¿ªÔÆÌåÓý

Date

E5500 licence

 

I have bought a 70420A phase noise test set - the heart of E5500 phase noise measurement system and used it with HP3561A and, later, with HP89441A samplers.?
It needs a PC software package to be anywhere close to being useful.

Agilent E5500 software is available on Keysight website but requires a key tied to the serial number of the test set.
I have been advised that Keysight is kind enough to resend it to new owners of old E5500 equipment but few emails requesting the key went unanswered.
Do you have to make up a compassionate story of why you don't have the key or has Keysight simply finally stopped supporting these?

In the end I have simply programmed somebody else's S/N into my kit and am using their licence key but I am wondering whether I did not try hard enough with Keysight tech support?

Leo


Re: 5385A error 52

 

Reading the manual of the counter , I verified the command to get te ID is "ID" or "SI".
The instruments doesn't support the SCPI standard command syntax .
On Agilent I/O there is the possibility to disable ,during the serching instruments , the *IDN? command.
I not remember if using NI Max there is this facilities
Best regards
Maurizio


Re: 5385A error 52

 

Hi Mark , I suppose teh counter interface was setted on listen mode at first.
You can check if the NI Max is sending *IDN? command during the search for instruments.
May be the counter doesn't support this command.
Best regards

Maurizio


Agilent 16700B/16702B Manuals

 

I am looking for the User manual and also the service guide for the 16700B/16701B/16702B series mainframe. I'm pretty sure I have seen it, but cannot find it now... And I don't seem to be able to find them on the Keysight site...

Any place they are on the Web?

Maurice


Re: I'm officially lost now with this 8753ES

 

¿ªÔÆÌåÓý

Hi I have the original CLIP manual for 8753D and I have done a PDF and uploaded at ?

To download ?

?

Regards

Yves

De?: [email protected] <[email protected]> De la part de Colby Burkett
·¡²Ô±¹´Ç²â¨¦?: 16 mars 2019 21:41
??: [email protected]
Objet?: Re: [HP-Agilent-Keysight-equipment] I'm officially lost now with this 8753ES

?

It's available on eBay listed as 8753D CLIP. I bought mine that way. Artek doesn't seem to have it.

?

Not sure if the eBay seller is licensed to sell it or not, but it's been years since I bought, and it's still listed by the same seller.

?

On Sat, Mar 16, 2019, 6:07 PM Peter Gottlieb <hpnpilot@...> wrote:

Thanks.? I've been working off of the 8753B CLIP, not too much has seemed to
have changed.? What is the part number for A11 in the 8753D?? In my ES it is
08753-60162.

What I'm seeing is that the YTO is over 200 MHz too high on the lower
frequencies (under say 1-1.5 GHz) so the test 48, generate proper pretune
voltages, fails.? The output at P1-1 to P1-2 is somewhat higher at the low end
than the manual shows, being 4.0 volts in mine vs. 3.5 volts stated in the
manual.? U18-6 voltages (not shown on the schematic shown below), out of the
DAC, look good.? The drive signal from A11 is current rather than voltage so a
slight difference could be attributed to tolerance of my YTO coil resistance
(45.2 vs. 44.4 ohms?? Not enough for that difference).? I also tried swapping
the YTO for another and the replacement behaved exactly the same.

The instrument PLL locks nicely when using an external source.

Where did you find the CLIP you sent?? I'd like to see the rest of the A11
board, and maybe more, as it may be closer than the A/B CLIP I have.? And I'd
*love* to see a CLIP for the A3 source PCB!

Peter


On 3/16/2019 8:42 PM, pianovt via Groups.Io wrote:
> Maybe this will be helpful. It's from the 8753D, but should be the same for
> the ES.
>
> Vladan
>
>
>
>




Re: I'm officially lost now with this 8753ES

 

I ordered the CLIPs from the ebay seller who likely got them from Manuals Plus which is now gone.

Nothing is obviously wrong with the VCO drive to the YTO.? I say "obviously" because the waveform has the same shape as it shows in the manual, but yet the voltages are off, most especially at the low frequency end where the manual says the drive should be 3.5 volts it is actually 4.0.? But is this a normal difference or has the PLL card drifted out of spec?

If I get a hold of another A11 card I will know that answer.

So here's the thing:? is it possible something I did in the hybrid module could have caused this?? I would not have thought so as it is operating at high RF frequencies only and not directly controlling the YTO frequency, especially when the PLL loop is opened.

