¿ªÔÆÌåÓý

Date

Re: k&s 1470 Wire bonding reconstruction

 

Hello,

I was also interested in getting a K&S wire bonding machine for the same
purpose to fix the HP YIGs but the ones that I found were missing parts.

I was looking for any manuals that showed the parts list so I can try to
identify exactly the missing parts but I found it impossible to find any
K&S manuals.

I am interested if anyone has user manuals and especially service
manuals for the 412x and 452x series of wire bonding machines from K&S.

One of the parts that the ones I found was missing was the heated
working table and I found exactly zero on eBay.

Let us know how is your project developing. I am more than interested in
updates.

Regards,
Razvan

On 24/06/2024 23:08, iv3hed via groups.io wrote:
Thank you Jim for your helpfulness.

If anyone in the group has any questions about the project I am more
than happy to answer them.

Regards,
Massimo
_._,_._,_


Re: k&s 1470 Wire bonding reconstruction

 

No problem, Massimo, I am happy to help out.? I did get a response from one of my contacts to reach out to another person who may be able to find you a service manual.? I need to look up that person;
?probably won't have time until tomorrow evening.? ? Jim


Thank you Jim for your helpfulness.

If anyone in the group has any questions about the project I am more than happy to answer them.

Regards,
Massimo


Re: k&s 1470 Wire bonding reconstruction

 

Thank you Jim for your helpfulness.

If anyone in the group has any questions about the project I am more than happy to answer them.

Regards,
Massimo


Re: 8566B - Low signal on 2-22 GHz band

 

On Sun, Jun 23, 2024 at 02:34 PM, Don Bitters wrote:
Don Bitters

Hi Don, thank you for your reply. I'm not very versed at all in the 8566B; I recently picked this unit up locally for a very good price, but was not able to test the 2-22 GHz section until I got it home. So I'm coming into this was no experience on the 8566. I appreciate your suggestions and will give them a try. 6 GHz is the highest source I have to test with at the moment.?

I'll be sure to post my findings, though it might be a day or two until I can get to it. I also have a thread on the EEV blog that includes everything I've done up until this point.



Re: looking for anritsu sitemaster s331a circuit diagram

 

Thank you. ?I will give it a go !


Re: HP 3325

 

Hi Martin,

That is the right idea. Now to wait for one to come up
in the e-scrap...

-Chuck Harris


On Mon, 24 Jun 2024 11:06:37 +0200 "Martin via groups.io"
<musaeum@...> wrote:
Hi Chuck,

I've seen a special book-scanner in an office of a large contractor,
many years ago, that seemed to be quite efficient when it comes to
"flipping page-by-page" of binded books.

It looked a bit like this one:


cheers
Martin


Re: HP 3325

 

Interesting.? Those direct links give me download options whereas my previous searchings from their home page did not.

Peter

On 6/23/2024 11:49 PM, tgerbic via groups.io wrote:
Try pasting this in your browser:



Dozens of National databooks, all downloadable. Motorola is the same:



The amount is less off this search but everything is downloadable.

Here is what a normal search for databooks looks like.
Add national databook in the search box and press GO




The search produces lots of databooks. Pick one.



Download options are at the bottom right. Pick PDF


--

T. Gerbic
Central California


Re: HP 3325

 

Hi Chuck,

I've seen a special book-scanner in an office of a large contractor, many years ago, that seemed to be quite efficient when it comes to "flipping page-by-page" of binded books.

It looked a bit like this one:


cheers
Martin

On 23. Jun 2024, at 17:36, Chuck Harris <cfharris@...> wrote:
...
If I have to go through it page-by-page, I may as well
be hand flipping the pages.

-Chuck Harris


Re: HP 3325

 

On Sun, Jun 23, 2024 at 10:36 AM, Chuck Harris wrote:
I have been on the lookout for a book scanner that would allow me to scan the books, manually but accurately,
in their bound state.
Sounds like YouTube's CuriousMarc found exactly what you're looking for here:



It's the CZUR 3 IN 1, 4k Document Camera-Fancy S Pro, please get it at CZUR's official store:
Watch the video for details.


Re: HP 3325

 

So Joe White how is your repair going?

