Re: BRIGHT Replacement LCD for 8753/8720 VNAs with dim LCD displays
Hello Ed,
I would like to have information about the keyboard and display you
have used with the 8753D. Thanks.
Steve, K8JQ
toggle quoted message
Show quoted text
Another option with
the HP8753D is to use an external display, a fairly common PC
display.
I have used one for
several years, and an external keyboard as well.? Both are
easily available
as PC
items, older and inexpensive.? The PC keyboard takes most of
the work out of labeling
the display
so the printed display is much better for your record keeping.
Back to the
PC video display.? Look at your 8753D manual and note the
frequency of the sweep.
It is lower
than the spec's of many video displays so match this when you
choose one.? Also check
which of
the three colors carries the sync signal.? That is critical.
Naturally
you have to match the connectors for all external
accessories.? Connect everything and look
at the
display.? It probably it needs the Horizontal width adjusted.?
Display a Smith Chart and adjust
the width
until it is round.? Easy!
I can go
check which display and keyboard I am using if there is any
interest in this addition.?
?
Ed
Lawrence WA5SWD
Physical Address: 93 Spring Haven Dr.,
Oldtown, ID 83822
Mailing Address: Box 313 Priest River, ID 83856
Cell 970-481-4491 Home 208-437-1378
?
Unfortunately the SimmConn kits are not
designed to replace the 8753D/E displays.
|
Re: BRIGHT Replacement LCD for 8753/8720 VNAs with dim LCD displays
Dr. David Kirkby (Kirkby Microwave Ltd)
|
Re: HP 16505a and 16500b/c scsi interface
On Wed, Feb 24, 2016 at 9:20 PM, Glen Slick <glen.slick@...> wrote: Just grab the FuturePlus FS2230 / FS2232 files:
(16700) (16500)
The 16700 files (DISK1.DAT, INDEX.TXT, INFO.TXT) need to be copied to a floppy and then you need to do a software installation from the files on the floppy.
This is an example of what you can see in a listing window using the FuturePlus config and IA tools with an E2423A SCSI Bus Preprocessor on a 16700 series logic analyzer. I truncated most of the columns from the listing window and the output is probably still too wide to display correctly in email. I captured this today from the start of a bus trace while looking into some SCSI controller/device issues. State Number Time FUTUREPLUS SYSTEMS c 1999 DATA PHASE Decimal Absolute Phase SCSI mnemonic U3/160M REV 1.0 Hex Symbols ____________ ___________ ______________________________________ ____ ___________ 0 0 s Arbitrate ID=None 0000 ST_DATA_OUT 1 584.000 ns Arbitrate ID=07 0080 ST_DATA_OUT 2 600.000 ns Arbitrate ID=07 0080 ST_DATA_OUT 3 4.364 ms Select ID=00 0081 ST_DATA_OUT 4 4.405 ms SDat_O 81H DBP=1 0081 ST_DATA_OUT 5 4.406 ms Commd Start/Stop 001B COMMAND 6 4.408 ms Commd Lu=0 0000 COMMAND 7 4.409 ms Commd 0000 COMMAND 8 4.410 ms Commd 0000 COMMAND 9 4.411 ms Commd Make Unit Ready 0001 COMMAND 10 5.154 ms Commd 0000 COMMAND 11 5.277 ms Status Check Condition 0002 STAT 12 5.489 ms Msg_in Command Complete 0000 MSG IN 13 5.490 ms Bus Free 0000 MSG IN 14 6.448 ms Bus Free 0000 ST_DATA_OUT 15 6.452 ms Arbitrate ID=None 0000 ST_DATA_OUT 16 6.452 ms Arbitrate ID=07 0080 ST_DATA_OUT 17 6.452 ms Arbitrate ID=07 0080 ST_DATA_OUT 18 6.453 ms Select ID=00 0081 ST_DATA_OUT 19 6.455 ms SDat_O 81H DBP=1 0081 ST_DATA_OUT 20 6.456 ms Commd Start/Stop 001B COMMAND 21 6.457 ms Commd Lu=0 0000 COMMAND 22 6.459 ms Commd 0000 COMMAND 23 6.460 ms Commd 0000 COMMAND 24 6.461 ms Commd Make Unit Ready 0001 COMMAND 25 7.144 ms Commd 0000 COMMAND 26 7.267 ms Status Good 0000 STAT 27 7.480 ms Msg_in Command Complete 0000 MSG IN 28 7.480 ms Bus Free 0000 MSG IN
|
Re: BRIGHT Replacement LCD for 8753/8720 VNAs with dim LCD displays
Another option with the HP8753D is to use an external display, a fairly common PC display.