I am thinking I will try and draw out a diagram of the PCB on the source module.

Peter

On 3/16/2019 10:28 PM, pianovt via Groups.Io wrote:
The schematic is for 08753-60162, so it matches what you have. I would monitor the voltage across R43 (the 40 Ohm power resistor) as you tune the YTO. It should be a fairly linear function of frequency with some d.c. offset.

As far as I know, the A3 source board schematic has never been published in a CLIP. I once saw it, and to my eye it didn't look nice enough for official publication. It was probably just good enough to extract the netlist and lay out the board. I suspect they never intended to publish it, so they didn't spend the money on making it look nice.

Vladan


Re: I'm officially lost now with this 8753ES

 

The schematic is for 08753-60162, so it matches what you have. I would monitor the voltage across R43 (the 40 Ohm power resistor) as you tune the YTO. It should be a fairly linear function of frequency with some d.c. offset.

As far as I know, the A3 source board schematic has never been published in a CLIP. I once saw it, and to my eye it didn't look nice enough for official publication. It was probably just good enough to extract the netlist and lay out the board. I suspect they never intended to publish it, so they didn't spend the money on making it look nice.

Vladan


5385A error 52

 

Hi all. So, hooked up my acquired 5385A to my HP-IB (GPIB) network and it crashed! The unit has "Error 52" on the display.
Tried it all by itself and same issue. (This is with NI MAX... search for instruments) Have NEVER seen this before. Even NI MAX
has problems when the counter is connected. Google search reveals that others have problems with LabView but not helpful.

I have NOT tried using BASIC (hp 85 for example) with the unit.

Any advice on this please?

Thank you!

Mark in Oregon
(previous HP employee)

PS - this error indicates "First character of command is illegal"


Re: I'm officially lost now with this 8753ES

 

Thanks, just ordered one.

On 3/16/2019 9:41 PM, Colby Burkett wrote:
It's available on eBay listed as 8753D CLIP. I bought mine that way. Artek doesn't seem to have it.

Not sure if the eBay seller is licensed to sell it or not, but it's been years since I bought, and it's still listed by the same seller.

On Sat, Mar 16, 2019, 6:07 PM Peter Gottlieb <hpnpilot@... <mailto:hpnpilot@...>> wrote:

Thanks. I've been working off of the 8753B CLIP, not too much has seemed to
have changed.? What is the part number for A11 in the 8753D? In my ES it is
08753-60162.

What I'm seeing is that the YTO is over 200 MHz too high on the lower
frequencies (under say 1-1.5 GHz) so the test 48, generate proper pretune
voltages, fails.? The output at P1-1 to P1-2 is somewhat higher at the low
end
than the manual shows, being 4.0 volts in mine vs. 3.5 volts stated in the
manual.? U18-6 voltages (not shown on the schematic shown below), out of the
DAC, look good.? The drive signal from A11 is current rather than voltage
so a
slight difference could be attributed to tolerance of my YTO coil resistance
(45.2 vs. 44.4 ohms?? Not enough for that difference).? I also tried swapping
the YTO for another and the replacement behaved exactly the same.

The instrument PLL locks nicely when using an external source.

Where did you find the CLIP you sent?? I'd like to see the rest of the A11
board, and maybe more, as it may be closer than the A/B CLIP I have.? And I'd
*love* to see a CLIP for the A3 source PCB!

Peter


On 3/16/2019 8:42 PM, pianovt via Groups.Io wrote:
> Maybe this will be helpful. It's from the 8753D, but should be the same for
> the ES.
>
> Vladan
>
>
>
>





Re: I'm officially lost now with this 8753ES

 

It's available on eBay listed as 8753D CLIP. I bought mine that way. Artek doesn't seem to have it.

Not sure if the eBay seller is licensed to sell it or not, but it's been years since I bought, and it's still listed by the same seller.

On Sat, Mar 16, 2019, 6:07 PM Peter Gottlieb <hpnpilot@...> wrote:
Thanks.? I've been working off of the 8753B CLIP, not too much has seemed to
have changed.? What is the part number for A11 in the 8753D?? In my ES it is
08753-60162.