On Thu, Jun 20, 2024 at 8:03?PM Joe White via groups.io
<skinnershorse@...> wrote:

These are from the power supply board of a 3325. Per OSM they are p/n 03325-66901, 03325-66902, 03325-66903 (left to right), described as "XSTR ASSEMBLY". The manufacturer is listed as HP. I have not found the part numbers in any cross-reference lists and I have been unable to identify the parts by their markings. Any help on identifying the parts and/or replacements is greatly appreciated.
--
Joe White
KW4YW


--


-----BEGIN PGP PUBLIC KEY BLOCK-----
Version: GnuPG v2

mQENBFYy4RYBCAC183JomLtbdAlcKiaPDoVHq52LDmVmH75aiEc69m7YxDt54/ai
VtYCAobbGVIyn3Hlz3uhF6LnPl/6Lm1VdnCfpwu3KQhCO6ds10ow2C30X4ohCqOd
hCVg5C+ILmQkEffFrFODy3ji+PYTF4pADvHCWsTMv0hf0llwFOJsBCK6cl02IffE
JPqy4PjM1nZ9HpzT84JBaG/4OGvTZ8SQ2yFUl265jagvygPTf88H1xpZHH1r8dB1
stjUHLmPH8AOyDgKxFchgGeDc3p/vJtgDDIXAFfDXG0NSRovLmtaQdGxe47Zf/go
bXiEM7YL2WqQe5zfEA919JxkEwlDKYniOSVzABEBAAG0N0V2YW4gRm9zcyAoVGhp
cyBpcyBteSBwdWJsaWMga2V5LikgPGV2YW5mb3NzQGdtYWlsLmNvbT6JATkEEwEC
ACMFAlYy4RYCGwMHCwkIBwMCAQYVCAIJCgsEFgIDAQIeAQIXgAAKCRCIpQTcE8nN
bbBaCACAm8pU5lG1ev2Fsw68Axtcl57SJrYieqX96c3YuYH9JpqMqJRnd9nDKw9X
tQuvuH7tUk0VbOaDqReOYJVI/4c5wb9AaOFp6K2DUcupq6XhgXpvz3HzoPwjAdIj
XuQzdRUx5+innTJrSkGuBYW/CZ2zqEx4xfLlq4rO0hoTUMR8QVp2cCrkw6BT0m86
APIw/ZnjoxM8IEzr7MxfRIg3qpzrZk28rmhx+k78Jyk61UhwcCPGIm/pjUopTwYJ
3YBdRB2cYD2aN7A1JVf5cRmSQYooHBGpH0kYvomGk97PKqypVuJ7OpG9xM58wUcC
qUVt9hKlePLzP8csYjt8onqI7qIIuQENBFYy4RYBCADlH8spG3WkCx62vB5mr5Z0
SCDd/RcyA4A5y5EOj5KurQkrSWpgi9Ho1yKruMJ6blQR2qkc66KqH9pnXDm/ZI1M
K/wdW3ngETxBmXoozzFMT89aEWIVR5/PFodWK1elekE9iJxACuR98Zg2QttTD3x8
A9w8VEyMLOXcDTrPFpHegMKswFBg5iuMulAdXAoGejWTI3n+qKFpabHm2Lfs6wjk
5rjucpTdeFK6UeWF1xAvNxXibuu5BlGwv53930qIXRwO/Gn2Rh5DXWxKU2fEIme/
xgQQmIsDeUoWbfybdjw/x7Q0LW4mINiLDQcGHHRQKFIxbAJCT3USPLGh5xwE9/Er
ABEBAAGJAR8EGAECAAkFAlYy4RYCGwwACgkQiKUE3BPJzW0uYAf9Hf30n8tM3mR2
Zo6ESE0ivgdgjaJtAWrBUx7JzAzPjBnBOlNnu5Y9lVEqetvUPH6e3PvaHYUuaUU8
0HwxuKBW9nUprgV6uIu1DZmlcp+SxpbuCy7RDpNocRLNWWFMaYYzznmTgfnTgD4D
gCq8Mf1mcfrluTkOAo+QNqbMfl1GISClopRqxVuAo59ewgMnFujwgd8w12BwWl24
CzqOs5HqcUslePj+LzcjSNgVCklYwKl+0dsb/fctMOCtHodwqm2CBJ+zydvNmYkD
fxda/J91Z1xrah5ec++FL0L4vs+jCiIWJeupJFKlr1hCMZiiGH7W554loK5l4jv3
EY347EidAw==
=Ta4p
-----END PGP PUBLIC KEY BLOCK-----


Re: HP 3325

 

Try pasting this in your browser:



Dozens of National databooks, all downloadable. Motorola is the same:



The amount is less off this search but everything is downloadable.