I have used one for several years, and an external keyboard as well.? Both are easily available as PC items, older and inexpensive.? The PC keyboard takes most of the work out of labeling
the display so the printed display is much better for your record keeping.
Back to the PC video display.? Look at your 8753D manual and note the frequency of the sweep. It is lower than the spec's of many video displays so match this when you choose one.? Also check
which of the three colors carries the sync signal.? That is critical.
Naturally you have to match the connectors for all external accessories.? Connect everything and look
at the display.? It probably it needs the Horizontal width adjusted.? Display a Smith Chart and adjust
the width until it is round.? Easy!
I can go check which display and keyboard I am using if there is any interest in this addition.?
? Ed Lawrence WA5SWD Physical Address:
93 Spring Haven Dr., Oldtown, ID 83822
Mailing Address:
Box 313 Priest River, ID 83856
Cell 970-481-4491 Home 208-437-1378
On Wednesday, February 24, 2016 7:54 PM, "ovidiu.popa@... [hp_agilent_equipment]" wrote:
?
Unfortunately the SimmConn kits are not designed to replace the 8753D/E displays.
|
Re: Weinschel step attenuator gets stuck
Hello Tom, I also have a HP8593E - a little over a year now.? I really like it.? About a month after purchase it went blind - backup battery died.? I'm just glad it was not a blown front end.? Anyway I had to re-accomplish the leveling calibration points - over 100 points.? In retrospect, when I first got it , I should have gone through the procedure that records all these points then it just would have been a matter of reentering them from a hard-copy record.?
Regards rastro
On Thursday, February 25, 2016 8:45 AM, "iphi@... [hp_agilent_equipment]" wrote:
?
Hi Ian,
many thanks for the warm welcome. Now it's over with being incognito ;-) I'm trying to get into the high GHz field using high notch second hand HP equipment. Here's my complete HP inventory for the records: HP425A (excellent uV meter with valves, that I could rescue from junk and fix it successfully) HP3430A (DVM, also rescued from junk, nice, but a bit bulky, not used lately) HP34401A (impressive 6.5 digit multimeter, purchased recently) HP8350B+HP83529A (cool sweeper 10MHz...20GHz, purchased recently. I'm using it a lot. Quite instructive to see how various coax cables behave at 18GHz.) HP8593E (my latest purchase, 26.5GHz SA to replace a borrowed HP8591A). I might need advice on the HP8593E soon when I have time to take care. I have the impression that the highest band (19GHz...26GHz) shows 10dB to low amplitudes, verified with the sweeper as signal source up to 20.4GHz and double-checked against a Gigatronics power meter, which nicely matches the HP83529A output power display within a few tenths of a dB). The noise floor of that band displays 5dB lower that that of the previous band. I would have expected it to be 5dB higher than that of the previous band as harmonic mixing gets increasingly less efficient at higher harmonics. I wonder if this can be fixed with a software setting.
Best regards, Tom DG8SAQ
|
Joe, drop me a private post. If you can wait 3+ weeks until I can access my lab. Ill look to see if I have any AVANTEK GPDs of the correct spec. Ill send you one free if I do....Im really not certain if I have any of them left.
Im on holiday (vacation) in Malaysia at present and home is UK
Pete G4GJL
|
Wanted: spare internal 10 dB pad for HP 8562A A9 input attenuator
Anyone knows how to get?at decent price?a working order 22 or 26.5 GHz 10 dB glass replacement pad (so-called "Attenuation Card"), that's found in the 70 dB A9 input step attenuator of my 8562A SA? Several part numbers, electrically matching A9 may be found: 5086-7371 5086-7783 5086-5783 5086-7796 5086-7822 5086-7842 8340-60175 8673-60043 8673-60044 33321-60038 33321-60039 33321-60057 33321-60058 33321-60061 33321-60080 83595-60019 85660-60301 HP8495K HP 33321D/K/RF/RG/RU/RV/T HP 33323K N4413-80004
| So I'm just looking for that 10 dB pad, not necessarily a $100-300+ complete assembly. |
Please note: 18 GHz 10 dB pads mechanically don't fit.
