wanted clip for E4432B output board E4400-60188

 

¿ªÔÆÌåÓý

Hello Can anyone help me with that. I have a Fet fault where bias circuit is not working correct either. I must the version that uses the H1 Fet (SHF-189)
Best regards Peter OZ1LPR


Re: HP 8566B - Can my high voltage transformer be faulty?

 

I have replaced several in the 85662A many, many years ago. There are 2 types depending whether you have a long CRT (PN 5083-5791, IIRC), or a shorter CRT newer type dot raster scan CRT (Panasonic). The HV transformer is inside the metal box at the rear, left of the CRT (A1A6 circuit). They usually fail open. You should see a shrunk image on the CRT when the HV transformer fails, both horizontally and vertically, or a dim dot. You will still have a good display if you check the video and horizontal outputs on an oscilloscope.

Don Bitters


Re: HP free e-book

 

Gianni, this is a work of art.? Very nicely done!

Peter

On 7/12/2024 5:23 PM, Gianni Becattini via groups.io wrote:
A new edition of the e-book "The Great HP" has been released and can be freely downloaded from my page www.k100.biz <>. I removed the Advantest R3132 (now in a separate ebook) and added various instruments. It has now more than 1,000 pages.

To make easier the download, I split the file in three parts which can be downloaded separately.

Tell me if you have problems - any suggestion welcome.


Re: Made one out of workshop scrap to enable the correct tightening of SMA 's

 

On July 12, 2024 9:35:13 PM "Steve Hendrix" <SteveHx@...> wrote:
That is what happens when you try to use the incorrect word for a spanner.
...or was that spammer? <couldn't resist>
You're a mean, bad man Steve. ;)

-Dave

--
Dave McGuire, AK4HZ
New Kensington, PA


Re: Made one out of workshop scrap to enable the correct tightening of SMA 's

 

On 2024-07-12 6:09 PM, george edmonds via groups.io wrote:
That is what happens when you try to use the incorrect word for a spanner.
...or was that spammer? <couldn't resist>


Re: Bubble Memory Reseed

 

Thanks for the responses. I did see the Hackaday project to build a Bubble Seed Module, but it was for a SBC-85 board. Evidently, HP built or had built their own module. Volume I of the service manual shows a (grainy on my copy) photo of the device. You plug it in and run Test Routine 169.

Ed, the Boot Loop routine you referred to can be run with loading a program from the manual with a HP-1B controller. Maybe, that is all I need. The program would have to be hand typed in, and I'd have to find a way to emulate the controller. I have a Prologix USP Controller. I'll have to figure out how to use it.

The history of Bubble Memory is interesting. Turned out to be a blind alley, but it was a creative idea.

Thanks again,
John


Re: Made one out of workshop scrap to enable the correct tightening of SMA 's

 

That is what happens when you try to use the incorrect word for a spanner.

G Edmonds
On Friday 12 July 2024 at 22:06:05 BST, Tom Lee <tomlee@...> wrote:


She must've been very strong...

(sorry, couldn't resist)


On 7/12/2024 10:10 AM, george edmonds via groups.io wrote:
> Hi John
>
> Used a wench of Chinese origin






HP free e-book

 

A new edition of the e-book "The Great HP" has been released and can be freely downloaded from my page . I removed the Advantest R3132 (now in a separate ebook) and added various instruments. It has now more than 1,000 pages.

To make easier the download, I split the file in three parts which can be downloaded separately.

Tell me if you have problems - any suggestion welcome.


Re: Made one out of workshop scrap to enable the correct tightening of SMA 's

 

soda -> keyboard

On July 12, 2024 5:06:11 PM "Tom Lee" <tomlee@...> wrote:
She must've been very strong...

(sorry, couldn't resist)

On 7/12/2024 10:10 AM, george edmonds via groups.io wrote:
Hi John

Used a wench of Chinese origin


--
Dave McGuire, AK4HZ
New Kensington, PA


Re: Made one out of workshop scrap to enable the correct tightening of SMA 's

 

She must've been very strong...

(sorry, couldn't resist)

On 7/12/2024 10:10 AM, george edmonds via groups.io wrote:
Hi John

Used a wench of Chinese origin


Re: VNA calibration kit to 50 GHz

 

I have an old 85056A that I bought 15 years ago from a company that has never used it since the first day they bought it new from HP. With it I also bought their 8510B 40 GHz that I converted into an 8510C. 20 years ago I had bought an 8510C 26.5 GHz with its 85052B still coming from an other company that was looking to modernize with a PNA.

Up to 26.5 GHz it's possible to do a fairly simple calibration with fixed loads and open and short circuits. But if you want to be precise, it's better to do a full calibration with the same elements plus sliding loads. Then I stocked up on the elements needed for TRL calibrations, but I've never had the courage to use them yet, as they're relatively more complicated.

I also had an 85053A individually calibrated for over $3000 less than 10 years ago at Keysight Roseville (the only place where they know how to do it according to Keysight), so I was able to check several times the difference between with and without sliding loads.

On the other hand, I was able to verify that for 40 GHz, it's really mandatory to use sliding loads because fixed loads aren't good enough to really know what we're doing. That's why I have also one (unfortunately) 2.4mm air line.

I think we should bear in mind that up to 50 GHz this is even truer, and that without sliding loads, without air lines, and without an individually calibrated verification kit (which is almost never the case) to check the result of what we're doing, we can only be in the fog.