Here is what a normal search for databooks looks like.
Add national databook in the search box and press GO




The search produces lots of databooks. Pick one.



Download options are at the bottom right. Pick PDF


--

T. Gerbic
Central California


Re: HP 3325

 

Ok, I found some.? But they aren't downloadable; they can only be "borrowed" for an hour or viewed in their online viewer.

I guess I'm keeping my hundreds of data books.

On 6/23/2024 9:00 PM, tgerbic via groups.io wrote:
Peter,
Go to archive.org and just use the search box in the white rectangular area and press Go. Don't select Books, or anything else on the top line.
--
T. Gerbic
Central California


Re: HP 3325

 

Peter,
Go to archive.org and just use the search box in the white rectangular area and press Go. Don't select Books, or anything else on the top line.
--
T. Gerbic
Central California


Re: HP 3852a extended memory modules info

 

On 6/21/24 20:17, Maciej Kawalkowski wrote:
Mistakes have been made and 1 bodge wire is required. But it's working.
Niiiiice! Excellent work. I definitely want one of these.

-Dave

--
Dave McGuire, AK4HZ
New Kensington, PA


Re: HP attenuator 08662-60082

 

I thought I'd follow up on my own post. I made a suitable PCB layout in 0.6mm FR4 and had some PCBs assembled with ENIG finish.? They use minicircuits LAT series attenuators (for the first section 3 LAT-20s since the ratios are 20dB and 40dB).



(original HP can be seen here after most of the gold had fallen off )

The gold contacts needed some cleaning with deoxit gold (I guess the plating is very thin and the green corrosion comes through).? The ENIG gold is a lot thinner than that on the original HP board and I put a 0.05mm shim underneath the PCB to compensate.? The result on the VNA seems reasonable over 10MHz to 1.25GHz.? I used a board with ground plane. Still not sure this was the best choice, but it was the one most familiar to me.

Now I have to figure out how to update the leveling EPROM, assuming I can measure the attenuators sufficiently accurately on the VNA. PCB layout is here if anyone is interested ?).? I don't suppose anyone has any documentation?


Alan


Re: 8566B - Low signal on 2-22 GHz band

 

Pete, How versed are you with the operation of the 8566B?
Try this first, there is a key on the RF section (it might be a shifted key) call PSP. This is preselector peaking. Do this at 2.6 GHz or 3 GHz, then at 6GHz (this is close to the band edges of the 1st preselected band), if you can see a signal after that, that is much higher in amplitude then you may have an alignment problem. Look up how you do manual preselector peaking and for each of the upper frequency bands (2-6.2GHz, 5.8-12.5GHz, 12.5-18.6GHz, 18.6-22GHz). You should check the low, middle, and high end of each preselected band. Ideally the DAC numbers should all be 32 +/- 10 (range is 0-63), anything outside of this range indicates a frequency realignment may be needed. If you still cannot see a signal that is close to the correct amplitude then you may have a component failure in the preselected RF/uWave path.
Pleas post your findings so that we can point you in the right direction .
Don Bitters


Re: Interfacing the HP3490A with AR488

 

Hi Chuck and Dave,

Thank you for your responses. I'm quite a newbie to HP-IB, and I'm learning !

At this time I am able to use the AR488 successfully with a 34401A and a 3457A, in both cases sending instruction codes (to select DCV, ohms, ACV, ranges, etc ...) and receiving measurement results. So far so good.

I did some further reading of the 3490A manual, and table 3-6 on page 3-9 would imply that the 3490A must use two different addresses. For example, it is shipped from the factory with the jumper configuration shown in figure 3-7. This sets an HP-IB address of 66 octal (54 decimal) which is for listening. But this same jumper configuration sets the talk address to 126 octal (86 decimal), which I understand is the address from which the 3490A will talk from, when sending results.