|
Re: Weinschel step attenuator gets stuck
Hi Ian,
many thanks for the warm welcome. Now it's over with being incognito ;-) I'm trying to get into the high GHz field using high notch second hand HP equipment. Here's my complete HP inventory for the records: HP425A (excellent uV meter with valves, that I could rescue from junk and fix it successfully) HP3430A (DVM, also rescued from junk, nice, but a bit bulky, not used lately) HP34401A (impressive 6.5 digit multimeter, purchased recently) HP8350B+HP83529A (cool sweeper 10MHz...20GHz, purchased recently. I'm using it a lot. Quite instructive to see how various coax cables behave at 18GHz.) HP8593E (my latest purchase, 26.5GHz SA to replace a borrowed HP8591A). I might need advice on the HP8593E soon when I have time to take care. I have the impression that the highest band (19GHz...26GHz) shows 10dB to low amplitudes, verified with the sweeper as signal source up to 20.4GHz and double-checked against a Gigatronics power meter, which nicely matches the HP83529A output power display within a few tenths of a dB). The noise floor of that band displays 5dB lower that that of the previous band. I would have expected it to be 5dB higher than that of the previous band as harmonic mixing gets increasingly less efficient at higher harmonics. I wonder if this can be fixed with a software setting.
Best regards, Tom DG8SAQ
|
Re: Weinschel step attenuator gets stuck
Hi Rastro,
many thanks for the very informative links! From what I can tell, the Weinschel attenuators are mechanically absolutely identical with the HP ones, either one builts in license from the other or all come from the same source.
Best regards, Tom
|
Re: Need HP34401A GPIB help
Many thanks for helping me on the syntax. Works great! Somehow, I thought nothing ought to follow the question mark...
|
Hi Gary,
I had a similar problem with a HP5335A counter. What I did was to design a circuit board and place 10 series resistors on it, each 15 ohm, 0.25 W. These resistors were placed in series with the DC fan. I mounted the board inside the unit on the right side with four screws. Perfect, nice and quiet.
G?ran
|
Re: HP 3709B Constellation Analyzer
toggle quoted message
Show quoted text
----- Original Message ----- From: "f1gwr f1gwr@... [hp_agilent_equipment]" <hp_agilent_equipment@...> To: <hp_agilent_equipment@...> Sent: Thursday, February 25, 2016 9:28 PM Subject: Re: [hp_agilent_equipment] HP 3709B Constellation Analyzer A bit late but if you didn't find yet here's one: A the related HP Journal: Link is valid on a 30 days since last download basis. Later just drop a request direct to my email. Christian Le 21 nov. 2015 ??? 17:48, starmux64@... [hp_agilent_equipment] a ???crit : I have acquired one of these in good condition and apparently working. I have not been able to find a Manual for it online. Before I try powering it up, has anyone got a .pdf Manual that they can send me please?
It was used to display constellation patterns of digital radios with distortion measurements and is described in HP Journal of July 1987. It needed a high impedance interface kit 15709A which I don't have. It looks too good to scrap so I wondered if it had any other general purpose use.
Steve
|
Re: HP 3709B Constellation Analyzer
Bonjour Christian Merci beaucoup pour le manual et journal. Cordialement Steve
toggle quoted message
Show quoted text
From: "f1gwr f1gwr@... [hp_agilent_equipment]" To: hp_agilent_equipment@... Sent: Thursday, February 25, 2016 8:28 AM Subject: Re: [hp_agilent_equipment] HP 3709B Constellation Analyzer
?
A bit late but if you didn't find yet here's one: A the related HP Journal: Link is valid on a 30 days since last download basis. Later just drop a request direct to my email.
Christian
Le 21 nov. 2015 ¨¤ 17:48, starmux64@... [hp_agilent_equipment] a ¨¦crit :
?
I have acquired one of these in good condition and apparently working. I have not been able to find a Manual for it online.? Before I try powering it up, has anyone got a .pdf Manual that they can send me please?
It was used to display constellation patterns of digital radios with distortion measurements and is described in HP Journal of July 1987.? It needed a high impedance interface kit 15709A which?I don't have.? It looks too good to scrap so I wondered if it had any other general purpose use.
Steve?
|
Re: HP 3709B Constellation Analyzer
A bit late but if you didn't find yet here's one: A the related HP Journal: Link is valid on a 30 days since last download basis. Later just drop a request direct to my email.
Christian
toggle quoted message
Show quoted text
Le 21 nov. 2015 ¨¤ 17:48, starmux64@... [hp_agilent_equipment] a ¨¦crit :
?
I have acquired one of these in good condition and apparently working. I have not been able to find a Manual for it online.? Before I try powering it up, has anyone got a .pdf Manual that they can send me please?
