote: Yes, PNG is great for screen shots. The problem with the 8753D/ES graphics is that the firmware guys didn't do the best job. They added a graphics dump, but they only implemented jpg. To make it worse, you can't download it via GPIB, it has to be stored to a floppy. At least, I don't see a command to download the JPG via GPIB. If anyone here knows how, please let me know. Wow, I can't imagine that any engineer who was able to get hired by HP wouldn't understand why JPG isn't appropriate for screen shots or other line art. (or anything else that isn't a continuous-tone natural scene, which is what JPG was designed for, and the only thing for which it works well) -Dave -- Dave McGuire, AK4HZ New Kensington, PA
|
Re: Understanding the 8753ES frequency loop
This is beginning to line up.?
3.5 volts at 300 kHz, 3.5 volts across a 40 ohm resistor = 87.5 mA, 87.5 mA divided by 23 mA/GHz = 3.804 GHz, Right in the correct 1-6 MHz high stated range.?
toggle quoted message
Show quoted text
On Mar 21, 2019, at 4:17 PM, pianovt via Groups.Io < pianovt@...> wrote: It's a straight line, so pick any frequency you want. There is a non-linearity starting above 5.5 GHz or so. But, you are having problems at the low end, below 1 GHz rf out. That means the YTO is tuned to ~4.8 GHz when the rf output is 1 GHz.. Let's say you pick 4 GHz as a test point. The current should be 4 x 23 = 92 mA. If the actual coil sensitivity is at the upper limit (24 mA/GHz), then 96 mA.
Vladan
|
Re: I'm officially lost now with this 8753ES
On 3/18/19 3:32 PM, pianovt via Groups.Io wrote: Yes, PNG is great for screen shots. The problem with the 8753D/ES graphics is that the firmware guys didn't do the best job. They added a graphics dump, but they only implemented jpg. To make it worse, you can't download it via GPIB, it has to be stored to a floppy. At least, I don't see a command to download the JPG via GPIB. If anyone here knows how, please let me know. Wow, I can't imagine that any engineer who was able to get hired by HP wouldn't understand why JPG isn't appropriate for screen shots or other line art. (or anything else that isn't a continuous-tone natural scene, which is what JPG was designed for, and the only thing for which it works well) -Dave -- Dave McGuire, AK4HZ New Kensington, PA
|
Re: Understanding the 8753ES frequency loop
Ah, so it starts at zero mA, although we are not interested (and likely not specced)below 3.8 GHz. 3.8 GHz output would then correspond to 3.8 * 23 mA on the main coil. Should the FM input be zero? ?I haven¡¯t yet checked that. I wonder how far the FM input can move it if for example one if those transistors are bad.?
This would be an easy bench test of the YTO, or even easier in-instrument.?
toggle quoted message
Show quoted text
On Mar 21, 2019, at 4:17 PM, pianovt via Groups.Io < pianovt@...> wrote: It's a straight line, so pick any frequency you want. There is a non-linearity starting above 5.5 GHz or so. But, you are having problems at the low end, below 1 GHz rf out. That means the YTO is tuned to ~4.8 GHz when the rf output is 1 GHz.. Let's say you pick 4 GHz as a test point. The current should be 4 x 23 = 92 mA. If the actual coil sensitivity is at the upper limit (24 mA/GHz), then 96 mA.
Vladan
|
Re: Understanding the 8753ES frequency loop
It's a straight line, so pick any frequency you want. There is a non-linearity starting above 5.5 GHz or so. But, you are having problems at the low end, below 1 GHz rf out. That means the YTO is tuned to ~4.8 GHz when the rf output is 1 GHz.. Let's say you pick 4 GHz as a test point. The current should be 4 x 23 = 92 mA. If the actual coil sensitivity is at the upper limit (24 mA/GHz), then 96 mA.
Vladan
|
Re: NVRAM 28-Pin Encapsulated Package DS1230Y and 6811B
You are thinking of the Dallas SmartSocket series, and it has no identifying marks on it. It was used with a normal low power CMOS RAM chip.
The OP's wording was somewhat awkward, but he doesn't have a smart socket.
The OP has already said that his is an encapsulated Dallas NVRAM that is installed in a socket.
He is safe to remove it, and read it with his EPROM programmer.
It will read just like a 6116, RAM chip. Different programmers have different ways of reading the Dallas chip.
-Chuck Harris
Alan Melia via Groups.Io wrote:
toggle quoted message
Show quoted text
Maybe not!! some of these Dallas devices have a battery in the socket to maintain the contents. It may not be one of those but do check before removing anything. The "socket" may actually be a header and the whole thing unplugged and read in a programmer as if it was an EPROM. I last came across these in the 1990s using prototyping/development kits. Others are rather "fat" 28-pin (0.6in.) IC shapes and have an internal battery I think.
Alan G3NYK
|
Looking to connect with other 4396B owners
This is my first post to the group, so hello everyone.? You've been a great resource when I've needed to search for information on prospective or actual TE purchases; thanks :)
I recently acquired an Agilent 4396B spectrum/network/impedance analyzer with all of the options (1D5 oven, 1D6 time-gated measurements, 010 impedance measurement) in non-working condition.? The main problem was some stressed SMD electrolytics on the A50 DC-DC converter board, as discussed in service bulletin 11.? There is also something odd in the source assembly, but it seems to be after the ALC power detector; current theory is the semi-rigid coax link from the output attenuator to the bulkhead N connector, or perhaps the N connector itself.? I'm documenting the repair here:
Looking ahead I have a few questions (and will undoubtedly have more as I dig into the analyzer's capabilities):
1) The calibration constants are stored in an EEPROM and updated over GPIB, but as far as I can tell the actual commands and format of the data are not documented.? The only approved way to update the calibration is to use an HP-UX program called "ADJ4396B" running on an ancient HP controller.? Is this likely to be a problem in the long run?? Should I consider sending the instrument out for calibration (assuming the cal lab even has access to this magic software)?? This is for home / hobby use only, so it likely doesn't matter, but I wonder if a window of opportunity is closing as these units get older and the last support is dropped.
2) Has anyone actually laid eyes (or camera) on the insides of the 43961A impedance test fixture?? I have no need for it except curiosity, but it would be fun to leverage the 010 software in my unit.? It could probably be a worthwhile learning experience to build something equivalent at lower frequencies.
3) Not a 4396B question per se, but what is the currently recommended GPIB interface solution in 2019?? I would prefer not to use any NI stuff or the VISA library, but if that's the price for being able to run useful third-party software (like KE5FX's phase-noise tool), I will.? Although in that case I would still prefer options which *can* work with the VISA library but can also work without them.? Cross-platform compatibility with Linux and Windows would be a plus.
Regards, Mark
|
Re: NVRAM 28-Pin Encapsulated Package DS1230Y and 6811B
Maybe not!! some of these Dallas devices have a battery in the socket to maintain the contents. It may not be one of those but do check before removing anything. The "socket" may actually be a header and the whole thing unplugged and read in a programmer as if it was an EPROM. I last came across these in the 1990s using prototyping/development kits. Others are rather "fat" 28-pin (0.6in.) IC shapes and have an internal battery I think.
Alan G3NYK
|