This dual-addressing mode is somewhat puzzling ...

Could anyone with experience of the 3490A on the HP-IB confirm my understanding ?

For me the next step would then be to use this "uncommon" addressing scheme with the AR488.

Thanks to all,

Joel


Re: 8753C Sampler Replacement?

 

I have replaced bad samplers in the past (on HP8753ES instruments).? And yes, when you replace a sampler or a port coupler you do have to run through a calibration proceedure to get everything just right.? The calibration proceedure isn't very intense and can be run in about 30 minutes, but it does require a few insturments (power meters, special cables, etc) to do it.? The power meter requirements vary depending on whether you have a 0-3 GHz or the 6 GHz option, but that might be only on "D" and up that you have to worry about this.

But if you fail to do the calibration then all that really happens is that your baseline response prior to doing a cal is a bit off.? Once you run through a normal 50 ohm cal (which of course you would do prior to any measurement), you are going to get pretty decent results.? So, long story short if you can do the recalibration, great - that's recommended.? If you can't, not ideal, but as long as you do a proper cal prior to making measurements you should be fine.

Depending on where you are at,? one of us might be able to loan you the equiptment needed to do a cal.? It's really not that extensive.


Re: HP 3325

 

As others have said, linear supply pass transistors usually aren't super critical and you can frequently figure out something to replace them with that will be at least as good.? I also look at manuals for similar equipment if I get stuck, as you suggest.

On 6/23/2024 11:54 AM, george edmonds via groups.io wrote:
Hi

To go back to the original topic.

Some time ago I repaired a HP3325B that had suffered a fan failure whilst housed in a rack, this resulted in a seriously damaged power supply from over heating, From memory, which now may not be as good as it was thanks to a Covid infection, the pass transistors were easy to come by as they al had commercial type numbers on them, the only odd thing was the use of heat shrink tubing to make connections to the transistor legs.

Suggest looking at the HP3325B service manual for type numbers for the HP3325A pass transistor type numbers as the power supply is similar.

G Edmonds

On Sunday 23 June 2024 at 16:49:10 BST, Dave McGuire <mcguire@...> wrote:



? At LSSM we use a Czur ET25-Pro for this sort of thing. The guys
doing the scanning say it takes a bit of getting used to, but it
generally does a pretty good job.

? ? ? ? ? ? -Dave

On 6/23/24 11:36, Chuck Harris wrote:
I thought so enough to save mine, and started to scan
them.? To that end I acquired a paper slicer that can
cut the binding off in a single stroke, but the project
stalled, as the paper used is very thin, and doesn't
auto feed well at all.? It got too laborious to have
to go through the scanned books a page at a time to
correct all of the pages that got fed together by the
scanner's feed rollers.

I have been on the lookout for a book scanner that
would allow me to scan the books, manually but accurately,
in their bound state.

If I have to go through it page-by-page, I may as well
be hand flipping the pages.

-Chuck Harris


On Sun, 23 Jun 2024 11:11:06 -0400 "Peter Gottlieb"
<hpnpilot@...> wrote:
Perhaps these hard copies of data books will be worth their weight in
gold when all the libraries are shut down.? At least worth something
to engineers like us.


On 6/23/2024 11:01 AM, Chuck Harris via groups.io wrote:
On March 26, 2023, a federal judge sided in favor of the four
leading publishers in the US, who sued the Internet Archive for
scanning and lending out numerous digital copies of their
published works.

In April, The internet Archive submitted a final appellate
reply brief fighting the judge's ruling...And the fight goes
on...

I can't help but note that Google does exactly the same thing,
but seems to be inured against legal harm...

-Chuck Harris


On Sun, 23 Jun 2024 10:03:17 -0400 "Peter Gottlieb"
<hpnpilot@...> wrote:
I went to archive.org, selected "books" then entered "Motorola data
book" and get 16 responses.? 13 are "not in library" and 3 are
available to "borrow" if I log in.

How specifically are you getting to where data books can be
downloaded?

I have shelves of data books.? If I could get searchable pdfs of
them I would give away or recycle the books, which would give me
more space for actual test equipment.