For some years now, as I've run out of space, I've stored the two 8510Cs at my country house, and all I have at home is a small (but fantastic) 8720ES.

The 8510C is a slow, impractical instrument, with no computing power, no possibility of storing tons of configurations in internal memory, and no ability to interpolate if the frequency limits are changed. However 8510C is still a very accurate instrument that I was able to compare to several from R&S and HPAK... It's only PNA-X that was better in term of stability along time.

For better verification of all calibration parameters, option 010 is practically mandatory. Fortunately, I have it on all three of my VNAs.

F1EKU.


Re: HP1743A scope

 

I have a todo list of things already very full but I begin to realize that indeed, as you said, it is worth the effort.
Of course all your tips are invaluable Raymond.
As I said I got with the scope, in the pouch, the "Operating and service manual" in an as-new condition. This is maybe silly but it's nearly as much pleasure than the scope itself! Aren't we all nuts?
Thanks,
Renaud


Re: Made one out of workshop scrap to enable the correct tightening of SMA 's

 

We torque our stainless steel SMAs to 8 in-lbs, not 10.

Regarding 5/16 inch wrenches with the slot to slip over coax, Mini-Circuits part number HT-4-SMA is about $37.? Nice knurled handle for good grip, too.? Found one in the lab this morning.? ? HTH.? ? Jim Ford?


On Fri, Jul 12, 2024 at 10:10 AM, george edmonds via groups.io
<G6HIG@...> wrote:
Hi John

Used a wench of Chinese origin, was very lightly case hardened so was possible to drill with a normal HSS drill bit in a battery drill. Would never attempt to do so with one of US manufacture.

I repair a lot of kit that use SMA and SMC connectors and being a nerd I like to torque the connectors up correctly which in the case of SMA is 5 in-lbs for brass ones and 10 in-lbs for stainless steel ones..

G Edmonds

On Friday 12 July 2024 at 16:04:49 BST, John Griessen via groups.io <john@...> wrote:


On 7/12/24 07:13, george edmonds via groups.io wrote:
> Made one out of workshop scrap

Good going!

Did you drill that wrench handle as-is, or after heating and annealing so it drilled more easily?? Or is it not a drill hole I see
there?






Re: How to remove the output cable of the isolator in the 8672A

 

¿ªÔÆÌåÓý

Flare wrench set available from Harbor Freight for <$20. 5/16 should work for SMA.

Just a thought.

On Jul 12, 2024, at 11:41?AM, Erik Kaashoek <erik@...> wrote:

?Its a bit overkill but it may work if there is sufficient space
<inline.0.part>


Re: VNA calibration kit to 50 GHz

 

many thanks for advice, Neil


Re: Made one out of workshop scrap to enable the correct tightening of SMA 's

 

Hi John

Used a wench of Chinese origin, was very lightly case hardened so was possible to drill with a normal HSS drill bit in a battery drill. Would never attempt to do so with one of US manufacture.

I repair a lot of kit that use SMA and SMC connectors and being a nerd I like to torque the connectors up correctly which in the case of SMA is 5 in-lbs for brass ones and 10 in-lbs for stainless steel ones..

G Edmonds

On Friday 12 July 2024 at 16:04:49 BST, John Griessen via groups.io <john@...> wrote:


On 7/12/24 07:13, george edmonds via groups.io wrote:
> Made one out of workshop scrap

Good going!

Did you drill that wrench handle as-is, or after heating and annealing so it drilled more easily?? Or is it not a drill hole I see
there?






Re: VNA calibration kit to 50 GHz

 

I don't think you want to do SOLT calibration at 50GHz with elements that you do not really know a whole lot about.

Get yourself a decent Thru element, an arbitrary (but good mechanical quality) reflect and a few decent 50 Ohms air lines.
Then you can do a TRL cal and you do not have to rely on the quality of reference data. At low frequencies TRM is a good alternative if you have a good low frequency match element.

I have experimented a lot with not very well known cal elements up to 27GHz and TRL was the only way to achieve a decent calibration, not to speak of 50GHz.

Best regards,
Tom


Re: VNA calibration kit to 50 GHz

 

On Fri, 12 Jul 2024 at 12:41, Marcus Gustafsson via <mankan=[email protected]> wrote:
Hi all,

I've have no idea of their quality but these are one of the cheapest I've found that still seems trustworthy.


However I've been looking at 40GHz, 2.92mm connectors and then you can get away even cheaper


Best regards
Marcus

The specifications don't make sense to me.?


VSWR of load < 1.17
VSWR of thru < 1.15
VSWR < 1.065

So what is the last VSWR? I don't see any mention of coefficients, only specs on the open (+/- 6 deg) and short (+/- 4 deg). Normally that would be quoted as some deviation from a model, but that's not stated.


--
Dr David Kirkby Ph.D
Kirkby Microwave Ltd
Email: drkirkby@...
Web:
Telephone 07910 441670 (UK) or +44 7910 441670 (international)
Registered in England and Wales, company number 08914892.?
Registered office: Stokes Hall Lodge, Burnham Rd, Althorne, Essex, CM3 6DT



Re: How to remove the output cable of the isolator in the 8672A

 

Its a bit overkill but it may work if there is sufficient space


Re: Made one out of workshop scrap to enable the correct tightening of SMA 's

 

On 7/12/24 07:13, george edmonds via groups.io wrote:
Made one out of workshop scrap
Good going!

Did you drill that wrench handle as-is, or after heating and annealing so it drilled more easily? Or is it not a drill hole I see there?