What I'm seeing is that the YTO is over 200 MHz too high on the lower
frequencies (under say 1-1.5 GHz) so the test 48, generate proper pretune
voltages, fails.? The output at P1-1 to P1-2 is somewhat higher at the low end
than the manual shows, being 4.0 volts in mine vs. 3.5 volts stated in the
manual.? U18-6 voltages (not shown on the schematic shown below), out of the
DAC, look good.? The drive signal from A11 is current rather than voltage so a
slight difference could be attributed to tolerance of my YTO coil resistance
(45.2 vs. 44.4 ohms?? Not enough for that difference).? I also tried swapping
the YTO for another and the replacement behaved exactly the same.

The instrument PLL locks nicely when using an external source.

Where did you find the CLIP you sent?? I'd like to see the rest of the A11
board, and maybe more, as it may be closer than the A/B CLIP I have.? And I'd
*love* to see a CLIP for the A3 source PCB!

Peter


On 3/16/2019 8:42 PM, pianovt via Groups.Io wrote:
> Maybe this will be helpful. It's from the 8753D, but should be the same for
> the ES.
>
> Vladan
>
>
>
>





Re: I'm officially lost now with this 8753ES

 

Thanks.? I've been working off of the 8753B CLIP, not too much has seemed to have changed.? What is the part number for A11 in the 8753D?? In my ES it is 08753-60162.

What I'm seeing is that the YTO is over 200 MHz too high on the lower frequencies (under say 1-1.5 GHz) so the test 48, generate proper pretune voltages, fails.? The output at P1-1 to P1-2 is somewhat higher at the low end than the manual shows, being 4.0 volts in mine vs. 3.5 volts stated in the manual.? U18-6 voltages (not shown on the schematic shown below), out of the DAC, look good.? The drive signal from A11 is current rather than voltage so a slight difference could be attributed to tolerance of my YTO coil resistance (45.2 vs. 44.4 ohms?? Not enough for that difference).? I also tried swapping the YTO for another and the replacement behaved exactly the same.

The instrument PLL locks nicely when using an external source.

Where did you find the CLIP you sent?? I'd like to see the rest of the A11 board, and maybe more, as it may be closer than the A/B CLIP I have.? And I'd *love* to see a CLIP for the A3 source PCB!

Peter

On 3/16/2019 8:42 PM, pianovt via Groups.Io wrote:
Maybe this will be helpful. It's from the 8753D, but should be the same for the ES.

Vladan




Re: I'm officially lost now with this 8753ES

 

Maybe this will be helpful. It's from the 8753D, but should be the same for the ES.

Vladan




Re: HP/Keysight/Agilent 34401A DMM power transformer

 

Siggi

Again, thank you. Amazing (to me) the Keysight part is only 10 dollars more than the Chinese eBay replacements. So no contest there....And the video you linked shows the transformer hanging on the side rail of the chassis, so that should be easy to rig.

I will cross my fingers and buy the Keysight part.

Vic


Re: HP/Keysight/Agilent 34401A DMM power transformer

 

On Fri, Mar 15, 2019 at 7:37 AM victor smith <vcs3@...> wrote:
Hello Siggi

Thank you for your quick response!?

Thanks for the info from the service manual. I see that there are a few on eBay (all from China) listed for the 34401A, but the part number is 9100-5608...
From the photo it does look to be a compatible part, at least in terms of the winding terminations. At $70.00 (US) each, I guess it might be worth it if it brings my meter to life! I just hate buying from Chinese vendors (not because of politics, simply because delivery is slow and unpredictable)

Then buy it from Keysight []. The registration process is a bit of a PITA, but the one time I bought a part from them it was Fedexed overnight at zero charge.
?
So I was hoping maybe someone from the group has one for sale (or a junk unit I could cannibalize).

Has anyone replaced one of these? I wonder how the transformer is physically attached to the chassis?

There's a lot of info on the internet, see e.g. here [], or you might want to look at the parts diagram in the service manual.?


Re: HP/Keysight/Agilent 34401A DMM power transformer

 

¿ªÔÆÌåÓý

Some machines use a special transformer with a?shield between primary and secondary. ?The secondary bobbin might even be air-gapped. ?You can get along with an ordinary part but your Guard function may be compromised.


Dave Wise


From: [email protected] <[email protected]> on behalf of saipan59 <[email protected]>
Sent: Friday, March 15, 2019 9:31 AM
To: [email protected]
Subject: Re: [HP-Agilent-Keysight-equipment] HP/Keysight/Agilent 34401A DMM power transformer
?
Suggestion:
Use any xfrmr that is electrically compatible, and mount it outside the DMM inside it's own little metal box - sorta like a "wall-wart" PS setup.