Peter


On 6/23/2024 12:49 AM, tgerbic via groups.io wrote:
Not sure how you are looking these up but I just? go to
archive.org, enter something like "Motorola data book" and I get a
databook I can download in PDF or PDF with OCR.? Searching for
"transistor cross reference" brought up four books that can be
downloaded.? Perhaps it is the way you search for them. --
T. Gerbic
Central California














--
Dave McGuire, AK4HZ
New Kensington, PA







Re: HP 3325

 

Hi

To go back to the original topic.

Some time ago I repaired a HP3325B that had suffered a fan failure whilst housed in a rack, this resulted in a seriously damaged power supply from over heating, From memory, which now may not be as good as it was thanks to a Covid infection, the pass transistors were easy to come by as they al had commercial type numbers on them, the only odd thing was the use of heat shrink tubing to make connections to the transistor legs.

Suggest looking at the HP3325B service manual for type numbers for the HP3325A pass transistor type numbers as the power supply is similar.

G Edmonds

On Sunday 23 June 2024 at 16:49:10 BST, Dave McGuire <mcguire@...> wrote:



? At LSSM we use a Czur ET25-Pro for this sort of thing.? The guys
doing the scanning say it takes a bit of getting used to, but it
generally does a pretty good job.

? ? ? ? ? ? -Dave


On 6/23/24 11:36, Chuck Harris wrote:
> I thought so enough to save mine, and started to scan
> them.? To that end I acquired a paper slicer that can
> cut the binding off in a single stroke, but the project
> stalled, as the paper used is very thin, and doesn't
> auto feed well at all.? It got too laborious to have
> to go through the scanned books a page at a time to
> correct all of the pages that got fed together by the
> scanner's feed rollers.
>
> I have been on the lookout for a book scanner that
> would allow me to scan the books, manually but accurately,
> in their bound state.
>
> If I have to go through it page-by-page, I may as well
> be hand flipping the pages.
>
> -Chuck Harris
>
>
> On Sun, 23 Jun 2024 11:11:06 -0400 "Peter Gottlieb"
> <hpnpilot@...> wrote:
>> Perhaps these hard copies of data books will be worth their weight in
>> gold when all the libraries are shut down.? At least worth something
>> to engineers like us.
>>
>>
>> On 6/23/2024 11:01 AM, Chuck Harris via groups.io wrote:
>>> On March 26, 2023, a federal judge sided in favor of the four
>>> leading publishers in the US, who sued the Internet Archive for
>>> scanning and lending out numerous digital copies of their
>>> published works.
>>>
>>> In April, The internet Archive submitted a final appellate
>>> reply brief fighting the judge's ruling...And the fight goes
>>> on...
>>>
>>> I can't help but note that Google does exactly the same thing,
>>> but seems to be inured against legal harm...
>>>
>>> -Chuck Harris
>>>
>>>
>>> On Sun, 23 Jun 2024 10:03:17 -0400 "Peter Gottlieb"
>>> <hpnpilot@...> wrote:
>>>> I went to archive.org, selected "books" then entered "Motorola data
>>>> book" and get 16 responses.? 13 are "not in library" and 3 are
>>>> available to "borrow" if I log in.
>>>>
>>>> How specifically are you getting to where data books can be
>>>> downloaded?
>>>>
>>>> I have shelves of data books.? If I could get searchable pdfs of
>>>> them I would give away or recycle the books, which would give me
>>>> more space for actual test equipment.
>>>>
>>>> Peter
>>>>
>>>>
>>>> On 6/23/2024 12:49 AM, tgerbic via groups.io wrote:
>>>>> Not sure how you are looking these up but I just? go to
>>>>> archive.org, enter something like "Motorola data book" and I get a
>>>>> databook I can download in PDF or PDF with OCR.? Searching for
>>>>> "transistor cross reference" brought up four books that can be
>>>>> downloaded.? Perhaps it is the way you search for them. --
>>>>> T. Gerbic
>>>>> Central California
>>>>>? ? ?
>>>>
>>>>
>>>>
>>>>
>>>>?
>>>
>>>
>>>
>>>
>>>?
>>
>>
>>
>>
>>
>>
>
>
>
>
>
>

--
Dave McGuire, AK4HZ
New Kensington, PA