It was used to display constellation patterns of digital radios with distortion measurements and is described in HP Journal of July 1987.? It needed a high impedance interface kit 15709A which?I don't have.? It looks too good to scrap so I wondered if it had any other general purpose use.
Steve?
|
Re: OT Rod terminating resistor
Yes, I asking for 3 samples and few days later was contacted by guy from France, company representing them for Europe. Typical story, what's project, annual prevision....and at end he come with small quantity "sample" price of about 25USD for BeO and 20USD for Alumina substrate!
BTW, I have very good experience with many other companies regarding the free samples of semiconductors, IC's and optoelectronics.
Milan
toggle quoted message
Show quoted text
?
Hi, Have you tried asking them for a sample? you would be surprised what results a nice letter or even email can get.
Robert G8RPI
|
Re: Spring loaded Old Style Front Panel Buttons
Ok that makes more sense! Thanks Chuck.
?
Gary? K4FMX
?
From:
hp_agilent_equipment@...
[mailto:hp_agilent_equipment@...]
Sent: Wednesday, February 24, 2016
11:19 PM
To:
hp_agilent_equipment@...
Subject: Re:
[hp_agilent_equipment] Spring loaded Old Style Front Panel Buttons
?
?
Oh, one
thing I didn't make clear, you have to remove
the panel circuit board from the panel bezel and overlay.
You do not have to, nor should you, remove the key tops
from the switches.
-Chuck Harris
Chuck Harris cfharris@... [hp_agilent_equipment] wrote:
> Hi Gary,
>
> What I do is remove the front panel from the instrument, and set
> it on a pan to catch the drippings. Then I take a little pump
> sprayer, like is used for windex, and liberally hose down the
> entire switch area with IPA. While I do this, I push the buttons
> multiple times to get the IPA under the springs, and wash out the
> remnants of the grease. I then hose it down a couple of times more
> with more IPA.
>
> After that is done, I let the panel drip, for a few minutes, and
> then blow the IPA off gently with compressed air. The more IPA
> you can remove mechanically, the cleaner the springs and switches
> will be.
>
> After blowing the IPA off, put the panel in a warm place, and let
> the remaining IPA evaporate, and you are back in business.
>
> -Chuck Harris
>
> 'Gary Schafer' garyschafer@... [hp_agilent_equipment] wrote:
>> Hi Chuck,
>>
>>
>>
>> Are you saying that you can clean the switches with IPA without taking
the
>> keyboard apart or removing it from the instrument? Do you just spray
it in
>> along side of the keys?
>>
>>
>>
>> I have an HP 3326A with keys that are kind of hard to push.
>>
>>
>>
>> Thanks
>>
>> Gary
K4FMX
>>
>>
>>
>> _____
>>
>> From: hp_agilent_equipment@...
>> [mailto:hp_agilent_equipment@...]
>> Sent: Wednesday, February 24, 2016 4:56 PM
>> To: hp_agilent_equipment@...
>> Subject: Re: [hp_agilent_equipment] Spring loaded Old Style Front
Panel
>> Buttons
>>
>>
>>
>>
>>
>> I have advocated washing the assembled keyboards with
>> 91 or 99% IPA before on this board. It changes the
>> character of the keys from something that feels like
>> you are going to do damage, to something that feels
>> nice.
>>
>> I think that HP put the grease on the springs to modify
>> the feel.... perhaps preferring a strong click? The grease
>> will stick the spring to the plunger until it abruptly
>> gives way, and the key pushed down.
>>
>> I think it was a bad idea given the adverse long term
>> effects of dried grease.
>>
>> I have left all of my keyboards dry, and they seem
>> perfect.
>>
>> -Chuck Harris
>>
>> Barry n4buq@... [hp_agilent_equipment] wrote:
>>> Hi Adrian,
>>>
>>> Okay - I may give that a try. I wonder if there was any grease
originally
>> in the
>>> mechanisms that has dried up and possibly should be replaced after
>> cleaning?
>>>
>>> Thanks, Barry - N4BUQ
>>>
toggle quoted message
Show quoted text
>>> ----- Original Message -----
>>>> From: "Adrian Nicol fenland787@...