Pete


Re: HP/Keysight/Agilent 34401A DMM power transformer

 

Suggestion:
Use any xfrmr that is electrically compatible, and mount it outside the DMM inside it's own little metal box - sorta like a "wall-wart" PS setup.

Pete


Re: HP/Keysight/Agilent 34401A DMM power transformer

 

Hello Siggi

Thank you for your quick response!?

Thanks for the info from the service manual. I see that there are a few on eBay (all from China) listed for the 34401A, but the part number is 9100-5608...
From the photo it does look to be a compatible part, at least in terms of the winding terminations. At $70.00 (US) each, I guess it might be worth it if it brings my meter to life! I just hate buying from Chinese vendors (not because of politics, simply because delivery is slow and unpredictable) So I was hoping maybe someone from the group has one for sale (or a junk unit I could cannibalize).

Has anyone replaced one of these? I wonder how the transformer is physically attached to the chassis?

Vic


Re: Maintaining good Firmware EPROM, PROM, EEPROM etc.

 

Hello,

I can also recommend Batronix programmers. I have used BX48 Batego II almost two years without any problems. This model is quite pricy but worth of every euro. Batronix support is excellent, if you have older chips not mentioned in their chip support list, they will tell you which settings you should use .

BTW: Does anyone have extra HP 8751A CPU card? I don't have 3-pin row euro card extension cards or cables. Repair without pulling the CPU out from the frame is almost impossible. I have checked EPROMS and EEROMS and they are ok. I swapped CPU card from working HP 8751A and it works perfectly. With faulty CPU card it looks like working, but the source frequency is not sweeping. Internal tests don't give any errors, but step oscillator and FN oscillator tests returns "not assigned".

Regards
Ilkka, OH1AZM

-----Original Message-----
From: [email protected] [mailto:[email protected]] On Behalf Of magnustoelle via Groups.Io
Sent: torstai 14. maaliskuuta 2019 19.15
To: [email protected]
Subject: Re: [HP-Agilent-Keysight-equipment] Maintaining good Firmware EPROM, PROM, EEPROM etc.

[EXTERNAL Sender]


Good Day,

I have allowed myself a Bagero II programmer from Batronix many moons ago and I am very happy with it. Very stable software, which also supports Linux and MAC OS.

See

Cheers,

Magnus

I got a secondhand - sorry, pre-loved - TL866 off ebay for hardly any
$, downloaded the current software (I'm running Win7) and it's done
everything I've needed so far in terms of reading/programming backup
EPROMS for old HP gear.

On 3/13/2019 4:10 PM, Eric Haskell [KC4YOE] wrote:
I have a number of ageing instruments (mostly HP RF) and I am
concerned about "bit rot" for loss of program or cal data integrity.
What kind of programmer do you folks recommend to archive and restore
these devices from the 80's, 90's and 2000's? (It would also be nice
to work with more modern parts as a hobbyist.)

I have seen TL866 mentioned here. The GQ-4X Willem Programmer+ looks
interesting. Others? I am looking for an inexpensive programmer
that works with a modern laptop (USB) and supports older devices.

I think I remember reading that one of the group members offers a
service for providing programmed devices for some instruments?

Is this type of service available anywhere for a reasonable cost?

Regards,
Eric Haskell (KC4YOE)


Re: E8357A upgrade questions

 

I guess I should have also mentioned that I can provide the CPU and hard drive at no charge.? I would rather see instruments being properly used rather than being dumped in a landfill.

You will need to purchase a #2335 lithium button battery ($5) since the existing one is probably dead. This is just for the real time clock.?


Re: E8357A upgrade questions

 

I have bad news, bad news, bad news, bad news, bad news, and good news.

1. The 266MHz CPU is not worth using...nothing is supported with that.? And Win 2000 is no longer supported either.? And it is dog slow.
2. The max memory on that CPU is 128MB.
3. The 10GB drive is not worth using and is unreliable after ll this time, but you could clone it if you wish.
4.? It is unlikely that you could get the PNA to work at 9 GHz.? My memory is a bit hazy but I think the 6 and 9 GHz units are identical (or maybe it was the 3 and 6GHz?? or maybe all 3???), but without the keyword to unlock it, you are out of luck.
5.? While I could generate the proper keyword for it, I won't. Sorry.

The good news:? I have a few used 500MHz CPU boards available (with 512MB of RAM) along with a 40GB XP hard drive.? A 1100 MHz CPU would be even better but I do not have any and those are hard to come by.

You should send me your contact info/address.

CV