[hp_agilent_equipment]"
>>>> To: "hp
agilent equipment"
>>>> Sent: Wednesday,
February 24, 2016
>>>> 9:52:43 AM Subject: Re: [hp_agilent_equipment] Spring loaded
Old Style
>> Front
>>>> Panel Buttons
>>>>
>>>> Hi Barry, short answer is yes, that is the reason and the fix
is dead
>> easy and
>>>> really effective, if my experience is anything to go by. There
is a
>> parallel
>>>> thread running where I and others have posted their solutions
- I just
>> did a
>>>> 3456A and 3455A as well as an 8566A and other gear - but this
is what I
>> did:
>>>>
>>>> Remove the switch PCB from the panel, flood spray the switches
with IPA
>> (Isoprop
>>>> alcohol) and operate each a few times, within just three or four
>> operations they
>>>> free up like magic going from clunky
>> 'feels-like-something-is-about-to-bust' to
>>>> almost silky smooth! Then blow it off with an air-gun and give
it another
>> dose
>>>> for luck. Works for me anyway!
>>>>
>>>>
>>>> On Wednesday, February 24, 2016 3:09 PM, "Chuck Harris
cfharris@...
>>>> [hp_agilent_equipment]"
>>
>>
>>
>>
>
>
> ------------------------------------
> Posted by: Chuck Harris
> ------------------------------------
>
>
> ------------------------------------
>
> Yahoo Groups Links
>
>
>
>
|
Re: HP 16505a and 16500b/c scsi interface
On Wed, Feb 24, 2016 at 7:00 PM, Marc Howard marc.howard@... [hp_agilent_equipment] <hp_agilent_equipment@...> wrote: I had the opposite problem. I had the disk but not the adapter. But they pop up often on eBay without the software.
I wish there was a 16700 port of that software.
Anyway, I'll try to pull it off my 16500 and upload it to the files section this weekend.
Marc
But there is. By coincidence I just installed the 16700 native config and IA files on a 16700B for an E2423A SCSI Bus Preprocessor. Just grab the FuturePlus FS2230 / FS2232 files: (16700) (16500) The 16700 files (DISK1.DAT, INDEX.TXT, INFO.TXT) need to be copied to a floppy and then you need to do a software installation from the files on the floppy. -Glen
|
Re: Spring loaded Old Style Front Panel Buttons
Oh, one thing I didn't make clear, you have to remove the panel circuit board from the panel bezel and overlay. You do not have to, nor should you, remove the key tops from the switches.
-Chuck Harris
Chuck Harris cfharris@... [hp_agilent_equipment] wrote:
toggle quoted message
Show quoted text
Hi Gary,
What I do is remove the front panel from the instrument, and set it on a pan to catch the drippings. Then I take a little pump sprayer, like is used for windex, and liberally hose down the entire switch area with IPA. While I do this, I push the buttons multiple times to get the IPA under the springs, and wash out the remnants of the grease. I then hose it down a couple of times more with more IPA.
After that is done, I let the panel drip, for a few minutes, and then blow the IPA off gently with compressed air. The more IPA you can remove mechanically, the cleaner the springs and switches will be.
After blowing the IPA off, put the panel in a warm place, and let the remaining IPA evaporate, and you are back in business.
-Chuck Harris
'Gary Schafer' garyschafer@... [hp_agilent_equipment] wrote:
Hi Chuck,
Are you saying that you can clean the switches with IPA without taking the keyboard apart or removing it from the instrument? Do you just spray it in along side of the keys?
I have an HP 3326A with keys that are kind of hard to push.
Thanks
Gary K4FMX
_____
From: hp_agilent_equipment@... [mailto:hp_agilent_equipment@...] Sent: Wednesday, February 24, 2016 4:56 PM To: hp_agilent_equipment@... Subject: Re: [hp_agilent_equipment] Spring loaded Old Style Front Panel Buttons
I have advocated washing the assembled keyboards with 91 or 99% IPA before on this board. It changes the character of the keys from something that feels like you are going to do damage, to something that feels nice.
I think that HP put the grease on the springs to modify the feel.... perhaps preferring a strong click? The grease will stick the spring to the plunger until it abruptly gives way, and the key pushed down.
I think it was a bad idea given the adverse long term effects of dried grease.
I have left all of my keyboards dry, and they seem perfect.
-Chuck Harris
Barry n4buq@... [hp_agilent_equipment] wrote:
Hi Adrian,
Okay - I may give that a try. I wonder if there was any grease originally in the
mechanisms that has dried up and possibly should be replaced after cleaning?
Thanks, Barry - N4BUQ
----- Original Message -----
From: "Adrian Nicol fenland787@... [hp_agilent_equipment]" <hp_agilent_equipment@...> To: "hp agilent equipment" <hp_agilent_equipment@...> Sent: Wednesday, February 24, 2016 9:52:43 AM Subject: Re: [hp_agilent_equipment] Spring loaded Old Style Front
Panel Buttons
Hi Barry, short answer is yes, that is the reason and the fix is dead easy and
really effective, if my experience is anything to go by. There is a parallel
thread running where I and others have posted their solutions - I just did a
3456A and 3455A as well as an 8566A and other gear - but this is what I did:
Remove the switch PCB from the panel, flood spray the switches with IPA
(Isoprop
alcohol) and operate each a few times, within just three or four operations they
free up like magic going from clunky 'feels-like-something-is-about-to-bust' to
almost silky smooth! Then blow it off with an air-gun and give it another dose
for luck. Works for me anyway!
On Wednesday, February 24, 2016 3:09 PM, "Chuck Harris cfharris@... [hp_agilent_equipment]" <hp_agilent_equipment@...> wrote:
------------------------------------ Posted by: Chuck Harris <cfharris@...> ------------------------------------
------------------------------------
Yahoo Groups Links
|
Re: Spring loaded Old Style Front Panel Buttons
Hi Gary,
What I do is remove the front panel from the instrument, and set it on a pan to catch the drippings. Then I take a little pump sprayer, like is used for windex, and liberally hose down the entire switch area with IPA. While I do this, I push the buttons multiple times to get the IPA under the springs, and wash out the remnants of the grease. I then hose it down a couple of times more with more IPA.
After that is done, I let the panel drip, for a few minutes, and then blow the IPA off gently with compressed air. The more IPA you can remove mechanically, the cleaner the springs and switches will be.
After blowing the IPA off, put the panel in a warm place, and let the remaining IPA evaporate, and you are back in business.
-Chuck Harris
'Gary Schafer' garyschafer@... [hp_agilent_equipment] wrote:
toggle quoted message
Show quoted text
Hi Chuck,
Are you saying that you can clean the switches with IPA without taking the keyboard apart or removing it from the instrument? Do you just spray it in along side of the keys?
I have an HP 3326A with keys that are kind of hard to push.
Thanks
Gary K4FMX
_____
From: hp_agilent_equipment@... [mailto:hp_agilent_equipment@...] Sent: Wednesday, February 24, 2016 4:56 PM To: hp_agilent_equipment@... Subject: Re: [hp_agilent_equipment] Spring loaded Old Style Front Panel Buttons
I have advocated washing the assembled keyboards with 91 or 99% IPA before on this board. It changes the character of the keys from something that feels like you are going to do damage, to something that feels nice.
I think that HP put the grease on the springs to modify the feel.... perhaps preferring a strong click? The grease will stick the spring to the plunger until it abruptly gives way, and the key pushed down.
I think it was a bad idea given the adverse long term effects of dried grease.
I have left all of my keyboards dry, and they seem perfect.
-Chuck Harris
Barry n4buq@... [hp_agilent_equipment] wrote:
Hi Adrian,
Okay - I may give that a try. I wonder if there was any grease originally in the
mechanisms that has dried up and possibly should be replaced after cleaning?
Thanks, Barry - N4BUQ
----- Original Message -----
From: "Adrian Nicol fenland787@... [hp_agilent_equipment]" <hp_agilent_equipment@...> To: "hp agilent equipment" <hp_agilent_equipment@...> Sent: Wednesday, February 24, 2016 9:52:43 AM Subject: Re: [hp_agilent_equipment] Spring loaded Old Style Front
Panel Buttons
Hi Barry, short answer is yes, that is the reason and the fix is dead easy and
really effective, if my experience is anything to go by. There is a parallel
thread running where I and others have posted their solutions - I just did a
3456A and 3455A as well as an 8566A and other gear - but this is what I did:
Remove the switch PCB from the panel, flood spray the switches with IPA
(Isoprop
alcohol) and operate each a few times, within just three or four operations they
free up like magic going from clunky 'feels-like-something-is-about-to-bust' to
almost silky smooth! Then blow it off with an air-gun and give it another dose
for luck. Works for me anyway!
On Wednesday, February 24, 2016 3:09 PM, "Chuck Harris cfharris@... [hp_agilent_equipment]" <hp_agilent_equipment@...> wrote:
|
Re: HP 16505a and 16500b/c scsi interface
toggle quoted message
Show quoted text
?
Thanks, that would be great to have the?E2423A?files online.? That's a bit of sad news that they don't work with 16700.? Though it justifies keeping a 16500 around.
|